>Methinks Gary's statistical/analytical approach is relevant. After 
>desperation, despair, and panic have set in, you may want to consider building 
>a small microcontroller module with an SD >card to constantly log stuff. Have 
>resorted to this several times with customers' field failures that made no 
>sense. And in one case, we identified a very hazardous condition that could 
>have >trashed the customer's site.

Or just buy a microcontroller data logger, I think there are lots of them out 
there.

-Dave

-----Original Message-----
From: Brian O'Connell [mailto:[email protected]] 
Sent: Thursday, September 17, 2015 1:34 PM
To: [email protected]
Subject: Re: [PSES] Unexplained High Fallout of Power Supplies

>From one Brian to another, we gotta get a root cause for your problem.

Methinks Gary's statistical/analytical approach is relevant. After desperation, 
despair, and panic have set in, you may want to consider building a small 
microcontroller module with an SD card to constantly log stuff. Have resorted 
to this several times with customers' field failures that made no sense. And in 
one case, we identified a very hazardous condition that could have trashed the 
customer's site.

Easy to do with some of those amazing chips from Atmel, TI, Freescale, etc (do 
not recommend Microchip stuff). Favorite chip is the Cortex M4 (MK20 series) 
from Freescale. Very versatile, powerful, and much I/O.

Wishing much luck to the other Brian

Brian

-----Original Message-----
From: Gary McInturff [mailto:[email protected]]
Sent: Thursday, September 17, 2015 8:45 AM
To: [email protected]
Subject: Re: [PSES] Unexplained High Fallout of Power Supplies

Brian

Have you tracked the time to failures and put it on a Weibull analysis to see 
if these are truly random failures. I'm wondering if the power supply 
components are not properly derated and are overstressed. Maybe they are 
failing not from a transient event but components are being overdrive. I 
presume a similar set of components are failing - could be a few rather than a 
single one - in these units. I know you could reproduce the failures in-house. 
I would, if you haven't, get a hold of the manufacturers reliability folks and 
talk about their anticipated failure modes.

-----Original Message-----
From: Kunde, Brian [mailto:[email protected]]
Sent: Thursday, September 17, 2015 8:16 AM
To: [email protected]
Subject: Re: [PSES] Unexplained High Fallout of Power Supplies

James,

No. We are seeing a high fallout of purchased power supplies across most (but 
not all) product lines regardless of the furnace type.


John,

The power supplies have an internal fuse. We add a supplementary over-current 
protector upstream of the power supply but it is sized to protect the wiring, 
connectors, and traces on the distritrabution board; not the power supply 
itself. So, yes, the OCP can be 2x or 3x the size of the fuse in the power 
supply.

When these power supplies fail (BAM!), the black soot is a 2" diameter area 
between the AC input connector and the bridge rectifier. The input traces are 
often vaporized off the PC card and the rectifier has failed shorted and of 
course the fuse(s) on the power supply are open. It is hard to tell if the 
damage was caused by high voltage arcing or by high current turning the traces 
into a flash bulb.

These power supplies are power-factor corrected. I will check with engineering 
if such components are damaged.

The large line filters are internal to our instruments so yes, they are always 
in the circuit ahead of all AC within the instrument.

Thanks,
The Other Brian

-----Original Message-----
From: Pawson, James [mailto:[email protected]]
Sent: Thursday, September 17, 2015 10:10 AM
To: Kunde, Brian; [email protected]
Subject: RE: [PSES] Unexplained High Fallout of Power Supplies

Hi Brian,

Is the furnace type common to all the failed pieces of equipment? If not 
common, similar?

Reading on with interest
James



-----Original Message-----
From: Kunde, Brian [mailto:[email protected]]
Sent: 17 September 2015 14:37
To: [email protected]
Subject: Re: [PSES] Unexplained High Fallout of Power Supplies

Bob and all.

Good question. Before I answer your questions, let me describe the typical 
architecture of the products that we make.

Our instruments (laboratory equipment) are single phase 230VAC 50/60hz which 
powers two separate internal circuits; one being a relatively low current 
electronics powered by a purchased over-the-shelf 24Vdc power supply, and 
second, a high current furnace of some type; inductive, electrode, resistive 
heating elements, etc.. Because of the high overall current of the instruments, 
typically between 30A and 50A, the instrument is plugged into a high current 
branch circuit with a huge high current line filter. Internally, we usually 
have a smaller supplementary over-current protector (5, 10 or 15 amp) driving 
the lower current non-furnace electronics such as the dc power supply, cooling 
fans and blowers, etc..

The issues we are having are with these purchased power supplies blowing up. 
And because we pre-test power supplies and our finished products so extensively 
and we are not able to cause a power supply failure with the same damage 
pattern as we are seeing in the field, we believe that in the real world our 
products are seeing some kind of condition that we are not able to simulate in 
our EMC Lab. Identifying and understanding such conditions is our goal at this 
time.

We use no additional inrush limiting other that what is built into the power 
supplies themselves. Other than a thermistor of some kind, how can the inrush 
be limited?

We have also seen on several occasions in the field where a power supply will 
blow up in one instrument which causes the power supply in a nearby instrument 
to also fail.

Two weeks ago we had an "event" occur right here at our own campus in our 
Application Lab where 40-50 instruments of different models and ages are being 
used daily to develop test methods. An instrument was power on but was not 
running an analysis (high current furnace wasn't running in what we call 
Standby-mode). BAM!! The 24Vdc power supply blew up. We sent two R&D engineers 
who have been working on our fallout problem in the field to investigate. They 
found the power supply had failed in the same way as those in the field. No 
other failed component in the instrument was found. The power supply was 
replaced and the instrument was once again functional. AS THE Engineers turned 
to walk back to their office, BAM!! The power supply blew up in the instrument 
installed NEXT to the one they just repaired. This second instrument is a 
different model with a different manufacturer of power supply. An AC Power Line 
monitor/analyzer was installed on the AC Mains circuit and has been c!
 hecked every morning since. No unusual transients or power dropouts have been 
detected.

This lab area has surge protection, so again, we believe the problem is some 
kind of low voltage transient, voltage dropout, or waveform distortion that we 
have been unable to detect and simulate.

Thanks to all.

The Other Brian

-----Original Message-----
From: Bob LaFrance [mailto:[email protected]]
Sent: Thursday, September 17, 2015 8:46 AM
To: [email protected]
Subject: Re: [PSES] Unexplained High Fallout of Power Supplies

I am curious to what you are using for inrush limiting.  Sorry if you already 
told us but I missed that.

Regards,
Bob LaFrance
N9NEO
Design Engineer
Creare Inc.
16 Great Hollow Road
Hanover, NH
603-640-2539


-----Original Message-----
From: CR [mailto:[email protected]]
Sent: Wednesday, September 16, 2015 2:56 PM
To: [email protected]
Subject: Re: [PSES] Unexplained High Fallout of Power Supplies

On 9/15/2015 4:53 PM, Kunde, Brian wrote:
> we have not been able to find the reason for the fallout. The power 
> supplies have all shown arc damage to the AC front end, signs of 
> arcing and traces burned or vaporized, blown fuses, and shorted FETs 
> and/or Rectifiers. These failures have occurred on several different 
> locations, on different power supply models, different manufactures 
> and on different instruments. Some instruments have been in service 
> for years; some for only a few weeks before they fail. Some 
> instruments even have surge suppression modules installed and though 
> the power supplies fail the surge modules tested out fine. The 
> failures did not occur during any known lightning storm or any other 
> known transient.
Brian,

The circumstances you describe -- arcing with no external transient seen
-- point to an on-board occurrence, not an external one.  I'd suspect (in no 
particular order and on little information -- heh) ) PWB material with 
insufficient dielectric withstand, and/or oscillation in a power FET due to its 
characteristics (or another device's) having changed enough to make the control 
loop unstable.  You might have an oscillator the controller can't even detect.

I'm on LinkedIn if you need help, eh?

Cortland Richmond

-
----------------------------------------------------------------
This message is from the IEEE Product Safety Engineering Society emc-pstc 
discussion list. To post a message to the list, send your e-mail to 
<[email protected]>

All emc-pstc postings are archived and searchable on the web at:
http://www.ieee-pses.org/emc-pstc.html

Attachments are not permitted but the IEEE PSES Online Communities site at 
http://product-compliance.oc.ieee.org/ can be used for graphics (in well-used 
formats), large files, etc.

Website:  http://www.ieee-pses.org/
Instructions:  http://www.ieee-pses.org/list.html (including how to 
unsubscribe) List rules: http://www.ieee-pses.org/listrules.html

For help, send mail to the list administrators:
Scott Douglas <[email protected]>
Mike Cantwell <[email protected]>

For policy questions, send mail to:
Jim Bacher:  <[email protected]>
David Heald: <[email protected]>

-
----------------------------------------------------------------
This message is from the IEEE Product Safety Engineering Society emc-pstc 
discussion list. To post a message to the list, send your e-mail to 
<[email protected]>

All emc-pstc postings are archived and searchable on the web at:
http://www.ieee-pses.org/emc-pstc.html

Attachments are not permitted but the IEEE PSES Online Communities site at 
http://product-compliance.oc.ieee.org/ can be used for graphics (in well-used 
formats), large files, etc.

Website:  http://www.ieee-pses.org/
Instructions:  http://www.ieee-pses.org/list.html (including how to unsubscribe)
List rules: http://www.ieee-pses.org/listrules.html

For help, send mail to the list administrators:
Scott Douglas <[email protected]>
Mike Cantwell <[email protected]>

For policy questions, send mail to:
Jim Bacher:  <[email protected]>
David Heald: <[email protected]>

Reply via email to