[time-nuts] Oscilloquartz 3210 Cesium Standard

2014-09-29 Thread Chris

Another update on progress:
--
Now at the stage where two of the three 3210's are working. The fault 
that was blowing the line fuse turned out to be a shorted reservoir cap. 
Managed to find an Ebay seller in Greece with exactly the same voltage, 
capacity and outline, though different manufacturer. The caps have a 
stud mount at the base of the can, so not easy to find, but bought four 
at just over 3 ukp each and replaced both caps in the psu, leaving two 
for spares.


From cold, the spec is up to 90 minutes warmup time, but if the OCXO 
and pump have been running for some time, both units lock up within 4 
minutes. Have done some setup. For example, normalised the gains and 
offsets in the preamps as per the manual and both units show the 
expected preamp level of ~165mV and 2nd harmonic amplitude of 9-10 on 
the meter, so assume both tubes are in good condition.  None of the 
settings were very far out.


Took a set of measurements from one of the good units and the faulty for 
comparison, with similar results, Good unit as follows, but faulty unit 
more or less the same:


Ion pump = EHT ok, meter = 0
Electron multiplier EHT voltage = 1800 volts, (100Mohm hv probe on o/p 
wire)

Cesium oven = Not open circuit, voltage = 5.5 volts
Ioniser = not open circuit, 1 volt p-p square wave at ~26KHz
Synthesiser output = 12.6317715, follows variation in OCXO and in lock
Backplane test points, tpb, tpc = 8.2 volts p-p square wave, ~137 Hz.
Multiplier 12.6... MHz input = 2v p-p (Scope)
Multiplier 137Hz input = 2.25v p-p (Scope)
Multiplier 180MHz output = +26DBm (HP 3406A rf voltmeter + 20DB attenuator)
Microwave tap on waveguide = -14.5DBm, threaded attenuator works 
(HP432A, 478A)


The 3406 and 432 are quite old and the sensor head is even older, but 
should be reasonably accurate. The 478A head was coupled to the 
microwave tap using a ~3 inch length of rigid coax, so not much 
attenuation. Hopefully, these figures may help others trying to debug 
these standards, but if there's anything i've missed, please let me know.


For the faulty unit, even with all the levels as expected, there is 
still no signal, nor 2nd harmonic, even with the preamp gain turned up 
to max. Connected a Fluke electrometer in series with the preamp input 
socket, but the current is at least 100 times down on 1nA. Looks like 
the tube is completely lifed, or has an internal fault. /FX:  Have 
visions of a sad, faithful and slowly dying 3210 left for decades in a 
rack,  in the dark, powered up, but long forgotten after the accurate 
reference it provided ceased to be used anywhere in the organisation. A 
ship adrift for thousands of years, but all systems still at least 
partially working and waiting to be discovered :-).


So what else am I missing in this puzzle ?...

Regards,

Chris



___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Oscilloquartz 3210 Cesium Standard

2014-09-29 Thread Björn
Hi Chris,

Nice work! 

I can't help you with the faulty unit. There are others on the list with deeper 
insight in that OSA Cs core. Hope they have time to chime in.

But I do have a telco variant of the same Cs core running in the basement. Its 
running continously (ocxo and ion pump) but its not burning the last fumes of 
Cs in the tube. This is since I don't see access or money available for a 
replacement tube.

As you noted. With ocxo warm and cosy and the ion pump running it does not take 
long for Cs lock when needed.

A mode where you can use the unit as premium ocxo (bva :-) ) standard that can 
be calibrated against the cs tube periodically as needed would suit my 
situation.

However there are (at least) two mods needed. Because unless the cs tube 
burnes, there is no power to the

 1)  output buffer card. And thus no output signal.
 2) potentiometer adjusting the bva frequency. Thus without cs tube running the 
bva is off freq.

This design might have been a safe one in the original market.

But I think modding the unit to let it run mostly as a very stable crystal 
standard is better for hobbyist use.

Have others done these kind of mods to prolong the Cs tube life? 

Kind regards,

          Björn



div Originalmeddelande /divdivFrån: Chris 
syseng.greenfi...@btconnect.com /divdivDatum:2014-09-29  22:43  
(GMT+01:00) /divdivTill: time-nuts@febo.com /divdivRubrik: [time-nuts]  
Oscilloquartz 3210 Cesium Standard /divdiv
/divAnother update on progress:
--
Now at the stage where two of the three 3210's are working. The fault 
that was blowing the line fuse turned out to be a shorted reservoir cap. 
Managed to find an Ebay seller in Greece with exactly the same voltage, 
capacity and outline, though different manufacturer. The caps have a 
stud mount at the base of the can, so not easy to find, but bought four 
at just over 3 ukp each and replaced both caps in the psu, leaving two 
for spares.

From cold, the spec is up to 90 minutes warmup time, but if the OCXO 
and pump have been running for some time, both units lock up within 4 
minutes. Have done some setup. For example, normalised the gains and 
offsets in the preamps as per the manual and both units show the 
expected preamp level of ~165mV and 2nd harmonic amplitude of 9-10 on 
the meter, so assume both tubes are in good condition.  None of the 
settings were very far out.

Took a set of measurements from one of the good units and the faulty for 
comparison, with similar results, Good unit as follows, but faulty unit 
more or less the same:

Ion pump = EHT ok, meter = 0
Electron multiplier EHT voltage = 1800 volts, (100Mohm hv probe on o/p 
wire)
Cesium oven = Not open circuit, voltage = 5.5 volts
Ioniser = not open circuit, 1 volt p-p square wave at ~26KHz
Synthesiser output = 12.6317715, follows variation in OCXO and in lock
Backplane test points, tpb, tpc = 8.2 volts p-p square wave, ~137 Hz.
Multiplier 12.6... MHz input = 2v p-p (Scope)
Multiplier 137Hz input = 2.25v p-p (Scope)
Multiplier 180MHz output = +26DBm (HP 3406A rf voltmeter + 20DB attenuator)
Microwave tap on waveguide = -14.5DBm, threaded attenuator works 
(HP432A, 478A)

The 3406 and 432 are quite old and the sensor head is even older, but 
should be reasonably accurate. The 478A head was coupled to the 
microwave tap using a ~3 inch length of rigid coax, so not much 
attenuation. Hopefully, these figures may help others trying to debug 
these standards, but if there's anything i've missed, please let me know.

For the faulty unit, even with all the levels as expected, there is 
still no signal, nor 2nd harmonic, even with the preamp gain turned up 
to max. Connected a Fluke electrometer in series with the preamp input 
socket, but the current is at least 100 times down on 1nA. Looks like 
the tube is completely lifed, or has an internal fault. /FX:  Have 
visions of a sad, faithful and slowly dying 3210 left for decades in a 
rack,  in the dark, powered up, but long forgotten after the accurate 
reference it provided ceased to be used anywhere in the organisation. A 
ship adrift for thousands of years, but all systems still at least 
partially working and waiting to be discovered :-).

So what else am I missing in this puzzle ?...

Regards,

Chris



___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Oscilloquartz 3210 Cesium Standard

2014-09-29 Thread Magnus Danielson

Hi,

We have already hacked a DC-DC converter to feed the output board with 
separate power. It was relatively easy.


Haven't had the time too look at doing the same to the external trimmer.

Cheers,
Magnus


On 09/30/2014 02:09 AM, Björn wrote:

Hi Chris,

Nice work!

I can't help you with the faulty unit. There are others on the list with deeper 
insight in that OSA Cs core. Hope they have time to chime in.

But I do have a telco variant of the same Cs core running in the basement. Its 
running continously (ocxo and ion pump) but its not burning the last fumes of 
Cs in the tube. This is since I don't see access or money available for a 
replacement tube.

As you noted. With ocxo warm and cosy and the ion pump running it does not take 
long for Cs lock when needed.

A mode where you can use the unit as premium ocxo (bva :-) ) standard that can 
be calibrated against the cs tube periodically as needed would suit my 
situation.

However there are (at least) two mods needed. Because unless the cs tube 
burnes, there is no power to the

  1)  output buffer card. And thus no output signal.
  2) potentiometer adjusting the bva frequency. Thus without cs tube running 
the bva is off freq.

This design might have been a safe one in the original market.

But I think modding the unit to let it run mostly as a very stable crystal 
standard is better for hobbyist use.

Have others done these kind of mods to prolong the Cs tube life?

Kind regards,

   Björn



div Originalmeddelande /divdivFrån: Chris syseng.greenfi...@btconnect.com 
/divdivDatum:2014-09-29  22:43  (GMT+01:00) /divdivTill: time-nuts@febo.com /divdivRubrik: [time-nuts]  
Oscilloquartz 3210 Cesium Standard /divdiv
/divAnother update on progress:
--
Now at the stage where two of the three 3210's are working. The fault
that was blowing the line fuse turned out to be a shorted reservoir cap.
Managed to find an Ebay seller in Greece with exactly the same voltage,
capacity and outline, though different manufacturer. The caps have a
stud mount at the base of the can, so not easy to find, but bought four
at just over 3 ukp each and replaced both caps in the psu, leaving two
for spares.

 From cold, the spec is up to 90 minutes warmup time, but if the OCXO
and pump have been running for some time, both units lock up within 4
minutes. Have done some setup. For example, normalised the gains and
offsets in the preamps as per the manual and both units show the
expected preamp level of ~165mV and 2nd harmonic amplitude of 9-10 on
the meter, so assume both tubes are in good condition.  None of the
settings were very far out.

Took a set of measurements from one of the good units and the faulty for
comparison, with similar results, Good unit as follows, but faulty unit
more or less the same:

Ion pump = EHT ok, meter = 0
Electron multiplier EHT voltage = 1800 volts, (100Mohm hv probe on o/p
wire)
Cesium oven = Not open circuit, voltage = 5.5 volts
Ioniser = not open circuit, 1 volt p-p square wave at ~26KHz
Synthesiser output = 12.6317715, follows variation in OCXO and in lock
Backplane test points, tpb, tpc = 8.2 volts p-p square wave, ~137 Hz.
Multiplier 12.6... MHz input = 2v p-p (Scope)
Multiplier 137Hz input = 2.25v p-p (Scope)
Multiplier 180MHz output = +26DBm (HP 3406A rf voltmeter + 20DB attenuator)
Microwave tap on waveguide = -14.5DBm, threaded attenuator works
(HP432A, 478A)

The 3406 and 432 are quite old and the sensor head is even older, but
should be reasonably accurate. The 478A head was coupled to the
microwave tap using a ~3 inch length of rigid coax, so not much
attenuation. Hopefully, these figures may help others trying to debug
these standards, but if there's anything i've missed, please let me know.

For the faulty unit, even with all the levels as expected, there is
still no signal, nor 2nd harmonic, even with the preamp gain turned up
to max. Connected a Fluke electrometer in series with the preamp input
socket, but the current is at least 100 times down on 1nA. Looks like
the tube is completely lifed, or has an internal fault. /FX:  Have
visions of a sad, faithful and slowly dying 3210 left for decades in a
rack,  in the dark, powered up, but long forgotten after the accurate
reference it provided ceased to be used anywhere in the organisation. A
ship adrift for thousands of years, but all systems still at least
partially working and waiting to be discovered :-).

So what else am I missing in this puzzle ?...

Regards,

Chris



___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


___
time

Re: [time-nuts] Oscilloquartz 3210 Cesium Standard

2014-09-09 Thread Adrian Constantinescu
Hello,
I uploaded some photos, here are the links:

http://www.datafilehost.com/d/c81e3461
http://www.datafilehost.com/d/95ac3833
http://www.datafilehost.com/d/b7707109
http://www.datafilehost.com/d/809a2ef8
http://www.datafilehost.com/d/6c3381df
http://www.datafilehost.com/d/c0684d5b

I double checked the capacitors and they are mounted accordingly. No diode
seems to be shorted.
I have one cap that doesn't seem to be connected to the ground, i will check
it later to see if it should or not.

On first picture you can see the connectors on the small board. Two of them
are the outputs, but i can't figure out what the other connectors are.

Thank you!

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Re: [time-nuts] Oscilloquartz 3210 Cesium Standard

2014-09-09 Thread Adrian Constantinescu
Did a major breakthrough! 
Obviously it was my fault, by mistake i swapped two connectors.
 My luck is that the power board has some safety elements. 
OK,  now with the connectors in correct position 
I have 8.5 and 17 volts on the outputs. 

Getting closer 

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Re: [time-nuts] Oscilloquartz 3210 Cesium Standard

2014-09-09 Thread paul swed
Ohhh that hurts and you were lucky. I always mark connectors for that exact
reason. I was going to download the pictures but it kept trying to download
some executable. So will not be able to look at your pictures.
Regards
Paul.

On Tue, Sep 9, 2014 at 4:35 AM, Adrian Constantinescu 
adrian.valentin.constantine...@gmail.com wrote:

 Did a major breakthrough!
 Obviously it was my fault, by mistake i swapped two connectors.
  My luck is that the power board has some safety elements.
 OK,  now with the connectors in correct position
 I have 8.5 and 17 volts on the outputs.

 Getting closer

 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to
 https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Re: [time-nuts] Oscilloquartz 3210 Cesium Standard

2014-09-09 Thread Jason Ball
Replacing the caps could have increased voltages and exposed a fault that
caused the caps to fail...

As always step 1 - check voltages/regulators.   Remove bridging components
(diodes, power resistors) if necessary to stop the smoke and trace the
circuit as any number of components could have gone open circuit.  Double
check your soldering just in case there is an accidental bridge in there.

Good luck.

Cheers
J.


On Tue, Sep 9, 2014 at 8:47 AM, Pete Lancashire p...@petelancashire.com
wrote:

 If not check for shorted diodes.

 -pete
 On Sep 8, 2014 2:43 PM, paul swed paulsw...@gmail.com wrote:

  Adrian please copy me on the pictures also.
  Whats odd is the caps were bulging. You replaced them and now things are
  smoking. Might one have been put in backwards?
  Regards
  Paul.
  WB8TSL
 
  On Mon, Sep 8, 2014 at 4:49 PM, Adrian Constantinescu 
  adrian.valentin.constantine...@gmail.com wrote:
 
   Bert Kehren via time-nuts time-nuts@... writes:
  
   
or send it off list if you can not compress it.
Bert
   
   
  
  
  
   Thank you for your quick reply, i will take some pictures in the
 morning
   and
   post them.
  
  
   ___
   time-nuts mailing list -- time-nuts@febo.com
   To unsubscribe, go to
   https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
   and follow the instructions there.
  ___
  time-nuts mailing list -- time-nuts@febo.com
  To unsubscribe, go to
  https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
  and follow the instructions there.
 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to
 https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.




-- 
--
Teach your kids Science, or somebody else will :/

ja...@ball.net
vk2...@google.com vk2f...@google.com
callsign: vk2vjb
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Re: [time-nuts] Oscilloquartz 3210 Cesium Standard

2014-09-09 Thread Alexander Pummer


yes I got these executable too, and therefore did not continued
73
Alex
On 9/9/2014 5:59 AM, paul swed wrote:

Ohhh that hurts and you were lucky. I always mark connectors for that exact
reason. I was going to download the pictures but it kept trying to download
some executable. So will not be able to look at your pictures.
Regards
Paul.

On Tue, Sep 9, 2014 at 4:35 AM, Adrian Constantinescu 
adrian.valentin.constantine...@gmail.com wrote:


Did a major breakthrough!
Obviously it was my fault, by mistake i swapped two connectors.
  My luck is that the power board has some safety elements.
OK,  now with the connectors in correct position
I have 8.5 and 17 volts on the outputs.

Getting closer

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Re: [time-nuts] Oscilloquartz 3210 Cesium Standard

2014-09-09 Thread Adrian Constantinescu
I forgot to tell you to uncheck the download manager thing, that's why it is 
downloading .exe.

Anyway, i ran some test today, i unplugged the high voltage driver and found 
out that instead of 10 volts on one output i now have 13-15 volts, which means 
that either it is a bad filter, or the pwm  feedback is bad. On the other hand, 
the 24 volts output remains 17 volts no matter what, which would probably 
indicate either a bad pwm controller or a bad converter.
I figured out the pwm controller and the mosfet for the 24 volts voltage, and i 
will put a scope tomorrow to see what happens, but i was unable to figure out 
where the 10 volts voltage is coming from.

I also found a technical manual for osa 5585, but it is rather ambiguous.
From it i found out that the input of the power board is on the db15 connector, 
but it seems there are some missing voltages from it. 

But one thing at the time. I have 23.5V on the input which is exactly what is 
written on the power supply. So i must be able to obtain the 10 and 24 
voltages.

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Re: [time-nuts] Oscilloquartz 3210 Cesium Standard

2014-09-09 Thread Raj
Untick the box and the download button will apear!

Raj

At 09-09-2014, you wrote:
 I was going to download the pictures but it kept trying to download some 
 executable. So will not be able to look at your pictures. Regards Paul. 

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Re: [time-nuts] Oscilloquartz 3210 Cesium Standard

2014-09-08 Thread Adrian Constantinescu
Hello,
I have a OSA-5585 cesium PRS and it is not working.
On the local manager i have a cesium module alarm. I opened it and found out 
that the high power module that drives the cesium oven is not working because 
it doesn't receive the correct voltages. Instead of 10 and 24 volts it only 
gets 8 and 13 volts. So i pulled out the power board and found out all the 
capacitors were bulged. I replaced all of them with the corresponding values 
and now i have 0 volts on its output and some smoke coming from a diode on the 
board.

Does anyone have some schematics for this unit, or could at least help me with 
the connectors on the power board? I found the output connectors that should 
deliver the 10 and 24 volts but i don't know what the inputs are. If i can find 
out what the inputs are then i could debug the board easier.

Please excuse my writing, english is not my native language.

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Re: [time-nuts] Oscilloquartz 3210 Cesium Standard

2014-09-08 Thread Bert Kehren via time-nuts
Can you post a picture of the board
Bert Miami
 
 
In a message dated 9/8/2014 2:20:37 P.M. Eastern Daylight Time,  
adrian.valentin.constantine...@gmail.com writes:

Hello,
I have a OSA-5585 cesium PRS and it is not working.
On  the local manager i have a cesium module alarm. I opened it and found 
out  
that the high power module that drives the cesium oven is not working  
because 
it doesn't receive the correct voltages. Instead of 10 and 24  volts it 
only 
gets 8 and 13 volts. So i pulled out the power board and  found out all the 
capacitors were bulged. I replaced all of them with the  corresponding 
values 
and now i have 0 volts on its output and some smoke  coming from a diode on 
the 
board.

Does anyone have some schematics  for this unit, or could at least help me 
with 
the connectors on the power  board? I found the output connectors that 
should 
deliver the 10 and 24  volts but i don't know what the inputs are. If i can 
find 
out what the  inputs are then i could debug the board easier.

Please excuse my  writing, english is not my native  language.

___
time-nuts  mailing list -- time-nuts@febo.com
To unsubscribe, go to  
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the  instructions there.
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Re: [time-nuts] Oscilloquartz 3210 Cesium Standard

2014-09-08 Thread Bert Kehren via time-nuts
or send it off list if you can not compress it.
Bert
 
 
In a message dated 9/8/2014 2:20:37 P.M. Eastern Daylight Time,  
adrian.valentin.constantine...@gmail.com writes:

Hello,
I have a OSA-5585 cesium PRS and it is not working.
On  the local manager i have a cesium module alarm. I opened it and found 
out  
that the high power module that drives the cesium oven is not working  
because 
it doesn't receive the correct voltages. Instead of 10 and 24  volts it 
only 
gets 8 and 13 volts. So i pulled out the power board and  found out all the 
capacitors were bulged. I replaced all of them with the  corresponding 
values 
and now i have 0 volts on its output and some smoke  coming from a diode on 
the 
board.

Does anyone have some schematics  for this unit, or could at least help me 
with 
the connectors on the power  board? I found the output connectors that 
should 
deliver the 10 and 24  volts but i don't know what the inputs are. If i can 
find 
out what the  inputs are then i could debug the board easier.

Please excuse my  writing, english is not my native  language.

___
time-nuts  mailing list -- time-nuts@febo.com
To unsubscribe, go to  
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the  instructions there.
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Re: [time-nuts] Oscilloquartz 3210 Cesium Standard

2014-09-08 Thread paul swed
Adrian please copy me on the pictures also.
Whats odd is the caps were bulging. You replaced them and now things are
smoking. Might one have been put in backwards?
Regards
Paul.
WB8TSL

On Mon, Sep 8, 2014 at 4:49 PM, Adrian Constantinescu 
adrian.valentin.constantine...@gmail.com wrote:

 Bert Kehren via time-nuts time-nuts@... writes:

 
  or send it off list if you can not compress it.
  Bert
 
 



 Thank you for your quick reply, i will take some pictures in the morning
 and
 post them.


 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to
 https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Re: [time-nuts] Oscilloquartz 3210 Cesium Standard

2014-09-08 Thread Pete Lancashire
If not check for shorted diodes.

-pete
On Sep 8, 2014 2:43 PM, paul swed paulsw...@gmail.com wrote:

 Adrian please copy me on the pictures also.
 Whats odd is the caps were bulging. You replaced them and now things are
 smoking. Might one have been put in backwards?
 Regards
 Paul.
 WB8TSL

 On Mon, Sep 8, 2014 at 4:49 PM, Adrian Constantinescu 
 adrian.valentin.constantine...@gmail.com wrote:

  Bert Kehren via time-nuts time-nuts@... writes:
 
  
   or send it off list if you can not compress it.
   Bert
  
  
 
 
 
  Thank you for your quick reply, i will take some pictures in the morning
  and
  post them.
 
 
  ___
  time-nuts mailing list -- time-nuts@febo.com
  To unsubscribe, go to
  https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
  and follow the instructions there.
 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to
 https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

[time-nuts] Oscilloquartz 3210 Cesium Standard

2014-09-05 Thread Chris

Hi Tom,

Just catching up and time for an update on progress. Had had to put the 
faulty 3210 to one side this week to get some work done, but a couple of 
developments meantime: First: Put an enquiry on the Oscilloquartz web 
enquiry form a week or so ago and they sent me the full 3210 pdf manual. 
Theory, adjustments and full schematics. Needless to say very grateful 
and sent back an email thanking them. Although there were no conditions 
attached, I have to assume it was sent in confidence, so can't upload to 
a file share site, but if anyone needs a copy, please contact me off 
list. Thanks to Tom for scanning his manual as well. It looks like an 
earlier version than I have and there seem to be minor differences, so a 
scan of the whole thing would be good, if possible. I have a Panasonic 
doc scanner, A4 A3, long and user defined pages, colour etc (Ebay, 20 
ukp :-) and could scan the whole volume if you are in the UK. Second: I 
had a possible line on two more of these 3210's and offer accepted, 
collected them yesterday. One blows the line fuse, which should be 
simple fix, but the other one actually works. Ion current way off scale 
to start, with one psu rail cycling on and off, but after an hour or 
two, lock light on and integrator starting to fall back as the OCXO 
warmed up. The 2nd harmonic, which was zero on the original unit, 
started creeping up from just visible indication, to 4 of 10 after ~5 
hours, then after running all night, shows full scale at 10, Utopia :-). 
Valid range from the manual is 2-10, so this looks like a very good 
tube. Am going to leave it running for a month, then go right through 
the setup procedure.


Both units collected yesterday were each in a small subrack, with Racal 
badged standbye psu + batteries and the whole rack marked as Portable 
Frequency Standard. That is, take to a site, power up, calibrate 
whatever, then power down and put back in store. They may have done very 
few hours, but that may be wishful thinking. Two still to fix then, but 
now have a working unit for comparison and at last, a properly working 
cesium standard for the lab. Thanks to everyone who replied here along 
the way, the response and encouragement have been amazing...


Regards,

Chris

cc this to tvb, in case this doesn't reach the list...
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Re: [time-nuts] Oscilloquartz 3210 Cesium Standard

2014-09-04 Thread Bert Kehren via time-nuts
Tom
Thank you for the information on the OSA 3210
This is what I know.  OSA 3000 is the Cs module only with three OCXO  
versions B-5400, 8602 and 8612. All the pictures and information I have seen 
and  
the modules that I have, show identical boards and assemblies.
I am not aware of a later version in my opinion the 3000 is a repackaged  
FTS 4000/5000.
Maybe they came up with a later version but only time nuts could answer  
that
OSA 32XX all are complete units just like FTS offered 4060 etc. all have a  
OSA 3000 inside. Same goes for RS-XSC
OSA 3020 and OSA 3100 no idea.
Bert Kehren
 
 
In a message dated 9/3/2014 9:02:52 P.M. Eastern Daylight Time,  
t...@leapsecond.com writes:

http://www.leapsecond.com/museum/osa3200/

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Re: [time-nuts] Oscilloquartz 3210 Cesium Standard

2014-09-04 Thread paul swed
Nice scan Tom. Clean. I most likely will never run across either unit. But
always good to read about them.
Regards
Paul
WB8TSL


On Wed, Sep 3, 2014 at 9:01 PM, Tom Van Baak t...@leapsecond.com wrote:

 Hi Chris,

 Sorry for a delayed reply. The OSA 3210 is one of the only cesium
 standards I don't have in the museum. I've seen one or two on eBay over
 the past decade but not in worthwhile condition. I hope you can get yours
 working. If not, let me know...

 The good news is that I have an original Oscilloquartz 3200 cesium manual.
 Today I scanned the first part of it for you. See:
 http://www.leapsecond.com/museum/osa3200/

 If this is helpful to your efforts, I'll try to scan the schematics as
 well. The manual is an inch thick.

 I don't know the model differences. I've seen 3000, 3020 and 3200, 3210,
 3230, 3235 mentioned on the web. Maybe 3120. I'm also not clear about the
 differences in series-3000 vs. series-3100 vs. series 3200. If anyone
 knows, please post details.

 Thanks,
 /tvb

 - Original Message -
 From: Chris syseng.greenfi...@btconnect.com
 To: time-nuts@febo.com
 Sent: Wednesday, August 27, 2014 2:09 PM
 Subject: [time-nuts] Oscilloquartz 3210 Cesium Standard


  Hi,
 
  I recently bought an Oscilloquartz 3210 cesium standard and although it
 does
  power up, it does not lock. All the parameters on the meter look
  sensible, other
  than the 2nd harmonic, #9 parameter, which shows zero. Quart oscillator
  output
  looks almost spot on in comparison with the lab Z3816. I have no info on
  this
  unit at all, so this may either be a dead tube, or other electronic
 fault on
  one of the many pcb's in the system. Ion pump had a small reading on
  power up,
  but now dropped back to zero.
 
  I am looking for a user guide for this unit, or even better, a manual
  with theory
  of operation and setup info. Can anyone help with this, or perhaps
 suggest a
  source ?. Year of manufacture estimated 1984/85, from date code on ic's.
 
  I notice that the tube is from FTS, the same construction as in the FTS
  4060, but
  different part number and wonder if this was original fitment, or has
 been
  replaced at some stage...
 
  Regards  Thanks
 
  Chris Quayle
 
  Oxford, England


 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to
 https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Re: [time-nuts] Oscilloquartz 3210 Cesium Standard

2014-09-04 Thread Magnus Danielson

Bert,

On 09/04/2014 12:56 PM, Bert Kehren via time-nuts wrote:

Tom
Thank you for the information on the OSA 3210
This is what I know.  OSA 3000 is the Cs module only with three OCXO
versions B-5400, 8602 and 8612.


Later versions of the OSA 3000 also had 8613. I have seen 8613 and 
8602.11 myself.



All the pictures and information I have seen and
the modules that I have, show identical boards and assemblies.


Depending on oscillator, time-constants was different, but that is just 
mounting-cases of the same board.


Cheers,
Magnus
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Re: [time-nuts] Oscilloquartz 3210 Cesium Standard

2014-09-04 Thread Bert Kehren via time-nuts
Good evening Magnus
You should post it on time nuts. Do you know the specs on these? I have not 
 seen a data sheet on the 8602 , 8612 or B-6400. now 8613 and what about  
11?
Bert
 
 
In a message dated 9/4/2014 1:27:12 P.M. Eastern Daylight Time,  
mag...@rubidium.dyndns.org writes:

Bert,

On 09/04/2014 12:56 PM, Bert Kehren via time-nuts  wrote:
 Tom
 Thank you for the information on the OSA  3210
 This is what I know.  OSA 3000 is the Cs module only with  three OCXO
 versions B-5400, 8602 and 8612.

Later versions of  the OSA 3000 also had 8613. I have seen 8613 and 
8602.11  myself.

 All the pictures and information I have seen and
  the modules that I have, show identical boards and  assemblies.

Depending on oscillator, time-constants was different, but  that is just 
mounting-cases of the same  board.

Cheers,
Magnus
___
time-nuts  mailing list -- time-nuts@febo.com
To unsubscribe, go to  
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the  instructions there.
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Re: [time-nuts] Oscilloquartz 3210 Cesium Standard

2014-09-04 Thread Magnus Danielson

Bert,

I hope this will satisfy your curiosity somewhat:
http://rubidium.se/~magnus/time/oscilloquartz/20140904_215635.jpg

I don't have any info on the B-5400 unfortunately.

As you see, the 8602 comes in three variants, 8602, 8602.11 ans 
8602.11.05. The later two have some differences in setability range, but 
nothing major, I don't have enough detail.


Some other 8600/8601 data is here:
http://rubidium.se/~magnus/time/oscilloquartz/8601/

Wish I had more to share, the little more I have is not that interesting.

Cheers,
Magnus

On 09/04/2014 08:31 PM, Bert Kehren via time-nuts wrote:

Good evening Magnus
You should post it on time nuts. Do you know the specs on these? I have not
  seen a data sheet on the 8602 , 8612 or B-6400. now 8613 and what about
11?
Bert


In a message dated 9/4/2014 1:27:12 P.M. Eastern Daylight Time,
mag...@rubidium.dyndns.org writes:

Bert,

On 09/04/2014 12:56 PM, Bert Kehren via time-nuts  wrote:

Tom
Thank you for the information on the OSA  3210
This is what I know.  OSA 3000 is the Cs module only with  three OCXO
versions B-5400, 8602 and 8612.


Later versions of  the OSA 3000 also had 8613. I have seen 8613 and
8602.11  myself.


All the pictures and information I have seen and
  the modules that I have, show identical boards and  assemblies.


Depending on oscillator, time-constants was different, but  that is just
mounting-cases of the same  board.

Cheers,
Magnus
___
time-nuts  mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the  instructions there.
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Re: [time-nuts] Oscilloquartz 3210 Cesium Standard

2014-09-04 Thread Tom Van Baak
 Good evening Magnus
 You should post it on time nuts.

Hi Bert,

It looks like Magnus did in fact post his message to time-nuts. I'd like to 
figure out why it appeared as a private email to you instead of a normal 
time-nuts posting. Give me a call or email.

His and your postings came through just fine. See:
https://www.febo.com/pipermail/time-nuts/2014-September/date.html
https://www.febo.com/pipermail/time-nuts/2014-September/086435.html
https://www.febo.com/pipermail/time-nuts/2014-September/086436.html



A note to everyone -- the past few weeks John and I are trying to figure out 
how to get the time-nuts list sort of back to the way it was.

The new version of mailman he's using has features to accommodate domains using 
DMARC (e.g., AOL, Yahoo, ATT, Microsoft (hotmail, outlook, live), Comcast). 
That should bring back the hundred or so members who keep getting bounced off 
time-nuts at febo.com

If you have technical questions or suggestions about this let me 
(tvb@leapsecond) and John (j...@febo.com) know.

Thanks,
/tvb


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Re: [time-nuts] Oscilloquartz 3210 Cesium Standard

2014-09-04 Thread Adrian

Magnus,

here is at least some information on the B-5400:
http://tf.boulder.nist.gov/general/pdf/26.pdf
That is the paper mentioned in the 3200 manual.

Btw. do you have any information on the B-1325?

Cheers,
Adrian


Magnus Danielson schrieb:

Bert,

I hope this will satisfy your curiosity somewhat:
http://rubidium.se/~magnus/time/oscilloquartz/20140904_215635.jpg

I don't have any info on the B-5400 unfortunately.

As you see, the 8602 comes in three variants, 8602, 8602.11 ans 
8602.11.05. The later two have some differences in setability range, 
but nothing major, I don't have enough detail.


Some other 8600/8601 data is here:
http://rubidium.se/~magnus/time/oscilloquartz/8601/

Wish I had more to share, the little more I have is not that interesting.

Cheers,
Magnus

On 09/04/2014 08:31 PM, Bert Kehren via time-nuts wrote:

Good evening Magnus
You should post it on time nuts. Do you know the specs on these? I 
have not
  seen a data sheet on the 8602 , 8612 or B-6400. now 8613 and what 
about

11?
Bert


In a message dated 9/4/2014 1:27:12 P.M. Eastern Daylight Time,
mag...@rubidium.dyndns.org writes:

Bert,

On 09/04/2014 12:56 PM, Bert Kehren via time-nuts  wrote:

Tom
Thank you for the information on the OSA  3210
This is what I know.  OSA 3000 is the Cs module only with three OCXO
versions B-5400, 8602 and 8612.


Later versions of  the OSA 3000 also had 8613. I have seen 8613 and
8602.11  myself.


All the pictures and information I have seen and
  the modules that I have, show identical boards and assemblies.


Depending on oscillator, time-constants was different, but  that is just
mounting-cases of the same  board.

Cheers,
Magnus
___
time-nuts  mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the  instructions there.
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to 
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts

and follow the instructions there.


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to 
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts

and follow the instructions there.


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Re: [time-nuts] Oscilloquartz 3210 Cesium Standard

2014-09-03 Thread Tom Van Baak
Hi Chris,

Sorry for a delayed reply. The OSA 3210 is one of the only cesium standards I 
don't have in the museum. I've seen one or two on eBay over the past decade 
but not in worthwhile condition. I hope you can get yours working. If not, let 
me know...

The good news is that I have an original Oscilloquartz 3200 cesium manual. 
Today I scanned the first part of it for you. See:
http://www.leapsecond.com/museum/osa3200/

If this is helpful to your efforts, I'll try to scan the schematics as well. 
The manual is an inch thick.

I don't know the model differences. I've seen 3000, 3020 and 3200, 3210, 3230, 
3235 mentioned on the web. Maybe 3120. I'm also not clear about the differences 
in series-3000 vs. series-3100 vs. series 3200. If anyone knows, please post 
details.

Thanks,
/tvb

- Original Message - 
From: Chris syseng.greenfi...@btconnect.com
To: time-nuts@febo.com
Sent: Wednesday, August 27, 2014 2:09 PM
Subject: [time-nuts] Oscilloquartz 3210 Cesium Standard


 Hi,
 
 I recently bought an Oscilloquartz 3210 cesium standard and although it does
 power up, it does not lock. All the parameters on the meter look 
 sensible, other
 than the 2nd harmonic, #9 parameter, which shows zero. Quart oscillator 
 output
 looks almost spot on in comparison with the lab Z3816. I have no info on 
 this
 unit at all, so this may either be a dead tube, or other electronic fault on
 one of the many pcb's in the system. Ion pump had a small reading on 
 power up,
 but now dropped back to zero.
 
 I am looking for a user guide for this unit, or even better, a manual 
 with theory
 of operation and setup info. Can anyone help with this, or perhaps suggest a
 source ?. Year of manufacture estimated 1984/85, from date code on ic's.
 
 I notice that the tube is from FTS, the same construction as in the FTS 
 4060, but
 different part number and wonder if this was original fitment, or has been
 replaced at some stage...
 
 Regards  Thanks
 
 Chris Quayle
 
 Oxford, England


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Re: [time-nuts] Oscilloquartz 3210 Cesium Standard

2014-08-30 Thread Magnus Danielson

I have one of those modules but in a Telecom Solutions wrapping.

It seems like for your purposes, at least bypassing the output synthesis 
would be a good thing.


Cheers,
Magnus

On 08/29/2014 10:06 AM, Javier Herrero wrote:

It seems that later, they decided to shameless use the
FTS/Datum/Symmetricom FTS-5045 module
http://www.gigatest.net/datum/5045txt2.pdf

The OSA-5585 I've has one inside, labeled Symmetricom everywere, and the
Oscilloquartz contribution is a subrack containing the DC-input and
AC-input power supplies, a controller that manages the FTS-5045 through
its serial port, and some clock synthesis and distribution cards to
provide PPS, 10MHz and 2.048MHz, with a spectral quality a lot worse
than the output from the FTS-5045. I find the Oscilloquartz part of the
equipment not very good nor very usefult to my purposes, to a point I'm
thinking on to remove it completely an control/monitor directly the
FTS-5045 with whatever thing with a serial port and a display (my
Blackfin module, a Beaglebone o whatever similar)

Regards,

Javier

On 29/08/2014 1:23, Magnus Danielson wrote:

FTS had a patent on microcontroller steered cesium, which could
naturally have limited the spreading time of that technology.

Oscilloquartz at the time where more focused on the telecommunication
market and meeting the ITU-T G.811 PRC quality requirement, keeping
within +/. 1E-11 in frequency, and that is achievable with the analog
design, so no rush changing it.

FTS and HP where more into time-keeping, so therefore improving the
design made more market sense for them.

Anyway, that's about how I have perceived the market at the time.

Cheers,
Magnus

On 08/28/2014 11:47 PM, Chris wrote:

On 08/28/14 19:53, Javier Herrero wrote:

Hello,

Then it is a quite different beast to the EUDICS 3120, that they also
call OSA-3120... I note now that yours is a 3210, not 3120 :)

Regards,

Javier


The 3210 looks like a much earlier design, prior to the inclusion of
microprocessor control. Date codes look mid eighties, by which time
companies like FTS did have microprocessor control. Maybe their later
design products ran in parallel because of gov or esa contracts. In
theory, the more straightforward hardware design should make it easier
to keep running, tube life permitting, assuming the info can be found.

Thanks for sending it anyway - it all adds to the sum of knowledge...

Regards,

Chris

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Re: [time-nuts] Oscilloquartz 3210 Cesium Standard

2014-08-29 Thread paul swed
Magnus is right if there is any tube life at all and I do mean fumes. (odd
spacing all of the sudden) In HP 5060/5061 Frankenstein (A combo of the two
systems) I built a new heater controller to drive the few fumes off of the
5060 tube. Amazingly the darn thing works. The i meter barely barely moves.
But yet it locks.
Good luck.
Paul
WB8TSL


On Thu, Aug 28, 2014 at 12:47 PM, Magnus Danielson 
mag...@rubidium.dyndns.org wrote:

 If you have tube-life and not other issues, it's about the same.
 Also works for rubidiums, as the loop aspect here is essentially the same.

 There can be *other* issues. For the 5060A for instance, you might need to
 also adjust the crystal filter of the OCXO, as that too drifts out of
 range, so you get no signal out.

 What I write is not a fix-it-all but rather addresses that one issue.

 Cheers,
 Magnus


 On 08/28/2014 06:35 PM, Bob Bownes wrote:

 Is there a similar 'bring it back to life' procedure for the 5061A?


 On Thu, Aug 28, 2014 at 12:32 PM, Magnus Danielson 
 mag...@rubidium.dyndns.org wrote:

  Chris,

 Do you have a GPS clock?

 First turn the operational mode setting from off to second step (ocxo +
 ion pump) and let it stay there for a day or so.

 Then, as the oven have stabilized, switch it over to third step, the open
 loop mode, and tune the OCXO up against a GPS reference so that it is
 very
 near 5 MHz. You use the calibration whole on the front of the clock for
 that.

 Then, you turn the operation mode switch to the fourth and last step, the
 closed loop step, and see if it locks up. Let it just sit there and lock
 up, as it takes some time.

 It's quite common that OCXOs have drifted outside the capture range of
 the
 analogue loop, so loosing lock and not being able to attain it again is a
 typical response.

 Cheers,
 Magnus


 On 08/28/2014 04:33 PM, Chris wrote:

  On 08/28/14 05:03, Javier Herrero wrote:

  Hello,

 Here is the manual I've. I have also some other documentations, and
 some
 Oscilloquartz software for the OSA-5585, but I don't know if they are
 very useful.

 Regards,

 Javier


  Hi Javier,

 Thanks for that and for the other replies. The 3210 looks like quite an
 early design, with no sign of microprocessors at all. There's a 8 slot
 card cage with a load of discrete analog circuitry, 741 op amps etc and
 a couple of boards full of 14 / 16 pin ssi cmos / ttl devices, which I
 guess would be the synthesiser logic and perhaps a state machine style
 startup sequencer. Apart from that, the rest is power supply related and
 what looks like an alarm board with optoisolator discete outputs to a 25
 way D connector. The step recovery diode (?) multiplier into the
 microwave cavity is a really neat gold plated assembly with what looks
 like a 50r termination (setup tap for spectrum analyser ?) and an
 adjustment trimmer, but am not touching that or the many trimpots on the
 boards or any adjustments until I have more info. The tube is from FTS,
 part number / model 7101.

 It seems strange that the 2nd harmonic, meter #9, is zero, since even
 with a tube approaching eol, one would expect at least some indication,
 which is why I think there may be an electronic fault. Perhaps the hv
 power supply module feeding the electron multiplier. Will try to measure
 that, but the area around the tube is really heavily rivetted and
 screwed down in all directions. Looks like a lot of the left hand side
 of the case will need to be disassembled just to get at the tube
 connections. It also had the battery backup option, with 4 sets of 3 x
 cyclon type cells, but with a date code of 1984, are seriously dead and
 have been removed.

 This time nuts things seems to be a growing interest and wonder if there
 is a cure ? :-). Recently bought a 1970's era Tracor 304D rubidium
 standard. Again, no lock, but a very well engineered and screwed
 together piece of kit and should be fixable. Collection now includes the
 Z3816, from Ebay US around 7 years ago, a Z3815 currently being
 repackaged, an HP103 with open circuit oven heater elements and the
 3210...

 Regards,

 Chris



   ___

 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to
 https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.


   ___

 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to
 https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.

  ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to https://www.febo.com/cgi-bin/
 mailman/listinfo/time-nuts
 and follow the instructions there.

  ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to https://www.febo.com/cgi-bin/
 mailman/listinfo/time-nuts
 and follow the instructions there.

  

Re: [time-nuts] Oscilloquartz 3210 Cesium Standard

2014-08-29 Thread Javier Herrero
It seems that later, they decided to shameless use the 
FTS/Datum/Symmetricom FTS-5045 module 
http://www.gigatest.net/datum/5045txt2.pdf


The OSA-5585 I've has one inside, labeled Symmetricom everywere, and the 
Oscilloquartz contribution is a subrack containing the DC-input and 
AC-input power supplies, a controller that manages the FTS-5045 through 
its serial port, and some clock synthesis and distribution cards to 
provide PPS, 10MHz and 2.048MHz, with a spectral quality a lot worse 
than the output from the FTS-5045. I find the Oscilloquartz part of the 
equipment not very good nor very usefult to my purposes, to a point I'm 
thinking on to remove it completely an control/monitor directly the 
FTS-5045 with whatever thing with a serial port and a display (my 
Blackfin module, a Beaglebone o whatever similar)


Regards,

Javier

On 29/08/2014 1:23, Magnus Danielson wrote:
FTS had a patent on microcontroller steered cesium, which could 
naturally have limited the spreading time of that technology.


Oscilloquartz at the time where more focused on the telecommunication 
market and meeting the ITU-T G.811 PRC quality requirement, keeping 
within +/. 1E-11 in frequency, and that is achievable with the analog 
design, so no rush changing it.


FTS and HP where more into time-keeping, so therefore improving the 
design made more market sense for them.


Anyway, that's about how I have perceived the market at the time.

Cheers,
Magnus

On 08/28/2014 11:47 PM, Chris wrote:

On 08/28/14 19:53, Javier Herrero wrote:

Hello,

Then it is a quite different beast to the EUDICS 3120, that they also
call OSA-3120... I note now that yours is a 3210, not 3120 :)

Regards,

Javier


The 3210 looks like a much earlier design, prior to the inclusion of
microprocessor control. Date codes look mid eighties, by which time
companies like FTS did have microprocessor control. Maybe their later
design products ran in parallel because of gov or esa contracts. In
theory, the more straightforward hardware design should make it easier
to keep running, tube life permitting, assuming the info can be found.

Thanks for sending it anyway - it all adds to the sum of knowledge...

Regards,

Chris

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to 
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts

and follow the instructions there.

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Re: [time-nuts] Oscilloquartz 3210 Cesium Standard

2014-08-28 Thread Chris

On 08/28/14 05:03, Javier Herrero wrote:

Hello,

Here is the manual I've. I have also some other documentations, and some
Oscilloquartz software for the OSA-5585, but I don't know if they are
very useful.

Regards,

Javier



Hi Javier,

Thanks for that and for the other replies. The 3210 looks like quite an 
early design, with no sign of microprocessors at all. There's a 8 slot 
card cage with a load of discrete analog circuitry, 741 op amps etc and 
a couple of boards full of 14 / 16 pin ssi cmos / ttl devices, which I 
guess would be the synthesiser logic and perhaps a state machine style 
startup sequencer. Apart from that, the rest is power supply related and 
what looks like an alarm board with optoisolator discete outputs to a 25 
way D connector. The step recovery diode (?) multiplier into the 
microwave cavity is a really neat gold plated assembly with what looks 
like a 50r termination (setup tap for spectrum analyser ?) and an 
adjustment trimmer, but am not touching that or the many trimpots on the 
boards or any adjustments until I have more info. The tube is from FTS, 
part number / model 7101.


It seems strange that the 2nd harmonic, meter #9, is zero, since even 
with a tube approaching eol, one would expect at least some indication, 
which is why I think there may be an electronic fault. Perhaps the hv 
power supply module feeding the electron multiplier. Will try to measure 
that, but the area around the tube is really heavily rivetted and 
screwed down in all directions. Looks like a lot of the left hand side 
of the case will need to be disassembled just to get at the tube 
connections. It also had the battery backup option, with 4 sets of 3 x 
cyclon type cells, but with a date code of 1984, are seriously dead and 
have been removed.


This time nuts things seems to be a growing interest and wonder if there 
is a cure ? :-). Recently bought a 1970's era Tracor 304D rubidium 
standard. Again, no lock, but a very well engineered and screwed 
together piece of kit and should be fixable. Collection now includes the 
Z3816, from Ebay US around 7 years ago, a Z3815 currently being 
repackaged, an HP103 with open circuit oven heater elements and the 3210...


Regards,

Chris




___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.



___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Re: [time-nuts] Oscilloquartz 3210 Cesium Standard

2014-08-28 Thread bownes
Careful Chris, it sounds like you are developing the symptoms of the Vintage 
strain of the time nuts infection. Next thing you know, you will be looking at 
tall clocks. 

Bob, who is debating the wisdom of non invasively synchronizing the family 
heirloom tall clock to the new cesium clock...

 On Aug 28, 2014, at 10:33, Chris syseng.greenfi...@btconnect.com wrote:
 
 On 08/28/14 05:03, Javier Herrero wrote:
 Hello,
 
 Here is the manual I've. I have also some other documentations, and some
 Oscilloquartz software for the OSA-5585, but I don't know if they are
 very useful.
 
 Regards,
 
 Javier
 
 Hi Javier,
 
 Thanks for that and for the other replies. The 3210 looks like quite an early 
 design, with no sign of microprocessors at all. There's a 8 slot card cage 
 with a load of discrete analog circuitry, 741 op amps etc and a couple of 
 boards full of 14 / 16 pin ssi cmos / ttl devices, which I guess would be the 
 synthesiser logic and perhaps a state machine style startup sequencer. Apart 
 from that, the rest is power supply related and what looks like an alarm 
 board with optoisolator discete outputs to a 25 way D connector. The step 
 recovery diode (?) multiplier into the microwave cavity is a really neat gold 
 plated assembly with what looks like a 50r termination (setup tap for 
 spectrum analyser ?) and an adjustment trimmer, but am not touching that or 
 the many trimpots on the boards or any adjustments until I have more info. 
 The tube is from FTS, part number / model 7101.
 
 It seems strange that the 2nd harmonic, meter #9, is zero, since even with a 
 tube approaching eol, one would expect at least some indication, which is why 
 I think there may be an electronic fault. Perhaps the hv power supply module 
 feeding the electron multiplier. Will try to measure that, but the area 
 around the tube is really heavily rivetted and screwed down in all 
 directions. Looks like a lot of the left hand side of the case will need to 
 be disassembled just to get at the tube connections. It also had the battery 
 backup option, with 4 sets of 3 x cyclon type cells, but with a date code of 
 1984, are seriously dead and have been removed.
 
 This time nuts things seems to be a growing interest and wonder if there is a 
 cure ? :-). Recently bought a 1970's era Tracor 304D rubidium standard. 
 Again, no lock, but a very well engineered and screwed together piece of kit 
 and should be fixable. Collection now includes the Z3816, from Ebay US around 
 7 years ago, a Z3815 currently being repackaged, an HP103 with open circuit 
 oven heater elements and the 3210...
 
 Regards,
 
 Chris
 
 
 
 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to
 https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.
 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Re: [time-nuts] Oscilloquartz 3210 Cesium Standard

2014-08-28 Thread paul swed
I have the same disease. I have the clock, but the wife thinks that is
insane. I don't get her concern at all??? Especially when she says NO!.
Regards
Paul
WB8TSL


On Thu, Aug 28, 2014 at 10:39 AM, bownes bow...@gmail.com wrote:

 Careful Chris, it sounds like you are developing the symptoms of the
 Vintage strain of the time nuts infection. Next thing you know, you will be
 looking at tall clocks.

 Bob, who is debating the wisdom of non invasively synchronizing the family
 heirloom tall clock to the new cesium clock...

  On Aug 28, 2014, at 10:33, Chris syseng.greenfi...@btconnect.com
 wrote:
 
  On 08/28/14 05:03, Javier Herrero wrote:
  Hello,
 
  Here is the manual I've. I have also some other documentations, and some
  Oscilloquartz software for the OSA-5585, but I don't know if they are
  very useful.
 
  Regards,
 
  Javier
 
  Hi Javier,
 
  Thanks for that and for the other replies. The 3210 looks like quite an
 early design, with no sign of microprocessors at all. There's a 8 slot card
 cage with a load of discrete analog circuitry, 741 op amps etc and a couple
 of boards full of 14 / 16 pin ssi cmos / ttl devices, which I guess would
 be the synthesiser logic and perhaps a state machine style startup
 sequencer. Apart from that, the rest is power supply related and what looks
 like an alarm board with optoisolator discete outputs to a 25 way D
 connector. The step recovery diode (?) multiplier into the microwave cavity
 is a really neat gold plated assembly with what looks like a 50r
 termination (setup tap for spectrum analyser ?) and an adjustment trimmer,
 but am not touching that or the many trimpots on the boards or any
 adjustments until I have more info. The tube is from FTS, part number /
 model 7101.
 
  It seems strange that the 2nd harmonic, meter #9, is zero, since even
 with a tube approaching eol, one would expect at least some indication,
 which is why I think there may be an electronic fault. Perhaps the hv power
 supply module feeding the electron multiplier. Will try to measure that,
 but the area around the tube is really heavily rivetted and screwed down in
 all directions. Looks like a lot of the left hand side of the case will
 need to be disassembled just to get at the tube connections. It also had
 the battery backup option, with 4 sets of 3 x cyclon type cells, but with a
 date code of 1984, are seriously dead and have been removed.
 
  This time nuts things seems to be a growing interest and wonder if there
 is a cure ? :-). Recently bought a 1970's era Tracor 304D rubidium
 standard. Again, no lock, but a very well engineered and screwed together
 piece of kit and should be fixable. Collection now includes the Z3816, from
 Ebay US around 7 years ago, a Z3815 currently being repackaged, an HP103
 with open circuit oven heater elements and the 3210...
 
  Regards,
 
  Chris
 
 
 
  ___
  time-nuts mailing list -- time-nuts@febo.com
  To unsubscribe, go to
  https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
  and follow the instructions there.
  ___
  time-nuts mailing list -- time-nuts@febo.com
  To unsubscribe, go to
 https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
  and follow the instructions there.
 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to
 https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Re: [time-nuts] Oscilloquartz 3210 Cesium Standard

2014-08-28 Thread Magnus Danielson

Chris,

Do you have a GPS clock?

First turn the operational mode setting from off to second step (ocxo + 
ion pump) and let it stay there for a day or so.


Then, as the oven have stabilized, switch it over to third step, the 
open loop mode, and tune the OCXO up against a GPS reference so that it 
is very near 5 MHz. You use the calibration whole on the front of the 
clock for that.


Then, you turn the operation mode switch to the fourth and last step, 
the closed loop step, and see if it locks up. Let it just sit there and 
lock up, as it takes some time.


It's quite common that OCXOs have drifted outside the capture range of 
the analogue loop, so loosing lock and not being able to attain it again 
is a typical response.


Cheers,
Magnus

On 08/28/2014 04:33 PM, Chris wrote:

On 08/28/14 05:03, Javier Herrero wrote:

Hello,

Here is the manual I've. I have also some other documentations, and some
Oscilloquartz software for the OSA-5585, but I don't know if they are
very useful.

Regards,

Javier



Hi Javier,

Thanks for that and for the other replies. The 3210 looks like quite an
early design, with no sign of microprocessors at all. There's a 8 slot
card cage with a load of discrete analog circuitry, 741 op amps etc and
a couple of boards full of 14 / 16 pin ssi cmos / ttl devices, which I
guess would be the synthesiser logic and perhaps a state machine style
startup sequencer. Apart from that, the rest is power supply related and
what looks like an alarm board with optoisolator discete outputs to a 25
way D connector. The step recovery diode (?) multiplier into the
microwave cavity is a really neat gold plated assembly with what looks
like a 50r termination (setup tap for spectrum analyser ?) and an
adjustment trimmer, but am not touching that or the many trimpots on the
boards or any adjustments until I have more info. The tube is from FTS,
part number / model 7101.

It seems strange that the 2nd harmonic, meter #9, is zero, since even
with a tube approaching eol, one would expect at least some indication,
which is why I think there may be an electronic fault. Perhaps the hv
power supply module feeding the electron multiplier. Will try to measure
that, but the area around the tube is really heavily rivetted and
screwed down in all directions. Looks like a lot of the left hand side
of the case will need to be disassembled just to get at the tube
connections. It also had the battery backup option, with 4 sets of 3 x
cyclon type cells, but with a date code of 1984, are seriously dead and
have been removed.

This time nuts things seems to be a growing interest and wonder if there
is a cure ? :-). Recently bought a 1970's era Tracor 304D rubidium
standard. Again, no lock, but a very well engineered and screwed
together piece of kit and should be fixable. Collection now includes the
Z3816, from Ebay US around 7 years ago, a Z3815 currently being
repackaged, an HP103 with open circuit oven heater elements and the 3210...

Regards,

Chris




___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.



___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Re: [time-nuts] Oscilloquartz 3210 Cesium Standard

2014-08-28 Thread Chris

On 08/28/14 14:39, bownes wrote:

Careful Chris, it sounds like you are developing the symptoms of the Vintage 
strain of the time nuts infection. Next thing you know, you will be looking at 
tall clocks.

Bob, who is debating the wisdom of non invasively synchronizing the family 
heirloom tall clock to the new cesium clock...



It all started with the need for an accurate standard for the home lab. 
I had one of the Racal double oven standards for years, but a pot core 
transformer on the inner oven control board went open circuit on one 
winding and was unable to fix it at the time. Then bought the Z3816 on 
Ebay, which was working fine for years until the oscillator EFC became 
unstable and of course, then one thing leads to another.


As for clocks, have a collection of synchronous motor clocks awaiting 
restoration and a couple of pendulum clocks - A gents pulsynetic and an 
IBM clock from the 1940's, quarter sawn oak case and all. I do embedded 
systems here and the IBM master clock is appropriate - an example of how 
IBM didn't only build computers. It's a work of art, with thick brass 
plates for the mechanism and much of it gold plated to reduce corrosion. 
The self winding mechanism had a broken bracket on the solenoid 
armature, metal fatigue over the years I guess. That's been completely 
stripped, cleaned, sparingly relubed and now working as it should.


So, not really an obsession, but definately a growing interest area :-)...



___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

.


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Re: [time-nuts] Oscilloquartz 3210 Cesium Standard

2014-08-28 Thread Bob Bownes
Is there a similar 'bring it back to life' procedure for the 5061A?


On Thu, Aug 28, 2014 at 12:32 PM, Magnus Danielson 
mag...@rubidium.dyndns.org wrote:

 Chris,

 Do you have a GPS clock?

 First turn the operational mode setting from off to second step (ocxo +
 ion pump) and let it stay there for a day or so.

 Then, as the oven have stabilized, switch it over to third step, the open
 loop mode, and tune the OCXO up against a GPS reference so that it is very
 near 5 MHz. You use the calibration whole on the front of the clock for
 that.

 Then, you turn the operation mode switch to the fourth and last step, the
 closed loop step, and see if it locks up. Let it just sit there and lock
 up, as it takes some time.

 It's quite common that OCXOs have drifted outside the capture range of the
 analogue loop, so loosing lock and not being able to attain it again is a
 typical response.

 Cheers,
 Magnus


 On 08/28/2014 04:33 PM, Chris wrote:

 On 08/28/14 05:03, Javier Herrero wrote:

 Hello,

 Here is the manual I've. I have also some other documentations, and some
 Oscilloquartz software for the OSA-5585, but I don't know if they are
 very useful.

 Regards,

 Javier


 Hi Javier,

 Thanks for that and for the other replies. The 3210 looks like quite an
 early design, with no sign of microprocessors at all. There's a 8 slot
 card cage with a load of discrete analog circuitry, 741 op amps etc and
 a couple of boards full of 14 / 16 pin ssi cmos / ttl devices, which I
 guess would be the synthesiser logic and perhaps a state machine style
 startup sequencer. Apart from that, the rest is power supply related and
 what looks like an alarm board with optoisolator discete outputs to a 25
 way D connector. The step recovery diode (?) multiplier into the
 microwave cavity is a really neat gold plated assembly with what looks
 like a 50r termination (setup tap for spectrum analyser ?) and an
 adjustment trimmer, but am not touching that or the many trimpots on the
 boards or any adjustments until I have more info. The tube is from FTS,
 part number / model 7101.

 It seems strange that the 2nd harmonic, meter #9, is zero, since even
 with a tube approaching eol, one would expect at least some indication,
 which is why I think there may be an electronic fault. Perhaps the hv
 power supply module feeding the electron multiplier. Will try to measure
 that, but the area around the tube is really heavily rivetted and
 screwed down in all directions. Looks like a lot of the left hand side
 of the case will need to be disassembled just to get at the tube
 connections. It also had the battery backup option, with 4 sets of 3 x
 cyclon type cells, but with a date code of 1984, are seriously dead and
 have been removed.

 This time nuts things seems to be a growing interest and wonder if there
 is a cure ? :-). Recently bought a 1970's era Tracor 304D rubidium
 standard. Again, no lock, but a very well engineered and screwed
 together piece of kit and should be fixable. Collection now includes the
 Z3816, from Ebay US around 7 years ago, a Z3815 currently being
 repackaged, an HP103 with open circuit oven heater elements and the
 3210...

 Regards,

 Chris



  ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to
 https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.


  ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to
 https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.

 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to https://www.febo.com/cgi-bin/
 mailman/listinfo/time-nuts
 and follow the instructions there.

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Re: [time-nuts] Oscilloquartz 3210 Cesium Standard

2014-08-28 Thread Magnus Danielson

If you have tube-life and not other issues, it's about the same.
Also works for rubidiums, as the loop aspect here is essentially the same.

There can be *other* issues. For the 5060A for instance, you might need 
to also adjust the crystal filter of the OCXO, as that too drifts out of 
range, so you get no signal out.


What I write is not a fix-it-all but rather addresses that one issue.

Cheers,
Magnus

On 08/28/2014 06:35 PM, Bob Bownes wrote:

Is there a similar 'bring it back to life' procedure for the 5061A?


On Thu, Aug 28, 2014 at 12:32 PM, Magnus Danielson 
mag...@rubidium.dyndns.org wrote:


Chris,

Do you have a GPS clock?

First turn the operational mode setting from off to second step (ocxo +
ion pump) and let it stay there for a day or so.

Then, as the oven have stabilized, switch it over to third step, the open
loop mode, and tune the OCXO up against a GPS reference so that it is very
near 5 MHz. You use the calibration whole on the front of the clock for
that.

Then, you turn the operation mode switch to the fourth and last step, the
closed loop step, and see if it locks up. Let it just sit there and lock
up, as it takes some time.

It's quite common that OCXOs have drifted outside the capture range of the
analogue loop, so loosing lock and not being able to attain it again is a
typical response.

Cheers,
Magnus


On 08/28/2014 04:33 PM, Chris wrote:


On 08/28/14 05:03, Javier Herrero wrote:


Hello,

Here is the manual I've. I have also some other documentations, and some
Oscilloquartz software for the OSA-5585, but I don't know if they are
very useful.

Regards,

Javier



Hi Javier,

Thanks for that and for the other replies. The 3210 looks like quite an
early design, with no sign of microprocessors at all. There's a 8 slot
card cage with a load of discrete analog circuitry, 741 op amps etc and
a couple of boards full of 14 / 16 pin ssi cmos / ttl devices, which I
guess would be the synthesiser logic and perhaps a state machine style
startup sequencer. Apart from that, the rest is power supply related and
what looks like an alarm board with optoisolator discete outputs to a 25
way D connector. The step recovery diode (?) multiplier into the
microwave cavity is a really neat gold plated assembly with what looks
like a 50r termination (setup tap for spectrum analyser ?) and an
adjustment trimmer, but am not touching that or the many trimpots on the
boards or any adjustments until I have more info. The tube is from FTS,
part number / model 7101.

It seems strange that the 2nd harmonic, meter #9, is zero, since even
with a tube approaching eol, one would expect at least some indication,
which is why I think there may be an electronic fault. Perhaps the hv
power supply module feeding the electron multiplier. Will try to measure
that, but the area around the tube is really heavily rivetted and
screwed down in all directions. Looks like a lot of the left hand side
of the case will need to be disassembled just to get at the tube
connections. It also had the battery backup option, with 4 sets of 3 x
cyclon type cells, but with a date code of 1984, are seriously dead and
have been removed.

This time nuts things seems to be a growing interest and wonder if there
is a cure ? :-). Recently bought a 1970's era Tracor 304D rubidium
standard. Again, no lock, but a very well engineered and screwed
together piece of kit and should be fixable. Collection now includes the
Z3816, from Ebay US around 7 years ago, a Z3815 currently being
repackaged, an HP103 with open circuit oven heater elements and the
3210...

Regards,

Chris



  ___

time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.



  ___

time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/
mailman/listinfo/time-nuts
and follow the instructions there.


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Re: [time-nuts] Oscilloquartz 3210 Cesium Standard

2014-08-28 Thread Javier Herrero

Hello,

Then it is a quite different beast to the EUDICS 3120, that they also 
call OSA-3120... I note now that yours is a 3210, not 3120 :)


Regards,

Javier

P.S. no, there is no known cure to the time nuts things interest. It 
becomes chronic, and only gets worse ;)


On 28/08/2014 16:33, Chris wrote:

On 08/28/14 05:03, Javier Herrero wrote:

Hello,

Here is the manual I've. I have also some other documentations, and some
Oscilloquartz software for the OSA-5585, but I don't know if they are
very useful.

Regards,

Javier



Hi Javier,

Thanks for that and for the other replies. The 3210 looks like quite 
an early design, with no sign of microprocessors at all. There's a 8 
slot card cage with a load of discrete analog circuitry, 741 op amps 
etc and a couple of boards full of 14 / 16 pin ssi cmos / ttl devices, 
which I guess would be the synthesiser logic and perhaps a state 
machine style startup sequencer. Apart from that, the rest is power 
supply related and what looks like an alarm board with optoisolator 
discete outputs to a 25 way D connector. The step recovery diode (?) 
multiplier into the microwave cavity is a really neat gold plated 
assembly with what looks like a 50r termination (setup tap for 
spectrum analyser ?) and an adjustment trimmer, but am not touching 
that or the many trimpots on the boards or any adjustments until I 
have more info. The tube is from FTS, part number / model 7101.


It seems strange that the 2nd harmonic, meter #9, is zero, since even 
with a tube approaching eol, one would expect at least some 
indication, which is why I think there may be an electronic fault. 
Perhaps the hv power supply module feeding the electron multiplier. 
Will try to measure that, but the area around the tube is really 
heavily rivetted and screwed down in all directions. Looks like a lot 
of the left hand side of the case will need to be disassembled just to 
get at the tube connections. It also had the battery backup option, 
with 4 sets of 3 x cyclon type cells, but with a date code of 1984, 
are seriously dead and have been removed.


This time nuts things seems to be a growing interest and wonder if 
there is a cure ? :-). Recently bought a 1970's era Tracor 304D 
rubidium standard. Again, no lock, but a very well engineered and 
screwed together piece of kit and should be fixable. Collection now 
includes the Z3816, from Ebay US around 7 years ago, a Z3815 currently 
being repackaged, an HP103 with open circuit oven heater elements and 
the 3210...


Regards,

Chris




___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.






___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Re: [time-nuts] Oscilloquartz 3210 Cesium Standard

2014-08-28 Thread Chris

On 08/28/14 19:53, Javier Herrero wrote:

Hello,

Then it is a quite different beast to the EUDICS 3120, that they also
call OSA-3120... I note now that yours is a 3210, not 3120 :)

Regards,

Javier


The 3210 looks like a much earlier design, prior to the inclusion of 
microprocessor control. Date codes look mid eighties, by which time 
companies like FTS did have microprocessor control. Maybe their later 
design products ran in parallel because of gov or esa contracts. In 
theory, the more straightforward hardware design should make it easier 
to keep running, tube life permitting, assuming the info can be found.


Thanks for sending it anyway - it all adds to the sum of knowledge...

Regards,

Chris

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Re: [time-nuts] Oscilloquartz 3210 Cesium Standard

2014-08-28 Thread Magnus Danielson
FTS had a patent on microcontroller steered cesium, which could 
naturally have limited the spreading time of that technology.


Oscilloquartz at the time where more focused on the telecommunication 
market and meeting the ITU-T G.811 PRC quality requirement, keeping 
within +/. 1E-11 in frequency, and that is achievable with the analog 
design, so no rush changing it.


FTS and HP where more into time-keeping, so therefore improving the 
design made more market sense for them.


Anyway, that's about how I have perceived the market at the time.

Cheers,
Magnus

On 08/28/2014 11:47 PM, Chris wrote:

On 08/28/14 19:53, Javier Herrero wrote:

Hello,

Then it is a quite different beast to the EUDICS 3120, that they also
call OSA-3120... I note now that yours is a 3210, not 3120 :)

Regards,

Javier


The 3210 looks like a much earlier design, prior to the inclusion of
microprocessor control. Date codes look mid eighties, by which time
companies like FTS did have microprocessor control. Maybe their later
design products ran in parallel because of gov or esa contracts. In
theory, the more straightforward hardware design should make it easier
to keep running, tube life permitting, assuming the info can be found.

Thanks for sending it anyway - it all adds to the sum of knowledge...

Regards,

Chris

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

[time-nuts] Oscilloquartz 3210 Cesium Standard

2014-08-27 Thread Chris

Hi,

I recently bought an Oscilloquartz 3210 cesium standard and although it does
power up, it does not lock. All the parameters on the meter look 
sensible, other
than the 2nd harmonic, #9 parameter, which shows zero. Quart oscillator 
output
looks almost spot on in comparison with the lab Z3816. I have no info on 
this

unit at all, so this may either be a dead tube, or other electronic fault on
one of the many pcb's in the system. Ion pump had a small reading on 
power up,

but now dropped back to zero.

I am looking for a user guide for this unit, or even better, a manual 
with theory

of operation and setup info. Can anyone help with this, or perhaps suggest a
source ?. Year of manufacture estimated 1984/85, from date code on ic's.

I notice that the tube is from FTS, the same construction as in the FTS 
4060, but

different part number and wonder if this was original fitment, or has been
replaced at some stage...

Regards  Thanks

Chris Quayle

Oxford, England
.

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Re: [time-nuts] Oscilloquartz 3210 Cesium Standard

2014-08-27 Thread Javier Herrero

Hello,

I've some manuals from an OSA-5585, one of them for the EUDICS 3120. The 
thing that Symmetricom calls EUDICS-3120 really is a FTS-5045 module 
(Symmetricom in my case), that I think that it is the same that is 
inside the FTS-4060. It seems that one common problem with these, 
discuted some times in the list, is an STEL-1173 IC that has a failure 
mode in which the outputs stop working (one after another, not all at 
same time :) ).


Regards,

Javier

On 27/08/2014 23:09, Chris wrote:

Hi,

I recently bought an Oscilloquartz 3210 cesium standard and although 
it does
power up, it does not lock. All the parameters on the meter look 
sensible, other
than the 2nd harmonic, #9 parameter, which shows zero. Quart 
oscillator output
looks almost spot on in comparison with the lab Z3816. I have no info 
on this
unit at all, so this may either be a dead tube, or other electronic 
fault on
one of the many pcb's in the system. Ion pump had a small reading on 
power up,

but now dropped back to zero.

I am looking for a user guide for this unit, or even better, a manual 
with theory
of operation and setup info. Can anyone help with this, or perhaps 
suggest a

source ?. Year of manufacture estimated 1984/85, from date code on ic's.

I notice that the tube is from FTS, the same construction as in the 
FTS 4060, but
different part number and wonder if this was original fitment, or has 
been

replaced at some stage...

Regards  Thanks

Chris Quayle

Oxford, England
.

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to 
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts

and follow the instructions there.

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.