Rob...

Are you and Adam using the same Perseus receiver for your measurements?

Have you looked at another similar type (as in small PCB configuration) and 
powered (as in from your HP supply) oscillator for comparison?

Not to beat the PS question to death, but does a battery supply change the 
outcome?

Tom Holmes, N8ZM

-----Original Message-----
From: time-nuts [mailto:[email protected]] On Behalf Of Rob Sherwood.
Sent: Saturday, February 27, 2016 7:20 PM
To: Discussion of precise time and frequency measurement <[email protected]>
Subject: Re: [time-nuts] Inside a CTS 1960017 OCXO

Both  Adam and I measure the same spurs.  Rob, NC0B

-----Original Message-----
From: time-nuts [mailto:[email protected]] On Behalf Of Graham / KE9H
Sent: Saturday, February 27, 2016 2:08 PM
To: Discussion of precise time and frequency measurement
Subject: Re: [time-nuts] Inside a CTS 1960017 OCXO



Rob:
Are you sure you were running it with a clean power supply?
Any 60 Hz ripple on the supply could show up as AM/PM modulation.
You might want to try another 5V power supply before you give up on the 
oscillator.
--- Graham




_______________________________________________
time-nuts mailing list -- [email protected]
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Reply via email to