Ed,
 
the calculation is the same, however the numbers are 100ps for 53132A  
versus 350ps, and I have not seen an average systemic offset being displayed on 
 
any of the 3x 53132A units I use, and I see one on the SR-620. That's  why 
I sent it into SRS for calibration, paid the $$$ and got it back with the  
same exact offset and a statement that it is operating within specifications 
so  no adjustment is necessary.
 
HP manages to show zero error on average, with the digits bouncing back and 
 forth. The SRS unit manages to show a hard frequency offset. If I remember 
 correctly the SR-620 even shows this offset with it's own reference 
connected to  the inputs, the HP does not.
 
bye,
Said
 
 
In a message dated 3/17/2013 11:26:16 Pacific Daylight Time,  
[email protected] writes:

Hi  Said,

That equation looks similar in form to the specs for any counter.  What 
are the comparable equations for the  53132A or the 5370(A or  B)?

Ed

_______________________________________________
time-nuts mailing list -- [email protected]
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Reply via email to