I suggest a similar message thread could be started on thermal measurement, or 
just about any other measurement made under carefully controlled conditions.   
The theory of error propagation, measurement uncertainty, rounding of results 
and standard deviation are great topics at university but rarely put into 
correct practice in industry, so it seems.
_____________________________________________________________________________________
 

Ralph McDiarmid  |   Schneider Electric   |  Renewable Energies Business  |   
CANADA  |   Regulatory Engineer




From:   "ce-test, qualified testing bv - Gert Gremmen" <[email protected]> 
To:     [email protected] 
List-Post: [email protected]
List-Post: [email protected]
List-Post: [email protected]
Date:   08/11/2010 12:43 PM 
Subject:        Re: [PSES] Calibration supplier for signal generator with pulse 
modulation

________________________________




The one and only reason why EMI is “ not precise”, or “uncertain” is exactly 
one 
piece of equipment in the measurement chain…. 
  
The EUT and it’s physical set-up. 
  
*All* other aspects can very well and relatively easy be controlled under a MU 
system, 
allowing customers to compare quality and performance of different labs and 
and judge and compare the quality of measurement alternatives. 
  
  
Defining a MU figure also implies that a round robin test using a well defined 
EUT will get identical results 
everywhere +/- the margin defined by the labs uncertainty figure. 
  
It also makes it possible by accurately reproducing a real EUT setup in another 
lab 
to get comparable results. 
  
I agree that this round robin test is not the definition of daily EMI work. 
In real life you WILL have an undefined EUT with as many degrees of freedom 
of operation and setup as you can recognize, resulting in an uncertainty well 
beyond the MU figures 
this discussion is about. 
But that is no reason to neglect the accuracy specifications of the defined 
chain. 
  
There is also an additional skill required in EMI work, that is finding and 
characterizing 
cause of emissions or lack of immunity. Often a lack of EMC is caused by a 
loophole 
in the design, and a simple fix may virtually make the emissions disappear to 
such 
a degree that discussions about a few dB seem ridiculous.  That is real 
engineering 
work and should not be confused with the art of measuring for which MU is 
meant. 
  
In Europe the EMC requirements (read limits) in the standards have been 
raised to an importance level  (not by the European Commission but by the 
market stakeholders 
united in CENELEC/ISO/IEC) where an excess of 0.1 dB can make the difference 
between pass/fail and market-access or no-market-access. Hence a decision of 
enormous economic value. Lab shopping has been a popular method to gain access 
for equipment 
that marginally failed. An unwanted situation that disturbs a level playing 
field. 
That is why the EMC labs have been brought under the regime of ISO 17025 
(also a market driven standard) and metrology entered the labs to 
force them to create reproducible measurements . 
 Metrology  assessors never saw an EUT and do not understand the EMI 
measurement chain. 
And if you are used to calibrate VOLTS and meters (SI !!!!), the V/m 
do not seem so very different. So they imposed their methods upon us. 
  
And for each of the parts they recognized, they imposed us to make a MU. 
How right they were and how wrong they are. 
But one part cannot be subject to metrology methods:  the  EUT. 
(never available when an audit is made) 
  
But that does not imply that you cannot gain in quality and performance 
by applying a relatively simple procedure to the inaccuracy figures 
your equipment providers supply you with. 
And it makes you think about the impact of all components 
in the chain that affect accuracy. 
  
But that is not the only thing to be said about EMC , as Derek emphasizes. 
  
  
Gert 
  
  
Van: [email protected] [mailto:[email protected] <mailto:[email protected]> ] 
Namens [email protected]
Verzonden: woensdag 11 augustus 2010 18:00
Aan: [email protected]; [email protected]
Onderwerp: Re: Calibration supplier for signal generator with pulse modulation 
  
Hi Deniz,

this is where "real" EMI guys differ from Metrology guys. We really don't need 
to be piddling with little numbers.... with the exception of frequency. Most of 
these modulations are compromises, and measuring a compromise accurately is a 
silly thing to do: in reality. 
  
You may want to get on board with the proof of the pudding phrase, it's OK in 
EMI: otherwise we expend all sorts of effort for no VALUE in return. The 
compliance world is obsessed with meeting a number, when the REAL reason for 
doing this testing is to ensure successful operation in use.

I strongly disagree with Gerts statement MU is simple, it's not. Of the 140+ 
labs I've visited as an assessor, only a handful have a valid effort, and less 
than 1/2 doz believe it was of real value.

Precision and EMI do not go together. I for one like it like that. Whats really 
needed is a competent individual with adequate equipment.

If a committee would like MU, then as test labs we should isolate the cost that 
adds, and identify it on peoples invoices when that test is run. It's an 
effective way of making the standard unpopular. 

The reason I'm on a soap box about this is because unless silly requiremnts 
like this re CHALLENGED, and not just carte-blanche accepted, they become 
requirements. For calibration, I believe MU is useful, but for EMI, it has no 
place.

Sincerely,

Derek. 

From: Deniz Demirci <[email protected]>
To: Untitled <[email protected]>
Sent: Wed, Aug 11, 2010 10:35 am
Subject: RE: Calibration supplier for signal generator with pulse modulation 
Could you define how precise?




  





  
What is your acceptance criteria for the modulation dept? 




  
Let's say for basic 80 % AM modulation.  Is your tolerance 0.1 % or




  
between 60 % to 90 % is good enough. 




  
There is more than a dB peak difference in the signal for 60 % AM and 80




  
% AM modulations. You won't realize the difference with a field probe in




  
radiated immunity tests with modulated signal. (Simple  math;




  
20*Log(1.6/1.8) = -1.02 dB) It seems not very insignificant to me




  





  
Another case;




  
If you are using 6 meter cable for radiated emission at 10 GHz, your




  
cable characterization uncertainty is more than 1.5 dB even with a quite




  
high grade cable (Experimental measurements). Check the cable vswr




  
figures in their specs measured in ideal conditions. Even a simple cable




  
is a significant uncertainty contributor.




  
Your measurement antenna is not exactly 50 Ohm at 10 GHz also. Site




  
imperfection is another story. Those figures are not accounted for in




  
the MIL-STD-461 RE102 verification. Everything seems to be very good




  
when you terminate the measurement cable with a signal generator (50




  
ohm) output.  I don't agree with this <quote> an ultimate




  
"proof-of-the-pudding," </quote>




  





  
You can use all your engineering skills for the tests but you have to




  
quantify your risk using type B measurement uncertainty analysis and do




  
the Type A if you can afford to see where you stand.




  





  
As Mr. Gremmen said " MU is not difficult"




  
No need to be afraid,




  





  
OOO (Own opinions only)




  





  
Best regards,




  





  
Deniz Demirci 




  
National Technical Systems (NTS Canada)




  
Phone: 403-568-6605 ext 244




  
fax: 403-568-6970




  
email:[email protected] <mailto:[email protected]> 




  
web: http://www.ntscorp.com/about/locations 
<http://www.ntscorp.com/about/locations> 




  





  





  





  





  
From: [email protected] <mailto:[email protected]>  [mailto:[email protected] 
<mailto:[email protected]?> ] On Behalf Of Ken




  
Javor




  
Sent: Wednesday, August 11, 2010 7:55 AM




  
To: Untitled




  
Subject: Re: Calibration supplier for signal generator with pulse




  
modulation




  





  
Precisely.




  
 




  
Ken Javor




  





  
Phone: (256) 650-5261




  





  





  
> From: Cortland Richmond <[email protected] <mailto:[email protected]> >




  
> Reply-To: <[email protected] <mailto:[email protected]> >




  
> Date: Wed, 11 Aug 2010 08:27:34 -0400




  
> To: emc-pstc <[email protected] <mailto:[email protected]> >




  
> Subject: re: Calibration supplier for signal generator with pulse




  
modulation




  
> 




  
> Not quite what you need to know but I've used signal taps or




  
directional




  
> couplers with a 'scope to watch the RF waveform and set modulation




  
depth,




  
> and a calibrated counter, or even a receiver or analyzer will for




  
> frequency. One always has recourse to calibrated devices to monitor




  
> another, uncalibrated one. It's a good idea anyway! If a detector must




  
be




  
> used whose linearity is unknown, one can use calibrated attenuators to




  
find




  
> out. 




  
> 




  
> 




  
> Cortland Richmond




  
> KA5S 




  
> 




  
> 




  
>> [Original Message]




  
>> From: Wendy Nya <[email protected] <mailto:[email protected]> >




  
>> To: <[email protected] <mailto:[email protected]> >




  
>> Date: 8/10/2010 4:23:29 AM




  
>> Subject: Calibration supplier for signal generator with pulse




  
modulation




  
>> 




  
>> Dear All,




  
>> 




  
>> I am looking for a calibration supplier that can provide accredited




  
> service for 




  
>> AR SG6000. It has built-in pulse modulation option (for Radiated




  
Immunity




  
> use).




  
>> 




  
>> By the way - Is anyone using this model? It seems to be OEM from




  
Agilent




  
> (it 




  
>> came in an agilent box) but the equipment is marked AR




  
>> and works with Agilent N5181A driver. Agilent is saying that it is




  
not




  
> able to 




  
>> provide accredited calibration service for this model.




  
>> 




  
>> Thanks & Regards,




  
>> Wendy Nya




  
> 




  
> -




  
> ----------------------------------------------------------------




  
> This message is from the IEEE Product Safety Engineering Society




  
emc-pstc




  
> discussion list. To post a message to the list, send your e-mail to




  
> <[email protected] <mailto:[email protected]> >




  
> 




  
> All emc-pstc postings are archived and searchable on the web at:




  
> http://www.ieeecommunities.org/emc-pstc 
> <http://www.ieeecommunities.org/emc-pstc> 




  
> Graphics (in well-used formats), large files, etc. can be posted to




  
that URL.




  
> 




  
> Website:  http://www.ieee-pses.org/ <http://www.ieee-pses.org/> 




  
> Instructions:  http://listserv.ieee.org/request/user-guide.html 
> <http://listserv.ieee.org/request/user-guide.html> 




  
> List rules: http://www.ieee-pses.org/listrules.html 
> <http://www.ieee-pses.org/listrules.html> 




  
> 




  
> For help, send mail to the list administrators:




  
> Scott Douglas <[email protected] <mailto:[email protected]> >




  
> Mike Cantwell <[email protected] <mailto:[email protected]> >




  
> 




  
> For policy questions, send mail to:




  
> Jim Bacher:  <[email protected] <mailto:[email protected]> >




  
> David Heald: <[email protected] <mailto:[email protected]> >




  





  
-




  





  
This message is from the IEEE Product Safety Engineering Society




  
emc-pstc discussion list. To post a message to the list, send your




  
e-mail to <[email protected] <mailto:[email protected]> >




  





  
All emc-pstc postings are archived and searchable on the web at:




  
http://www.ieeecommunities.org/emc-pstc 
<http://www.ieeecommunities.org/emc-pstc> 




  
Graphics (in well-used formats), large files, etc. can be posted to that




  
URL.




  





  
Website:  http://www.ieee-pses.org/ <http://www.ieee-pses.org/> 




  
Instructions:  http://listserv.ieee.org/request/user-guide.html 
<http://listserv.ieee.org/request/user-guide.html> 




  
List rules: http://www.ieee-pses.org/listrules.html 
<http://www.ieee-pses.org/listrules.html> 




  





  
For help, send mail to the list administrators:




  
Scott Douglas <[email protected] <mailto:[email protected]> >




  
Mike Cantwell <[email protected] <mailto:[email protected]> >




  





  
For policy questions, send mail to:




  
Jim Bacher:  <[email protected] <mailto:[email protected]> >




  
David Heald: <[email protected] <mailto:[email protected]> >




  





  
-




  





  
This message is from the IEEE Product Safety Engineering Society emc-pstc 




  
discussion list. To post a message to the list, send your e-mail to 




  
<[email protected] <mailto:[email protected]> >




  





  
All emc-pstc postings are archived and searchable on the web at:




  
http://www.ieeecommunities.org/emc-pstc 
<http://www.ieeecommunities.org/emc-pstc> 




  
Graphics (in well-used formats), large files, etc. can be posted to that URL.




  





  
Website:  http://www.ieee-pses.org/ <http://www.ieee-pses.org/> 




  
Instructions:  http://listserv.ieee.org/request/user-guide.html 
<http://listserv.ieee.org/request/user-guide.html> 




  
List rules: http://www.ieee-pses.org/listrules.html 
<http://www.ieee-pses.org/listrules.html> 




  





  
For help, send mail to the list administrators:




  
Scott Douglas <[email protected] <mailto:[email protected]> >




  
Mike Cantwell <[email protected] <mailto:[email protected]> >




  





  
For policy questions, send mail to:




  
Jim Bacher:  <[email protected] <mailto:[email protected]> >




  
David Heald: <[email protected] <mailto:[email protected]> >




  
-

This message is from the IEEE Product Safety Engineering Society emc-pstc 
discussion list. To post a message to the list, send your e-mail to 
<[email protected] <mailto:[email protected]> >

All emc-pstc postings are archived and searchable on the web at 
http://www.ieeecommunities.org/emc-pstc 
<http://www.ieeecommunities.org/emc-pstc> 
Graphics (in well-used formats), large files, etc. can be posted to that URL. 

Website: http://www.ieee-pses.org/ <http://www.ieee-pses.org/> 
Instructions: http://listserv.ieee.org/request/user-guide.html 
<http://listserv.ieee.org/request/user-guide.html> 
List rules: http://www.ieee-pses.org/listrules.html 
<http://www.ieee-pses.org/listrules.html>  

For help, send mail to the list administrators:
Scott Douglas <[email protected] <mailto:[email protected]> >
Mike Cantwell <[email protected] <mailto:[email protected]> > 

For policy questions, send mail to:
Jim Bacher <[email protected] <mailto:[email protected]> >
David Heald <[email protected] <mailto:[email protected]> > 


________________________________________________________________________
This email has been scanned for SPAM content and Viruses by the MessageLabs 
Email Security System.
________________________________________________________________________ 
-

This message is from the IEEE Product Safety Engineering Society emc-pstc 
discussion list. To post a message to the list, send your e-mail to 
<[email protected] <mailto:[email protected]> >

All emc-pstc postings are archived and searchable on the web at 
http://www.ieeecommunities.org/emc-pstc 
<http://www.ieeecommunities.org/emc-pstc> 
Graphics (in well-used formats), large files, etc. can be posted to that URL. 

Website: http://www.ieee-pses.org/ <http://www.ieee-pses.org/> 
Instructions: http://listserv.ieee.org/request/user-guide.html 
<http://listserv.ieee.org/request/user-guide.html> 
List rules: http://www.ieee-pses.org/listrules.html 
<http://www.ieee-pses.org/listrules.html>  

For help, send mail to the list administrators:
Scott Douglas <[email protected] <mailto:[email protected]> >
Mike Cantwell <[email protected] <mailto:[email protected]> > 

For policy questions, send mail to:
Jim Bacher <[email protected] <mailto:[email protected]> >
David Heald <[email protected] <mailto:[email protected]> > 

-

This message is from the IEEE Product Safety Engineering Society emc-pstc 
discussion list. To post a message to the list, send your e-mail to 
<[email protected]>

All emc-pstc postings are archived and searchable on the web at 
http://www.ieeecommunities.org/emc-pstc
Graphics (in well-used formats), large files, etc. can be posted to that URL. 

Website: http://www.ieee-pses.org/
Instructions: http://listserv.ieee.org/request/user-guide.html
List rules: http://www.ieee-pses.org/listrules.html 

For help, send mail to the list administrators:
Scott Douglas <[email protected]>
Mike Cantwell <[email protected]> 

For policy questions, send mail to:
Jim Bacher <[email protected]>
David Heald <[email protected]> 


Reply via email to