Hello group!
I've got a customer in our EMC lab that makes a laboratory instrument
incorporating heaters with a typical 3-4 amp input current at 230vac. He
is failing the flicker test due to the design of the heater control. They
are undertaking a redesign but would like to keep selling product in the
interim.
They have come up with an interesting interpretation of the scope of EN
61000-3-3: the standard specifies that the tests are applicable to
products connected to the "public low voltage distribution systems" but if
you take a typical industrial location or even a typical commercial
location, isn't it true that these locations are fed by a dedicated
transformer at the building that isolates them from the public network? I
guess the question becomes - what is considered the public network? What
if the marketing of the product was restricted to "industrial" locations?
My initial reaction to the question was that anyplace you can plug or wire
into at typical 230vac is considered the "public network", especially
considering the mixed residential / commercial buildings common in Europe.
Anyone care to comment?
Bruce Fagley
Sr. Specialist, EMC
Industrial Products
TUV Rheinland
12 Commerce Rd.
Newtown CT. 06470
203-426-0888 ext 119
203-426-4009 fax
-------------------------------------------
This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.
To cancel your subscription, send mail to:
[email protected]
with the single line:
unsubscribe emc-pstc
For help, send mail to the list administrators:
Jim Bacher: [email protected]
Michael Garretson: [email protected]
For policy questions, send mail to:
Richard Nute: [email protected]