The Z3805A has two serial ports, both of which are RS-232.  Serial Port 2 
broadcasts time of day only and will not accept scpi commands.  Serial Port 1 
shoudl be used for scpi control.
 
Chuck Zabilski
BD Systems, Inc.
 

________________________________
 From: "[email protected]" <[email protected]>
To: [email protected] 
Sent: Saturday, October 13, 2012 11:31 PM
Subject: time-nuts Digest, Vol 99, Issue 62
  
Send time-nuts mailing list submissions to
    [email protected]

To subscribe or unsubscribe via the World Wide Web, visit
    https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
or, via email, send a message with subject or body 'help' to
    [email protected]

You can reach the person managing the list at
    [email protected]

When replying, please edit your Subject line so it is more specific
than "Re: Contents of time-nuts digest..."


Today's Topics:

   1. Re: HP 10811A failure ([email protected])
   2. Re: PRS10 dying from old age... (paul swed)
   3. Re: HP 10811A failure (Adrian)
   4. Re: HP 10811A failure ([email protected])
   5. HP Z3805A com port and monitoring questions (Edgardo Molina)
   6. Re: PRS10 dying from old age... (Magnus Danielson)
   7. Seistek pn3048 (David Hooke)


----------------------------------------------------------------------

Message: 1
Date: Sat, 13 Oct 2012 16:31:24 -0400 (EDT)
From: [email protected]
To: [email protected]
Subject: Re: [time-nuts] HP 10811A failure
Message-ID: <[email protected]>
Content-Type: text/plain; charset="US-ASCII"

Perhaps a silly question, but as I get the impression that both seem  to 
have failed simultaneously with the same fault I'm just wondering  if you're 
sure it isn't what you're checking them with that's  developed a problem?

Regards

Nigel
GM8PZR


In a message dated 13/10/2012 13:52:46 GMT Daylight Time, [email protected]  
writes:

Hi  All,

both of my double oven 10811A's have been running flawlessly until  now 
when I noticed that the output power has dropped by about 3 dB  (measures 
only 4...4.5 dBm at 50 Ohm load), and the noise floor has gone  up by 
about 20 dB, while the frequency is still spot on.

I used  them for various mesurements, so I can unfortunately not remember 
what  might have caused the failure. Only the 12V linear lab power supply 
that  fed the heaters had developed a completely dead filter cap, so the 
feeding  voltage was a 100 Hz sawtooth rather than DC. The heaters appear 
to be  still working though. I'm only using the internal heater. They 
draw some  700 mA at power up and drop to around 150 mA  each when warm. 
So the  oscillators are unlikely to be cooked inside. Something must have 
happened  to the 10 MHz output, possibly an ESD issue?

Before I start taking them  apart, can anyone tell what has happened or 
what to look for  first?

Regards,
Adrian



_______________________________________________
time-nuts  mailing list -- [email protected]
To unsubscribe, go to  
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the  instructions there.



------------------------------

Message: 2
Date: Sat, 13 Oct 2012 17:19:09 -0400
From: paul swed <[email protected]>
To: Discussion of precise time and frequency measurement
    <[email protected]>
Subject: Re: [time-nuts] PRS10 dying from old age...
Message-ID:
    <CAD2JfAhnLv1q4=otfmmbaiwneg84nrbvejdzihfhvx_v7-1...@mail.gmail.com>
Content-Type: text/plain; charset=ISO-8859-1

Poul
If it is dying there should maybe be a lamp voltage you can check and see
what it indicates. If you consider it dead then you can always checK to see
is it something going on like low RF or maybe the RB truly is an issue. I
have recovered some olf FRS by reheating the rb bulb and essentially
getting the RB that collects on the bulb back into circulation. Heck once
its dead you can have real fun.
Regards
Paul
WB8TSL




On Fri, Oct 12, 2012 at 4:48 PM, Poul-Henning Kamp <[email protected]>wrote:

>
> One of my PRS10s is dying from old age by the looks of it, here is some
> data I have
> collected from it over the last 800 days:
>
> http://phk.freebsd.dk/hacks/prs10_death/index.html
>
> --
> Poul-Henning Kamp       | UNIX since Zilog Zeus 3.20
> [email protected]         | TCP/IP since RFC 956
> FreeBSD committer       | BSD since 4.3-tahoe
> Never attribute to malice what can adequately be explained by incompetence.
>
> _______________________________________________
> time-nuts mailing list -- [email protected]
> To unsubscribe, go to
> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.
>


------------------------------

Message: 3
Date: Sun, 14 Oct 2012 01:26:35 +0200
From: Adrian <[email protected]>
To: Discussion of precise time and frequency measurement
    <[email protected]>
Subject: Re: [time-nuts] HP 10811A failure
Message-ID: <[email protected]>
Content-Type: text/plain; charset=ISO-8859-1; format=flowed

A very good point!

I checked the output amplitude with a spectrum analyzer, a power meter 
and a scope, the latter with a 50 ohm load to the input.
I re-checked the power supply connections and can confirm they are the 
same as before.
I used three different power supplies for the oscillators and two for 
the heaters, still no difference.
But...

To make it short, your comments helped to get me back on track. Both 
beauties escaped unnecessary dismantling and are now working as they should.

Actually, the problem was caused by a long-term misunderstanding. I was 
always wondering why 'HP used simple stranded wire' and not coaxial 
cable on the 10 MHz output and EFC.
As long as I connected them on the bench, I replaced the 'missing' 
ground connection with a short wired croc clamp between coax and ground. 
Today I realized that the thin blue wire IS actually coax cable, so I 
should have connected the coax shield to the BNC ground on my newly 
built 2x 10811 enclosure. With the new wiring, the output ground had 
just become much more inductive, up to a point where the nominal source 
impedance of 50 ohms had increased to over 120 ohms, causing the 
amplitude loss of some 3 dB and capturing noise.

Btw. the Sprague 6800 uF / 40 V from my R&S NGA power supply has indeed 
died, but independently of the osc. problem.

Adrian


[email protected] schrieb:
> Perhaps a silly question, but as I get the impression that both seem  to
> have failed simultaneously with the same fault I'm just wondering  if you're
> sure it isn't what you're checking them with that's  developed a problem?
>  
> Regards
>  
> Nigel
> GM8PZR
>  
>  
> In a message dated 13/10/2012 13:52:46 GMT Daylight Time, [email protected]
> writes:
>
> Hi  All,
>
> both of my double oven 10811A's have been running flawlessly until  now
> when I noticed that the output power has dropped by about 3 dB  (measures
> only 4...4.5 dBm at 50 Ohm load), and the noise floor has gone  up by
> about 20 dB, while the frequency is still spot on.
>
> I used  them for various mesurements, so I can unfortunately not remember
> what  might have caused the failure. Only the 12V linear lab power supply
> that  fed the heaters had developed a completely dead filter cap, so the
> feeding  voltage was a 100 Hz sawtooth rather than DC. The heaters appear
> to be  still working though. I'm only using the internal heater. They
> draw some  700 mA at power up and drop to around 150 mA  each when warm.
> So the  oscillators are unlikely to be cooked inside. Something must have
> happened  to the 10 MHz output, possibly an ESD issue?
>
> Before I start taking them  apart, can anyone tell what has happened or
> what to look for  first?
>
> Regards,
> Adrian
>
>
>
> _______________________________________________
> time-nuts  mailing list -- [email protected]
> To unsubscribe, go to
> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the  instructions there.
>
> _______________________________________________
> time-nuts mailing list -- [email protected]
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.
>





------------------------------

Message: 4
Date: Sat, 13 Oct 2012 20:58:17 -0400 (EDT)
From: [email protected]
To: [email protected]
Subject: Re: [time-nuts] HP 10811A failure
Message-ID: <[email protected]>
Content-Type: text/plain; charset="US-ASCII"

Hi Adrian

Glad to hear you're back on track and that all is now well

I can understand that very thin coax being quite  deceptive, it's certainly 
amongst the thinnest I've ever come across.

I've got a couple of double oven10811s bought from one of the usual Chinese 
Ebay sellers a few years ago as potential spares for my Z3801As and both 
have  one of the coax connectors chopped off, can't remember now whether it's 
the  10MHz output or the EFC input.
Either way, should they ever be needed in anger I'm keeping my fingers  
crossed that the leads can be swapped over, cos fitting another connector sure  
don't look to be a very user friendly option:-)

Regards

Nigel
GM8PZR


In a message dated 14/10/2012 00:27:23 GMT Daylight Time, [email protected]  
writes:

A very  good point!

I checked the output amplitude with a spectrum analyzer, a  power meter 
and a scope, the latter with a 50 ohm load to the input.
I  re-checked the power supply connections and can confirm they are the 
same  as before.
I used three different power supplies for the oscillators and  two for 
the heaters, still no difference.
But...

To make it  short, your comments helped to get me back on track. Both 
beauties escaped  unnecessary dismantling and are now working as they 
should.

Actually,  the problem was caused by a long-term misunderstanding. I was 
always  wondering why 'HP used simple stranded wire' and not coaxial 
cable on the  10 MHz output and EFC.
As long as I connected them on the bench, I replaced  the 'missing' 
ground connection with a short wired croc clamp between coax  and ground. 
Today I realized that the thin blue wire IS actually coax  cable, so I 
should have connected the coax shield to the BNC ground on my  newly 
built 2x 10811 enclosure. With the new wiring, the output ground had  
just become much more inductive, up to a point where the nominal source  
impedance of 50 ohms had increased to over 120 ohms, causing the  
amplitude loss of some 3 dB and capturing noise.

Btw. the Sprague  6800 uF / 40 V from my R&S NGA power supply has indeed 
died, but  independently of the osc. problem.

Adrian


[email protected]  schrieb:
> Perhaps a silly question, but as I get the impression that  both seem  to
> have failed simultaneously with the same fault I'm  just wondering  if 
you're
> sure it isn't what you're checking them  with that's  developed a problem?
>  
>  Regards
>  
> Nigel
> GM8PZR
>    
>  
> In a message dated 13/10/2012 13:52:46 GMT  Daylight Time, [email protected]
> writes:
>
> Hi   All,
>
> both of my double oven 10811A's have been running  flawlessly until  now
> when I noticed that the output power has  dropped by about 3 dB  (measures
> only 4...4.5 dBm at 50 Ohm  load), and the noise floor has gone  up by
> about 20 dB, while the  frequency is still spot on.
>
> I used  them for various  mesurements, so I can unfortunately not remember
> what  might have  caused the failure. Only the 12V linear lab power supply
> that   fed the heaters had developed a completely dead filter cap, so the
>  feeding  voltage was a 100 Hz sawtooth rather than DC. The heaters  
appear
> to be  still working though. I'm only using the internal  heater. They
> draw some  700 mA at power up and drop to around 150  mA  each when warm.
> So the  oscillators are unlikely to be  cooked inside. Something must have
> happened  to the 10 MHz  output, possibly an ESD issue?
>
> Before I start taking  them  apart, can anyone tell what has happened or
> what to look  for  first?
>
> Regards,
>  Adrian
>
>
>
>  _______________________________________________
> time-nuts   mailing list -- [email protected]
> To unsubscribe, go to
>  https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow  the  instructions there.
>
>  _______________________________________________
> time-nuts mailing list  -- [email protected]
> To unsubscribe, go to  
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the  instructions  there.
>



_______________________________________________
time-nuts  mailing list -- [email protected]
To unsubscribe, go to  
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the  instructions there.



------------------------------

Message: 5
Date: Sat, 13 Oct 2012 21:18:17 -0500
From: Edgardo Molina <[email protected]>
To: Discussion of precise time and frequency measurement
    <[email protected]>
Subject: [time-nuts] HP Z3805A com port and monitoring questions
Message-ID: <[email protected]>
Content-Type: text/plain;    charset=iso-8859-1

Dear Group,

Good evening. I bought two HP Z3805A units a few weeks ago. They were bare 
units without antenna or power supply. After struggling to find a suitable 24V 
3A noise free power supply for each of them, I mounted a couple of Symmetricom 
antennas and just fired them up. 

After around 25 minutes, the GPS Lock light lighted up. As the manual for this 
particular model is unobtainable in several searches through the Internet, I 
found that the readily available manual for the Z3801A would cover much of the 
Z3905A operation. I also got a software for monitoring the operation from BD 
Systems Inc. In Colorado. Now that I am sure both are operating and locked only 
by looking at the front panel lights, I would like to ask a couple of questions 
regarding their communication to the PC.

1. Is the DB-25 serial port carrying RS-232 signals? I see the port of the 
Z380A1 is different: RS-422.
2. If the port is plain RS-232 is it possible to use a Serial to USB adapter? I 
have heard about the inconsistencies that such an adapter could cause when used 
with a TBolt. Should I assume this is the same case with the Z3805A?
3. There was once a project for building a TBolt state monitor on a 2 line LCD 
screen. I own three of those monitors and they are just great.  Is there an 
equivalent for these HP GPS units?

Any suggestions for the operation of these units and your ever expert and kind 
advise is always welcome. You all have a nice weekend!

Kindest regards,



Edgardo Molina
Direcci?n IPTEL

www.iptel.net.mx

T : 55 55 55202444
M : 04455 20501854

Piensa en Bits SA de CV



Informaci?n anexa:




CONFIDENCIALIDAD DE INFORMACION

Este mensaje tiene car?cter confidencial. Si usted no es el destinarario de 
este mensaje, le suplicamos se lo notifique al remitente mediante un correo 
electr?nico y que borre el presente mensaje y sus anexos de su computadora sin 
retener una copia de los mismos. Queda estrictamente prohibido copiar este 
mensaje o hacer usode el para cualquier prop?sito o divulgar su en forma 
parcial o total su contenido. Gracias.


NON-DISCLOSURE OF INFORMATION

This email is strictly confidential and may also be privileged. If you are not 
the intended recipient please immediately advise the sender by replying to this 
e-mail and then deleting the message and its attachments from your computer 
without keeping a copy. It is strictly forbidden to copy it or use it for any 
purpose or disclose its contents to any third party. Thank you.







------------------------------

Message: 6
Date: Sun, 14 Oct 2012 04:26:02 +0200
From: Magnus Danielson <[email protected]>
To: [email protected]
Subject: Re: [time-nuts] PRS10 dying from old age...
Message-ID: <[email protected]>
Content-Type: text/plain; charset=ISO-8859-1; format=flowed

On 10/13/2012 11:19 PM, paul swed wrote:
> Poul
> If it is dying there should maybe be a lamp voltage you can check and see
> what it indicates. If you consider it dead then you can always checK to see
> is it something going on like low RF or maybe the RB truly is an issue. I
> have recovered some olf FRS by reheating the rb bulb and essentially
> getting the RB that collects on the bulb back into circulation.

What I did was that I heated it up with an air gun. First try, it 
depleted onto the glass, but then I realized that I had left the 
collection dot to the side, so it had to deplete on the colder glass, 
but when I turned the collection dot to the top of the lamp and then 
heated, as the hot atom raised to the top, they would deplete into the 
rubidium dot and the glass surface would be free of them. The later is 
important, because if they deplete on the glass, then they will absorb 
the wavelengths of rubidium and this is the energy you want to get out 
of the lamp.

I was fortunate as the R&S rubidium is built such that you can get the 
rubidium lamp out of the clock without even turning it off. The special 
tool you need is part of the assembly.

Looking at the graphs, you have had some intermittent operations with 
glimpses of operation. I wonder if this is the lamp. You can open it up 
and check the lamp, it may be part of your problem. I would look for 
contact problems, like bad solder joint or something.

Cheers,
Magnus



------------------------------

Message: 7
Date: Sun, 14 Oct 2012 15:30:42 +1000
From: David Hooke <[email protected]>
To: [email protected]
Subject: [time-nuts] Seistek pn3048
Message-ID: <[email protected]>
Content-Type: text/plain; charset=ISO-8859-1; format=flowed

Hi All,

Does anyone have a download link for the Seistek PN3048 phase noise 
software, or details about contacting the author?

I'd like to use it to configure my 11848A, although I don't know how far 
I'll get without having a 3561A (just an 8568B).

Cheers,

davidh



------------------------------

_______________________________________________
time-nuts mailing list
[email protected]
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts

End of time-nuts Digest, Vol 99, Issue 62
*****************************************
_______________________________________________
time-nuts mailing list -- [email protected]
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Reply via email to