Re: UL 60950 preliminary inspection / design review

2008-11-07 Thread peter merguerian
Dear Mr Balasubramanian,
 
Looks like you need to try and use another National Recognized Test Laboratory
(NRTL). Have you tried TUV Rheinland - the North American office is NRTL
accredited by OSHA and Standards Council of Canada - Out of 19 countries in
Asia, TUV Rheinland has 8 offices in India with test laboratory capabilities
for world markets.
 
http://www.tuv.com/in/en/regional_offices.html
 
 
Best Regards,
 
Peter Merguerian
 


--- On Thu, 11/6/08, kbalasubraman...@scmmicro.co.in
kbalasubraman...@scmmicro.co.in wrote:


From: kbalasubraman...@scmmicro.co.in kbalasubraman...@scmmicro.co.in
Subject: UL 60950 preliminary inspection / design review
To: emc-p...@ieee.org
Cc: EMC-PSTC@LISTSERV.IEEE.ORG EMC-PSTC@LISTSERV.IEEE.ORG
Date: Thursday, November 6, 2008, 6:45 AM


Dear Experts,

  We submit most of products to UL for 60950 listing. Almost 95 % of
our products are UL listed. We find it very difficult to complete the
cycle. Many times we come across situations that the parts used are not
meeting UL spec. and a redesign is inevitable. This further complicates 
the
things and delay the cycle. We look for some consultants who can help 
us by
participating from the design review stage itself.
  The most sad part is many times we submit the product to UL for
preliminary inspection during the proto stage and UL (India office 
located
at Bangalore) gives pass report. Then we go for pilot production
(simultaneously we initiate procurement for mass production also) and
submit the pilot samples to the same UL office for final certification. 
At
this time UL is coming with new findings and ultimately product fails to
get UL listing. All the times I am sure that there is no difference 
between
the proto samples and pilot samples. Mostly some resistor values get
changed between proto and pilot.
  It is not clear to me whether it is the competence issue at UL 
India
or UL is greedy in money making. I look for advise from the experts of 
this
forum. I have been told that UL Taiwan is more proficient in 60950
standard. Is this true?

Sincerely  
SCM
Microsystems (India) Pvt. Ltd.
K. Balasubramanian  
Modules 0506,
0507  0508, D – Block,
Manager – Hardware 
South
Wing,
TIDEL Park,
Tel : +91 44 42931391
   #4, Rajiv
Gandhi Road,(formerly Canal Bank Road)
Fax :+91 44
22540029   
Taramani,
Chennai – 600113. India
E-mail:kbalasubraman...@scmmicro.co.in  Web :
http://scmmicro.com
‘This email and any attachment may contain confidential information and 
are
intended solely for the use of the named recipient(s). If you have 
received
this information in error, you are prohibited from reading, copying,
distributing and using the information. If you are not a named 
addressee or
otherwise an intended recipient you are requested to immediately notify 
the
sender and to delete this email and all attachments from your system.’

-

This message is from the IEEE Product Safety Engineering Society
emc-pstc discussion list.Website:  http://www.ieee-pses.org/

To post a message to the list, send your e-mail to emc-p...@ieee.org

Instructions:  http://listserv.ieee.org/request/user-guide.html

List rules: http://www.ieee-pses.org/listrules.html

For help, send mail to the list administrators:

 Scott Douglas   emcp...@ptcnh.net
 Mike Cantwell   mcantw...@ieee.org

For policy questions, send mail to:

 Jim Bacher: j.bac...@ieee.org
 David Heald:   dhe...@gmail.com

All emc-pstc postings are archived and searchable on the web at:

http://www.ieeecommunities.org/emc-pstc


-



This message is from the IEEE Product Safety Engineering Society emc-pstc
discussion list. Website: http://www.ieee-pses.org/ 

To post a message to the list, send your e-mail to emc-p...@ieee.org 

Instructions: http://listserv.ieee.org/request/user-guide.html 

List rules: http://www.ieee-pses.org/listrules.html 

For help, send mail to the list administrators:

Scott Douglas emcp...@ptcnh.net
Mike Cantwell mcantw...@ieee.org 

For policy questions, send mail to:

Jim Bacher: j.bac...@ieee.org

Re: 13.56 MHz RFID for India

2008-11-07 Thread peter merguerian
Not allowed but special permission may be granted by the WPC depending on the
power levels. Each case is considered separately.
 
Peter Merguerian

--- On Mon, 11/3/08, rehel...@mmm.com rehel...@mmm.com wrote:


From: rehel...@mmm.com rehel...@mmm.com
Subject: 13.56 MHz RFID for India
To: emc-p...@ieee.org
Date: Monday, November 3, 2008, 2:43 AM


I am aware that 865-867 MHz RFID is allowed in India and it does not
require a license. Does anyone know if 13.56 MHz is an allowable RFID
frequency for India?

Thank you,
Bob Heller
3M EMC Laboratory, 76-1-01
St. Paul, MN 55107-1208
Tel:  651- 778-6336
Fax:  651-778-6252

-

This message is from the IEEE Product Safety Engineering Society
emc-pstc discussion list.Website:  http://www.ieee-pses.org/

To post a message to the list, send your e-mail to emc-p...@ieee.org

Instructions:  http://listserv.ieee.org/request/user-guide.html

List rules: http://www.ieee-pses.org/listrules.html

For help, send mail to the list administrators:

 Scott Douglas   emcp...@ptcnh.net
 Mike Cantwell   mcantw...@ieee.org

For policy questions, send mail to:

 Jim Bacher: j.bac...@ieee.org
 David Heald:   dhe...@gmail.com

All emc-pstc postings are archived and searchable on the web at:

http://www.ieeecommunities.org/emc-pstc


-



This message is from the IEEE Product Safety Engineering Society emc-pstc
discussion list. Website: http://www.ieee-pses.org/ 

To post a message to the list, send your e-mail to emc-p...@ieee.org 

Instructions: http://listserv.ieee.org/request/user-guide.html 

List rules: http://www.ieee-pses.org/listrules.html 

For help, send mail to the list administrators:

Scott Douglas emcp...@ptcnh.net
Mike Cantwell mcantw...@ieee.org 

For policy questions, send mail to:

Jim Bacher: j.bac...@ieee.org
David Heald: dhe...@gmail.com 

All emc-pstc postings are archived and searchable on the web at:

http://www.ieeecommunities.org/emc-pstc 




RE: UL 60950 preliminary inspection / design review

2008-11-07 Thread Pete Perkins
K. Balasubramanian  PSNet,

It seems that because of the described trouble you are having with
your certification supplier you are making the process more difficult for
yourself by the way you approach the process.  

Working with the test house in steps opens the door for additional
review of the design for compliance.  

You have probably started this 2 step method because you weren't
sure if you understood the requirements and the standard well enough to get
the equipment to pass all at once.  

Most manufacturers that I know don't utilize a multiple pass
certification process.  

Once you have developed confidence (maybe you are there already) why
don't you just approach them once to get the certification you need and move
into production of the unit?  It would seem that you would minimize your
trouble and probably save a little money in the process.  

For each new product there are only a few new issues that arise and
you should be able to discuss those with a certification engineer that you
have worked with recently to get assurance that you are moving ahead in a
mutually agreeable way.  This will pave the way for a smooth certification
process.  

As with any business process with a supplier the relationship that
you develop with the folks with which you work regularly is important.  

:) br, Pete
 
Peter E Perkins, PE
Principal Product Safety Engineer
PO Box 23427
Tigard, ORe  97281-3427
 
503/452-1201 fone/fax
p.perk...@ieee.org
 


From: emc-p...@ieee.org [mailto:emc-p...@ieee.org] On Behalf Of
kbalasubraman...@scmmicro.co.in
Sent: Thursday, November 06, 2008 6:45 AM
To: emc-p...@ieee.org
Cc: EMC-PSTC@LISTSERV.IEEE.ORG
Subject: UL 60950 preliminary inspection / design review

Dear Experts,

  We submit most of products to UL for 60950 listing. Almost 95 % of
our products are UL listed. We find it very difficult to complete the
cycle. Many times we come across situations that the parts used are not
meeting UL spec. and a redesign is inevitable. This further complicates the
things and delay the cycle. We look for some consultants who can help us by
participating from the design review stage itself.
  The most sad part is many times we submit the product to UL for
preliminary inspection during the proto stage and UL (India office located
at Bangalore) gives pass report. Then we go for pilot production
(simultaneously we initiate procurement for mass production also) and
submit the pilot samples to the same UL office for final certification. At
this time UL is coming with new findings and ultimately product fails to
get UL listing. All the times I am sure that there is no difference between
the proto samples and pilot samples. Mostly some resistor values get
changed between proto and pilot.
  It is not clear to me whether it is the competence issue at UL India
or UL is greedy in money making. I look for advise from the experts of this
forum. I have been told that UL Taiwan is more proficient in 60950
standard. Is this true?

Sincerely   SCM
Microsystems (India) Pvt. Ltd.
K. Balasubramanian       Modules 0506,
0507  0508, D  Block,
Manager  Hardware      South Wing,
TIDEL Park,
Tel : +91 44 42931391    #4, Rajiv
Gandhi Road,(formerly Canal Bank Road)
Fax :+91 44 22540029    Taramani,
Chennai  600113. India
E-mail:kbalasubraman...@scmmicro.co.in  Web : http://scmmicro.com
This email and any attachment may contain confidential information and are
intended solely for the use of the named recipient(s). If you have received
this information in error, you are prohibited from reading, copying,
distributing and using the information. If you are not a named addressee or
otherwise an intended recipient you are requested to immediately notify the
sender and to delete this email and all attachments from your system.

-

This message is from the IEEE Product Safety Engineering Society
emc-pstc discussion list.Website:  http://www.ieee-pses.org/

To post a message to the list, send your e-mail to emc-p...@ieee.org

Instructions:  http://listserv.ieee.org/request/user-guide.html

List rules: http://www.ieee-pses.org/listrules.html

For help, send mail to the list administrators:

 Scott Douglas   emcp...@ptcnh.net
 Mike Cantwell   mcantw...@ieee.org

For policy questions, send mail to:

 Jim Bacher: j.bac...@ieee.org
 David Heald:   dhe...@gmail.com

All emc-pstc postings are archived and searchable on the web at:

http://www.ieeecommunities.org/emc-pstc

-

This message is from the IEEE Product Safety Engineering Society
emc-pstc discussion list.Website:  http://www.ieee-pses.org/

To post a message to the list, send your e-mail to emc-p...@ieee.org

Instructions:  

Re: Deviation of Performance Criteria - other question.

2008-11-07 Thread Piotr Galka

- Original Message - 
From: John Woodgate j...@jmwa.demon.co.uk

 In the early days of the old Directive, it was considered that educational 
 products did not have to be tested but that Article 4 applied: if 
 interference occurred it must be prevented.

 I don't see anything about this in a quick scan of the current Directive 
 or its guidelines, but I think it should be obvious that educational 
 products have to be treated as a special case. Maybe they are not 
 'apparatus' as defined in the Directive, as they are not 'finished 
 appliances'.

Thank you John.

In 2002 my conclusion about old Directive was that _ONLY_ products destined 
for EMC education can be not tested, as they must violate EMC to show EMC 
problems. All others should be tested (I don't know where from my conclusion 
comes). It is why I gave up those days.
I am mainly a designer. My target is to spend 90% time designing, and 10% 
reading all about EMC, LVD, ROHS,. and standards. Because of this I 
can't be well oriented in any of them.
I have read the ROHS/WEEE interpretation where 'finished appliances' were 
discussed based on the question if they have end user needed function. In my 
understanding education PCB has end user needed function - education.
It has not its special case, but it will newer have such case (it is not the 
module used as part of other apparatus).
My problem is if it is in the EMC scope.
The basestation for this PCB models has its case and I think is 'apparatus'. 
The generator I have described is in that basestation. But the 15cm 
unshielded wire is used only when PCB model is used.
The basestation will be well ESD protected.
Do I have to use this 15cm wire when testing basestation, or test it with 
model not needing generator, or needing 10Hz sinus, or without any model ?
Selecting test that way is certainly not the right way.
There are such systems on the market, and I'm sure use unshielded cables. 
What is the way I don't know.
May be 3MHz TTL at such wire is well under the emission levels, but I don't 
think so.

Where should I ask my questions to have the problem solved once for always?

Piotr Galka

-

This message is from the IEEE Product Safety Engineering Society
emc-pstc discussion list.Website:  http://www.ieee-pses.org/

To post a message to the list, send your e-mail to emc-p...@ieee.org

Instructions:  http://listserv.ieee.org/request/user-guide.html

List rules: http://www.ieee-pses.org/listrules.html

For help, send mail to the list administrators:

 Scott Douglas   emcp...@ptcnh.net
 Mike Cantwell   mcantw...@ieee.org

For policy questions, send mail to:

 Jim Bacher: j.bac...@ieee.org
 David Heald:   dhe...@gmail.com

All emc-pstc postings are archived and searchable on the web at:

http://www.ieeecommunities.org/emc-pstc





Re: Deviation of Performance Criteria - other question.

2008-11-07 Thread John Woodgate

In message EB68941A997545109E32136A716A995D@MmPc21, dated Fri, 7 Nov 
2008, Piotr Galka piotr.ga...@micromade.pl writes:


Where should I ask my questions to have the problem solved once for always?

You should take the problem to the Polish National Standards body and 
ask the relevant committee to submit your case to CENELEC TC210 for an 
official 'interpretation'. It may be that TC210 would also need to refer 
your case to the Commission's 'EMC Working Party'.

Some people associated with those bodies read the mailing list. Maybe 
further assistance can be offered. Maybe not.
-- 
OOO - Own Opinions Only. Try www.jmwa.demon.co.uk and www.isce.org.uk
Either we are causing global warming, in which case we may be able to stop it,
or natural variation is causing it, and we probably can't stop it. You choose!
John Woodgate, J M Woodgate and Associates, Rayleigh, Essex UK

-

This message is from the IEEE Product Safety Engineering Society
emc-pstc discussion list.Website:  http://www.ieee-pses.org/

To post a message to the list, send your e-mail to emc-p...@ieee.org

Instructions:  http://listserv.ieee.org/request/user-guide.html

List rules: http://www.ieee-pses.org/listrules.html

For help, send mail to the list administrators:

 Scott Douglas   emcp...@ptcnh.net
 Mike Cantwell   mcantw...@ieee.org

For policy questions, send mail to:

 Jim Bacher: j.bac...@ieee.org
 David Heald:   dhe...@gmail.com

All emc-pstc postings are archived and searchable on the web at:

http://www.ieeecommunities.org/emc-pstc





Re: Deviation of Performance Criteria - other question.

2008-11-07 Thread Piotr Galka

From: John Woodgate j...@jmwa.demon.co.uk
 
 You should take the problem to the Polish National Standards body and 
 ask the relevant committee to submit your case to CENELEC TC210 for an 
 official 'interpretation'. It may be that TC210 would also need to refer 
 your case to the Commission's 'EMC Working Party'.
 
Do you know that way takes few weeks/months/years ?

Piotr Galka

-

This message is from the IEEE Product Safety Engineering Society
emc-pstc discussion list.Website:  http://www.ieee-pses.org/

To post a message to the list, send your e-mail to emc-p...@ieee.org

Instructions:  http://listserv.ieee.org/request/user-guide.html

List rules: http://www.ieee-pses.org/listrules.html

For help, send mail to the list administrators:

 Scott Douglas   emcp...@ptcnh.net
 Mike Cantwell   mcantw...@ieee.org

For policy questions, send mail to:

 Jim Bacher: j.bac...@ieee.org
 David Heald:   dhe...@gmail.com

All emc-pstc postings are archived and searchable on the web at:

http://www.ieeecommunities.org/emc-pstc





Re: Deviation of Performance Criteria - other question.

2008-11-07 Thread John Woodgate

In message 074D7671D8D74339973279276937D262@MmPc21, dated Fri, 7 Nov 
2008, Piotr Galka piotr.ga...@micromade.pl writes:


Do you know that way takes few weeks/months/years ?

You asked 'how to have the problem solved once for always'. There is no 
other way than the one I described to do that. It may be that there is 
on record a previous case in which special treatment was seen to be 
necessary for products of the type you are concerned with, in which case 
the time-scale could be quite short.

You could try the 'EMC assessment and Notified Body report' route for 
demonstrating compliance with the Directive without applying standards, 
but it's uncertain and costly in many cases. It, too, can take time and 
it takes money as well. And you seem to think that the emissions would 
be excessive anyway, so it's doubtful that you would get a positive 
report from the Notified Body.
-- 
OOO - Own Opinions Only. Try www.jmwa.demon.co.uk and www.isce.org.uk
Either we are causing global warming, in which case we may be able to stop it,
or natural variation is causing it, and we probably can't stop it. You choose!
John Woodgate, J M Woodgate and Associates, Rayleigh, Essex UK

-

This message is from the IEEE Product Safety Engineering Society
emc-pstc discussion list.Website:  http://www.ieee-pses.org/

To post a message to the list, send your e-mail to emc-p...@ieee.org

Instructions:  http://listserv.ieee.org/request/user-guide.html

List rules: http://www.ieee-pses.org/listrules.html

For help, send mail to the list administrators:

 Scott Douglas   emcp...@ptcnh.net
 Mike Cantwell   mcantw...@ieee.org

For policy questions, send mail to:

 Jim Bacher: j.bac...@ieee.org
 David Heald:   dhe...@gmail.com

All emc-pstc postings are archived and searchable on the web at:

http://www.ieeecommunities.org/emc-pstc





Re: UL 60950 preliminary inspection / design review

2008-11-07 Thread Kbalasubramanian
Dear Peter,

  Thanks for the feedback. For one product we went through TUV and
found that for everything they follow UL standards. They demand UL file
number for all the critical parts. In that way we felt UL is the master and
leading in defining the standard hence continuing with UL.

Sincerely                                                          
 SCM
Microsystems (India) Pvt. Ltd.
K. Balasubramanian                   
                        Modules 0506,
0507  0508, D – Block,

Manager – Hardware      
                                    South 
Wing,

TIDEL Park,
Tel : +91 44 42931391                 
                       #4, Rajiv
Gandhi Road,(formerly Canal Bank Road)
Fax :+91 44 
22540029                                       
 Taramani,
Chennai – 600113. India

E-mail:kbalasubraman...@scmmicro.co.in          Web : 
http://scmmicro.com
‘This email and any attachment may contain confidential information and are
intended solely for the use of the named recipient(s). If you have received
this information in error, you are prohibited from reading, copying,
distributing and using the information. If you are not a named addressee or
otherwise an intended recipient you are requested to immediately notify the
sender and to delete this email and all attachments from your system.’


   
 peter merguerian  
 pmerguerian2001@ 
 yahoo.com To 
   emc-p...@ieee.org,  
 11/07/2008 11:03  kbalasubraman...@scmmicro.co.in 
 AM cc 
   EMC-PSTC@LISTSERV.IEEE.ORG
   EMC-PSTC@LISTSERV.IEEE.ORG
 Please respond to Subject 
 pmerguerian2001@y Re: UL 60950 preliminary inspection 
 ahoo.com  / design review 
   
   
   
   
   
   





 Dear Mr Balasubramanian,   

 Looks like you need to try and use another National Recognized Test
 Laboratory (NRTL). Have you tried TUV Rheinland - the North American   
 office is NRTL accredited by OSHA and Standards Council of Canada - Out of 
 19 countries in Asia, TUV Rheinland has 8 offices in India with test   
 laboratory capabilities for world markets. 

 http://www.tuv.com/in/en/regional_offices.html 


 Best Regards,  

 Peter Merguerian   



 --- On Thu, 11/6/08, kbalasubraman...@scmmicro.co.in   
 kbalasubraman...@scmmicro.co.in wrote:   
  From: kbalasubraman...@scmmicro.co.in kbalasubraman...@scmmicro.co.in   
  Subject: UL 60950 preliminary inspection / design review  
  To: emc-p...@ieee.org 
  Cc: EMC-PSTC@LISTSERV.IEEE.ORG EMC-PSTC@LISTSERV.IEEE.ORG 
  Date: Thursday, November 6, 2008, 6:45 AM 

  Dear Experts, 
 

Re: UL 60950 preliminary inspection / design review

2008-11-07 Thread John Woodgate

In message 
ofca9d98c8.f412f212-on652574fa.0040cae4-652574fa.0041c...@scmmicro.co.in
 , dated Fri, 7 Nov 2008, kbalasubraman...@scmmicro.co.in writes:


For one product we went through TUV and found that for everything they 
follow UL standards. They demand UL file number for all the critical 
parts.

A test-house isn't in a position to 'demand' anything. And it certainly 
has no authority to insist on certification of parts by one particular 
organization.
-- 
OOO - Own Opinions Only. Try www.jmwa.demon.co.uk and www.isce.org.uk
Either we are causing global warming, in which case we may be able to stop it,
or natural variation is causing it, and we probably can't stop it. You choose!
John Woodgate, J M Woodgate and Associates, Rayleigh, Essex UK

-

This message is from the IEEE Product Safety Engineering Society
emc-pstc discussion list.Website:  http://www.ieee-pses.org/

To post a message to the list, send your e-mail to emc-p...@ieee.org

Instructions:  http://listserv.ieee.org/request/user-guide.html

List rules: http://www.ieee-pses.org/listrules.html

For help, send mail to the list administrators:

 Scott Douglas   emcp...@ptcnh.net
 Mike Cantwell   mcantw...@ieee.org

For policy questions, send mail to:

 Jim Bacher: j.bac...@ieee.org
 David Heald:   dhe...@gmail.com

All emc-pstc postings are archived and searchable on the web at:

http://www.ieeecommunities.org/emc-pstc





FW: Deviation of Performance Criteria

2008-11-07 Thread John Harrington
 

Hi Group

 

While we all have our copies of EN 61326 close at hand I’d like to see what
your consensus of opinion is on the meaning of paragraph 6.2.102 of
EN61326-2-1.

 

‘ 6.2.102  Tests with continuously present electromagnetic phenomenon

 

No visual degradation of parameters of the EUT is allowed during application
of the test, except as specified by the manufacturer.”

 

The standard calls it an “addition” to the requirements of EN 61326-1
section 6.2 but does not indicate whether it replaces the tables there or
modifies the Criterion A requirements.

 

What do you think?

 

Thanks

 

John Harrington



From: emc-p...@ieee.org [mailto:emc-p...@ieee.org] On Behalf Of Kunde, Brian
Sent: Thursday, November 06, 2008 8:12 AM
To: EMC-PSTC@listserv.ieee.org
Subject: RE: Deviation of Performance Criteria

 

Thanks to all. Your comments were greatly appreciated and helpful.

The Other Brian

 



From: Martin, Charles R (GE EntSol, SensInsp) [mailto:charles.mar...@ge.com] 
Sent: Wednesday, November 05, 2008 6:25 PM
To: Kunde, Brian; EMC-PSTC@LISTSERV.IEEE.ORG
Subject: RE: Deviation of Performance Criteria

 

Hi  Brian,

 

I agree with you, I don't see any other statements.

 

Table 2 evaluation of immunity test results notes possible combinations of
device function, phenomenon and performance criteria. This has been removed in
EN 61326-1:2006 and the performance criteria has been added to the immunity
test requirements tables for each port and phenomenon.

 

Thanks,

 

Charlie Martin

 



From: emc-p...@ieee.org [mailto:emc-p...@ieee.org] On Behalf Of Kunde, Brian
Sent: Wednesday, November 05, 2008 2:24 PM
To: Ted Eckert; EMC-PSTC@LISTSERV.IEEE.ORG
Subject: RE: Deviation of Performance Criteria

As Doug mentioned in his email, the allowance of deviations appear to have
been removed in the 2006 version of the EN 61326-1 standard which I believe
becomes mandatory in February 2009.  Can anyone else confirm this or is there
some other statement in the 2006 version that I’m not seeing that allows the
deviation? 

 

Thanks,

The Other Brian

 



From: Ted Eckert [mailto:ted.eck...@microsoft.com] 
Sent: Wednesday, November 05, 2008 1:53 PM
To: Kunde, Brian; EMC-PSTC@LISTSERV.IEEE.ORG
Subject: RE: Deviation of Performance Criteria

 

As I read EN 61326, it allows the manufacturer to specify the performance
criteria for each test.  The standard only requires that equipment shall not
become dangerous or unsafe as a result of the application of the tests.  Table
2 is only an example of evaluation of immunity test results.  The note states
“…performance criteria B and/or C may be accepted provided that both the
specification and the test report highlight such deviation(s) for the relevant
combination(s) of function and test.”

 

The standard does require that the deviations be listed in the specifications
in addition to the test report.  If the published specifications do not
describe the deviations, the manufacturer may be considered in noncompliance
with the standard.  If they have declared compliance via the standard, this
could be an issue.  If the manufacturer has chosen not to use the standards
route but claims compliance with the essential requirements of the directive,
you may have an argument if the peripheral is not suitable for its intended
use because of the deviation.

 

I agree with David Spencer; you are the customer and the vendor needs to meet
your requirements if they want you to buy their products.

 

Ted Eckert

Compliance Engineer

Microsoft Corporation

ted.eck...@microsoft.com

 

The opinions expressed are my own and do not necessarily reflect those of my
employer.

 

From: Kunde, Brian [mailto:brian_ku...@lecotc.com] 
Sent: Wednesday, November 05, 2008 9:59 AM
To: EMC-PSTC@LISTSERV.IEEE.ORG
Subject: Deviation of Performance Criteria

 

Greetings Compliance Experts.

 

Something disturbing came across my desk today I thought I would get your
opinion on it.

 

On a CE marked peripheral we buy/sell as part of our laboratory equipment
system, failed several immunity tests when we tested it as part of our system.
 When we notified the peripheral's manufacturer of the problem (yes, I'm being
purposely vague) they said the failures were ok and sent us a Certificate of
Compliance by a very very well know compliance lab with the following
statement:

 

Snip

 

EMC Immunity:

EN 61326-1:1997/A1:1998/A2:2001 EMC requirements for Electrical equipment for
measurement, control, and laboratory use.

- General Use for the following test with deviation of performance criteria to
Criteria C instead of B.

EN 61000-4-2

EN 61000-4-3

EN 61000-4-4

 

Unsnip

 

 

The peripheral manufacture said the EMC test lab told them they can put the CE
marking on their product as long as they included the above deviation
statement in their documentation and DOC.  

 

Is 

Re: FW: Deviation of Performance Criteria

2008-11-07 Thread reheller
John, one of the problems with defining criteria is that you cannot
define criteria that will encompass all of the past, present, and future
equipment in the world. Only the manufacturer (keeping in mind his
customers and competitors) can definitively define what is or is not
acceptable.

Bob Heller
3M EMC Laboratory, 76-1-01
St. Paul, MN 55107-1208
Tel:  651- 778-6336
Fax:  651-778-6252




   
 John Harrington 
 jharrington@keit 
 hley.com  To 
 Sent by:  emc-p...@ieee.org 
 emc-p...@ieee.org  cc 
   
   Subject 
 11/07/2008 07:21  FW: Deviation of Performance
 AMCriteria
   
   
   
   
   
   





Hi Group

While we all have our copies of EN 61326 close at hand I’d like to see what
your consensus of opinion is on the meaning of paragraph 6.2.102 of
EN61326-2-1.

‘ 6.2.102  Tests with continuously present electromagnetic phenomenon

No visual degradation of parameters of the EUT is allowed during
application of the test, except as specified by the manufacturer.”

The standard calls it an “addition” to the requirements of EN 61326-1
section 6.2 but does not indicate whether it replaces the tables there or
modifies the Criterion A requirements.

What do you think?

Thanks

John Harrington

From: emc-p...@ieee.org [mailto:emc-p...@ieee.org] On Behalf Of Kunde,
Brian
Sent: Thursday, November 06, 2008 8:12 AM
To: EMC-PSTC@listserv.ieee.org
Subject: RE: Deviation of Performance Criteria

Thanks to all. Your comments were greatly appreciated and helpful.
The Other Brian


From: Martin, Charles R (GE EntSol, SensInsp)
[mailto:charles.mar...@ge.com]
Sent: Wednesday, November 05, 2008 6:25 PM
To: Kunde, Brian; EMC-PSTC@LISTSERV.IEEE.ORG
Subject: RE: Deviation of Performance Criteria

Hi  Brian,

I agree with you, I don't see any other statements.

Table 2 evaluation of immunity test results notes possible combinations
of device function, phenomenon and performance criteria. This has been
removed in EN 61326-1:2006 and the performance criteria has been added to
the immunity test requirements tables for each port and phenomenon.

Thanks,

Charlie Martin


From: emc-p...@ieee.org [mailto:emc-p...@ieee.org] On Behalf Of Kunde,
Brian
Sent: Wednesday, November 05, 2008 2:24 PM
To: Ted Eckert; EMC-PSTC@LISTSERV.IEEE.ORG
Subject: RE: Deviation of Performance Criteria
As Doug mentioned in his email, the allowance of deviations appear to have
been removed in the 2006 version of the EN 61326-1 standard which I believe
becomes mandatory in February 2009.  Can anyone else confirm this or is
there some other statement in the 2006 version that I’m not seeing that
allows the deviation?

Thanks,
The Other Brian


From: Ted Eckert [mailto:ted.eck...@microsoft.com]
Sent: Wednesday, November 05, 2008 1:53 PM
To: Kunde, Brian; EMC-PSTC@LISTSERV.IEEE.ORG
Subject: RE: Deviation of Performance Criteria

As I read EN 61326, it allows the manufacturer to specify the performance
criteria for each test.  The standard only requires that equipment shall
not become dangerous or unsafe as a result of the application of the tests.
Table 2 is only an example of evaluation of immunity test results.  The
note states “…performance criteria B and/or C may be accepted provided that

both the specification and the test report highlight such deviation(s) for
the relevant combination(s) of function and test.”

The standard does require that the deviations be listed in the
specifications in addition to the test report.  If the published
specifications do not describe the deviations, the manufacturer may be
considered in noncompliance with the standard.  If they have declared
compliance via the standard, this could be an issue.  If the manufacturer
has chosen not to use the standards route but claims compliance with the
essential requirements of the directive, you may have an argument if the
peripheral is not suitable for its intended use because of the deviation.

I agree with David Spencer; you 

Re: FW: Deviation of Performance Criteria

2008-11-07 Thread John Woodgate

In message F8D22487033A4C4490CE53694BB9D2D7@harrington09269, dated 
Fri, 7 Nov 2008, John Harrington jharring...@keithley.com writes:


Hi Group

 

While we all have our copies of EN 61326 close at hand I?d like to see 
what your consensus of opinion is on the meaning of paragraph 6.2.102 
of EN61326-2-1.

 

? 6.2.102  Tests with continuously present electromagnetic phenomenon

 

No visual degradation of parameters of the EUT is allowed during 
application of the test, except as specified by the manufacturer.?

 

The standard calls it an ?addition? to the requirements of EN 61326-1 
section 6.2 but does not indicate whether it replaces the tables there 
or modifies the Criterion A requirements.

It modifies the Criterion A requirements, for philosophical reasons, 
which, as very rarely happens (;-), are also practical reasons.

It must mean to apply to Criterion A, because that's the one most 
associated with continuously-present disturbances.

What does 'no degradation' mean? 0%, 1%, 0.001%, 10%?  It can only mean 
0%: zilch, nada, nowt! How can that possibly be tested? The 'except' 
text allows the manufacturer to specify a finite degradation that is not 
exceeded.
-- 
OOO - Own Opinions Only. Try www.jmwa.demon.co.uk and www.isce.org.uk
Either we are causing global warming, in which case we may be able to stop it,
or natural variation is causing it, and we probably can't stop it. You choose!
John Woodgate, J M Woodgate and Associates, Rayleigh, Essex UK

-

This message is from the IEEE Product Safety Engineering Society
emc-pstc discussion list.Website:  http://www.ieee-pses.org/

To post a message to the list, send your e-mail to emc-p...@ieee.org

Instructions:  http://listserv.ieee.org/request/user-guide.html

List rules: http://www.ieee-pses.org/listrules.html

For help, send mail to the list administrators:

 Scott Douglas   emcp...@ptcnh.net
 Mike Cantwell   mcantw...@ieee.org

For policy questions, send mail to:

 Jim Bacher: j.bac...@ieee.org
 David Heald:   dhe...@gmail.com

All emc-pstc postings are archived and searchable on the web at:

http://www.ieeecommunities.org/emc-pstc





Re: Deviation of Performance Criteria - other question.

2008-11-07 Thread Piotr Galka

From: John Woodgate j...@jmwa.demon.co.uk

 And you seem to think that the emissions would be excessive anyway, so 
 it's doubtful that you would get a positive report from the Notified Body.

My main problem is how to make evident in papers that it is CE OK if 
education circuits can be ESD damaged because I don't see any solution to 
that.

I should be able to limit the emission just making square being not square 
and persuade the pupil that it is really square what they see ;-)

Regards

Piotr Galka 

-

This message is from the IEEE Product Safety Engineering Society
emc-pstc discussion list.Website:  http://www.ieee-pses.org/

To post a message to the list, send your e-mail to emc-p...@ieee.org

Instructions:  http://listserv.ieee.org/request/user-guide.html

List rules: http://www.ieee-pses.org/listrules.html

For help, send mail to the list administrators:

 Scott Douglas   emcp...@ptcnh.net
 Mike Cantwell   mcantw...@ieee.org

For policy questions, send mail to:

 Jim Bacher: j.bac...@ieee.org
 David Heald:   dhe...@gmail.com

All emc-pstc postings are archived and searchable on the web at:

http://www.ieeecommunities.org/emc-pstc





Replacement of UK Statutory Instruments 1994 1768

2008-11-07 Thread Scott Xe
There was an consultation to replace SI 1994 1768 in 2006.  Is there any
progress in the replacement?

Thanks and regards,

Scott

-

This message is from the IEEE Product Safety Engineering Society
emc-pstc discussion list.Website:  http://www.ieee-pses.org/

To post a message to the list, send your e-mail to emc-p...@ieee.org

Instructions:  http://listserv.ieee.org/request/user-guide.html

List rules: http://www.ieee-pses.org/listrules.html

For help, send mail to the list administrators:

 Scott Douglas   emcp...@ptcnh.net
 Mike Cantwell   mcantw...@ieee.org

For policy questions, send mail to:

 Jim Bacher: j.bac...@ieee.org
 David Heald:   dhe...@gmail.com

All emc-pstc postings are archived and searchable on the web at:

http://www.ieeecommunities.org/emc-pstc




RE: FW: Deviation of Performance Criteria

2008-11-07 Thread John Harrington
Hi Bob

I get your point but criterion A comes with some guidelines at least.  I'd
like peoples' opinion on whether this clause widens, negates or has no
effect on those guidelines.

Thanks

John

John Harrington

From: rehel...@mmm.com [mailto:rehel...@mmm.com] 
Sent: Friday, November 07, 2008 8:41 AM
To: John Harrington
Cc: emc-p...@ieee.org
Subject: Re: FW: Deviation of Performance Criteria

John, one of the problems with defining criteria is that you cannot
define criteria that will encompass all of the past, present, and future
equipment in the world. Only the manufacturer (keeping in mind his
customers and competitors) can definitively define what is or is not
acceptable.

Bob Heller
3M EMC Laboratory, 76-1-01
St. Paul, MN 55107-1208
Tel:  651- 778-6336
Fax:  651-778-6252




   
 John Harrington 
 jharrington@keit 
 hley.com  To 
 Sent by:  emc-p...@ieee.org 
 emc-p...@ieee.org  cc 
   
   Subject 
 11/07/2008 07:21  FW: Deviation of Performance
 AMCriteria
   
   
   
   
   
   





Hi Group

While we all have our copies of EN 61326 close at hand I'd like to see what
your consensus of opinion is on the meaning of paragraph 6.2.102 of
EN61326-2-1.

' 6.2.102  Tests with continuously present electromagnetic phenomenon

No visual degradation of parameters of the EUT is allowed during
application of the test, except as specified by the manufacturer.

The standard calls it an addition to the requirements of EN 61326-1
section 6.2 but does not indicate whether it replaces the tables there or
modifies the Criterion A requirements.

What do you think?

Thanks

John Harrington

From: emc-p...@ieee.org [mailto:emc-p...@ieee.org] On Behalf Of Kunde,
Brian
Sent: Thursday, November 06, 2008 8:12 AM
To: EMC-PSTC@listserv.ieee.org
Subject: RE: Deviation of Performance Criteria

Thanks to all. Your comments were greatly appreciated and helpful.
The Other Brian


From: Martin, Charles R (GE EntSol, SensInsp)
[mailto:charles.mar...@ge.com]
Sent: Wednesday, November 05, 2008 6:25 PM
To: Kunde, Brian; EMC-PSTC@LISTSERV.IEEE.ORG
Subject: RE: Deviation of Performance Criteria

Hi  Brian,

I agree with you, I don't see any other statements.

Table 2 evaluation of immunity test results notes possible combinations
of device function, phenomenon and performance criteria. This has been
removed in EN 61326-1:2006 and the performance criteria has been added to
the immunity test requirements tables for each port and phenomenon.

Thanks,

Charlie Martin


From: emc-p...@ieee.org [mailto:emc-p...@ieee.org] On Behalf Of Kunde,
Brian
Sent: Wednesday, November 05, 2008 2:24 PM
To: Ted Eckert; EMC-PSTC@LISTSERV.IEEE.ORG
Subject: RE: Deviation of Performance Criteria
As Doug mentioned in his email, the allowance of deviations appear to have
been removed in the 2006 version of the EN 61326-1 standard which I believe
becomes mandatory in February 2009.  Can anyone else confirm this or is
there some other statement in the 2006 version that I'm not seeing that
allows the deviation?

Thanks,
The Other Brian


From: Ted Eckert [mailto:ted.eck...@microsoft.com]
Sent: Wednesday, November 05, 2008 1:53 PM
To: Kunde, Brian; EMC-PSTC@LISTSERV.IEEE.ORG
Subject: RE: Deviation of Performance Criteria

As I read EN 61326, it allows the manufacturer to specify the performance
criteria for each test.  The standard only requires that equipment shall
not become dangerous or unsafe as a result of the application of the tests.
Table 2 is only an example of evaluation of immunity test results.  The
note states .performance criteria B and/or C may be accepted provided that
both the specification and the test report highlight such deviation(s) for
the relevant combination(s) of function and test.

The standard does require that the deviations be listed in the
specifications in addition to the test report.  If the published
specifications do not describe the deviations, the manufacturer may be
considered in 

RE: UL 60950 preliminary inspection / design review

2008-11-07 Thread Kbalasubramanian
Dear Perkins,

  Thanks for the valuable feedback, certainly it is very useful in many
practical aspects. The condition prevailing in ITE industry today does not
allow slips in schedule. I am sure that not only SCM every one is waking in
a tight rope. If I tell my supply chain to start procurement after I
complete all certification they will laugh at me.

  I thank to everyone who replied to my query through the forum or to
my personal ID.

Sincerely                                                          
 SCM
Microsystems (India) Pvt. Ltd.
K. Balasubramanian                   
                        Modules 0506,
0507  0508, D – Block,

Manager – Hardware      
                                    South 
Wing,

TIDEL Park,
Tel : +91 44 42931391                 
                       #4, Rajiv
Gandhi Road,(formerly Canal Bank Road)
Fax :+91 44 
22540029                                       
 Taramani,
Chennai – 600113. India

E-mail:kbalasubraman...@scmmicro.co.in          Web : 
http://scmmicro.com
‘This email and any attachment may contain confidential information and are
intended solely for the use of the named recipient(s). If you have received
this information in error, you are prohibited from reading, copying,
distributing and using the information. If you are not a named addressee or
otherwise an intended recipient you are requested to immediately notify the
sender and to delete this email and all attachments from your system.’


   
 Pete Perkins
 peperkinspe@cs.c 
 omTo 
 Sent by:  kbalasubraman...@scmmicro.co.in,  
 emc-p...@ieee.org emc-p...@ieee.org 
cc 
   EMC-PSTC@LISTSERV.IEEE.ORG
 11/07/2008 01:51  Subject 
 PMRE: UL 60950 preliminary
   inspection / design review  
   
   
   
   
   
   




K. Balasubramanian  PSNet,

It seems that because of the described trouble you are having with
your certification supplier you are making the process more difficult for
yourself by the way you approach the process.

Working with the test house in steps opens the door for additional
review of the design for compliance.

You have probably started this 2 step method because you weren't
sure if you understood the requirements and the standard well enough to get
the equipment to pass all at once.

Most manufacturers that I know don't utilize a multiple pass
certification process.

Once you have developed confidence (maybe you are there already) why
don't you just approach them once to get the certification you need and
move
into production of the unit?  It would seem that you would minimize your
trouble and probably save a little money in the process.

For each new product there are only a few new issues that arise and
you should be able to discuss those with a certification engineer that you
have worked with recently to get assurance that you are moving ahead in a
mutually agreeable way.  This will pave the way for a smooth certification
process.

As with any business process with a supplier the relationship that
you develop with the folks with which you work regularly is important.

:) br, Pete

Peter E Perkins, PE
Principal Product Safety Engineer
PO Box 23427
Tigard, ORe  97281-3427

503/452-1201 fone/fax
p.perk...@ieee.org



From: emc-p...@ieee.org [mailto:emc-p...@ieee.org] On Behalf Of
kbalasubraman...@scmmicro.co.in
Sent: Thursday, November 06, 2008 6:45 AM
To: emc-p...@ieee.org
Cc: EMC-PSTC@LISTSERV.IEEE.ORG
Subject: UL 60950 preliminary inspection / design review

Dear Experts,

  We submit most of products to UL for 60950 listing. Almost 95 % of
our products are UL listed. We find it very difficult to complete the
cycle. Many times we come across situations that the parts used are not
meeting UL spec. and a redesign is 

UL flammability marking

2008-11-07 Thread Kim Boll Jensen
Hi all

I have been told that the UL marking on PCB (UR 94V-0 etc.) can also be
something like E4. Does anyone know if the marking requirements have been
changed and if so where I can get more information.

Best regards,


Mr. Kim Boll Jensen
Bolls Rådgivning
Ved Gadekæret 11F
DK-3660 Stenløse
Denmark

T: +45 48 18 35 66
F: +45 48 18 35 30
k...@bolls.dk
www.bolls.dk 

-

This message is from the IEEE Product Safety Engineering Society
emc-pstc discussion list.Website:  http://www.ieee-pses.org/

To post a message to the list, send your e-mail to emc-p...@ieee.org

Instructions:  http://listserv.ieee.org/request/user-guide.html

List rules: http://www.ieee-pses.org/listrules.html

For help, send mail to the list administrators:

 Scott Douglas   emcp...@ptcnh.net
 Mike Cantwell   mcantw...@ieee.org

For policy questions, send mail to:

 Jim Bacher: j.bac...@ieee.org
 David Heald:   dhe...@gmail.com

All emc-pstc postings are archived and searchable on the web at:

http://www.ieeecommunities.org/emc-pstc




RE: FW: Deviation of Performance Criteria

2008-11-07 Thread John Harrington
John


What does 'no degradation' mean? 0%, 1%, 0.001%, 10%?  It can only mean 
0%: zilch, nada, nowt! How can that possibly be tested? The 'except' 
text allows the manufacturer to specify a finite degradation that is not 
exceeded.

So can that finite degradation exceed the normal performance within
specification limits of the general criterion A degradation? 

John 

John Harrington


From: emc-p...@ieee.org [mailto:emc-p...@ieee.org] On Behalf Of John
Woodgate
Sent: Friday, November 07, 2008 8:46 AM
To: emc-p...@ieee.org
Subject: Re: FW: Deviation of Performance Criteria

In message F8D22487033A4C4490CE53694BB9D2D7@harrington09269, dated 
Fri, 7 Nov 2008, John Harrington jharring...@keithley.com writes:


Hi Group

 

While we all have our copies of EN 61326 close at hand I?d like to see 
what your consensus of opinion is on the meaning of paragraph 6.2.102 
of EN61326-2-1.

 

? 6.2.102  Tests with continuously present electromagnetic phenomenon

 

No visual degradation of parameters of the EUT is allowed during 
application of the test, except as specified by the manufacturer.?

 

The standard calls it an ?addition? to the requirements of EN 61326-1 
section 6.2 but does not indicate whether it replaces the tables there 
or modifies the Criterion A requirements.

It modifies the Criterion A requirements, for philosophical reasons, 
which, as very rarely happens (;-), are also practical reasons.

It must mean to apply to Criterion A, because that's the one most 
associated with continuously-present disturbances.

What does 'no degradation' mean? 0%, 1%, 0.001%, 10%?  It can only mean 
0%: zilch, nada, nowt! How can that possibly be tested? The 'except' 
text allows the manufacturer to specify a finite degradation that is not 
exceeded.
-- 
OOO - Own Opinions Only. Try www.jmwa.demon.co.uk and www.isce.org.uk
Either we are causing global warming, in which case we may be able to stop
it,
or natural variation is causing it, and we probably can't stop it. You
choose!
John Woodgate, J M Woodgate and Associates, Rayleigh, Essex UK

-

This message is from the IEEE Product Safety Engineering Society
emc-pstc discussion list.Website:  http://www.ieee-pses.org/

To post a message to the list, send your e-mail to emc-p...@ieee.org

Instructions:  http://listserv.ieee.org/request/user-guide.html

List rules: http://www.ieee-pses.org/listrules.html

For help, send mail to the list administrators:

 Scott Douglas   emcp...@ptcnh.net
 Mike Cantwell   mcantw...@ieee.org

For policy questions, send mail to:

 Jim Bacher: j.bac...@ieee.org
 David Heald:   dhe...@gmail.com

All emc-pstc postings are archived and searchable on the web at:

http://www.ieeecommunities.org/emc-pstc


_
Scanned by IBM Email Security Management Services powered by MessageLabs.
For more information please visit http://www.ers.ibm.com

_



_
Scanned by IBM Email Security Management Services powered by MessageLabs. For 
more information please visit http://www.ers.ibm.com
_

-

This message is from the IEEE Product Safety Engineering Society
emc-pstc discussion list.Website:  http://www.ieee-pses.org/

To post a message to the list, send your e-mail to emc-p...@ieee.org

Instructions:  http://listserv.ieee.org/request/user-guide.html

List rules: http://www.ieee-pses.org/listrules.html

For help, send mail to the list administrators:

 Scott Douglas   emcp...@ptcnh.net
 Mike Cantwell   mcantw...@ieee.org

For policy questions, send mail to:

 Jim Bacher: j.bac...@ieee.org
 David Heald:   dhe...@gmail.com

All emc-pstc postings are archived and searchable on the web at:

http://www.ieeecommunities.org/emc-pstc




RE: Replacement of UK Statutory Instruments 1994 1768

2008-11-07 Thread James, Chris
Scott - see current position para near bottom of BERR page below and related
documents on right side of page:

 

http://www.berr.gov.uk/whatwedo/sectors
sustainability/regulations/ecdirect/page12568.html

 

 

Chris 

 

 


From: emc-p...@ieee.org [mailto:emc-p...@ieee.org] On Behalf Of Scott Xe
Sent: 07 November 2008 14:01
To: emc-p...@ieee.org
Subject: Replacement of UK Statutory Instruments 1994 1768

 

There was an consultation to replace SI 1994 1768 in 2006.  Is there any

progress in the replacement?

 

Thanks and regards,

 

Scott

 

-



This message is from the IEEE Product Safety Engineering Society

emc-pstc discussion list.Website:  http://www.ieee-pses.org/

 

To post a message to the list, send your e-mail to emc-p...@ieee.org

 

Instructions:  http://listserv.ieee.org/request/user-guide.html

 

List rules: http://www.ieee-pses.org/listrules.html

 

For help, send mail to the list administrators:

 

 Scott Douglas   emcp...@ptcnh.net

 Mike Cantwell   mcantw...@ieee.org

 

For policy questions, send mail to:

 

 Jim Bacher: j.bac...@ieee.org

 David Heald:   dhe...@gmail.com

 

All emc-pstc postings are archived and searchable on the web at:

 

http://www.ieeecommunities.org/emc-pstc



This message (including any attachments) may contain confidential information
intended for a specific individual and purpose. If you are not the intended
recipient, delete this message. If you are not the intended recipient,
disclosing, copying, distributing, or taking any action based on this message
is strictly prohibited. 

-



This message is from the IEEE Product Safety Engineering Society emc-pstc
discussion list. Website: http://www.ieee-pses.org/ 

To post a message to the list, send your e-mail to emc-p...@ieee.org 

Instructions: http://listserv.ieee.org/request/user-guide.html 

List rules: http://www.ieee-pses.org/listrules.html 

For help, send mail to the list administrators:

Scott Douglas emcp...@ptcnh.net
Mike Cantwell mcantw...@ieee.org 

For policy questions, send mail to:

Jim Bacher: j.bac...@ieee.org
David Heald: dhe...@gmail.com 

All emc-pstc postings are archived and searchable on the web at:

http://www.ieeecommunities.org/emc-pstc 




Re: Deviation of Performance Criteria - other question.

2008-11-07 Thread John Woodgate

In message 8D6DAA21BC8A460FBCA6AE9A5229C9DA@MmPc21, dated Fri, 7 Nov 
2008, Piotr Galka piotr.ga...@micromade.pl writes:


My main problem is how to make evident in papers that it is CE OK if 
education circuits can be ESD damaged because I don't see any solution 
to that.

No possibility of prevention, indeed, so there MUST be a special case.

I should be able to limit the emission just making square being not 
square and persuade the pupil that it is really square what they see ;-)

Instead of work-arounds, there should be an EMC standard (or a pair, for 
emissions and immunity) for such educational products. However, neither 
IEC or ISO has a committee with responsibility for hardware (of any 
sort) for education and training. It's a very surprising omission; that 
doesn't help you, though.
-- 
OOO - Own Opinions Only. Try www.jmwa.demon.co.uk and www.isce.org.uk
Either we are causing global warming, in which case we may be able to stop it,
or natural variation is causing it, and we probably can't stop it. You choose!
John Woodgate, J M Woodgate and Associates, Rayleigh, Essex UK

-

This message is from the IEEE Product Safety Engineering Society
emc-pstc discussion list.Website:  http://www.ieee-pses.org/

To post a message to the list, send your e-mail to emc-p...@ieee.org

Instructions:  http://listserv.ieee.org/request/user-guide.html

List rules: http://www.ieee-pses.org/listrules.html

For help, send mail to the list administrators:

 Scott Douglas   emcp...@ptcnh.net
 Mike Cantwell   mcantw...@ieee.org

For policy questions, send mail to:

 Jim Bacher: j.bac...@ieee.org
 David Heald:   dhe...@gmail.com

All emc-pstc postings are archived and searchable on the web at:

http://www.ieeecommunities.org/emc-pstc





Re: FW: Deviation of Performance Criteria

2008-11-07 Thread John Woodgate

In message F766B5AD23EA4C37836CDA2A2379DA99@harrington09269, dated 
Fri, 7 Nov 2008, John Harrington jharring...@keithley.com writes:


So can that finite degradation exceed the normal performance within 
specification limits of the general criterion A degradation?

It depends what individual standards say: not all have exactly the same 
wording for Criterion A.

In any case, the manufacturer has to consider carefully how to word the 
statement. If the manufacturer uses those words then he must widen his 
spec limits to allow for the degradation, otherwise a product at the 
unfavourable performance limit would violate the criterion in the 
presence of the disturbance.  For example, 'Signal-to noise ratio better 
than 80 dB'. If that can be degraded by 3 dB in the presence of the 
disturbance, then the statement must say '77 dB', not '80 dB'.
-- 
OOO - Own Opinions Only. Try www.jmwa.demon.co.uk and www.isce.org.uk
Either we are causing global warming, in which case we may be able to stop it,
or natural variation is causing it, and we probably can't stop it. You choose!
John Woodgate, J M Woodgate and Associates, Rayleigh, Essex UK

-

This message is from the IEEE Product Safety Engineering Society
emc-pstc discussion list.Website:  http://www.ieee-pses.org/

To post a message to the list, send your e-mail to emc-p...@ieee.org

Instructions:  http://listserv.ieee.org/request/user-guide.html

List rules: http://www.ieee-pses.org/listrules.html

For help, send mail to the list administrators:

 Scott Douglas   emcp...@ptcnh.net
 Mike Cantwell   mcantw...@ieee.org

For policy questions, send mail to:

 Jim Bacher: j.bac...@ieee.org
 David Heald:   dhe...@gmail.com

All emc-pstc postings are archived and searchable on the web at:

http://www.ieeecommunities.org/emc-pstc





RE: FW: Deviation of Performance Criteria

2008-11-07 Thread John Harrington
So can that finite degradation exceed the normal performance within 
specification limits of the general criterion A degradation?

It depends what individual standards say: not all have exactly the same 
wording for Criterion A.

I think this is specific to EN 61326-1 and EN 61326-2-1

In any case, the manufacturer has to consider carefully how to word the 
statement. If the manufacturer uses those words then he must widen his 
spec limits to allow for the degradation, otherwise a product at the 
unfavourable performance limit would violate the criterion in the 
presence of the disturbance.  For example, 'Signal-to noise ratio better 
than 80 dB'. If that can be degraded by 3 dB in the presence of the 
disturbance, then the statement must say '77 dB', not '80 dB'.

But where is this statement?  Not on the DofC.  Not on the spec sheet.  May
be in the test report, may be on some justification document at the bottom
of the filing cabinet.  My point being that the manufacturer can now claim
compliance to criterion A regardless of actual performance and the customer
(like Brian's original issue) has even less opportunity to determine what
that means.

This appears to be worse than the previous version of the standard.


_
Scanned by IBM Email Security Management Services powered by MessageLabs.
For more information please visit http://www.ers.ibm.com

_



_
Scanned by IBM Email Security Management Services powered by MessageLabs. For 
more information please visit http://www.ers.ibm.com
_

-

This message is from the IEEE Product Safety Engineering Society
emc-pstc discussion list.Website:  http://www.ieee-pses.org/

To post a message to the list, send your e-mail to emc-p...@ieee.org

Instructions:  http://listserv.ieee.org/request/user-guide.html

List rules: http://www.ieee-pses.org/listrules.html

For help, send mail to the list administrators:

 Scott Douglas   emcp...@ptcnh.net
 Mike Cantwell   mcantw...@ieee.org

For policy questions, send mail to:

 Jim Bacher: j.bac...@ieee.org
 David Heald:   dhe...@gmail.com

All emc-pstc postings are archived and searchable on the web at:

http://www.ieeecommunities.org/emc-pstc




RE: UL 60950 preliminary inspection / design review

2008-11-07 Thread Brian O'Connell
Perhaps the test house was confusing the requirement for
components to be certified to a UL/CSA standard vs UL recognition
of the component.

from UL60950-1
1.5.1
...In this standard, certain IEC component standard requirements
are replaced by the relevant requirements of component standards
listed in Annex P.1.

P.1 UL and CSA Component Requirements (mandatory)
...The following components shall comply with the requirements
specified below. All IEC standard requirements in this standard
are either replaced or modified, as noted, by the relevant
requirements of either CSA or UL or both component standards as
listed in this annex.



From: emc-p...@ieee.org [mailto:emc-p...@ieee.org]On Behalf Of
John
Woodgate
Sent: Friday, November 07, 2008 5:18 AM
To: emc-p...@ieee.org
Subject: Re: UL 60950 preliminary inspection / design review

A test-house isn't in a position to 'demand' anything. And it
certainly
has no authority to insist on certification of parts by one
particular
organization.

-

This message is from the IEEE Product Safety Engineering Society
emc-pstc discussion list.Website:  http://www.ieee-pses.org/

To post a message to the list, send your e-mail to emc-p...@ieee.org

Instructions:  http://listserv.ieee.org/request/user-guide.html

List rules: http://www.ieee-pses.org/listrules.html

For help, send mail to the list administrators:

 Scott Douglas   emcp...@ptcnh.net
 Mike Cantwell   mcantw...@ieee.org

For policy questions, send mail to:

 Jim Bacher: j.bac...@ieee.org
 David Heald:   dhe...@gmail.com

All emc-pstc postings are archived and searchable on the web at:

http://www.ieeecommunities.org/emc-pstc




Re: Deviation of Performance Criteria - other question.

2008-11-07 Thread Cortland Richmond
IMO if you can show no problem with emissions and safety, ESD is not a big
problem.
1) Use less than 9 KHz for a clock.  
2) Note that the equipment is not complete (a breadboard isn't, after all)
and so not subject to the standards for completed equipment. 
3) Use a SELV power limited supply.


Cortland
KA5S

 [Original Message]
 From: Piotr Galka piotr.ga...@micromade.pl
 To: EMC-PSTC emc-p...@ieee.org; John Woodgate j...@jmwa.demon.co.uk
 Date: 11/7/2008 8:59:55 AM
 Subject: Re: Deviation of Performance Criteria - other question.

 From: John Woodgate j...@jmwa.demon.co.uk

  And you seem to think that the emissions would be excessive anyway, so 
  it's doubtful that you would get a positive report from the Notified
Body.

 My main problem is how to make evident in papers that it is CE OK if 
 education circuits can be ESD damaged because I don't see any solution to 
 that.

 I should be able to limit the emission just making square being not
square 
 and persuade the pupil that it is really square what they see ;-)

 Regards

 Piotr Galka 

-

This message is from the IEEE Product Safety Engineering Society
emc-pstc discussion list.Website:  http://www.ieee-pses.org/

To post a message to the list, send your e-mail to emc-p...@ieee.org

Instructions:  http://listserv.ieee.org/request/user-guide.html

List rules: http://www.ieee-pses.org/listrules.html

For help, send mail to the list administrators:

 Scott Douglas   emcp...@ptcnh.net
 Mike Cantwell   mcantw...@ieee.org

For policy questions, send mail to:

 Jim Bacher: j.bac...@ieee.org
 David Heald:   dhe...@gmail.com

All emc-pstc postings are archived and searchable on the web at:

http://www.ieeecommunities.org/emc-pstc




RE: UL flammability marking

2008-11-07 Thread Ted Eckert
Strictly speaking, the flame rating does not need to be marked on the circuit 
board.  The following is from the UL guide for category code ZPMV2 for Printed 
Wiring Boards.

Printed wiring boards Recognized under UL's Component Recognition Program are 
identified by significant markings consisting of the Recognized company's 
identification, the factory identification (if the printed wiring board is 
produced at more than one location), and the Recognized type designation that 
correspond with the marking specified in UL's published records. Only those 
components that actually bear the Marking shown in the individual 
Recognitions should be considered as being covered under the Component 
Recognition Program.

As noted, the board must have the following items.
1. Company Identification; this can be a name or logo.
2. Recognized Type Identification; this is normally a code issued by the 
circuit board manufacturer.
3. Factory identification, if made at more than one factory.

You will need to go to UL's on-line certification database.
http://database.ul.com/cgi-bin/XYV/template/LISEXT/1FRAME/index.htm
If you enter the category code ZPMV2, many pages of manufacturers will show up. 
 I will take the first listing as an example, ZPMV2.E46545 for A  C 
Electronics.  (They are only first alphabetically and I am in no way endorsing 
them as a vendor.)  At the bottom of the page is listed the required marking 
for this vendor.  It is their name or logo (and a picture of the logo is 
shown), type designation corresponding to the table displayed and factory 
marking.  The flammability rating can only be determined from the table in the 
file.  The vendor may choose to mark the rating on the board, but it isn't 
required by UL.

Ted Eckert
Compliance Engineer
Microsoft Corporation
ted.eck...@microsoft.com

The opinions expressed are my own and do not necessarily reflect those of my 
employer.





From: Kim Boll Jensen [mailto:k...@bolls.dk]
Sent: Friday, November 07, 2008 6:14 AM
To: EMC-PSTC@LISTSERV.IEEE.ORG
Subject: UL flammability marking

Hi all

I have been told that the UL marking on PCB (UR 94V-0 etc.) can also be
something like E4. Does anyone know if the marking requirements have been
changed and if so where I can get more information.

Best regards,


Mr. Kim Boll Jensen
Bolls Rådgivning
Ved Gadekæret 11F
DK-3660 Stenløse
Denmark

T: +45 48 18 35 66
F: +45 48 18 35 30
k...@bolls.dk
www.bolls.dk

-

This message is from the IEEE Product Safety Engineering Society
emc-pstc discussion list.Website:  http://www.ieee-pses.org/

To post a message to the list, send your e-mail to emc-p...@ieee.org

Instructions:  http://listserv.ieee.org/request/user-guide.html

List rules: http://www.ieee-pses.org/listrules.html

For help, send mail to the list administrators:

 Scott Douglas   emcp...@ptcnh.net
 Mike Cantwell   mcantw...@ieee.org

For policy questions, send mail to:

 Jim Bacher: j.bac...@ieee.org
 David Heald:   dhe...@gmail.com

All emc-pstc postings are archived and searchable on the web at:

http://www.ieeecommunities.org/emc-pstc

-

This message is from the IEEE Product Safety Engineering Society
emc-pstc discussion list.Website:  http://www.ieee-pses.org/

To post a message to the list, send your e-mail to emc-p...@ieee.org

Instructions:  http://listserv.ieee.org/request/user-guide.html

List rules: http://www.ieee-pses.org/listrules.html

For help, send mail to the list administrators:

 Scott Douglas   emcp...@ptcnh.net
 Mike Cantwell   mcantw...@ieee.org

For policy questions, send mail to:

 Jim Bacher: j.bac...@ieee.org
 David Heald:   dhe...@gmail.com

All emc-pstc postings are archived and searchable on the web at:

http://www.ieeecommunities.org/emc-pstc




Re: FW: Deviation of Performance Criteria

2008-11-07 Thread John Woodgate

In message F9E8844C84FF4B8CB92370D766C07003@harrington09269, dated 
Fri, 7 Nov 2008, John Harrington jharring...@keithley.com writes:


But where is this statement?

If you need it, and the manufacturer hasn't supplied it, you have to ask 
for it. Some standards (such as EN 55103) are more explicit about this 
than others.
-- 
OOO - Own Opinions Only. Try www.jmwa.demon.co.uk and www.isce.org.uk
Either we are causing global warming, in which case we may be able to stop it,
or natural variation is causing it, and we probably can't stop it. You choose!
John Woodgate, J M Woodgate and Associates, Rayleigh, Essex UK

-

This message is from the IEEE Product Safety Engineering Society
emc-pstc discussion list.Website:  http://www.ieee-pses.org/

To post a message to the list, send your e-mail to emc-p...@ieee.org

Instructions:  http://listserv.ieee.org/request/user-guide.html

List rules: http://www.ieee-pses.org/listrules.html

For help, send mail to the list administrators:

 Scott Douglas   emcp...@ptcnh.net
 Mike Cantwell   mcantw...@ieee.org

For policy questions, send mail to:

 Jim Bacher: j.bac...@ieee.org
 David Heald:   dhe...@gmail.com

All emc-pstc postings are archived and searchable on the web at:

http://www.ieeecommunities.org/emc-pstc





Re: Deviation of Performance Criteria - other question.

2008-11-07 Thread Piotr Galka

- Original Message - 
From: John Woodgate j...@jmwa.demon.co.uk

I should be able to limit the emission just making square being not square 
and persuade the pupil that it is really square what they see ;-)

 Instead of work-arounds, there should be an EMC standard (or a pair, for 
 emissions and immunity) for such educational products.

I don't expect such standards will allow for higher emission then others. 
There is possibility that in next classroom there is TV working.

Except described (method 1) I see other methods of emission limiting:
2) - not allowing the pupil to make connections as they wont, but making 
them permanent when test board is installed (limiting the board flexibility 
of use).
3) - using shielded cables (BNC seems to me too big, but for smaller 
shielded connectors I don't give long live at school environment).
As in 2 and 3 signal wires are not accessible you must add something to 
allow for oscilloscope probe connection.

At the moment method 1 seems to me the easiest so the best one.

Now I believe there is nothing I just missed in these subjects. Thanks John.
But still don't know how existing at market education systems manage with 
ESD problem.

Regards

Piotr Galka
P.S. Live was easy and pleasant before 2004 ;-) 

-

This message is from the IEEE Product Safety Engineering Society
emc-pstc discussion list.Website:  http://www.ieee-pses.org/

To post a message to the list, send your e-mail to emc-p...@ieee.org

Instructions:  http://listserv.ieee.org/request/user-guide.html

List rules: http://www.ieee-pses.org/listrules.html

For help, send mail to the list administrators:

 Scott Douglas   emcp...@ptcnh.net
 Mike Cantwell   mcantw...@ieee.org

For policy questions, send mail to:

 Jim Bacher: j.bac...@ieee.org
 David Heald:   dhe...@gmail.com

All emc-pstc postings are archived and searchable on the web at:

http://www.ieeecommunities.org/emc-pstc





RE: UL flammability marking

2008-11-07 Thread Brian O'Connell
PCB flammability rating marks are specifically required when

...found that flammability classifications of different grades
of base
materials (previously identified by one printed-wiring type
designation) are different...

I have never seen an E4 flammability rating specified in UL94
or UL796.

PCB marking requirements are found in UL796, clause 31, the CCN
guide for 'ZPMV2', and the individual FUS file. And they are not
always consistent.



From: emc-p...@ieee.org [mailto:emc-p...@ieee.org]On Behalf Of
Ted
Eckert
Sent: Friday, November 07, 2008 7:29 AM
To: Kim Boll Jensen; EMC-PSTC@LISTSERV.IEEE.ORG
Subject: RE: UL flammability marking

Strictly speaking, the flame rating does not need to be marked on
the circuit board.  The following is from the UL guide for
category code ZPMV2 for Printed Wiring Boards.

Printed wiring boards Recognized under UL's Component
Recognition Program are identified by significant markings
consisting of the Recognized company's identification, the
factory identification (if the printed wiring board is produced
at more than one location), and the Recognized type designation
that correspond with the marking specified in UL's published
records. Only those components that actually bear the Marking
shown in the individual Recognitions should be considered as
being covered under the Component Recognition Program.

As noted, the board must have the following items.
1. Company Identification; this can be a name or logo.
2. Recognized Type Identification; this is normally a code issued
by the circuit board manufacturer.
3. Factory identification, if made at more than one factory.

You will need to go to UL's on-line certification database.
http://database.ul.com/cgi-bin/XYV/template/LISEXT/1FRAME/index.h
tm
If you enter the category code ZPMV2, many pages of manufacturers
will show up.  I will take the first listing as an example,
ZPMV2.E46545 for A  C Electronics.  (They are only first
alphabetically and I am in no way endorsing them as a vendor.)
At the bottom of the page is listed the required marking for this
vendor.  It is their name or logo (and a picture of the logo is
shown), type designation corresponding to the table displayed and
factory marking.  The flammability rating can only be determined
from the table in the file.  The vendor may choose to mark the
rating on the board, but it isn't required by UL.

Ted Eckert
Compliance Engineer
Microsoft Corporation
ted.eck...@microsoft.com


From: Kim Boll Jensen [mailto:k...@bolls.dk]
Sent: Friday, November 07, 2008 6:14 AM
To: EMC-PSTC@LISTSERV.IEEE.ORG
Subject: UL flammability marking

Hi all

I have been told that the UL marking on PCB (UR 94V-0 etc.) can
also be
something like E4. Does anyone know if the marking requirements
have been
changed and if so where I can get more information.

Best regards,

Mr. Kim Boll Jensen
Bolls Rådgivning
Ved Gadekæret 11F
DK-3660 Stenløse
Denmark

-

This message is from the IEEE Product Safety Engineering Society
emc-pstc discussion list.Website:  http://www.ieee-pses.org/

To post a message to the list, send your e-mail to emc-p...@ieee.org

Instructions:  http://listserv.ieee.org/request/user-guide.html

List rules: http://www.ieee-pses.org/listrules.html

For help, send mail to the list administrators:

 Scott Douglas   emcp...@ptcnh.net
 Mike Cantwell   mcantw...@ieee.org

For policy questions, send mail to:

 Jim Bacher: j.bac...@ieee.org
 David Heald:   dhe...@gmail.com

All emc-pstc postings are archived and searchable on the web at:

http://www.ieeecommunities.org/emc-pstc




Re: Deviation of Performance Criteria - other question.

2008-11-07 Thread Piotr Galka

- Original Message - 
From: Cortland Richmond k...@earthlink.net


 1) Use less than 9 KHz for a clock.
1MHz generator is standard in such systems.
AM and FM receivers are a kind of basic electronic in my opinion so I'd like 
to allow for taking frequency characteristics of 465kHz and 10.7MHz filters, 
and see AM and FM demodulators working.

 2) Note that the equipment is not complete (a breadboard isn't, after all)
 and so not subject to the standards for completed equipment.

It is used by end user and there is nothing planned to be added to it so why 
it is not complete?
Typical breadboard isn't complete as it is destined to be installed in some 
case.

 3) Use a SELV power limited supply.
Right.

Regards

Piotr Galka 

-

This message is from the IEEE Product Safety Engineering Society
emc-pstc discussion list.Website:  http://www.ieee-pses.org/

To post a message to the list, send your e-mail to emc-p...@ieee.org

Instructions:  http://listserv.ieee.org/request/user-guide.html

List rules: http://www.ieee-pses.org/listrules.html

For help, send mail to the list administrators:

 Scott Douglas   emcp...@ptcnh.net
 Mike Cantwell   mcantw...@ieee.org

For policy questions, send mail to:

 Jim Bacher: j.bac...@ieee.org
 David Heald:   dhe...@gmail.com

All emc-pstc postings are archived and searchable on the web at:

http://www.ieeecommunities.org/emc-pstc





RE: UL flammability marking

2008-11-07 Thread Pete Perkins
KBJ  PSNet,

As you probably know the requirements for flammability of PWBs come
down thru the chain of UL 94, UL 746 and UL 796.  There are IEC equivalents
for these requirements, but I don't have all the standard numbers at hand.  

To show conformity to requirements markings on the board are
needed/required.  Traceability is easily done once the manufacturer's mark
and type number are shown since all the PWB characteristics are listed in
the UL Component database under this mfgr  type.  

The inclusion of the UL V-X rating is superfluous once the mfgr 
type are identified.  I never look for the V rating on the board; I'm
examining the UL database info for proper ratings.  There are other
parameters which are of interest in meeting the needed requirements for the
product standard which are given there.  

So, your reference to 'E4' may be the manufacturers' type number
from which you can determine the V-rating.  

:) br, Pete
 
Peter E Perkins, PE
Principal Product Safety Engineer
PO Box 23427
Tigard, ORe  97281-3427
 
503/452-1201 fone/fax
p.perk...@ieee.org
 


From: emc-p...@ieee.org [mailto:emc-p...@ieee.org] On Behalf Of Kim Boll
Jensen
Sent: Friday, November 07, 2008 6:14 AM
To: EMC PSTC
Subject: UL flammability marking

Hi all

I have been told that the UL marking on PCB (UR 94V-0 etc.) can also be
something like E4. Does anyone know if the marking requirements have been
changed and if so where I can get more information.

Best regards,


Mr. Kim Boll Jensen
Bolls Rådgivning
Ved Gadekæret 11F
DK-3660 Stenløse
Denmark

T: +45 48 18 35 66
F: +45 48 18 35 30
k...@bolls.dk
www.bolls.dk 

-

This message is from the IEEE Product Safety Engineering Society
emc-pstc discussion list.Website:  http://www.ieee-pses.org/

To post a message to the list, send your e-mail to emc-p...@ieee.org

Instructions:  http://listserv.ieee.org/request/user-guide.html

List rules: http://www.ieee-pses.org/listrules.html

For help, send mail to the list administrators:

 Scott Douglas   emcp...@ptcnh.net
 Mike Cantwell   mcantw...@ieee.org

For policy questions, send mail to:

 Jim Bacher: j.bac...@ieee.org
 David Heald:   dhe...@gmail.com

All emc-pstc postings are archived and searchable on the web at:

http://www.ieeecommunities.org/emc-pstc

-

This message is from the IEEE Product Safety Engineering Society
emc-pstc discussion list.Website:  http://www.ieee-pses.org/

To post a message to the list, send your e-mail to emc-p...@ieee.org

Instructions:  http://listserv.ieee.org/request/user-guide.html

List rules: http://www.ieee-pses.org/listrules.html

For help, send mail to the list administrators:

 Scott Douglas   emcp...@ptcnh.net
 Mike Cantwell   mcantw...@ieee.org

For policy questions, send mail to:

 Jim Bacher: j.bac...@ieee.org
 David Heald:   dhe...@gmail.com

All emc-pstc postings are archived and searchable on the web at:

http://www.ieeecommunities.org/emc-pstc




RE: Deviation of Performance Criteria - other question.

2008-11-07 Thread Sterner, David (NY80)
Piotr

Under EMC directive 2004/108/EC, apparatus means any finished
appliance or combination thereof made commercially available as a single
functional unit, intended for the end user and liable to generate
electromagnetic disturbance, or the performance of which is liable to be
affected by such disturbance.

As your classromm lab equipment does not seem to fit this definition,
some provisions of the directive are therefore non-applicable.  Marking
is another issue.

David


From: emc-p...@ieee.org [mailto:emc-p...@ieee.org] On Behalf Of Piotr
Galka
Sent: Friday, November 07, 2008 10:44 AM
To: EMC-PSTC; John Woodgate
Subject: Re: Deviation of Performance Criteria - other question.


From: John Woodgate j...@jmwa.demon.co.uk

I should be able to limit the emission just making square being not 
square and persuade the pupil that it is really square what they see 
;-)

 Instead of work-arounds, there should be an EMC standard (or a pair, 
 for emissions and immunity) for such educational products.

I don't expect such standards will allow for higher emission then
others. 
There is possibility that in next classroom there is TV working.

Except described (method 1) I see other methods of emission limiting:
2) - not allowing the pupil to make connections as they wont, but making
them permanent when test board is installed (limiting the board
flexibility of use).
3) - using shielded cables (BNC seems to me too big, but for smaller
shielded connectors I don't give long live at school environment).
As in 2 and 3 signal wires are not accessible you must add something to
allow for oscilloscope probe connection.

At the moment method 1 seems to me the easiest so the best one.

Now I believe there is nothing I just missed in these subjects. Thanks
John.
But still don't know how existing at market education systems manage
with ESD problem.

Regards

Piotr Galka
P.S. Live was easy and pleasant before 2004 ;-) 

-

This message is from the IEEE Product Safety Engineering Society
emc-pstc discussion list.Website:  http://www.ieee-pses.org/

To post a message to the list, send your e-mail to emc-p...@ieee.org

Instructions:  http://listserv.ieee.org/request/user-guide.html

List rules: http://www.ieee-pses.org/listrules.html

For help, send mail to the list administrators:

 Scott Douglas   emcp...@ptcnh.net
 Mike Cantwell   mcantw...@ieee.org

For policy questions, send mail to:

 Jim Bacher: j.bac...@ieee.org
 David Heald:   dhe...@gmail.com

All emc-pstc postings are archived and searchable on the web at:

http://www.ieeecommunities.org/emc-pstc

-

This message is from the IEEE Product Safety Engineering Society
emc-pstc discussion list.Website:  http://www.ieee-pses.org/

To post a message to the list, send your e-mail to emc-p...@ieee.org

Instructions:  http://listserv.ieee.org/request/user-guide.html

List rules: http://www.ieee-pses.org/listrules.html

For help, send mail to the list administrators:

 Scott Douglas   emcp...@ptcnh.net
 Mike Cantwell   mcantw...@ieee.org

For policy questions, send mail to:

 Jim Bacher: j.bac...@ieee.org
 David Heald:   dhe...@gmail.com

All emc-pstc postings are archived and searchable on the web at:

http://www.ieeecommunities.org/emc-pstc




Re: Deviation of Performance Criteria - other question.

2008-11-07 Thread Cortland Richmond
Piotr Galka wrote:
 1MHz generator is standard in such systems.
 AM and FM receivers are a kind of basic electronic in my opinion so I'd
like 
 to allow for taking frequency characteristics of 465kHz and 10.7MHz
filters, 
 and see AM and FM demodulators working.

Ah, NOW I see your dilemma.

For digital systems, a lower frequency clock avoids EMI regulations. 

For the other, is there any reason why the radio must be sold and qualified
disassembled?   One doesn't have to meet the Automotive Directive to teach
mechanics how to rebuild an engine.  In the past, teaching electronics in
military schools, we would use an AM or FM reciever built on a sheet of
plywood. I agree this is not what one should sell the general public for a
kitchen radio, but it appears the bureacrats never considered that
education involves more than looking at schematic diagrams! Still, one
might get around their short-sighted blunder by building a radio set that,
*when assembled in its cabinet*, does comply with the regulations and CAN
be placed on the market but which, when opened up, may be spread out on a
workbench for students to learn from.

If connecting to an oscillator circuit with an oscilloscope violates the
EMC directive, we may all be out of business.



Cortland
KA5S

-

This message is from the IEEE Product Safety Engineering Society
emc-pstc discussion list.Website:  http://www.ieee-pses.org/

To post a message to the list, send your e-mail to emc-p...@ieee.org

Instructions:  http://listserv.ieee.org/request/user-guide.html

List rules: http://www.ieee-pses.org/listrules.html

For help, send mail to the list administrators:

 Scott Douglas   emcp...@ptcnh.net
 Mike Cantwell   mcantw...@ieee.org

For policy questions, send mail to:

 Jim Bacher: j.bac...@ieee.org
 David Heald:   dhe...@gmail.com

All emc-pstc postings are archived and searchable on the web at:

http://www.ieeecommunities.org/emc-pstc




Re: Deviation of Performance Criteria - other question.

2008-11-07 Thread John Woodgate

In message 380-220081157185821...@earthlink.net, dated Fri, 7 Nov 
2008, Cortland Richmond k...@earthlink.net writes:


If connecting to an oscillator circuit with an oscilloscope violates 
the EMC directive, we may all be out of business.

In the early days of the 89 Directive, people did ask such questions. 
They were quite convinced that the whole thing was insane.

It was later shown that the insanity primarily resided with some 
alarmist 'interpreters'. But some of the terminology used by officials 
was bizarre.
-- 
OOO - Own Opinions Only. Try www.jmwa.demon.co.uk and www.isce.org.uk
Either we are causing global warming, in which case we may be able to stop it,
or natural variation is causing it, and we probably can't stop it. You choose!
John Woodgate, J M Woodgate and Associates, Rayleigh, Essex UK

-

This message is from the IEEE Product Safety Engineering Society
emc-pstc discussion list.Website:  http://www.ieee-pses.org/

To post a message to the list, send your e-mail to emc-p...@ieee.org

Instructions:  http://listserv.ieee.org/request/user-guide.html

List rules: http://www.ieee-pses.org/listrules.html

For help, send mail to the list administrators:

 Scott Douglas   emcp...@ptcnh.net
 Mike Cantwell   mcantw...@ieee.org

For policy questions, send mail to:

 Jim Bacher: j.bac...@ieee.org
 David Heald:   dhe...@gmail.com

All emc-pstc postings are archived and searchable on the web at:

http://www.ieeecommunities.org/emc-pstc





Conversion between A/m and Gauss

2008-11-07 Thread Ravinder.Ajmani

Hi Experts, 

I know it is Friday afternoon, but I was wondering if anyone can help me
convert Gauss in to Amp/meter RMS.  I understand these two represent different
magnetic properties, but I have been testing our products in Gauss, and EN
61000-4-8 requires testing in Amp/meter. 

Thanks for the help. 

Regards, 
Ravinder Ajmani
Server PCB Development
Hitachi Global Storage Technologies
Voice: (408) 717-7956 Fax: (408) 717-9044

Email: ravinder.ajm...@hitachigst.com 

-



This message is from the IEEE Product Safety Engineering Society emc-pstc
discussion list. Website: http://www.ieee-pses.org/ 

To post a message to the list, send your e-mail to emc-p...@ieee.org 

Instructions: http://listserv.ieee.org/request/user-guide.html 

List rules: http://www.ieee-pses.org/listrules.html 

For help, send mail to the list administrators:

Scott Douglas emcp...@ptcnh.net
Mike Cantwell mcantw...@ieee.org 

For policy questions, send mail to:

Jim Bacher: j.bac...@ieee.org
David Heald: dhe...@gmail.com 

All emc-pstc postings are archived and searchable on the web at:

http://www.ieeecommunities.org/emc-pstc 




RE: Conversion between A/m and Gauss

2008-11-07 Thread John McAuley
Ravinder

 

1 Gauss = 1000 mGauss = 100 microTesla = 79.6 A/m

 

John McAuley
www.cei.ie
john.mcau...@cei.ie mailto:john.mcau...@cei.ie 

* 
DISCLAIMER:   The information contained in this e-mail may be confidential and
is intended solely for the use of the named addressee.  Access, copying or
re-use of the e-mail or any information contained therein by any other person
is not authorised.  If you are not the intended recipient please notify us
immediately by returning the e-mail to the originator 



 



From: emc-p...@ieee.org [mailto:emc-p...@ieee.org] On Behalf Of
ravinder.ajm...@hitachigst.com
Sent: 07 November 2008 23:04
To: emc-p...@ieee.org
Subject: Conversion between A/m and Gauss

 


Hi Experts, 

I know it is Friday afternoon, but I was wondering if anyone can help me
convert Gauss in to Amp/meter RMS.  I understand these two represent different
magnetic properties, but I have been testing our products in Gauss, and EN
61000-4-8 requires testing in Amp/meter. 

Thanks for the help. 

Regards, 
Ravinder Ajmani
Server PCB Development
Hitachi Global Storage Technologies
Voice: (408) 717-7956 Fax: (408) 717-9044

Email: ravinder.ajm...@hitachigst.com 

-


This message is from the IEEE Product Safety Engineering Society emc-pstc
discussion list. Website: http://www.ieee-pses.org/ 

To post a message to the list, send your e-mail to emc-p...@ieee.org 

Instructions: http://listserv.ieee.org/request/user-guide.html 

List rules: http://www.ieee-pses.org/listrules.html 

For help, send mail to the list administrators:

Scott Douglas emcp...@ptcnh.net
Mike Cantwell mcantw...@ieee.org 

For policy questions, send mail to:

Jim Bacher: j.bac...@ieee.org
David Heald: dhe...@gmail.com 

All emc-pstc postings are archived and searchable on the web at:

http://www.ieeecommunities.org/emc-pstc 

-



This message is from the IEEE Product Safety Engineering Society emc-pstc
discussion list. Website: http://www.ieee-pses.org/ 

To post a message to the list, send your e-mail to emc-p...@ieee.org 

Instructions: http://listserv.ieee.org/request/user-guide.html 

List rules: http://www.ieee-pses.org/listrules.html 

For help, send mail to the list administrators:

Scott Douglas emcp...@ptcnh.net
Mike Cantwell mcantw...@ieee.org 

For policy questions, send mail to:

Jim Bacher: j.bac...@ieee.org
David Heald: dhe...@gmail.com 

All emc-pstc postings are archived and searchable on the web at:

http://www.ieeecommunities.org/emc-pstc 




RE: Conversion between A/m and Gauss

2008-11-07 Thread rnute
 
 
Hi Ravinder:
 
 
Can't do it.  As you said, they are different magnetic
properties.
 
Gauss is a measure of magnetic flux density.
 
Amperes/meter is a measure of magnetic field strength.
 
 
Best regards,
Rich
 
 

-Original Message-
From: emc-p...@ieee.org [mailto:emc-p...@ieee.org] On Behalf Of
ravinder.ajm...@hitachigst.com
Sent: Friday, November 07, 2008 3:04 PM
To: emc-p...@ieee.org
Subject: Conversion between A/m and Gauss



Hi Experts, 

I know it is Friday afternoon, but I was wondering if anyone can help me
convert Gauss in to Amp/meter RMS.  I understand these two represent different
magnetic properties, but I have been testing our products in Gauss, and EN
61000-4-8 requires testing in Amp/meter. 

Thanks for the help. 

Regards, 
Ravinder Ajmani
Server PCB Development
Hitachi Global Storage Technologies
Voice: (408) 717-7956 Fax: (408) 717-9044

Email: ravinder.ajm...@hitachigst.com 

-



This message is from the IEEE Product Safety Engineering Society emc-pstc
discussion list. Website: http://www.ieee-pses.org/ 

To post a message to the list, send your e-mail to emc-p...@ieee.org 

Instructions: http://listserv.ieee.org/request/user-guide.html 

List rules: http://www.ieee-pses.org/listrules.html 

For help, send mail to the list administrators:

Scott Douglas emcp...@ptcnh.net
Mike Cantwell mcantw...@ieee.org 

For policy questions, send mail to:

Jim Bacher: j.bac...@ieee.org
David Heald: dhe...@gmail.com 

All emc-pstc postings are archived and searchable on the web at:

http://www.ieeecommunities.org/emc-pstc 




RE: UL flammability marking

2008-11-07 Thread rnute
Hello Kim Boll:


 I have been told that the UL marking on PCB (UR 94V-0 etc.) 
 can also be something like E4. Does anyone know if the 
 marking requirements have been changed and if so where I can 
 get more information.

UL 796 specifies marking requirements for PCBs.  Here is
a quote from UL 796:

27.6 A printed-wiring board that is identified by an 
individual type designation for each different base
material is not required to have an additional marking 
to identify its flammability rating.

I believe this answers your question.  If you need more
information, I suggest you obtain a copy of UL 796.

If you need to confirm or determine the flammability rating 
of a manufacturer's type designation, you will need to look
it up in UL's Component Directory.


Best regards,
Rich

-

This message is from the IEEE Product Safety Engineering Society
emc-pstc discussion list.Website:  http://www.ieee-pses.org/

To post a message to the list, send your e-mail to emc-p...@ieee.org

Instructions:  http://listserv.ieee.org/request/user-guide.html

List rules: http://www.ieee-pses.org/listrules.html

For help, send mail to the list administrators:

 Scott Douglas   emcp...@ptcnh.net
 Mike Cantwell   mcantw...@ieee.org

For policy questions, send mail to:

 Jim Bacher: j.bac...@ieee.org
 David Heald:   dhe...@gmail.com