I see that we are not communicating accurately.  To quote you in response to 
Alain message regarding this subject  several days ago, "I will not bother" 
with further communications.

Warm regards, 

Bob Cook 
  ----- Original Message ----- 
  From: Jed Rothwell 
  To: [email protected] 
  Sent: Friday, January 09, 2015 11:24 AM
  Subject: Re: [Vo]:"Report on Mizuno's Adiabatic Calorimetry" revised


  Bob Cook <[email protected]> wrote:

    I think you assume that the reaction did not change the differential 
pressure that the pump sees as the reaction occurs. (In other words the 
"baseline" energy is a constant during the reaction.)


  I did not "assume" that. I observed that during the 18 hour calibration. 
There is a huge difference between assuming X and running a test to prove X. 



     What I have suggested is that the differential pressure across the pump 
should be measured, and it should be constant, if your assumption is valid. 


  The 18-hour test measures this and any other factor that might change the 
amount of work done by the pump. It covers all bases. This proves that any 
hypothesis you can come up with about how or why the pump output might have 
changed must be wrong.



    If it is not, the pump energy input would have changed during the reaction.


  It did not change measurably. The kinds of effects you are describing would 
be at the milliwatt level I believe. We could not detect with this system.




    Local two phase flow in a water stream will change the pressure drop in 
that region where two phase flow occurs.  Changes in the water viscosity (and 
resulting pressure drop) with the small changes in temperature of the test 
should be negligible compared to potential two phase flow conditions--i.e., 
steam bubbles plus liquid water. 


  Steam bubbles? Have you looked at the data? The water temperature changes 
from 23.0°C to 25.5°C. There will be no steam at these temperatures. There is 
not the slightest chance such a small temperature change could have a 
measurable effect on the amount of work done by the pump. (Measurable with this 
system I mean. Perhaps a micro-calorimeter could detect it..)


  The water temperature did rise 0.6°C from the effect of the pump. I am not 
saying the pump had no measurable effect at all. I'm saying that the work done 
by the pump did not vary enough to be measured by the system.


  - Jed