Re: [PSES] Ground Bond Tests

2021-05-13 Thread Richard Nute
Hi Steve: First, one of my clients has been told by their 3rd party NRTL that a ground bond test is required as part of factory/production routine tests, even though it is not required by 61010-1. I have never run into this requirement in my work history and I would be interested in

Re: [PSES] Ground Bond Tests

2021-05-12 Thread Wiseman, Joshua
ohms. Properly tightening the screw corrected the issue. In this situation, it passed ground continuity but failed ground bond. Josh From: Brian Kunde Sent: Wednesday, May 12, 2021 3:06 PM To: EMC-PSTC@LISTSERV.IEEE.ORG Subject: Re: [PSES] Ground Bond Tests EXTERNAL SENDER: Verify links

Re: [PSES] Ground Bond Tests

2021-05-12 Thread Brian Kunde
KE SHERMAN > *Sent:* Wednesday, May 12, 2021 11:08 AM > *To:* EMC-PSTC@LISTSERV.IEEE.ORG > *Subject:* Re: [PSES] Ground Bond Tests > > > > Rich Nute has 3 Technically Speaking columns on this that can be found at > the In Compliance magazine website. The first is here >

Re: [PSES] Ground Bond Tests

2021-05-12 Thread John Cochran
400 x219 From: MIKE SHERMAN Sent: Wednesday, May 12, 2021 11:08 AM To: EMC-PSTC@LISTSERV.IEEE.ORG Subject: Re: [PSES] Ground Bond Tests Rich Nute has 3 Technically Speaking columns on this that can be found at the In Compliance magazine website. The first is here https://incompliancemag.com/arti

Re: [PSES] Ground Bond Tests

2021-05-12 Thread MIKE SHERMAN
m: Brian Kunde > Sent: Wednesday, May 12, 2021 8:42 AM > To: EMC-PSTC@LISTSERV.IEEE.ORG > Subject: Re: [PSES] Ground Bond Tests > > > > EXTERNAL SENDER: Verify links, attachments and sender before taking action > > > > 61010-1 § 4.1

Re: [PSES] Ground Bond Tests

2021-05-12 Thread Wiseman, Joshua
III devices or devices using an external power supply (brick or wall-wart styles.) Josh From: Brian Kunde Sent: Wednesday, May 12, 2021 8:42 AM To: EMC-PSTC@LISTSERV.IEEE.ORG Subject: Re: [PSES] Ground Bond Tests EXTERNAL SENDER: Verify links, attachments and sender before taking action 61010

Re: [PSES] Ground Bond Tests

2021-05-12 Thread Brian Kunde
61010-1 § 4.1 says, "Tests in this standard are TYPE TESTS to be carried out on samples of equipment or parts. Their only purpose is to check that the design and construction ensure conformity with this standard. In addition, *manufacturers shall perform the ROUTINE TESTS of Annex F on 100 % of

[PSES] Ground Bond Tests

2021-05-11 Thread Steve Brody
So here is my question, or actually two of them: First, one of my clients has been told by their 3rd party NRTL that a ground bond test is required as part of factory/production routine tests, even though it is not required by 61010-1. I have never run into this requirement in my work history