Use the Sample Size switches

Juerg Koegel

2012/10/16 Eric Haskell <eric_hask...@hotmail.com>:
> Thanks for all the responses.
>
> Well it was even simpler than I expected,  All I had to do was a cold reset 
> followed by an autocal. the unit now will power-up with the display active.  
> Apparently it was not a hardware error at all.
>
> I now want to calibrate the internal 10 MHz reference.  I read the cal 
> procedure several times. I need to set the value of calbyte 4 (and 50 to 
> minimize offset).  I found and enabled the cal jumper.  I accesses the cal 
> menu.  I displayed and can edit the calbyte 0 but I can't seem to figure out 
> how to navigate to the next calbyte.  How do I select the calbyte to operate 
> on?
>
> Eric Haskell
>