Hi Rich,

Thanks for confirming that isolation in doorlock can be only functional. This eliminates me one conception that 12V can be ELV, but second isolation is in doorlock. So now, as I'm sure that 12V have to be SELV/PELV I have to ask them if it is. I know that for other supplies they specify that output is SELV but here not, but they write that application can be access control so doorlocks....

There are TUV certificates at this supply page, but not report.

I don't know if I will be able to read with enough understanding the TUV report. I have never seen one.

Best Regards
Piotr


W dniu 2017-11-23 o 02:28, Richard Nute pisze:
Hi Piotr:

The easy way to answer your questions is to request a copy of the TUV report.

1. I suppose that 12 doorlock manufacturer can use only functional isolation 
betwean 12V and its metal casing as he can assume (I think) that he gets safe 
voltage.
If I'm wrong with it let me know.

Yes, only functional insulation between the 12 volts and the metal casing.

2. Then as doorlock metal casing can be touched by user than 12V have to be 
SELV circuit I think.

Yes.  SELV or PELV; makes no difference.

3. I'm not sure if the 12V output from this supply is SELV - is it enough to 
test the isolation with 3kV.

No, 3 kV alone is not sufficient.  The isolation must also meet creepage and 
clearance distances after being subject to a mechanical test of the isolation 
means and the conductors.

Best regards,
Rich

-
----------------------------------------------------------------
This message is from the IEEE Product Safety Engineering Society emc-pstc discussion 
list. To post a message to the list, send your e-mail to <emc-p...@ieee.org>

All emc-pstc postings are archived and searchable on the web at:
http://www.ieee-pses.org/emc-pstc.html

Attachments are not permitted but the IEEE PSES Online Communities site at 
http://product-compliance.oc.ieee.org/ can be used for graphics (in well-used 
formats), large files, etc.

Website:  http://www.ieee-pses.org/
Instructions:  http://www.ieee-pses.org/list.html (including how to unsubscribe)
List rules: http://www.ieee-pses.org/listrules.html

For help, send mail to the list administrators:
Scott Douglas <sdoug...@ieee.org>
Mike Cantwell <mcantw...@ieee.org>

For policy questions, send mail to:
Jim Bacher:  <j.bac...@ieee.org>
David Heald: <dhe...@gmail.com>


-
----------------------------------------------------------------
This message is from the IEEE Product Safety Engineering Society emc-pstc discussion 
list. To post a message to the list, send your e-mail to <emc-p...@ieee.org>

All emc-pstc postings are archived and searchable on the web at:
http://www.ieee-pses.org/emc-pstc.html

Attachments are not permitted but the IEEE PSES Online Communities site at 
http://product-compliance.oc.ieee.org/ can be used for graphics (in well-used 
formats), large files, etc.

Website:  http://www.ieee-pses.org/
Instructions:  http://www.ieee-pses.org/list.html (including how to unsubscribe)
List rules: http://www.ieee-pses.org/listrules.html

For help, send mail to the list administrators:
Scott Douglas <sdoug...@ieee.org>
Mike Cantwell <mcantw...@ieee.org>

For policy questions, send mail to:
Jim Bacher:  <j.bac...@ieee.org>
David Heald: <dhe...@gmail.com>

Reply via email to