On 03/17/2013 01:43 AM, Bob Camp wrote:

Hi

The 53132 is indeed a fine counter. It's got another flaw though - right at 10 
MHz the resolution takes a dive. If you are doing time nut stuff, that may be a 
significant issue.


The frequency averaging method can make use of the beating of the 
reference clock against the incomming clock, to get more information on 
the frequency difference. If the clocks are near each other, it takes 
longer averaging to get the same effective resolution. You can expect 
the same type of effect around frequency ratios like 3/2 etc.


This is expected as it is a systematic effect where most of the new 
samples does not bring any new info as they pop in systematically in the 
places where they don't give info. Adding noise (sloppy trigger) could 
actually improve the precission, as it would combat the quantization effect.


Cheers,
Magnus
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] HP53132A vs SR625

2013-03-17 Thread Bob Camp
Hi

If the gate time is fairly long, the notch in the resolution is quite narrow. 
You don't have to be very far off of 10 MHz to go back to fairly high 
resolution. Again, not a knock on this fine counter, just something to watch 
out for. 

Bob

On Mar 17, 2013, at 7:45 AM, Magnus Danielson mag...@rubidium.dyndns.org 
wrote:

 On 03/17/2013 01:43 AM, Bob Camp wrote:
 Hi
 
 The 53132 is indeed a fine counter. It's got another flaw though - right at 
 10 MHz the resolution takes a dive. If you are doing time nut stuff, that 
 may be a significant issue.
 
 The frequency averaging method can make use of the beating of the reference 
 clock against the incomming clock, to get more information on the frequency 
 difference. If the clocks are near each other, it takes longer averaging to 
 get the same effective resolution. You can expect the same type of effect 
 around frequency ratios like 3/2 etc.
 
 This is expected as it is a systematic effect where most of the new samples 
 does not bring any new info as they pop in systematically in the places where 
 they don't give info. Adding noise (sloppy trigger) could actually improve 
 the precission, as it would combat the quantization effect.
 
 Cheers,
 Magnus
 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] HP53132A vs SR625

2013-03-17 Thread Magnus Danielson

Hi Bob,

On 03/17/2013 02:08 PM, Bob Camp wrote:

Hi

If the gate time is fairly long, the notch in the resolution is quite narrow.


As expected. The gate-time controls the width of this notch.


You don't have to be very far off of 10 MHz to go back to fairly high 
resolution. Again, not a knock on this fine counter, just something to watch 
out for.


Indeed. This is know your tool stuff.

Cheers,
Magnus
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] HP53132A vs SR625

2013-03-17 Thread Said Jackson
Volker,

The error I have seen was in the high xE-011's to the low xE-010's. the only 
way around it was to turn on relative measurements, which then subtracted out 
this error. That error makes the unit almost useless to me.

The factory told me as long as it is within specs they will ship it after cal 
and not bother trying to improve this.

The acceptable specs are pretty crappy in tim-nuts terms: +/-350pico * 
frequency with a 1s gate time. Thats straight from the user manual and assuming 
no reference error. From the manual:

Frequency Accuracy:
 ± ((100ps typ [350 ps max])/Gate + Timebase Error ) x Frequency

That equates to up to +/-3.5E-010 if my math is right!

Not so impressive. I can confirm Ricks comment about HP 53132A units not 
showing that type of error.

Not knowing how big the error is without another non-SRS counter to compare to, 
and if it may actually be bigger than spec is a problem. I don't remember if it 
shows up when feeding the counter its own reference.

Bye
Said


Sent From iPhone

On Mar 16, 2013, at 16:06, Volker Esper ail...@t-online.de wrote:

 
 What small error are we speaking about? The statement that SRS users have 
 to tolerates a small error while HP users don't seems a little to general to 
 me. IMHO we might be a bit more precise. Anyone who's already done an error 
 analysis for - say - a 10MHz count and a comparison of the counters?
 
 In real life every type of equipment has it's domain, where it has it's 
 specific advantage. Could it be, that's the case for these counters, too?
 
 Cheers
 
 Volker
 
 
 Am 16.03.2013 19:57, schrieb Rick Karlquist:
 1) I paid quite a bit of money and I had it calibrated and fixed by
 SRS,
 and it still exhibits a significant frequency offset with a perfect
 reference  and perfect DUT!!!
   
 
 SRS says a small frequency error is normal, well that prevents me from
 using the unit as a frequency counter, for me it's only useful as a
 relative
 display frequency counter. HP doesn't have such a frequency error, so no
 worries there.
   
 I worked with the guy who designed the HP53132A.  He would
 never tolerate as normal a so-called small error.  The term
 frequency counter brings to mind something that digitally counts
 zero crossings and should never have an error.  First of all, even
 if that is all you do, it is still possible to screw it up.
 Secondly, counters have relied on analog interpolation even going
 back to the HP524 circa 1950.  There is no theoretical basis of having zero
 error in this case, but the idea is that you display the number of
 digits that are commensurate with the worst case accuracy of your
 interpolator.  Again, my colleague who designed the interpolator
 did very high quality work.  I am pleased to learn that our stuff
 is better than the stuff from the company up the road.
 
 Rick Karlquist N6RK
 HP Santa Clara Division 1979-1998
 (still working for Agilent!)
 
 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.
 
 
   
 
 
 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Re: [time-nuts] HP53132A vs SR625

2013-03-17 Thread Ed Palmer

Hi Said,

That equation looks similar in form to the specs for any counter. What 
are the comparable equations for the  53132A or the 5370(A or B)?


Ed

On 3/17/2013 10:41 AM, Said Jackson wrote:

Volker,

The error I have seen was in the high xE-011's to the low xE-010's. the only 
way around it was to turn on relative measurements, which then subtracted out 
this error. That error makes the unit almost useless to me.

The factory told me as long as it is within specs they will ship it after cal 
and not bother trying to improve this.

The acceptable specs are pretty crappy in tim-nuts terms: +/-350pico * 
frequency with a 1s gate time. Thats straight from the user manual and assuming no 
reference error. From the manual:

Frequency Accuracy:
 ± ((100ps typ [350 ps max])/Gate + Timebase Error ) x Frequency

That equates to up to +/-3.5E-010 if my math is right!

Not so impressive. I can confirm Ricks comment about HP 53132A units not 
showing that type of error.

Not knowing how big the error is without another non-SRS counter to compare to, 
and if it may actually be bigger than spec is a problem. I don't remember if it 
shows up when feeding the counter its own reference.

Bye
Said


Sent From iPhone

On Mar 16, 2013, at 16:06, Volker Esper ail...@t-online.de wrote:


What small error are we speaking about? The statement that SRS users have to 
tolerates a small error while HP users don't seems a little to general to me. IMHO we 
might be a bit more precise. Anyone who's already done an error analysis for - say - a 
10MHz count and a comparison of the counters?

In real life every type of equipment has it's domain, where it has it's 
specific advantage. Could it be, that's the case for these counters, too?

Cheers

Volker


Am 16.03.2013 19:57, schrieb Rick Karlquist:

1) I paid quite a bit of money and I had it calibrated and fixed by
SRS,
and it still exhibits a significant frequency offset with a perfect
reference  and perfect DUT!!!

SRS says a small frequency error is normal, well that prevents me from
using the unit as a frequency counter, for me it's only useful as a
relative
display frequency counter. HP doesn't have such a frequency error, so no
worries there.
   

I worked with the guy who designed the HP53132A.  He would
never tolerate as normal a so-called small error.  The term
frequency counter brings to mind something that digitally counts
zero crossings and should never have an error.  First of all, even
if that is all you do, it is still possible to screw it up.
Secondly, counters have relied on analog interpolation even going
back to the HP524 circa 1950.  There is no theoretical basis of having zero
error in this case, but the idea is that you display the number of
digits that are commensurate with the worst case accuracy of your
interpolator.  Again, my colleague who designed the interpolator
did very high quality work.  I am pleased to learn that our stuff
is better than the stuff from the company up the road.

Rick Karlquist N6RK
HP Santa Clara Division 1979-1998
(still working for Agilent!)


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Re: [time-nuts] HP53132A vs SR625

2013-03-17 Thread SAIDJACK
Ed,
 
the calculation is the same, however the numbers are 100ps for 53132A  
versus 350ps, and I have not seen an average systemic offset being displayed on 
 
any of the 3x 53132A units I use, and I see one on the SR-620. That's  why 
I sent it into SRS for calibration, paid the $$$ and got it back with the  
same exact offset and a statement that it is operating within specifications 
so  no adjustment is necessary.
 
HP manages to show zero error on average, with the digits bouncing back and 
 forth. The SRS unit manages to show a hard frequency offset. If I remember 
 correctly the SR-620 even shows this offset with it's own reference 
connected to  the inputs, the HP does not.
 
bye,
Said
 
 
In a message dated 3/17/2013 11:26:16 Pacific Daylight Time,  
ed_pal...@sasktel.net writes:

Hi  Said,

That equation looks similar in form to the specs for any counter.  What 
are the comparable equations for the  53132A or the 5370(A or  B)?

Ed

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] HP53132A vs SR625

2013-03-17 Thread Mike S

On 3/17/2013 1:56 PM, Ed Palmer wrote:

On 3/17/2013 10:41 AM, Said Jackson wrote:

The acceptable specs are pretty crappy in tim-nuts terms: +/-350pico
* frequency with a 1s gate time. Thats straight from the user manual
and assuming no reference error. From the manual:

Frequency Accuracy:
 ± ((100ps typ [350 ps max])/Gate + Timebase Error ) x Frequency


 That equation looks similar in form to the specs for any counter.
 What are the comparable equations for the  53132A or the 5370(A or B)?

5370B is similar

±Resolution ±(Time Base Error) x FREQ
± (100 ps Systematic/Gate Time) x FREQ
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Re: [time-nuts] HP53132A vs SR625

2013-03-17 Thread Volker Esper


Hi,

I just powered on my SR and looked for the offset, when the 10 MHz 
reference is connected to the input (at a gate time of 1s without 
further averaging). It shows an offset of 0 to 400uHz which should 
represent a mean error of 2E-11, while the manual predicts an error of 
about 1E-10 (as Said told us, and as my manual tells me). That's within 
the spec.


Unfortunately I don't have a 53132, but the manual of the HP predicts an 
error of E-10 - just the same value as with the SR.


If I was a manufacturer of such a counter, I wouldn't show the digits 
beneath the predicted error, but SR does: it shows 13 digits. How many 
digits does the HP show? Could it be, that the HP shows one or two 
digits less at this measurement? With only 11 digits displayed, the SR 
wouldn't show any offset, too.


By the way, HP's 100ps isn't the worst case value - that is 500ps.

So, what's the big difference beetween them?
- the predicted error is the same for both (or am I wrong?).
- my SR is within it's specification at 10 MHz (I did the calibration 
myself).
- the uncertain digits (below 1mHz, in this case) are within the error 
spec, and I guess they are uncertain because they come from an analog 
circuitry (namely the interpolation circuitry). The statement, that a 
counter only has to count zero crossings and nothing else, isn't right 
at that point, and that's the case for both,

- and they both have to deal with the analog stuff, such as noise and so on

Again, since I don't have a 53132 I can't compare the counters directly, 
I was just a little concerned about a discussion - no offense! - that 
compares apples and oranges.


Thank you for a still inspiring discussion!

Volker





Am 17.03.2013 20:05, schrieb saidj...@aol.com:

Ed,

the calculation is the same, however the numbers are 100ps for 53132A
versus 350ps, and I have not seen an average systemic offset being displayed on
any of the 3x 53132A units I use, and I see one on the SR-620. That's  why
I sent it into SRS for calibration, paid the $$$ and got it back with the
same exact offset and a statement that it is operating within specifications
so  no adjustment is necessary.

HP manages to show zero error on average, with the digits bouncing back and
  forth. The SRS unit manages to show a hard frequency offset. If I remember
  correctly the SR-620 even shows this offset with it's own reference
connected to  the inputs, the HP does not.

bye,
Said


In a message dated 3/17/2013 11:26:16 Pacific Daylight Time,
ed_pal...@sasktel.net writes:

Hi  Said,

That equation looks similar in form to the specs for any counter.  What
are the comparable equations for the  53132A or the 5370(A or  B)?

Ed

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


   



___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] HP53132A vs SR625

2013-03-17 Thread SAIDJACK
Hi Volker,
 
there are some issues here, first the worst case  frequency systematic 
uncertainty is 100ps for the 53132A, not 350ps as  on the SRS unit or 500ps as 
you stated. So they are not the same, they are 3.5x  different.
 
From the Agilent manual:
 
Systematic Uncertainty:
Agilent 53131A Agilent 53132A
tacc  tacc
typical 100 ps 10 ps
worst case 300 ps 100 ps 
 
Notice the 10ps typical error, and 100ps worst case error. That  compares 
to a 100ps typical error for the SR-620 or 10x worse typically than the  
53132A.
 
So we get 10x worse typically, and 3.5x less for the worst case -  in my 
opinion these units are not even in the same class.
 
Now for practical matters, I just measured the SR-620 we have with a  
randomly selected 53132A. Both connected to their own reference input.  
2-second 
samples on both, and here are the results:
 
The SR-620 shows a frequency error of -0.00203Hz consistently.That's  
2.03E-010. Within its specifications but making the unit useless to  me.
 
The 53132A showed an error of only 2E-012 to 8E-012. So about 25x better  
accuracy! And the 53132A is showing 12 digits on the front panel as well for 
2  second gate times at 10MHz. Nor does it require time-consuming and error 
prone  and annoying internal adjustments to achieve this.
 
What's even more damming for the SRS unit: as I increased the sample size  
(1s gate time is the max front panel selection, so I had to increase sample 
size  instead of gate time) the error stayed persistent independent of 
sample size or  thus measurement length.
 
On the HP unit however, increasing the gate time made the error get smaller 
 and smaller, and at 10+ seconds gate time I got 13 digits of resolution 
out of  the unit, and an error of only 1E-012 at that point.

So in summary, the SR-620 requires careful user adjustment of internal  
adjustment points. I don't have time to do that, so sent it in and paid the  
$600+ or so (if I remember correctly) for the standard calibration fee they  
charge. I got a unit back with the error unchanged, which was the original  
reason I sent it in to them in the first place. An error of 2E-010 makes the  
unit useless as we are in need of measuring xE-011 accurately. If I had 
time to  learn how to calibrate the unit myself, I may do so, but even then you 
showed a  2E-011 error on your carefully adjusted unit, whereas I measured 
a 2 to 8E-012  error on a random non-adjusted 53132A unit here. Still about 
3x to 10x  difference in performance.
 
If someone is interested in a swap of a working 53132A with input-c option  
for our SR-620 I would like to talk to you offline. I would even throw-in  
an FEI Rubidium reference in that swap, even though the SRS' sell for about  
$2300, and the 53132A'a go for about $1400.
 
bye,
Said
 
 
In a message dated 3/17/2013 13:02:46 Pacific Daylight Time,  
ail...@t-online.de writes:


I  just powered on my SR and looked for the offset, when the 10 MHz 
reference  is connected to the input (at a gate time of 1s without 
further  averaging). It shows an offset of 0 to 400uHz which should 
represent a  mean error of 2E-11, while the manual predicts an error of 
about 1E-10 (as  Said told us, and as my manual tells me). That's within 
the  spec.

Unfortunately I don't have a 53132, but the manual of the HP  predicts an 
error of E-10 - just the same value as with the  SR.


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] HP53132A vs SR625

2013-03-17 Thread Magnus Danielson

On 03/17/2013 08:05 PM, saidj...@aol.com wrote:

Ed,

the calculation is the same, however the numbers are 100ps for 53132A
versus 350ps, and I have not seen an average systemic offset being displayed on
any of the 3x 53132A units I use, and I see one on the SR-620. That's  why
I sent it into SRS for calibration, paid the $$$ and got it back with the
same exact offset and a statement that it is operating within specifications
so  no adjustment is necessary.

HP manages to show zero error on average, with the digits bouncing back and
  forth. The SRS unit manages to show a hard frequency offset. If I remember
  correctly the SR-620 even shows this offset with it's own reference
connected to  the inputs, the HP does not.


You can calibrate it out. SRS techs just being lazy, but it is not a bad 
design.


They could have included it in the autocal thought.

Cheers,
Magnus
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] HP53132A vs SR625

2013-03-17 Thread Bob Camp
Hi

Be very careful of what the 53132(1) reports with the ref out connected to the 
input. You are guaranteed to be in the dead zone on the counter when you do 
that. 

Bob

On Mar 17, 2013, at 5:33 PM, saidj...@aol.com wrote:

 Hi Volker,
 
 there are some issues here, first the worst case  frequency systematic 
 uncertainty is 100ps for the 53132A, not 350ps as  on the SRS unit or 500ps 
 as 
 you stated. So they are not the same, they are 3.5x  different.
 
 From the Agilent manual:
 
 Systematic Uncertainty:
 Agilent 53131A Agilent 53132A
 tacc  tacc
 typical 100 ps 10 ps
 worst case 300 ps 100 ps 
 
 Notice the 10ps typical error, and 100ps worst case error. That  compares 
 to a 100ps typical error for the SR-620 or 10x worse typically than the  
 53132A.
 
 So we get 10x worse typically, and 3.5x less for the worst case -  in my 
 opinion these units are not even in the same class.
 
 Now for practical matters, I just measured the SR-620 we have with a  
 randomly selected 53132A. Both connected to their own reference input.  
 2-second 
 samples on both, and here are the results:
 
 The SR-620 shows a frequency error of -0.00203Hz consistently.That's  
 2.03E-010. Within its specifications but making the unit useless to  me.
 
 The 53132A showed an error of only 2E-012 to 8E-012. So about 25x better  
 accuracy! And the 53132A is showing 12 digits on the front panel as well for 
 2  second gate times at 10MHz. Nor does it require time-consuming and error 
 prone  and annoying internal adjustments to achieve this.
 
 What's even more damming for the SRS unit: as I increased the sample size  
 (1s gate time is the max front panel selection, so I had to increase sample 
 size  instead of gate time) the error stayed persistent independent of 
 sample size or  thus measurement length.
 
 On the HP unit however, increasing the gate time made the error get smaller 
 and smaller, and at 10+ seconds gate time I got 13 digits of resolution 
 out of  the unit, and an error of only 1E-012 at that point.
 
 So in summary, the SR-620 requires careful user adjustment of internal  
 adjustment points. I don't have time to do that, so sent it in and paid the  
 $600+ or so (if I remember correctly) for the standard calibration fee they  
 charge. I got a unit back with the error unchanged, which was the original  
 reason I sent it in to them in the first place. An error of 2E-010 makes the  
 unit useless as we are in need of measuring xE-011 accurately. If I had 
 time to  learn how to calibrate the unit myself, I may do so, but even then 
 you 
 showed a  2E-011 error on your carefully adjusted unit, whereas I measured 
 a 2 to 8E-012  error on a random non-adjusted 53132A unit here. Still about 
 3x to 10x  difference in performance.
 
 If someone is interested in a swap of a working 53132A with input-c option  
 for our SR-620 I would like to talk to you offline. I would even throw-in  
 an FEI Rubidium reference in that swap, even though the SRS' sell for about  
 $2300, and the 53132A'a go for about $1400.
 
 bye,
 Said
 
 
 In a message dated 3/17/2013 13:02:46 Pacific Daylight Time,  
 ail...@t-online.de writes:
 
 
 I  just powered on my SR and looked for the offset, when the 10 MHz 
 reference  is connected to the input (at a gate time of 1s without 
 further  averaging). It shows an offset of 0 to 400uHz which should 
 represent a  mean error of 2E-11, while the manual predicts an error of 
 about 1E-10 (as  Said told us, and as my manual tells me). That's within 
 the  spec.
 
 Unfortunately I don't have a 53132, but the manual of the HP  predicts an 
 error of E-10 - just the same value as with the  SR.
 
 
 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] HP53132A vs SR625

2013-03-17 Thread Said Jackson
Bob,

Thats why the 53132A counter reduces the resolution to one digit less at that 
frequency, and why we use an external divide by 2 for 10MHz measurements to 
regain that digit.

I wanted to be fair and compare apples to apples. If i use our 5Mhz input, the 
53132A will be even better.

We are always measuring at the deadzone because our gpsdo's are phase aligned 
via gps. But I can guarantee that the counter can differentiate xE-011 
difference in frequencies, as we measure at this level all the time...

And I assume the zero offset error of the Sr-620 is also due to this deadzone 
issue.

Btw I was wrong the 53132A now sells for $999 on Ebay. The SR-620 is about 
$1450.

Bye
Said



Sent From iPhone

On Mar 17, 2013, at 16:22, Bob Camp li...@rtty.us wrote:

 Hi
 
 Be very careful of what the 53132(1) reports with the ref out connected to 
 the input. You are guaranteed to be in the dead zone on the counter when 
 you do that. 
 
 Bob
 
 On Mar 17, 2013, at 5:33 PM, saidj...@aol.com wrote:
 
 Hi Volker,
 
 there are some issues here, first the worst case  frequency systematic 
 uncertainty is 100ps for the 53132A, not 350ps as  on the SRS unit or 500ps 
 as 
 you stated. So they are not the same, they are 3.5x  different.
 
 From the Agilent manual:
 
 Systematic Uncertainty:
 Agilent 53131A Agilent 53132A
 tacc  tacc
 typical 100 ps 10 ps
 worst case 300 ps 100 ps 
 
 Notice the 10ps typical error, and 100ps worst case error. That  compares 
 to a 100ps typical error for the SR-620 or 10x worse typically than the  
 53132A.
 
 So we get 10x worse typically, and 3.5x less for the worst case -  in my 
 opinion these units are not even in the same class.
 
 Now for practical matters, I just measured the SR-620 we have with a  
 randomly selected 53132A. Both connected to their own reference input.  
 2-second 
 samples on both, and here are the results:
 
 The SR-620 shows a frequency error of -0.00203Hz consistently.That's  
 2.03E-010. Within its specifications but making the unit useless to  me.
 
 The 53132A showed an error of only 2E-012 to 8E-012. So about 25x better  
 accuracy! And the 53132A is showing 12 digits on the front panel as well for 
 2  second gate times at 10MHz. Nor does it require time-consuming and error 
 prone  and annoying internal adjustments to achieve this.
 
 What's even more damming for the SRS unit: as I increased the sample size  
 (1s gate time is the max front panel selection, so I had to increase sample 
 size  instead of gate time) the error stayed persistent independent of 
 sample size or  thus measurement length.
 
 On the HP unit however, increasing the gate time made the error get smaller 
 and smaller, and at 10+ seconds gate time I got 13 digits of resolution 
 out of  the unit, and an error of only 1E-012 at that point.
 
 So in summary, the SR-620 requires careful user adjustment of internal  
 adjustment points. I don't have time to do that, so sent it in and paid the  
 $600+ or so (if I remember correctly) for the standard calibration fee they  
 charge. I got a unit back with the error unchanged, which was the original  
 reason I sent it in to them in the first place. An error of 2E-010 makes the 
  
 unit useless as we are in need of measuring xE-011 accurately. If I had 
 time to  learn how to calibrate the unit myself, I may do so, but even then 
 you 
 showed a  2E-011 error on your carefully adjusted unit, whereas I measured 
 a 2 to 8E-012  error on a random non-adjusted 53132A unit here. Still about 
 3x to 10x  difference in performance.
 
 If someone is interested in a swap of a working 53132A with input-c option  
 for our SR-620 I would like to talk to you offline. I would even throw-in  
 an FEI Rubidium reference in that swap, even though the SRS' sell for about  
 $2300, and the 53132A'a go for about $1400.
 
 bye,
 Said
 
 
 In a message dated 3/17/2013 13:02:46 Pacific Daylight Time,  
 ail...@t-online.de writes:
 
 
 I  just powered on my SR and looked for the offset, when the 10 MHz 
 reference  is connected to the input (at a gate time of 1s without 
 further  averaging). It shows an offset of 0 to 400uHz which should 
 represent a  mean error of 2E-11, while the manual predicts an error of 
 about 1E-10 (as  Said told us, and as my manual tells me). That's within 
 the  spec.
 
 Unfortunately I don't have a 53132, but the manual of the HP  predicts an 
 error of E-10 - just the same value as with the  SR.
 
 
 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.
 
 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go 

Re: [time-nuts] HP53132A vs SR625

2013-03-17 Thread Bob Camp
Hi

This brings up the basic how bad is it question. Since the counter is 
fundamentally a 200 ps gizmo, a simple period measurement at 1 second will give 
you ~ 10 digits per second.  That's with no magic multiple sample stuff at all. 
At an offset / noise / what ever state where the multiple sample stuff works 
100%, you get ~ 12 digits per second.  I doubt you ever get everything so 
perfect that you are down to 10 digits / second. 

Bob


On Mar 17, 2013, at 7:41 PM, Said Jackson saidj...@aol.com wrote:

 Bob,
 
 Thats why the 53132A counter reduces the resolution to one digit less at that 
 frequency, and why we use an external divide by 2 for 10MHz measurements to 
 regain that digit.
 
 I wanted to be fair and compare apples to apples. If i use our 5Mhz input, 
 the 53132A will be even better.
 
 We are always measuring at the deadzone because our gpsdo's are phase 
 aligned via gps. But I can guarantee that the counter can differentiate 
 xE-011 difference in frequencies, as we measure at this level all the time...
 
 And I assume the zero offset error of the Sr-620 is also due to this deadzone 
 issue.
 
 Btw I was wrong the 53132A now sells for $999 on Ebay. The SR-620 is about 
 $1450.
 
 Bye
 Said
 
 
 
 Sent From iPhone
 
 On Mar 17, 2013, at 16:22, Bob Camp li...@rtty.us wrote:
 
 Hi
 
 Be very careful of what the 53132(1) reports with the ref out connected to 
 the input. You are guaranteed to be in the dead zone on the counter when 
 you do that. 
 
 Bob
 
 On Mar 17, 2013, at 5:33 PM, saidj...@aol.com wrote:
 
 Hi Volker,
 
 there are some issues here, first the worst case  frequency systematic 
 uncertainty is 100ps for the 53132A, not 350ps as  on the SRS unit or 500ps 
 as 
 you stated. So they are not the same, they are 3.5x  different.
 
 From the Agilent manual:
 
 Systematic Uncertainty:
 Agilent 53131A Agilent 53132A
 tacc  tacc
 typical 100 ps 10 ps
 worst case 300 ps 100 ps 
 
 Notice the 10ps typical error, and 100ps worst case error. That  compares 
 to a 100ps typical error for the SR-620 or 10x worse typically than the  
 53132A.
 
 So we get 10x worse typically, and 3.5x less for the worst case -  in my 
 opinion these units are not even in the same class.
 
 Now for practical matters, I just measured the SR-620 we have with a  
 randomly selected 53132A. Both connected to their own reference input.  
 2-second 
 samples on both, and here are the results:
 
 The SR-620 shows a frequency error of -0.00203Hz consistently.That's  
 2.03E-010. Within its specifications but making the unit useless to  me.
 
 The 53132A showed an error of only 2E-012 to 8E-012. So about 25x better  
 accuracy! And the 53132A is showing 12 digits on the front panel as well 
 for 
 2  second gate times at 10MHz. Nor does it require time-consuming and error 
 prone  and annoying internal adjustments to achieve this.
 
 What's even more damming for the SRS unit: as I increased the sample size  
 (1s gate time is the max front panel selection, so I had to increase sample 
 size  instead of gate time) the error stayed persistent independent of 
 sample size or  thus measurement length.
 
 On the HP unit however, increasing the gate time made the error get smaller 
 and smaller, and at 10+ seconds gate time I got 13 digits of resolution 
 out of  the unit, and an error of only 1E-012 at that point.
 
 So in summary, the SR-620 requires careful user adjustment of internal  
 adjustment points. I don't have time to do that, so sent it in and paid the 
  
 $600+ or so (if I remember correctly) for the standard calibration fee they 
  
 charge. I got a unit back with the error unchanged, which was the original  
 reason I sent it in to them in the first place. An error of 2E-010 makes 
 the  
 unit useless as we are in need of measuring xE-011 accurately. If I had 
 time to  learn how to calibrate the unit myself, I may do so, but even then 
 you 
 showed a  2E-011 error on your carefully adjusted unit, whereas I measured 
 a 2 to 8E-012  error on a random non-adjusted 53132A unit here. Still about 
 3x to 10x  difference in performance.
 
 If someone is interested in a swap of a working 53132A with input-c option  
 for our SR-620 I would like to talk to you offline. I would even throw-in  
 an FEI Rubidium reference in that swap, even though the SRS' sell for about 
  
 $2300, and the 53132A'a go for about $1400.
 
 bye,
 Said
 
 
 In a message dated 3/17/2013 13:02:46 Pacific Daylight Time,  
 ail...@t-online.de writes:
 
 
 I  just powered on my SR and looked for the offset, when the 10 MHz 
 reference  is connected to the input (at a gate time of 1s without 
 further  averaging). It shows an offset of 0 to 400uHz which should 
 represent a  mean error of 2E-11, while the manual predicts an error of 
 about 1E-10 (as  Said told us, and as my manual tells me). That's within 
 the  spec.
 
 Unfortunately I don't have a 53132, but the manual of the HP  predicts an 
 error of E-10 - just the same value as 

Re: [time-nuts] HP53132A vs SR625

2013-03-17 Thread Said Jackson
Hi Volcker, Bob,

I guess it depends on what one needs. The SR-620 is probably more of a gizmo to 
play with when one likes to manually adjust things or needs the better time 
interval resolution.

The HP unit is more of a fire-and-forget unit.

Me having the benefit to be able to chose, I would generally prefer the HP 
unit. It's cheaper too!

Also if HP says its a 12 digit per second unit, then it probably is doing that, 
albeit with some caveats as the manual states..

Bye,
Said

Sent From iPhone

On Mar 17, 2013, at 17:23, Bob Camp li...@rtty.us wrote:

 Hi
 
 This brings up the basic how bad is it question. Since the counter is 
 fundamentally a 200 ps gizmo, a simple period measurement at 1 second will 
 give you ~ 10 digits per second.  That's with no magic multiple sample stuff 
 at all. At an offset / noise / what ever state where the multiple sample 
 stuff works 100%, you get ~ 12 digits per second.  I doubt you ever get 
 everything so perfect that you are down to 10 digits / second. 
 
 Bob
 
 
 On Mar 17, 2013, at 7:41 PM, Said Jackson saidj...@aol.com wrote:
 
 Bob,
 
 Thats why the 53132A counter reduces the resolution to one digit less at 
 that frequency, and why we use an external divide by 2 for 10MHz 
 measurements to regain that digit.
 
 I wanted to be fair and compare apples to apples. If i use our 5Mhz input, 
 the 53132A will be even better.
 
 We are always measuring at the deadzone because our gpsdo's are phase 
 aligned via gps. But I can guarantee that the counter can differentiate 
 xE-011 difference in frequencies, as we measure at this level all the time...
 
 And I assume the zero offset error of the Sr-620 is also due to this 
 deadzone issue.
 
 Btw I was wrong the 53132A now sells for $999 on Ebay. The SR-620 is about 
 $1450.
 
 Bye
 Said
 
 
 
 Sent From iPhone
 
 On Mar 17, 2013, at 16:22, Bob Camp li...@rtty.us wrote:
 
 Hi
 
 Be very careful of what the 53132(1) reports with the ref out connected to 
 the input. You are guaranteed to be in the dead zone on the counter when 
 you do that. 
 
 Bob
 
 On Mar 17, 2013, at 5:33 PM, saidj...@aol.com wrote:
 
 Hi Volker,
 
 there are some issues here, first the worst case  frequency systematic 
 uncertainty is 100ps for the 53132A, not 350ps as  on the SRS unit or 
 500ps as 
 you stated. So they are not the same, they are 3.5x  different.
 
 From the Agilent manual:
 
 Systematic Uncertainty:
 Agilent 53131A Agilent 53132A
 tacc  tacc
 typical 100 ps 10 ps
 worst case 300 ps 100 ps 
 
 Notice the 10ps typical error, and 100ps worst case error. That  compares 
 to a 100ps typical error for the SR-620 or 10x worse typically than the  
 53132A.
 
 So we get 10x worse typically, and 3.5x less for the worst case -  in my 
 opinion these units are not even in the same class.
 
 Now for practical matters, I just measured the SR-620 we have with a  
 randomly selected 53132A. Both connected to their own reference input.  
 2-second 
 samples on both, and here are the results:
 
 The SR-620 shows a frequency error of -0.00203Hz consistently.That's  
 2.03E-010. Within its specifications but making the unit useless to  me.
 
 The 53132A showed an error of only 2E-012 to 8E-012. So about 25x better  
 accuracy! And the 53132A is showing 12 digits on the front panel as well 
 for 
 2  second gate times at 10MHz. Nor does it require time-consuming and 
 error 
 prone  and annoying internal adjustments to achieve this.
 
 What's even more damming for the SRS unit: as I increased the sample size  
 (1s gate time is the max front panel selection, so I had to increase 
 sample 
 size  instead of gate time) the error stayed persistent independent of 
 sample size or  thus measurement length.
 
 On the HP unit however, increasing the gate time made the error get 
 smaller 
 and smaller, and at 10+ seconds gate time I got 13 digits of resolution 
 out of  the unit, and an error of only 1E-012 at that point.
 
 So in summary, the SR-620 requires careful user adjustment of internal  
 adjustment points. I don't have time to do that, so sent it in and paid 
 the  
 $600+ or so (if I remember correctly) for the standard calibration fee 
 they  
 charge. I got a unit back with the error unchanged, which was the original 
  
 reason I sent it in to them in the first place. An error of 2E-010 makes 
 the  
 unit useless as we are in need of measuring xE-011 accurately. If I had 
 time to  learn how to calibrate the unit myself, I may do so, but even 
 then you 
 showed a  2E-011 error on your carefully adjusted unit, whereas I measured 
 a 2 to 8E-012  error on a random non-adjusted 53132A unit here. Still 
 about 
 3x to 10x  difference in performance.
 
 If someone is interested in a swap of a working 53132A with input-c option 
  
 for our SR-620 I would like to talk to you offline. I would even throw-in  
 an FEI Rubidium reference in that swap, even though the SRS' sell for 
 about  
 $2300, and the 53132A'a go for about $1400.
 
 bye,
 Said

Re: [time-nuts] HP53132A vs SR625

2013-03-17 Thread Richard (Rick) Karlquist



On 3/17/2013 4:54 PM, Volker Esper wrote:


The HP seems to be the more modern design. As I guess, the analog
circuits are to blame, maybe HP was able to make use of newer technologies.


FWIW, the 53132A design goes back 20 years

Rick
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] HP53132A vs SR625

2013-03-16 Thread Rick Karlquist
1) I paid quite a bit of money and I had it calibrated and fixed by
 SRS,
and it still exhibits a significant frequency offset with a perfect
reference  and perfect DUT!!!

SRS says a small frequency error is normal, well that prevents me from
using the unit as a frequency counter, for me it's only useful as a
 relative
display frequency counter. HP doesn't have such a frequency error, so no
worries there.

I worked with the guy who designed the HP53132A.  He would
never tolerate as normal a so-called small error.  The term
frequency counter brings to mind something that digitally counts
zero crossings and should never have an error.  First of all, even
if that is all you do, it is still possible to screw it up.
Secondly, counters have relied on analog interpolation even going
back to the HP524 circa 1950.  There is no theoretical basis of having zero
error in this case, but the idea is that you display the number of
digits that are commensurate with the worst case accuracy of your
interpolator.  Again, my colleague who designed the interpolator
did very high quality work.  I am pleased to learn that our stuff
is better than the stuff from the company up the road.

Rick Karlquist N6RK
HP Santa Clara Division 1979-1998
(still working for Agilent!)

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] HP53132A vs SR625

2013-03-16 Thread Volker Esper


What small error are we speaking about? The statement that SRS users 
have to tolerates a small error while HP users don't seems a little to 
general to me. IMHO we might be a bit more precise. Anyone who's already 
done an error analysis for - say - a 10MHz count and a comparison of the 
counters?


In real life every type of equipment has it's domain, where it has it's 
specific advantage. Could it be, that's the case for these counters, too?


Cheers

Volker


Am 16.03.2013 19:57, schrieb Rick Karlquist:

1) I paid quite a bit of money and I had it calibrated and fixed by
SRS,
and it still exhibits a significant frequency offset with a perfect
reference  and perfect DUT!!!
   
 

SRS says a small frequency error is normal, well that prevents me from
using the unit as a frequency counter, for me it's only useful as a
relative
display frequency counter. HP doesn't have such a frequency error, so no
worries there.
   

I worked with the guy who designed the HP53132A.  He would
never tolerate as normal a so-called small error.  The term
frequency counter brings to mind something that digitally counts
zero crossings and should never have an error.  First of all, even
if that is all you do, it is still possible to screw it up.
Secondly, counters have relied on analog interpolation even going
back to the HP524 circa 1950.  There is no theoretical basis of having zero
error in this case, but the idea is that you display the number of
digits that are commensurate with the worst case accuracy of your
interpolator.  Again, my colleague who designed the interpolator
did very high quality work.  I am pleased to learn that our stuff
is better than the stuff from the company up the road.

Rick Karlquist N6RK
HP Santa Clara Division 1979-1998
(still working for Agilent!)

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


   



___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] HP53132A vs SR625

2013-03-16 Thread Bob Camp
Hi

The whole how to align the SR-620's thing has been gone over in great length 
(and very precisely) on the list. The bottom line is that Stanford's alignment 
procedure is not quite as good as it should / could be. 

Oddly enough, I don't remember seeing the residual offset in any brand new 
SR-620. I've only seen it on repaired units. My guess is that the factory 
tech's did a bit better alignment than the repair manual outlines.

Bob

On Mar 16, 2013, at 7:06 PM, Volker Esper ail...@t-online.de wrote:

 
 What small error are we speaking about? The statement that SRS users have 
 to tolerates a small error while HP users don't seems a little to general to 
 me. IMHO we might be a bit more precise. Anyone who's already done an error 
 analysis for - say - a 10MHz count and a comparison of the counters?
 
 In real life every type of equipment has it's domain, where it has it's 
 specific advantage. Could it be, that's the case for these counters, too?
 
 Cheers
 
 Volker
 
 
 Am 16.03.2013 19:57, schrieb Rick Karlquist:
 1) I paid quite a bit of money and I had it calibrated and fixed by
 SRS,
 and it still exhibits a significant frequency offset with a perfect
 reference  and perfect DUT!!!
   
 
 SRS says a small frequency error is normal, well that prevents me from
 using the unit as a frequency counter, for me it's only useful as a
 relative
 display frequency counter. HP doesn't have such a frequency error, so no
 worries there.
   
 I worked with the guy who designed the HP53132A.  He would
 never tolerate as normal a so-called small error.  The term
 frequency counter brings to mind something that digitally counts
 zero crossings and should never have an error.  First of all, even
 if that is all you do, it is still possible to screw it up.
 Secondly, counters have relied on analog interpolation even going
 back to the HP524 circa 1950.  There is no theoretical basis of having zero
 error in this case, but the idea is that you display the number of
 digits that are commensurate with the worst case accuracy of your
 interpolator.  Again, my colleague who designed the interpolator
 did very high quality work.  I am pleased to learn that our stuff
 is better than the stuff from the company up the road.
 
 Rick Karlquist N6RK
 HP Santa Clara Division 1979-1998
 (still working for Agilent!)
 
 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.
 
 
   
 
 
 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] HP53132A vs SR625

2013-03-16 Thread Bob Camp
Hi

The 53132 is indeed a fine counter. It's got another flaw though - right at 10 
MHz the resolution takes a dive. If you are doing time nut stuff, that may be a 
significant issue.

Bob

On Mar 16, 2013, at 8:32 PM, Brooke Clarke bro...@pacific.net wrote:

 Hi Volker:
 
 The HP HP53132A can measure frequency at the rate of 12 digits per second, 
 that's way better than ordinary counters, but when measuring time interval 
 it's the same as any other counter.
 The big disadvantage of the HP53132A in my opinion is it's user hostile menu 
 system.
 http://www.prc68.com/I/TandFTE.shtml#HP51132
 If you're going to be measuring frequency then this counter may make more 
 sense than the SRS unit.
 
 The SR620 was designed to be a time interval counter, and that's what gets 
 measured when working with precision frequency or time signals.  It's great 
 for this because it has a large number of digits.
 In addition there's a way to make 1,000 measurements and average them to 
 increase the precision compared to a one shot measurement.  The front panel 
 is much much easier to use than the HP.
 http://www.prc68.com/I/TandFTE.shtml#SR620
 
 Have Fun,
 
 Brooke Clarke, N6GCE
 http://www.PRC68.com
 http://www.end2partygovernment.com/2012Issues.html
 
 Volker Esper wrote:
 
 What small error are we speaking about? The statement that SRS users have 
 to tolerates a small error while HP users don't seems a little to general to 
 me. IMHO we might be a bit more precise. Anyone who's already done an error 
 analysis for - say - a 10MHz count and a comparison of the counters?
 
 In real life every type of equipment has it's domain, where it has it's 
 specific advantage. Could it be, that's the case for these counters, too?
 
 Cheers
 
 Volker
 
 
 Am 16.03.2013 19:57, schrieb Rick Karlquist:
 1) I paid quite a bit of money and I had it calibrated and fixed by
 SRS,
 and it still exhibits a significant frequency offset with a perfect
 reference  and perfect DUT!!!
 SRS says a small frequency error is normal, well that prevents me from
 using the unit as a frequency counter, for me it's only useful as a
 relative
 display frequency counter. HP doesn't have such a frequency error, so no
 worries there.
 I worked with the guy who designed the HP53132A.  He would
 never tolerate as normal a so-called small error.  The term
 frequency counter brings to mind something that digitally counts
 zero crossings and should never have an error.  First of all, even
 if that is all you do, it is still possible to screw it up.
 Secondly, counters have relied on analog interpolation even going
 back to the HP524 circa 1950.  There is no theoretical basis of having zero
 error in this case, but the idea is that you display the number of
 digits that are commensurate with the worst case accuracy of your
 interpolator.  Again, my colleague who designed the interpolator
 did very high quality work.  I am pleased to learn that our stuff
 is better than the stuff from the company up the road.
 
 Rick Karlquist N6RK
 HP Santa Clara Division 1979-1998
 (still working for Agilent!)
 
 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to 
 https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.
 
 
 
 
 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.
 
 
 
 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] HP53132A vs SR625

2013-03-15 Thread Hui Zhang
Hello SAID:
  I am very sorry I almost missed this email in my inbox, and say sorry for 
reply so late, and I am very appreciate you wrote so many useful word for me.  
I answred your in quote part in this email. Thanks a lot.


  After a short consideration, finally I bought SR625, now it on my workbench, 
I am very satisfied with its work.


  I still have few quesiton:


1) When I use long gate time to mesure fequency (like as 10s or more), the STOP 
LED in channel B will light up, at moment in the reading result, the STOP will 
off few second, and light up again, is this normal?


2) When I use TI mesurement function, I connect GPS's 1PPS to CH1 and use a BNC 
tee split signal with 1.5m cable to CH2, in the TIME mode, I saw the 
0xx etc, if I changed cable length the reading will change, what mean 
of this reading?




Hui,
 
rent one of each if you can before you make your choice. I have both, and  
the HP unit is much easier to use once you know which button sequence to 
push to  get more than just Frequency/Time-Interval type measurements - 
these 
can be  single-button events on the HP unit.
 
Even offsetting and normalizing frequencies becomes very easy after a  
couple of days of using the unit, there is no setting that takes me longer 
than  
about 5 seconds to set up, so while not perfect, the user interface can be  
learned easily. I find the SR620 to have too many buttons(!) I always find  
myself searching for just that one button. Anyways, more buttons are just 
more  things that can fail. If you are a pilot, and have used a Garmin 430W  
GPS in your life, then the HP user interface is no challenge whatsoever and  
seems very easy to use..
 
The SR-620 has it's advantages, especially when you just do one single type 
 of measurement, but for me it has a huge number of disadvantages, and I 
mostly  use the 53132A for that reason:
 
1) I paid quite a bit of money and I had it calibrated and fixed by SRS,  
and it still exhibits a significant frequency offset with a perfect 
reference  and perfect DUT!!!


 
SRS says a small frequency error is normal, well that prevents me from  
using the unit as a frequency counter, for me it's only useful as a relative  
display frequency counter. HP doesn't have such a frequency error, so no  
worries there.


I will pay attention and observe those problem
 
2) The SRS unit is s loud that it's totally annoying and unacceptable  
for long measurements. Many folks reported this here before. It's just bad.  
Whining like crazy.


Indeed this is little problem but really troube me.
 
3) The SRS unit is 19 wide, huge, heavy, and clunky. I need my  counter 
portable, only the HP unit will do


No problem, I like huge and heavy device, they are usually reliabe.
 
4) The SRS unit has a much lower MTBF because of all the parts inside, and  
it needs finicky adjustments, see item 1) above. The HP unit either works, 
or is  just dead. Not much to adjust. Different technology generation. And 
the coolness  factor: a nice florescent tube display is so much more modern 
looking than those  clunky old 7-segment LED's..


Yes I agree you, the SR625 is 90's product, sort of old and vintage, I am very 
worry about his lift. The only thing I can do it pray god I can have good luck. 
 
5) The SRS unit is usually $1000 more than the HP unit, and you don't know  
how good the unit is you are buying because of all of the calibration 
stuff.  Usually there is no hit-or-miss issue with the HP units, they either 
work, or  are dead.


My SR625's price not bad, very close with HP53132, although it's old. I hope it 
can work five to ten years that I will be satisfied.
 
That said, the HP unit doesn't measure well at 10MHz, so I mostly use a  
divide-by-two to get one more digit of resolution out of it, and it's time  
interval resolution is not as good as the SR620. But for time interval  
measurements I use a Wavecrest DTS unit that blows the SR620 and the HP out  
of 
the water anyways..
 
Bye,
Said
 


Hui 


 
 
 
In a message dated 2/7/2013 16:39:04 Pacific Standard Time, ba6it at 163.com  
writes:


Hello  Dear Group:




I am very glad to see so many replies in  the morning, and I am very 
grateful to every time nuts gave me useful  information, your proposal has 
strengthened my determination, in fact, I am  also very like SR625, So I will 
to 
find and buy a good shape SR625 for my new  time interval measure instrument.




Thanks again for everyone's  advice, which is very useful to make a choice 
for me. Sorry for not reply  everyone's mail. 




Best Regards!




Hui  Zhang
[time-nuts] HP53132A vs SR625
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] HP53132A vs SR625

2013-02-07 Thread Bob Camp
Hi

The SR625 is a SR620 with a rubidium time base. If the Rb is in good shape,
and you don't have a house standard, then that's a significant plus. 

Bob

-Original Message-
From: time-nuts-boun...@febo.com [mailto:time-nuts-boun...@febo.com] On
Behalf Of Hui Zhang
Sent: Thursday, February 07, 2013 9:43 AM
To: time-nuts@febo.com
Subject: [time-nuts] HP53132A vs SR625

Dear Group:


I am intend to buy a second-hand counter, I have not decided yet between
SR625 and HP53132A, they have very close price in surplus market of here,
so, which is better choice? Any suggestion will be appreciated.


Best Regards!


Hui Zhang
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] HP53132A vs SR625

2013-02-07 Thread Azelio Boriani
The SR625 at the same price of the HP53132A: either the HP is overvalued or
the SR625 is a real bargain...

On Thu, Feb 7, 2013 at 5:36 PM, Bob Camp li...@rtty.us wrote:

 Hi

 The SR625 is a SR620 with a rubidium time base. If the Rb is in good shape,
 and you don't have a house standard, then that's a significant plus.

 Bob

 -Original Message-
 From: time-nuts-boun...@febo.com [mailto:time-nuts-boun...@febo.com] On
 Behalf Of Hui Zhang
 Sent: Thursday, February 07, 2013 9:43 AM
 To: time-nuts@febo.com
 Subject: [time-nuts] HP53132A vs SR625

 Dear Group:


 I am intend to buy a second-hand counter, I have not decided yet
 between
 SR625 and HP53132A, they have very close price in surplus market of here,
 so, which is better choice? Any suggestion will be appreciated.


 Best Regards!


 Hui Zhang
 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to
 https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.


 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to
 https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] HP53132A vs SR625

2013-02-07 Thread Volker Esper


If the SR is in good shape I would prefer the SR. I love my one not only 
for it's technical data but also for the way of operation. Less menues, 
more controls, logically arranged.


Volker


Am 07.02.2013 17:53, schrieb Azelio Boriani:

The SR625 at the same price of the HP53132A: either the HP is overvalued or
the SR625 is a real bargain...

On Thu, Feb 7, 2013 at 5:36 PM, Bob Campli...@rtty.us  wrote:

   

Hi

The SR625 is a SR620 with a rubidium time base. If the Rb is in good shape,
and you don't have a house standard, then that's a significant plus.

Bob

-Original Message-
From: time-nuts-boun...@febo.com [mailto:time-nuts-boun...@febo.com] On
Behalf Of Hui Zhang
Sent: Thursday, February 07, 2013 9:43 AM
To: time-nuts@febo.com
Subject: [time-nuts] HP53132A vs SR625

Dear Group:


 I am intend to buy a second-hand counter, I have not decided yet
between
SR625 and HP53132A, they have very close price in surplus market of here,
so, which is better choice? Any suggestion will be appreciated.


Best Regards!


Hui Zhang
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

 

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


   



___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] HP53132A vs SR625

2013-02-07 Thread paul swed
Hui
Well given what you have shared for information neither answer is right or
wrong better or worse. The question that we would need answered to be
helpful is how will you use it. What price are you willing to pay. There
are tradeoffs. I am not an expert. I have many HP counters that I like
well. I have SRS equipment thats equally very fine. I do not have a SRS 620
or 625. But if I ever see one at a really good price. (Really cheap) I
would buy it, even if broken.
Regards
Paul

On Thu, Feb 7, 2013 at 1:43 PM, Volker Esper ail...@t-online.de wrote:


 If the SR is in good shape I would prefer the SR. I love my one not only
 for it's technical data but also for the way of operation. Less menues,
 more controls, logically arranged.

 Volker


 Am 07.02.2013 17:53, schrieb Azelio Boriani:

  The SR625 at the same price of the HP53132A: either the HP is overvalued
 or
 the SR625 is a real bargain...

 On Thu, Feb 7, 2013 at 5:36 PM, Bob Campli...@rtty.us  wrote:



 Hi

 The SR625 is a SR620 with a rubidium time base. If the Rb is in good
 shape,
 and you don't have a house standard, then that's a significant plus.

 Bob

 -Original Message-
 From: time-nuts-boun...@febo.com 
 [mailto:time-nuts-bounces@**febo.comtime-nuts-boun...@febo.com]
 On
 Behalf Of Hui Zhang
 Sent: Thursday, February 07, 2013 9:43 AM
 To: time-nuts@febo.com
 Subject: [time-nuts] HP53132A vs SR625

 Dear Group:


  I am intend to buy a second-hand counter, I have not decided yet
 between
 SR625 and HP53132A, they have very close price in surplus market of here,
 so, which is better choice? Any suggestion will be appreciated.


 Best Regards!


 Hui Zhang
 __**_
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to
 https://www.febo.com/cgi-bin/**mailman/listinfo/time-nutshttps://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.


 __**_
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to
 https://www.febo.com/cgi-bin/**mailman/listinfo/time-nutshttps://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.



 __**_
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to https://www.febo.com/cgi-bin/**
 mailman/listinfo/time-nutshttps://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.






 __**_
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to https://www.febo.com/cgi-bin/**
 mailman/listinfo/time-nutshttps://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] HP53132A vs SR625

2013-02-07 Thread Brooke Clarke

Hi Hui:

The HP 53132A is called a Universal Frequency Counter can count frequency at 12 digits per second, but that is not the 
case when it is measuring time intervals.  The human interface might be called user hostile.


The front panel of the SR 620 is marked Universal Time Interval Counter and is great for making time interval 
measurements. That's why I traded my 53132 for the SR620. http://www.prc68.com/I/TandFTE.shtml#SR620


If you're doing Time Nuts measurements you probably have a good 10 MHz oscillator that you can feed into either of these 
counters so getting an optional time base in one of them serves little purpose.


Have Fun,

Brooke Clarke
http://www.PRC68.com
http://www.end2partygovernment.com/2012Issues.html

Hui Zhang wrote:

Dear Group:


 I am intend to buy a second-hand counter, I have not decided yet between 
SR625 and HP53132A, they have very close price in surplus market of here, so, 
which is better choice? Any suggestion will be appreciated.


Best Regards!


Hui Zhang
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.




___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] HP53132A vs SR625

2013-02-07 Thread Magnus Danielson

On 02/07/2013 07:43 PM, Volker Esper wrote:


If the SR is in good shape I would prefer the SR. I love my one not only
for it's technical data but also for the way of operation. Less menues,
more controls, logically arranged.


I tend to agree here.

The HP/Agilent 53132A lacks the ease of use aspect that older HP 
counters had. It has better performance than many of the older, but not 
all. The SR620/625 has (if in good condition) better performance than 
the 53132A.


These days I have both.

Cheers,
Magnus


Volker


Am 07.02.2013 17:53, schrieb Azelio Boriani:

The SR625 at the same price of the HP53132A: either the HP is
overvalued or
the SR625 is a real bargain...

On Thu, Feb 7, 2013 at 5:36 PM, Bob Campli...@rtty.us wrote:


Hi

The SR625 is a SR620 with a rubidium time base. If the Rb is in good
shape,
and you don't have a house standard, then that's a significant plus.

Bob

-Original Message-
From: time-nuts-boun...@febo.com [mailto:time-nuts-boun...@febo.com] On
Behalf Of Hui Zhang
Sent: Thursday, February 07, 2013 9:43 AM
To: time-nuts@febo.com
Subject: [time-nuts] HP53132A vs SR625

Dear Group:


I am intend to buy a second-hand counter, I have not decided yet
between
SR625 and HP53132A, they have very close price in surplus market of
here,
so, which is better choice? Any suggestion will be appreciated.


Best Regards!


Hui Zhang
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.





___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] HP53132A vs SR625

2013-02-07 Thread Hui Zhang
Hello Dear Group:


I am very glad to see so many replies in the morning, and I am very 
grateful to every time nuts gave me useful information, your proposal has 
strengthened my determination, in fact, I am also very like SR625, So I will to 
find and buy a good shape SR625 for my new time interval measure instrument.


Thanks again for everyone's advice, which is very useful to make a choice for 
me. Sorry for not reply everyone's mail. 


Best Regards!


Hui Zhang












At 2013-02-08 05:12:37,Brooke Clarke bro...@pacific.net wrote:
Hi Hui:

The HP 53132A is called a Universal Frequency Counter can count frequency at 
12 digits per second, but that is not the 
case when it is measuring time intervals.  The human interface might be called 
user hostile.

The front panel of the SR 620 is marked Universal Time Interval Counter and 
is great for making time interval 
measurements. That's why I traded my 53132 for the SR620. 
http://www.prc68.com/I/TandFTE.shtml#SR620

If you're doing Time Nuts measurements you probably have a good 10 MHz 
oscillator that you can feed into either of these 
counters so getting an optional time base in one of them serves little purpose.

Have Fun,

Brooke Clarke
http://www.PRC68.com
http://www.end2partygovernment.com/2012Issues.html

Hui Zhang wrote:
 Dear Group:


  I am intend to buy a second-hand counter, I have not decided yet 
 between SR625 and HP53132A, they have very close price in surplus market of 
 here, so, which is better choice? Any suggestion will be appreciated.


 Best Regards!


 Hui Zhang
 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.



___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] HP53132A vs SR625

2013-02-07 Thread SAIDJACK
Hui,
 
rent one of each if you can before you make your choice. I have both, and  
the HP unit is much easier to use once you know which button sequence to 
push to  get more than just Frequency/Time-Interval type measurements - these 
can be  single-button events on the HP unit.
 
Even offsetting and normalizing frequencies becomes very easy after a  
couple of days of using the unit, there is no setting that takes me longer than 
 
about 5 seconds to set up, so while not perfect, the user interface can be  
learned easily. I find the SR620 to have too many buttons(!) I always find  
myself searching for just that one button. Anyways, more buttons are just 
more  things that can fail. If you are a pilot, and have used a Garmin 430W  
GPS in your life, then the HP user interface is no challenge whatsoever and  
seems very easy to use..
 
The SR-620 has it's advantages, especially when you just do one single type 
 of measurement, but for me it has a huge number of disadvantages, and I 
mostly  use the 53132A for that reason:
 
1) I paid quite a bit of money and I had it calibrated and fixed by SRS,  
and it still exhibits a significant frequency offset with a perfect 
reference  and perfect DUT!!!
 
SRS says a small frequency error is normal, well that prevents me from  
using the unit as a frequency counter, for me it's only useful as a relative  
display frequency counter. HP doesn't have such a frequency error, so no  
worries there.
 
2) The SRS unit is s loud that it's totally annoying and unacceptable  
for long measurements. Many folks reported this here before. It's just bad.  
Whining like crazy.
 
3) The SRS unit is 19 wide, huge, heavy, and clunky. I need my  counter 
portable, only the HP unit will do
 
4) The SRS unit has a much lower MTBF because of all the parts inside, and  
it needs finicky adjustments, see item 1) above. The HP unit either works, 
or is  just dead. Not much to adjust. Different technology generation. And 
the coolness  factor: a nice florescent tube display is so much more modern 
looking than those  clunky old 7-segment LED's..
 
5) The SRS unit is usually $1000 more than the HP unit, and you don't know  
how good the unit is you are buying because of all of the calibration 
stuff.  Usually there is no hit-or-miss issue with the HP units, they either 
work, or  are dead.
 
That said, the HP unit doesn't measure well at 10MHz, so I mostly use a  
divide-by-two to get one more digit of resolution out of it, and it's time  
interval resolution is not as good as the SR620. But for time interval  
measurements I use a Wavecrest DTS unit that blows the SR620 and the HP out  of 
the water anyways..
 
Bye,
Said
 
 
 
 
In a message dated 2/7/2013 16:39:04 Pacific Standard Time, ba...@163.com  
writes:

Hello  Dear Group:


I am very glad to see so many replies in  the morning, and I am very 
grateful to every time nuts gave me useful  information, your proposal has 
strengthened my determination, in fact, I am  also very like SR625, So I will 
to 
find and buy a good shape SR625 for my new  time interval measure instrument.


Thanks again for everyone's  advice, which is very useful to make a choice 
for me. Sorry for not reply  everyone's mail. 


Best Regards!


Hui  Zhang


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] HP53132A vs SR625

2013-02-07 Thread Bob Camp
Hi

Normally a 625 is priced above a 620. Rubidium standards have a finite life. 
You are paying for the Rb when you buy a 625. If the Rb is worn out or nearly 
at end of life, the 625 is not a good idea.

Bob

On Feb 7, 2013, at 7:22 PM, Hui Zhang ba...@163.com wrote:

 Hello Dear Group:
 
 
I am very glad to see so many replies in the morning, and I am very 
 grateful to every time nuts gave me useful information, your proposal has 
 strengthened my determination, in fact, I am also very like SR625, So I will 
 to find and buy a good shape SR625 for my new time interval measure 
 instrument.
 
 
 Thanks again for everyone's advice, which is very useful to make a choice for 
 me. Sorry for not reply everyone's mail. 
 
 
 Best Regards!
 
 
 Hui Zhang
 
 
 
 
 
 
 
 
 
 
 
 
 At 2013-02-08 05:12:37,Brooke Clarke bro...@pacific.net wrote:
 Hi Hui:
 
 The HP 53132A is called a Universal Frequency Counter can count frequency 
 at 12 digits per second, but that is not the 
 case when it is measuring time intervals.  The human interface might be 
 called user hostile.
 
 The front panel of the SR 620 is marked Universal Time Interval Counter 
 and is great for making time interval 
 measurements. That's why I traded my 53132 for the SR620. 
 http://www.prc68.com/I/TandFTE.shtml#SR620
 
 If you're doing Time Nuts measurements you probably have a good 10 MHz 
 oscillator that you can feed into either of these 
 counters so getting an optional time base in one of them serves little 
 purpose.
 
 Have Fun,
 
 Brooke Clarke
 http://www.PRC68.com
 http://www.end2partygovernment.com/2012Issues.html
 
 Hui Zhang wrote:
 Dear Group:
 
 
 I am intend to buy a second-hand counter, I have not decided yet 
 between SR625 and HP53132A, they have very close price in surplus market of 
 here, so, which is better choice? Any suggestion will be appreciated.
 
 
 Best Regards!
 
 
 Hui Zhang
 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to 
 https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.
 
 
 
 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.
 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.