Re: [Vo]:Inexpensive steam/water calorimeter

2011-09-28 Thread Horace Heffner


On Sep 27, 2011, at 9:27 PM, Peter Gluck wrote:


Dear Horace,

The missing variable is cooling water flow- to be established by  
Rossi- water that carries the excess heat generated by the 52 (?)  
Fat Cats and is partially transformed  in steam- F1.


To achieve accuracy in delta T measuring the condensing water flow  
rate should be adjusted to the flow rate of the steam.  If the flow  
is too high the delta T is small and even very small errors in  
measuring T translate into very large errors in delta T.  If the  
device enthalpy varies rapidly then it is much easier to adjust the  
cooling water flow to a longer term moving average than to  
instantaneous measurements.


The flow of mixing water- condensing  the steam is say, 5-10 times  
greater than F1 see please the formula given in my paper.


Like most people I don't generally go looking for a URL if it is not  
provided in a reference.


What matters is not the mixer cooling water flow rate but its  
combined temperature and flow rate.  The flow has to be matched to  
the steam thermal power, mass flow, and cooling water temperature in  
order to achieve a significant delta T.  This problem does not exist  
when the steam is condensed into a very large thermal mass of water -  
provided the large mass is kept in a useful temperature range, and  
the thermal power from the secondary cooling circuit is matched to  
the device thermal power.   If the thermal mass is large enough such  
matching can take place gradually and even manually, provided it is  
properly recorded.



No peristaltic but other types of positive displacement pumps to be  
used,


I said, Unfortunately my two peristaltic pumps are too small for  
this power range.   This does not imply that I would even consider  
trying to buy large peristaltic pumps.  Perhaps we have a language  
barrier.  Also, the flow rate for the cooling water should ideally be  
adjustable to the thermal power output of the device if that is  
variable and unpredictable.  An adjustable flow rate pump, or a  
selection of pumps, would thus be useful for driving the secondary  
cooling circuit.



e,g. gear pumps- for which the flow is not influenced by  
counterpressure.


The flow rate of gear pumps is influenced by a pumping into a large  
pressure head, both due to rpm loss (slip)  for AC induction motors  
under load, and due to rotor seal leakage under high pressure.  In  
the case of the new Rossi device, it looked like perhaps the water  
flow was entirely blocked towards the end of the test. This would  
create as large a pump pressure head as required to terminate flow.   
The evidence for flow blockage was the high pressure the device was  
under at the end.



This system measures the enthalpy in any moment, Including the  
start up period and possibly the heat after death.


The mass flow measurement depends on measuring the mixer exit mass  
flow.  This flow likely contains bubbles, is not well thermally  
mixed, and has fast dynamics requiring fast sampling times.  Some  
degree of smoothing increases reliability of the numbers and reduces  
the required sampling rate.  A large degree of smoothing provides a  
first principle check on the flow calorimetry numbers.  Of course, in  
the case of Rossi's device any even low precision mass flow  
calorimetry is an improvement.  In the case of my own work I would  
like some degree of consistency checking.  A hybrid method provides  
this consistency check.





The formula for efficiency is actually O/3I because electrical  
energy is at least  3 times more valuable or expensive than

thermal energy


That is not a formula for efficiency but relative value.



Peter

On Wed, Sep 28, 2011 at 7:38 AM, Horace Heffner  
hheff...@mtaonline.net wrote:


On Sep 27, 2011, at 9:35 AM, Jouni Valkonen wrote:

2011/9/27 Peter Gluck peter.gl...@gmail.com:
The simplest solution is to use a Steam Water mixing valve,in which  
the
heated mixture coming out from the demo is mixed with a constant  
flow of

cold water, you can know the enthalpy performance in any moment.

Indeed, continuous experiments easiest way is to use enthalpy sensors,
that gives as total enthalpy for any given moment. Even more simple is
to measure the steam pressure inside E-Cat, because it gives directly
the total enthalpy, but of course we need to first calibrate this kind
of enthalpy sensors.

   –Jouni




You have again not specified the precise method you would use.

It would appear you have a case of missing variables.  The  
principle missing variable is mass flow, m dot, which is best to  
isolate and measure directly.



Best regards,

Horace Heffner
http://www.mtaonline.net/~hheffner/







--
Dr. Peter Gluck
Cluj, Romania
http://egooutpeters.blogspot.com



Best regards,

Horace Heffner
http://www.mtaonline.net/~hheffner/






Re: [Vo]:Inexpensive steam/water calorimeter

2011-09-28 Thread Horace Heffner
When I say precise method  I mean the inclusion of the specific  
data to be obtained, where it is obtained,  and the formulas applied.


You wrote: Indeed, continuous experiments easiest way is to use  
enthalpy sensors, that gives as total enthalpy for any given moment.  
Even more simple is to measure the steam pressure inside E-Cat,  
because it gives directly the total enthalpy, but of course we need  
to first calibrate this kind of enthalpy sensors.


There is no such thing as an actual enthalpy sensor.  Only specific  
enthalpy is sensed.  Only incremental enthalpies (delta H) of a  
system can be measured. To obtain energy of a mass of steam, relative  
to that mass at some temperature, you need to know the mass of the  
steam.   The mass of an army tank differs from the mass of a small  
car.   Measuring only pressure, or specific enthalpy,  provides an  
insufficient amount of information. To obtain thermal power you need  
to know the mass flow.   The water overflow is a significant part of  
the flow by volume, more than 2% in some cases by volume.   This  
means the specific enthalpy of the steam is almost insignificant in  
those cases.


If x is the liquid portion by volume, then x/((x+(1-x)*0.0006)) is  
the portion by mass. This gives the following table which I posted  
here last January:


Liquid LiquidGas
PortionPortion   Portion
by Volume  by Mass   by Mass
-  ---   ---
0.000  0. 100.00
0.001  0.6252 0.3747
0.002  0.7695 0.2304
0.003  0.8337 0.1662
0.004  0.8700 0.1299
0.005  0.8933 0.1066
0.006  0.9095 0.0904
0.007  0.9215 0.0784
0.008  0.9307 0.0692
0.009  0.9380 0.0619
0.010  0.9439 0.0560
0.011  0.9488 0.0511
0.012  0.9529 0.0470
0.013  0.9564 0.0435
0.014  0.9594 0.0405

I consider the big deal about the definition of steam quality to be  
a red herring, a diversion from the important issues of measurement  
of the thermal power carried by the mass flow of a water steam mixture.


Best regards,

Horace Heffner
http://www.mtaonline.net/~hheffner/


On Sep 27, 2011, at 9:16 PM, Jouni Valkonen wrote:

First I would add to my previous message, that I think that Peter's  
method is more accurate than measuring pressure. That is because in  
order to find out correlation between pressure and enthalpy we need  
to do very careful calibration. In short run high accuracy may be  
difficult to archieve, but if experiment lasts for example 10 years  
continuously, then of course calibrating pressure sensor for  
enthalpy calculations will give great pay off.


Horace wrote:
« You have again not specified the precise method you would use.

It would appear you have a case of missing variables. The principle  
missing variable is mass flow, m dot, which is best to isolate and  
measure directly. »


Actually I have defined but it is so simple that you have probably  
missed it. First of course, we need to know that system is at  
equilibrium, i.e. water massflow in and massflow out are both  
matching. If water inflow rate varies a lot then calculations and  
calibrations are difficult, if system is overflowing. That means  
that for sure massflow must be known and it must be measured in  
calibration.


But if system is a kettle boiler that does not overflow, then  
calibration is very easy. In industrial water boilers this is the  
most reasonable situation because this ensures high steam quality  
because we can easily superheat steam to remove that 1-2% natural  
wettness of steam. This reduces the corrosion. Superheating can  
also be considered in calculations so this does not reduce the  
accuracy of method.


Pressure can be measured either directly with pressure sensor  
(easiest and most reliable and it is always available in pressure  
boilers.) or in kettle boilers boiling water temperature can be  
measured or last method is to measure steam temperature (this works  
only if steam is not superheated and is thus wet. I.e. steam  
quality must be measured, therefore this method is not universal).


—Jouni

On Sep 28, 2011 7:41 AM, Horace Heffner hheff...@mtaonline.net  
wrote:


 On Sep 27, 2011, at 9:35 AM, Jouni Valkonen wrote:

 2011/9/27 Peter Gluck peter.gl...@gmail.com:
 The simplest solution is to use a Steam Water mixing valve,in
 which the
 heated mixture coming out from the demo is mixed with a constant
 flow of
 cold water, you can know the enthalpy performance in any moment.

 Indeed, continuous experiments easiest way is to use enthalpy  
sensors,
 that gives as total enthalpy for any given moment. Even more  
simple is
 to measure the steam pressure inside E-Cat, because it gives  
directly
 the total enthalpy, but of course we need to first calibrate  
this kind

 of enthalpy sensors.

 –Jouni




 You have again not specified the precise method you would use.

 It would appear you have a case 

Re: [Vo]:Inexpensive steam/water calorimeter

2011-09-27 Thread Peter Gluck
The simplest solution is to use a *Steam Water mixing* *valve*,in which the
heated mixture coming out from the demo is mixed with a constant flow of
cold water, you can know the enthalpy performance in any moment.
Peter

On Tue, Sep 27, 2011 at 7:41 PM, Horace Heffner hheff...@mtaonline.netwrote:

 A simple inexpensive continuously operating steam/water calorimeter can be
 obtained using a combined barrel and flow calorimetry.

 A water container, a barrel, or perhaps a trash can which is silicone
 sealed for leaks, can be used to condense steam via a submerged copper coil,
 preferably mostly located near the top of the barrel to avoid imposing a
 steam pressure head on the tested device.  This water container can be
 insulated cheaply using construction foam board and fiberglass.  A stirrer
 can be driven via a shaft through the foam board.

 A secondary coil can be used for pumped coolant.  A fixed flow rate pump
 can be used to deliver the coolant flow.  The coolant flow circuit can be
 open or closed. A closed secondary coolant temperature can be maintained via
 either water or air heat exchange or ice heat exchange.  The source of the
 coolant energy is not important if the Tin and Tout are measured close to
 the water container, and any tubing between the temperature measuring
 stations and the water container is insulated.  Ideally the secondary flow
 rate would be measured by a digital flow meter, and driven by a variable
 speed pump.  The coolant flow rate can then be adjusted to suit the coolant
 delta T and the thermal power of the device under test. Alternatively, an
 accurate fixed flow rate pump can be chosen with a flow rate approximately
 matching the expected thermal power of the device under test given the
 expected coolant delta T.  A reasonable goal for the water container
 temperature is the range 50°C to 70°C.

 Use of a large water container provides some degree of momentary heat pulse
 energy integration and confidence in the device thermal power measurements.
 It applies a significant time constant to the thermal data that reduces the
 frequency temperature data must be taken.  It even permits manual
 temperature reading if a modestly stable condition is established.  This is
 at the cost of being able to see instant response thermal and energy output
 curves. There is no need to see such fast response curves if the primary
 goal is to measure total energy in vs total energy out for a long run.

 The primary circuit water flow can be pumped directly from the water
 container. Ideally the primary water flow should be measured by digital flow
 meter. If a low pressure head is presented to the primary circuit flow pump,
 then a precision fixed flow rate pump can be used.  If precision digital
 flow meters are not used, and reliance is placed on precision flow rate
 pumps, then at minimum simple (flow integrating) water meters should be
 monitored periodically to verify assumed pump mean flow rates. Calibration
 runs on dummy devices should be used to verify the calorimeter over the
 thermal range expected.  A calibration control run should be used with the
 device under test to determine the water capacity of the device so the
 volume of water in the barrel is known in order to provide improved
 intermediate time thermal power  measurements.  At the conclusion of a run,
 the circuits should continue to be driven until thermal equilibrium is
 obtained and essentially all thermal energy is drained form the device under
 test. A water depth gage for the barrel may be of use, calibrated to depth
 vs volume, in order to keep track of the amount of water in the device under
 test.

 The secondary circuit input and output temperature should be recorded
 frequently.  Alternatively, a direct delta T can be measured frequently
 using an appropriate dual thermocouple arrangement, thus providing improved
 data quality and reducing data acquisition required. Flow stirrers should be
 used, if feasible, in the secondary circuit prior to the thermometer wells.
 Barrel water temperature should be monitored. Ideally primary circuit water
 input temperature and room temperature should be monitored as well.

 A thermal decline curve should be measured for the water container when
 there is no primary circuit flow, and the water is stirred.   The
 calorimeter constant C(dT) as a function of the difference between room
 temperature and water contained temperature (dT) should be determined. The
 curve C(dT) can be fit to a polynomial using regression analysis for
 convenient use in data analysis. Experience shows this method is not very
 accurate if the water container is not well insulated.  This is due to room
 drafts, variations in humidity and temperature during the day, etc.  Ideally
 active insulation could be used, whereby an extra envelope surrounds the
 water container insulation and the temperature there is maintained at the
 temperature of the water, thereby producing a dT = 0, and no heat loss.

Re: [Vo]:Inexpensive steam/water calorimeter

2011-09-27 Thread Jouni Valkonen
2011/9/27 Peter Gluck peter.gl...@gmail.com:
 The simplest solution is to use a Steam Water mixing valve,in which the
 heated mixture coming out from the demo is mixed with a constant flow of
 cold water, you can know the enthalpy performance in any moment.

Indeed, continuous experiments easiest way is to use enthalpy sensors,
that gives as total enthalpy for any given moment. Even more simple is
to measure the steam pressure inside E-Cat, because it gives directly
the total enthalpy, but of course we need to first calibrate this kind
of enthalpy sensors.

–Jouni



Re: [Vo]:Inexpensive steam/water calorimeter

2011-09-27 Thread Joe Catania

It might be nice to know the metal mass and temps as well.
- Original Message - 
From: Horace Heffner hheff...@mtaonline.net

To: Vortex-L vortex-l@eskimo.com
Sent: Tuesday, September 27, 2011 12:41 PM
Subject: [Vo]:Inexpensive steam/water calorimeter


A simple inexpensive continuously operating steam/water calorimeter
can be obtained using a combined barrel and flow calorimetry.

A water container, a barrel, or perhaps a trash can which is silicone
sealed for leaks, can be used to condense steam via a submerged
copper coil, preferably mostly located near the top of the barrel to
avoid imposing a steam pressure head on the tested device.  This
water container can be insulated cheaply using construction foam
board and fiberglass.  A stirrer can be driven via a shaft through
the foam board.

A secondary coil can be used for pumped coolant.  A fixed flow rate
pump can be used to deliver the coolant flow.  The coolant flow
circuit can be open or closed. A closed secondary coolant temperature
can be maintained via either water or air heat exchange or ice heat
exchange.  The source of the coolant energy is not important if the
Tin and Tout are measured close to the water container, and any
tubing between the temperature measuring stations and the water
container is insulated.  Ideally the secondary flow rate would be
measured by a digital flow meter, and driven by a variable speed
pump.  The coolant flow rate can then be adjusted to suit the coolant
delta T and the thermal power of the device under test.
Alternatively, an accurate fixed flow rate pump can be chosen with a
flow rate approximately matching the expected thermal power of the
device under test given the expected coolant delta T.  A reasonable
goal for the water container temperature is the range 50°C to 70°C.

Use of a large water container provides some degree of momentary heat
pulse energy integration and confidence in the device thermal power
measurements. It applies a significant time constant to the thermal
data that reduces the frequency temperature data must be taken.  It
even permits manual temperature reading if a modestly stable
condition is established.  This is at the cost of being able to see
instant response thermal and energy output curves. There is no need
to see such fast response curves if the primary goal is to measure
total energy in vs total energy out for a long run.

The primary circuit water flow can be pumped directly from the water
container. Ideally the primary water flow should be measured by
digital flow meter. If a low pressure head is presented to the
primary circuit flow pump, then a precision fixed flow rate pump can
be used.  If precision digital flow meters are not used, and reliance
is placed on precision flow rate pumps, then at minimum simple (flow
integrating) water meters should be monitored periodically to verify
assumed pump mean flow rates. Calibration runs on dummy devices
should be used to verify the calorimeter over the thermal range
expected.  A calibration control run should be used with the device
under test to determine the water capacity of the device so the
volume of water in the barrel is known in order to provide improved
intermediate time thermal power  measurements.  At the conclusion of
a run, the circuits should continue to be driven until thermal
equilibrium is obtained and essentially all thermal energy is drained
form the device under test. A water depth gage for the barrel may be
of use, calibrated to depth vs volume, in order to keep track of the
amount of water in the device under test.

The secondary circuit input and output temperature should be recorded
frequently.  Alternatively, a direct delta T can be measured
frequently using an appropriate dual thermocouple arrangement, thus
providing improved data quality and reducing data acquisition
required. Flow stirrers should be used, if feasible, in the secondary
circuit prior to the thermometer wells. Barrel water temperature
should be monitored. Ideally primary circuit water input temperature
and room temperature should be monitored as well.

A thermal decline curve should be measured for the water container
when there is no primary circuit flow, and the water is stirred.
The calorimeter constant C(dT) as a function of the difference
between room temperature and water contained temperature (dT) should
be determined. The curve C(dT) can be fit to a polynomial using
regression analysis for convenient use in data analysis. Experience
shows this method is not very accurate if the water container is not
well insulated.  This is due to room drafts, variations in humidity
and temperature during the day, etc.  Ideally active insulation could
be used, whereby an extra envelope surrounds the water container
insulation and the temperature there is maintained at the temperature
of the water, thereby producing a dT = 0, and no heat loss.  This is
excessive for this approach, however, the goals of which are cheap,
simple,  and good enough.

Re: [Vo]:Inexpensive steam/water calorimeter

2011-09-27 Thread Peter Gluck
Dear Jouni,
I have described this method long ago, for individual e-Cats
e.g. here:
http://egooutpeters.blogspot.com/2011/05/call-for-perfect-e-cat-experiment.html

See please: http://www.onlineconversion.com/mixing_water.htm

I have asked Rossii to use this method but he has ignored it

with hostility- I could never understand why he don't want

good correct measurements

Peter


On Tue, Sep 27, 2011 at 8:35 PM, Jouni Valkonen jounivalko...@gmail.comwrote:

 2011/9/27 Peter Gluck peter.gl...@gmail.com:
  The simplest solution is to use a Steam Water mixing valve,in which the
  heated mixture coming out from the demo is mixed with a constant flow of
  cold water, you can know the enthalpy performance in any moment.

 Indeed, continuous experiments easiest way is to use enthalpy sensors,
 that gives as total enthalpy for any given moment. Even more simple is
 to measure the steam pressure inside E-Cat, because it gives directly
 the total enthalpy, but of course we need to first calibrate this kind
 of enthalpy sensors.

–Jouni




-- 
Dr. Peter Gluck
Cluj, Romania
http://egooutpeters.blogspot.com


Re: [Vo]:Inexpensive steam/water calorimeter

2011-09-27 Thread Horace Heffner


On Sep 27, 2011, at 10:49 AM, Peter Gluck wrote:


Dear Jouni,
I have described this method long ago, for individual e-Cats


A key part of this idea is the reliability obtained by the averaging  
performed by the large thermal mass of the water container.  I am  
suggesting a hybrid design, a hybrid flow and partial isoperibolic  
method.  It would of course be feasible to employ a mixer and extra  
thermometer just prior to the water container which does the  
averaging, but that would also require an extra pump, and flow meter.


I should also note this idea was initially largely for my own use.  I  
have a potential use for calorimetry in the multiple kW range.  I  
optimize the  cheap variable when designing for my own purposes,  
with some constraints regarding reliability and accuracy.  This is  
because I am so tight with money the little birdies say cheep  
cheep when they fly over me.  8^)  I have a 5 thermometer system  
that should work OK with this approach even with manual recording and  
spreadsheet analysis.  In winter I have the advantage of practically  
unlimited cooling capacity here in Alaska. Unfortunately my two  
peristaltic pumps are too small for this power range.   I can readily  
afford the barrel, blue board insulation, copper pipe and hose,  
fittings etc.


Best regards,

Horace Heffner
http://www.mtaonline.net/~hheffner/






Re: [Vo]:Inexpensive steam/water calorimeter

2011-09-27 Thread Horace Heffner


On Sep 27, 2011, at 9:35 AM, Jouni Valkonen wrote:


2011/9/27 Peter Gluck peter.gl...@gmail.com:
The simplest solution is to use a Steam Water mixing valve,in  
which the
heated mixture coming out from the demo is mixed with a constant  
flow of

cold water, you can know the enthalpy performance in any moment.


Indeed, continuous experiments easiest way is to use enthalpy sensors,
that gives as total enthalpy for any given moment. Even more simple is
to measure the steam pressure inside E-Cat, because it gives directly
the total enthalpy, but of course we need to first calibrate this kind
of enthalpy sensors.

–Jouni





You have again not specified the precise method you would use.

It would appear you have a case of missing variables.  The principle  
missing variable is mass flow, m dot, which is best to isolate and  
measure directly.


Best regards,

Horace Heffner
http://www.mtaonline.net/~hheffner/






Re: [Vo]:Inexpensive steam/water calorimeter

2011-09-27 Thread Jouni Valkonen
First I would add to my previous message, that I think that Peter's method
is more accurate than measuring pressure. That is because in order to find
out correlation between pressure and enthalpy we need to do very careful
calibration. In short run high accuracy may be difficult to archieve, but if
experiment lasts for example 10 years continuously, then of course
calibrating pressure sensor for enthalpy calculations will give great pay
off.

Horace wrote:
« You have again not specified the precise method you would use.

It would appear you have a case of missing variables. The principle missing
variable is mass flow, m dot, which is best to isolate and measure directly.
»

Actually I have defined but it is so simple that you have probably missed
it. First of course, we need to know that system is at equilibrium, i.e.
water massflow in and massflow out are both matching. If water inflow rate
varies a lot then calculations and calibrations are difficult, if system is
overflowing. That means that for sure massflow must be known and it must be
measured in calibration.

But if system is a kettle boiler that does not overflow, then calibration is
very easy. In industrial water boilers this is the most reasonable situation
because this ensures high steam quality because we can easily superheat
steam to remove that 1-2% natural wettness of steam. This reduces the
corrosion. Superheating can also be considered in calculations so this does
not reduce the accuracy of method.

Pressure can be measured either directly with pressure sensor (easiest and
most reliable and it is always available in pressure boilers.) or in kettle
boilers boiling water temperature can be measured or last method is to
measure steam temperature (this works only if steam is not superheated and
is thus wet. I.e. steam quality must be measured, therefore this method is
not universal).

—Jouni
On Sep 28, 2011 7:41 AM, Horace Heffner hheff...@mtaonline.net wrote:

 On Sep 27, 2011, at 9:35 AM, Jouni Valkonen wrote:

 2011/9/27 Peter Gluck peter.gl...@gmail.com:
 The simplest solution is to use a Steam Water mixing valve,in
 which the
 heated mixture coming out from the demo is mixed with a constant
 flow of
 cold water, you can know the enthalpy performance in any moment.

 Indeed, continuous experiments easiest way is to use enthalpy sensors,
 that gives as total enthalpy for any given moment. Even more simple is
 to measure the steam pressure inside E-Cat, because it gives directly
 the total enthalpy, but of course we need to first calibrate this kind
 of enthalpy sensors.

 –Jouni




 You have again not specified the precise method you would use.

 It would appear you have a case of missing variables. The principle
 missing variable is mass flow, m dot, which is best to isolate and
 measure directly.

 Best regards,

 Horace Heffner
 http://www.mtaonline.net/~hheffner/






Re: [Vo]:Inexpensive steam/water calorimeter

2011-09-27 Thread Peter Gluck
Dear Horace,

The missing variable is cooling water flow- to be established by Rossi-
water that carries the excess heat generated by the 52 (?) Fat Cats and is
partially transformed  in steam- F1.
The flow of mixing water- condensing  the steam is say, 5-10 times greater
than F1 see please the formula given in my paper.
No peristaltic but other types of positive displacement pumps to be used,
e,g. gear pumps- for which the flow is not influenced by counterpressure.
This system measures the enthalpy in any moment, Including the start up
period and possibly the heat after death.
The formula for efficiency is actually O/3I because electrical energy is at
least  3 times more valuable or expensive than
thermal energy
Peter

On Wed, Sep 28, 2011 at 7:38 AM, Horace Heffner hheff...@mtaonline.netwrote:


 On Sep 27, 2011, at 9:35 AM, Jouni Valkonen wrote:

  2011/9/27 Peter Gluck peter.gl...@gmail.com:

 The simplest solution is to use a Steam Water mixing valve,in which the
 heated mixture coming out from the demo is mixed with a constant flow of
 cold water, you can know the enthalpy performance in any moment.


 Indeed, continuous experiments easiest way is to use enthalpy sensors,
 that gives as total enthalpy for any given moment. Even more simple is
 to measure the steam pressure inside E-Cat, because it gives directly
 the total enthalpy, but of course we need to first calibrate this kind
 of enthalpy sensors.

–Jouni




 You have again not specified the precise method you would use.

 It would appear you have a case of missing variables.  The principle
 missing variable is mass flow, m dot, which is best to isolate and measure
 directly.


 Best regards,

 Horace Heffner
 http://www.mtaonline.net/~**hheffner/http://www.mtaonline.net/~hheffner/







-- 
Dr. Peter Gluck
Cluj, Romania
http://egooutpeters.blogspot.com


Re: [Vo]:Inexpensive steam/water calorimeter

2011-09-27 Thread Peter Gluck
We have to measure HEAT OUT (enthalpy) and this is the unique relevant
variable, steam, water, pressure, dryness, even F1 (this HAS to be measured
anyway) are not relevant.

When I speak about steam water mixing it is because I have used the method
many times many years ago.

But I bet that Rossi who is attracted by complications will not use this
simplissim clear method.

Peter

On Wed, Sep 28, 2011 at 8:16 AM, Jouni Valkonen jounivalko...@gmail.comwrote:

 First I would add to my previous message, that I think that Peter's method
 is more accurate than measuring pressure. That is because in order to find
 out correlation between pressure and enthalpy we need to do very careful
 calibration. In short run high accuracy may be difficult to archieve, but if
 experiment lasts for example 10 years continuously, then of course
 calibrating pressure sensor for enthalpy calculations will give great pay
 off.

 Horace wrote:
 « You have again not specified the precise method you would use.

 It would appear you have a case of missing variables. The principle missing
 variable is mass flow, m dot, which is best to isolate and measure directly.
 »

 Actually I have defined but it is so simple that you have probably missed
 it. First of course, we need to know that system is at equilibrium, i.e.
 water massflow in and massflow out are both matching. If water inflow rate
 varies a lot then calculations and calibrations are difficult, if system is
 overflowing. That means that for sure massflow must be known and it must be
 measured in calibration.

 But if system is a kettle boiler that does not overflow, then calibration
 is very easy. In industrial water boilers this is the most reasonable
 situation because this ensures high steam quality because we can easily
 superheat steam to remove that 1-2% natural wettness of steam. This reduces
 the corrosion. Superheating can also be considered in calculations so this
 does not reduce the accuracy of method.

 Pressure can be measured either directly with pressure sensor (easiest and
 most reliable and it is always available in pressure boilers.) or in kettle
 boilers boiling water temperature can be measured or last method is to
 measure steam temperature (this works only if steam is not superheated and
 is thus wet. I.e. steam quality must be measured, therefore this method is
 not universal).

 —Jouni
 On Sep 28, 2011 7:41 AM, Horace Heffner hheff...@mtaonline.net wrote:
 
  On Sep 27, 2011, at 9:35 AM, Jouni Valkonen wrote:
 
  2011/9/27 Peter Gluck peter.gl...@gmail.com:
  The simplest solution is to use a Steam Water mixing valve,in
  which the
  heated mixture coming out from the demo is mixed with a constant
  flow of
  cold water, you can know the enthalpy performance in any moment.
 
  Indeed, continuous experiments easiest way is to use enthalpy sensors,
  that gives as total enthalpy for any given moment. Even more simple is
  to measure the steam pressure inside E-Cat, because it gives directly
  the total enthalpy, but of course we need to first calibrate this kind
  of enthalpy sensors.
 
  –Jouni
 
 
 
 
  You have again not specified the precise method you would use.
 
  It would appear you have a case of missing variables. The principle
  missing variable is mass flow, m dot, which is best to isolate and
  measure directly.
 
  Best regards,
 
  Horace Heffner
  http://www.mtaonline.net/~hheffner/
 
 
 
 




-- 
Dr. Peter Gluck
Cluj, Romania
http://egooutpeters.blogspot.com