Re: [PSES] Hipot with in-circuit voltage limiting devices IEC 61010-1 3rd Ed.

2012-05-14 Thread Ralph . McDiarmid
| CANADA | Regulatory Compliance Engineering From: Richard Nute ri...@ieee.org To: EMC-PSTC@LISTSERV.IEEE.ORG Date: 05/12/2012 05:33 PM Subject: Re: [PSES] Hipot with in-circuit voltage limiting devices IEC 61010-1 3rd Ed. Hi Mick: The problem is the one originally described

Re: [PSES] Hipot with in-circuit voltage limiting devices IEC 61010-1 3rd Ed.

2012-05-12 Thread Michael Maytum
Richard, To be clear, what problem and subject correctly are you referring to? The IEC 61010-1 3rd Ed problem is specifying a test voltage of 0.9xclamping voltage. From memory Edition 1 of IEC 62368-1 never specified a test voltage of this level. Yes IEC 62368-1 had a twice AC mains

Re: [PSES] Hipot with in-circuit voltage limiting devices IEC 61010-1 3rd Ed.

2012-05-12 Thread John Woodgate
In message 018DCF4B527741149F102EFC42E21D42@RichardHPdv6, dated Fri, 11 May 2012, Richard Nute ri...@bendbroadband.com writes: Would you like to check the latest draft of IEC 62368-1 to see if it treats the subject correctly? With respect to this problem, yes, IEC 62368-1 treats the

Re: [PSES] Hipot with in-circuit voltage limiting devices IEC 61010-1 3rd Ed.

2012-05-12 Thread Richard Nute
] Hipot with in-circuit voltage limiting devices IEC 61010-1 3rd Ed. Richard, To be clear, what problem and subject correctly are you referring to? The IEC 61010-1 3rd Ed problem is specifying a test voltage of 0.9xclamping voltage. From memory Edition 1 of IEC 62368-1 never

Re: [PSES] Hipot with in-circuit voltage limiting devices IEC 61010-1 3rd Ed.

2012-05-12 Thread Richard Nute
Hello John: With respect to this problem, yes, IEC 62368-1 treats the subject correctly. So it could be used as evidence to persuade TC66 to reconsider. Yes. The 2nd Edition of IEC 62368-1 refers to IEC 61051. TC66 should first identify the safety issue associated with varistors (fire)

Re: [PSES] Hipot with in-circuit voltage limiting devices IEC 61010-1 3rd Ed.

2012-05-11 Thread Michael Maytum
Doug, I have received a reply from TC 66 that they will consider the clause F.3.2 problem at their next meeting in October. They have acknowledge that the usage of the term “clamping voltage” is imprecise. They would be receptive to any further inputs on this matter. I used the IEC system

Re: [PSES] Hipot with in-circuit voltage limiting devices IEC 61010-1 3rd Ed.

2012-05-11 Thread John Woodgate
In message 5168965687470166.wa.mjmaytumgmail@listserv.ieee.org, dated Fri, 11 May 2012, Michael Maytum mjmay...@gmail.com writes: I have received a reply from TC 66 that they will consider the clause F.3.2 problem at their next meeting in October. They have acknowledge that the usage of

Re: [PSES] Hipot with in-circuit voltage limiting devices IEC 61010-1 3rd Ed.

2012-05-11 Thread Michael Maytum
John I would be interested to see the latest draft of IEC 62368-1. I assume this is Edition 2. On surge protective components the original IEC 62368 introduced a safety test run at twice the the nominal AC voltage with decreasing values of series resistance. I would like to see a

Re: [PSES] Hipot with in-circuit voltage limiting devices IEC 61010-1 3rd Ed.

2012-05-11 Thread Richard Nute
Would you like to check the latest draft of IEC 62368-1 to see if it treats the subject correctly? With respect to this problem, yes, IEC 62368-1 treats the subject correctly. Best regards, Richard Nute Product Safety Consultant Bend, Oregon, U.S.A. -

Re: [PSES] Hipot with in-circuit voltage limiting devices IEC 61010-1 3rd Ed.

2012-05-10 Thread Mick Maytum
Doug, My previous message gave the IEC rulings - terms clamping voltage and nominal varistor voltage - for two specific points on the MOV clamping (clipping) characteristic. For completeness IEC 61010-1, ed. 3.0 (2010-06) defines the following: *working voltage * highest r.m.s. value

Re: [PSES] Hipot with in-circuit voltage limiting devices IEC 61010-1 3rd Ed.

2012-05-10 Thread Doug Powell
Mick, Sorry I didn't review this email before I just sent my last. Clearly you see the same problem as I. I have to wonder if the committee in reviewing this clause only used data sheets and did not validate in the lab. Or is it possibly the committee members just missed this? I would like to

Re: [PSES] Hipot with in-circuit voltage limiting devices IEC 61010-1 3rd Ed.

2012-05-10 Thread John Woodgate
In message CAByvTVN7XfQWJQQFE0hzQv7QfDpzN9G6ykcTQ1gr=2ofguf...@mail.gmail.com, dated Thu, 10 May 2012, Doug Powell doug...@gmail.com writes: I would put a graphical clip of the curves in this email with areas of concern highlighted except I remember this is prohibited on the listserv. You

Re: [PSES] Hipot with in-circuit voltage limiting devices IEC 61010-1 3rd Ed.

2012-05-10 Thread Mick Maytum
Doug, I wouldn't credit the TC 66 has having MOV component knowledge, just perceptions. Since my last message I have sent messages to the Chair and Secretary of TC 66 wearing my IEC SC 37B chair hat. I was restrained for a change, merely stating they were breaking the laws of

Re: [PSES] Hipot with in-circuit voltage limiting devices IEC 61010-1 3rd Ed.

2012-05-10 Thread Kunde, Brian
...@ieee.org] On Behalf Of Mick Maytum Sent: Thursday, May 10, 2012 12:00 PM To: EMC-PSTC@LISTSERV.IEEE.ORG Cc: Doug Powell Subject: Re: [PSES] Hipot with in-circuit voltage limiting devices IEC 61010-1 3rd Ed. Doug, I wouldn't credit the TC 66 has having MOV component knowledge, just perceptions

Re: [PSES] Hipot with in-circuit voltage limiting devices IEC 61010-1 3rd Ed.

2012-05-10 Thread John Woodgate
In message 64D32EE8B9CBDD44963ACB076A5F6ABB0261DE06@Mailbox-Tech.lecotech.local, dated Thu, 10 May 2012, Kunde, Brian brian_ku...@lecotc.com writes: F.3.2 says, ...test of F.3.1 can be carried out.. Because it says can and not shall does this make this test optional? It seems, from the

Re: [PSES] Hipot with in-circuit voltage limiting devices IEC 61010-1 3rd Ed.

2012-05-10 Thread Doug Powell
, 2012 12:00 PM *To:* EMC-PSTC@LISTSERV.IEEE.ORG *Cc:* Doug Powell *Subject:* Re: [PSES] Hipot with in-circuit voltage limiting devices IEC 61010-1 3rd Ed. Doug, I wouldn't credit the TC 66 has having MOV component knowledge, just perceptions. Since my last message I have sent

Re: [PSES] Hipot with in-circuit voltage limiting devices IEC 61010-1 3rd Ed.

2012-05-10 Thread Doug Powell
The routine voltage test on mains is mandatory, but the method has options. This is how I read the standard. -- Thanks, -doug Douglas E Powell doug...@gmail.com http://www.linkedin.com/in/dougp01 On Thu, May 10, 2012 at 2:04 PM, John Woodgate j...@jmwa.demon.co.uk wrote: In message

[PSES] Hipot with in-circuit voltage limiting devices IEC 61010-1 3rd Ed.

2012-05-09 Thread Doug Powell
A change was made in 3rd Ed. for routine mains hipot tests while clamping devices are still in the circuit; specifically clause F.3.2 (Ed. 3). The standard states the test can be carried out at 0.9 times the clamping voltage of the device and not less than twice the working voltage. Edition 2

Re: [PSES] Hipot with in-circuit voltage limiting devices IEC 61010-1 3rd Ed.

2012-05-09 Thread Mick Maytum
Clamping voltage is the MOV voltage under specified surge conditions. Typically that will be an 8/20 impulse. IEC 61643-331, ed. 1.0 (2003-05) clamping voltage VC peak voltage across the MOV measured under conditions of a specified peak pulse current (IP) and specified waveform Your 1 mA