> Yes, you're on the right track... see my posting at 6/21 at 9:04pm.
> http://www.mail-archive.com/[email protected]/msg48242.html
>
> I would bet that Galantini is making an indirect measurement of the liquid
> water content as explained in my posting.
>
> -Mark


no, there is no way to make an indirect measurement of steam quality
using a humidity probe that is designed for *air*


HD 37AB1347
HD37AB1347 IAQ Monitor is a tool manufactured by Delta Ohm for the
analysis of air quality (Indoor Air Quality , IAQ). The instrument
simultaneously measures several parameters: Carbon Dioxide CO2, Carbon
monoxide CO, Temperature, Relative humidity and calculates Dew Point,
wet bulb temperature, absolute humidity, mixing ratio, enthalpy and
atmospheric pressure. All this with the P37AB147 SICRAM probe. The
probe SICRAM P37B147 does not measure the Carbon Monoxide CO. Also
combined temperature and humidity SICRAM probes, Hot wire Air speed
SICRAM probes, Vane air speed SICRAM probes and temperature SICRAM
probes can be connected to the instrument. The instrument, with proper
procedure, calculates the percentage of outdoor air intake (% Outside
Air) as a function of both carbon dioxide CO2 and temperature and the
Ventilation Rate. HD37AB1347 data logger has a storage capacity of
67,600 presets for each of the two inputs divided into 64 blocks. Use
the software DeltaLog10 version 0.1.5.0. The instrument is equipped
with a large dot matrix graphic display with a resolution of 160x160
points. Standards: ASHRAE 62.1-2004, Decree Law 81/2008. The rules
apply to all enclosed spaces that may be occupied by people. Should be
considered, depending on air quality, chemical contaminants, physical
and biological or outdoor air flow inside inadequately purified
(Ventilation Rate). Typical applications of the instrument with the
range of sensors mentioned above are: - Measure IAQ and comfort
conditions in schools, offices and indoor. - Analysis and study of
sick building syndrome (Sick Building Syndrome) and consequences. -
Verification of HVAC system. - Investigation of IAQ conditions in
factories to optimize the microclimate and improve productivity. -
Audits in Building Automation.
DELTA OHM, SIT, calibration centre

http://www.deltaohm.com/ver2010/uk/st_airQ.php?str=HD37AB1347
========================================


note the work enthalpy in the above text,  it measures the enthalpy of
the humid air!   not steam quality !!!