Ann Earhart
Margy Galloway
Brenda McKay
Gayle Eagan Sansivero
 
You do alot of PICC lines of all sizes
What is your measurement system on veins  vs. PICC size
I know your thrombosis rate Ann, Brenda and Margy are very low with the TL tapered PICC and you do over 2,000 lines a year in your facilities.  I am curious to your answers.  I know Ann, Brenda and Margy you also use both tapered and non tapered catheters.  Do you see a difference?
Kathy


 
 
 
Confidentiality Notice:  This e-mail and any attachments are intended only for the use of those to whom it is addressed and may contain information that is confidential and prohibited from further disclosure under law. If you have received this e-mail in error, its review, use, retention and/or distribution is strictly prohibited. If you are not the intended recipient, please contact the sender by reply e-mail and destroy all copies of the original message and any attachments.[v1.0]




From: [EMAIL PROTECTED] [mailto:[EMAIL PROTECTED] On Behalf Of Cheryl Kelley
Sent: Thursday, October 19, 2006 11:31 AM
To: Earhart, Ann; [EMAIL PROTECTED]; [EMAIL PROTECTED]
Cc: Tim Talbert; [EMAIL PROTECTED]
Subject: Re: reverse tapering

This area of discussion is not scientific mind you, because there has not been data collection on it yet, as far as I know.  But I would be happy to share the measurements with you that I use in my institution.
 
A 6F PICC is 2mm in size, and likely a 3F is 1mm in size.  (Our team has the I-Look by Sono-Site, which has the capability to provide these measurements.  The Site Rite 5 has recently been updated to include a visual interpretation of the vessel size)  I ball part the 4F at 1.3mm and the 5F at 1.7mm.  These are the base line measurements that we go from.  I then look at the hypercoagulability of the patient, based on Virchow's Triad of Stasis and have deducted the following.  If a patient has  no risk factors according to the triad, then my goal is to decrease the blood flow in the vein by no more than 50%, therefore a 6R (2mm) catheter would need a vessel size of at least 4mm.  The vessel is scanned to assess for narrowing on up the vein and if it is 4mm to the axillary, then I would place it.  As far as a 5F (1.7MM) the vessel would need to be at least 3.5mm in size.  When a tapered catheter is introduced, then as you can see, the size of the taper can be as much as 8-9F, or up to 3mm.  Is the vessel at least 3mm at the point of introduction of the catheter?  Maybe not, that is why I think that we are seeing some thrombus, beginning at the insertion site of the PICC.
 
If a patient is hypercoagulable, then the criteria changes.  For a 6F, I want to decrease the blood flow only by 1/3 or 33%, therefore a vessel would need to be 6mm. Realistically, not a lot of patients have a vessel that is 6mm in size unless  you go high in the arm, where the basilic and the brachials often join. 
 
As far as what meets the criteria for hypercoagulabiliy, I look at three things--the vein condition, blood flow thru the veins and the underlying medical condition that can naturally predispose the patient to thrombus formation.
 
Sorry to be so long winded, but I think that this is necessary information for clinicians to understand when placing large lumened catheters.  I have a great interest in this subject.  If you want to talk further, please email me off the list so as not to clog it up!
 
Before you say, show me the data.....I will remind you that I am basing this on what I think and what I have seen, not published data.
 
 
 
Cheryl Kelley RN
PICC Nurse and Infusion Consultant
                    and
PICC Nurse at West Virginia University Hospitals
304-823-3196    or     304-669-3061