_____________________________________________
From: [email protected] [mailto:[email protected]] On Behalf Of 
[email protected]
Sent: Wednesday, January 04, 2012 9:44 AM
To: [email protected]
Subject: Powernet: Beta Correction Factor for Handheld Ion Chamber Instruments


Fort Calhoun  responds:

Palo Verde requests the following information:

      What beta source is used to establish beta correction factors for 
handheld ion chambers (RO-20, RO-2/2A, RSO-50E) at your facility? Depleted 
Uranium slab

      Is a generic beta correction factor is assigned (such as 4 or 5) or is 
the specific value established during calibration assigned (such as 3.8)? A 
specific value is assigned at calibration for each instrument. Although 
typically for field measurement the generic value of 4.0 is used for ease of 
calculation and results in a conservative measurement.

Thanks

                              Gordon Nelson
                              Radiological Engineering
                              Palo Verde Nuclear Generating Station
                              Phone (623) 393-5200
                              Cell  (623) 238-0542
                              Pager (877) 466-7534
                              
[email protected]<mailto:[email protected]>




----------------------------------------------------------------------