F.3.2 says, "...test of F.3.1 can be carried out.....". Because it says "can" 
and not "shall" does this make this test optional?

Doesn't the previous version of this standard allow you to disconnect surge 
suppressors during the HiPot test?

The Other Brian

From: [email protected] [mailto:[email protected]] On Behalf Of Mick Maytum
Sent: Thursday, May 10, 2012 12:00 PM
To: [email protected]
Cc: Doug Powell
Subject: Re: [PSES] Hipot with in-circuit voltage limiting devices IEC 61010-1 
3rd Ed.

Doug,
    I wouldn't credit the TC 66 has having MOV component knowledge, just 
perceptions.

    Since my last message I have sent messages to the Chair and Secretary of TC 
66 wearing my IEC SC 37B chair hat.
    I was restrained for a change, merely stating they were breaking the laws 
of physics for MOVs rather than crazy! I suggested that if the 0.9 factor was 
to work it should be applied to the nominal voltage not the clamping voltage.

       I'll keep you informed of developments if any.

Regards
Mick

On 10/05/2012 15:15, Doug Powell wrote:
Mick,

Sorry I didn't review this email before I just sent my last.  Clearly you see 
the same problem as I.  I have to wonder if the committee in reviewing this 
clause only used data sheets and did not validate in the lab.  Or is it 
possibly the committee members just missed this?

I would like to suggest the Vclamp we should use this the 8 mS / 10 mS rating.  
That's mS, not uS.


--
Thanks, -doug

Douglas E Powell
[email protected]<mailto:[email protected]>
http://www.linkedin.com/in/dougp01



On Thu, May 10, 2012 at 6:58 AM, Mick Maytum 
<[email protected]<mailto:[email protected]>> wrote:
Doug,
    My previous message gave the IEC rulings - terms clamping voltage and 
nominal varistor voltage  - for two specific points on the MOV clamping 
(clipping) characteristic.

For completeness  IEC 61010-1, ed. 3.0 (2010-06) defines the following:
working voltage
highest r.m.s. value of the a.c. or d.c. voltage across any particular 
insulation which can occur when the equipment is supplied at rated voltage
NOTE 1 Transients and voltage fluctuations are not considered to be part of the 
working voltage.
NOTE 2 Both open-circuit conditions and normal operating conditions are taken 
into account.

I'm now giving an opinion.

Your (nominal) AC supply is 230V rms. The phase "not less than twice the 
working voltage." means the hipot test is done with an AC value of at least 460 
V rms or 650 V pk.

The hipot test voltage of 0.9 times the clamping voltage of the MOV is clearly 
crazy, as from your figures of 710 V  clamping voltage and 473 V nominal 
varistor voltage, testing at 0.9x710 = 630 V would cause substantial current in 
the MOV.

 One could skate round this and say that the 1 mA nominal varistor voltage is 
the clamping threshold voltage - a new term - and use an MOV or combination of 
MOVs to have a nominal voltage of 650/0.9 = 720 V.

Sounds like a letter to the Chair and Secretary of TC 66 is needed to resolve 
this matter.

Regards
Mick

-
----------------------------------------------------------------

This message is from the IEEE Product Safety Engineering Society emc-pstc 
discussion list. To post a message to the list, send your e-mail to 
<[email protected]<mailto:[email protected]>>

All emc-pstc postings are archived and searchable on the web at: 
http://www.ieee-pses.org/emc-pstc.html

Attachments are not permitted but the IEEE PSES Online Communities site at 
http://product-compliance.oc.ieee.org/ can be used for graphics (in well-used 
formats), large files, etc.

Website: http://www.ieee-pses.org/
Instructions: http://listserv.ieee.org/request/user-guide.html
List rules: http://www.ieee-pses.org/listrules.html

For help, send mail to the list administrators:
Scott Douglas <[email protected]<mailto:[email protected]>>
Mike Cantwell <[email protected]<mailto:[email protected]>>

For policy questions, send mail to:
Jim Bacher <[email protected]<mailto:[email protected]>>
David Heald <[email protected]<mailto:[email protected]>>

________________________________
LECO Corporation Notice: This communication may contain confidential 
information intended for the named recipient(s) only. If you received this by 
mistake, please destroy it and notify us of the error. Thank you.

-
----------------------------------------------------------------
This message is from the IEEE Product Safety Engineering Society emc-pstc 
discussion list. To post a message to the list, send your e-mail to 
<[email protected]>

All emc-pstc postings are archived and searchable on the web at:
http://www.ieee-pses.org/emc-pstc.html

Attachments are not permitted but the IEEE PSES Online Communities site at 
http://product-compliance.oc.ieee.org/ can be used for graphics (in well-used 
formats), large files, etc.

Website:  http://www.ieee-pses.org/
Instructions:  http://listserv.ieee.org/request/user-guide.html
List rules: http://www.ieee-pses.org/listrules.html

For help, send mail to the list administrators:
Scott Douglas <[email protected]>
Mike Cantwell <[email protected]>

For policy questions, send mail to:
Jim Bacher:  <[email protected]>
David Heald: <[email protected]>

Reply via email to