Great comments.
I can confirm the oven behaviors. The out is not needed. It is controlled
from the micro and warms slowly at about 17 volts. Gets to 150 F after a
solid hour as long as its insulated other wise it has a hard time keeping
up with ambient air. The 10811 does heat quickly to the 85C.
Hi
The “outer oven” on the Z3801 10811’s is simply a warmup heater to improve
performance when the unit
starts at -40C. When operated as designed, it is effectively “gone” in normal
operation. The 10811’s normal
oven circuit (“inner oven”) on the modified 10811 is still what does all of the
I had a reference problem on my Z3801A years ago, and ultimately found
that the opamp that controls the 10811 oven temperature was bad. I think
it turned out there was a bad batch of certain date codes. Replacing the
IC with an equivalent type fixed it right up, with no other changes or
Hi
If it is a crystal, a broken seal on the can is a pretty good guess. Often this
runs up
the resistance as well.
Bob
> On Dec 15, 2016, at 8:57 PM, paul swed wrote:
>
> Bob
> Took your advise and did check all of the caps in teh oscillator along with
> resistors. All
Bob
Took your advise and did check all of the caps in teh oscillator along with
resistors. All are as they should be. Though they do differ from the
schematic I have for the HP 10811. But thats normal reality on HP stuff. So
it is a sick 10811.
So at this stage I have jacked a external 10 MHz
Hi
Crystals do indeed fail. In the case of 40 Hz low, I’d look for a shorted cap
in series with the crystal.
Bob
> On Dec 14, 2016, at 5:23 PM, paul swed wrote:
>
> I have a hp z3801 that is spare parts it never worked correctly. Did some
> digging and at least one major
I have a hp z3801 that is spare parts it never worked correctly. Did some
digging and at least one major issue is the actually HP 10811 is 38 Hz low.
It needs to be within a Hz. Checked everything feeding voltage and such and
they are solid at 5.7V from the regulator. The varicap works as
I also added the HP10811 double oven board to Dirty PCB's shared projects. 10
boards will cost you a whopping $25 (They usually send you couple of extras if
they had good yields).
http://dirtypcbs.com/view.php?share=22492=f5a69b7619ce8f139bb9d1ea789ad45e
I got a couple of requests for the schematic... apparently the old link is
dead. Here is Warren's original drawing. Buy the man a beer... The only
change that I made was using a single high-beta (say 1000) transistor to drive
the heater. Warren used two transistors. All the part numbers
I put my circuit board layout (Eagle .BRD file) for Warren Sarkison's outer
oven controller for the HP-10811 double oven oscillator onto OSHPARK.COM's
shared projects. You can order boards from OSHPARK or download the .BRD file
and send it to your favorite board maker.
Warren,
What is the purpose of "Unseal the internal tuning nut by loosing it ~
1/2 turn"?
And which internal tuning nut are you referring to?
Cheers,
Corby
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
.
But if you want nut performance, best not to use the same supply for them.
I start with a 15V supply, so that there is no interaction between circuits.
ws
*
From: Richard (Rick) Subject: Re: [time-nuts] HP 10811
On 2/11/2016 2:56 PM, ws at Yahoo via time-nuts wrote:
>
Joe
The inner oven voltage needs to be stable! To better than 0.1V.
Unlike the single oven unit, the inner oven on the dual oven unit runs fine
at 15 Volts. It draws a couple hundred ma after warm-up.
One thing that improves the performance of a dual oven 10811 when being used
in a standalone
I have a spare 10811 that I want to setup as a standalone reference.
Since it is a double oven unit, should I put it and the power supply
in just a single walled plastic or metal box? Should I make air vents
or not? Going to extremes, can I put the OCXO inside a thick Styrofoam
box with no air
In message 6B73B186D5BB46E497D773717A4A03D1@pc52, Tom Van Baak writes:
Wonderfully, if you google for:
HP 10811-60165 pinouts
you get a bunch of time-nuts pages with the info you want.
Isn't it high time we create a time-nuts wiki somewhere ?
--
Poul-Henning Kamp | UNIX
Message -
From: KA2WEU--- via time-nuts time-nuts@febo.com
To: time-nuts@febo.com
Sent: Thursday, August 13, 2015 1:14 PM
Subject: [time-nuts] HP 10811-60165 Double Oven
Who can help me with the connection information ?
http://www.ebay.com/itm/HP-10811-60165-Double-Oven-CRYSTAL-OSCILLATOR
Who can help me with the connection information ?
http://www.ebay.com/itm/HP-10811-60165-Double-Oven-CRYSTAL-OSCILLATOR-10Mhz-
/281062658455
Ulrich
In a message dated 8/13/2015 4:10:33 P.M. Eastern Daylight Time,
paulsw...@gmail.com writes:
OK I can contribute at
Of Tom Van Baak
Sent: Thursday, August 13, 2015 4:36 PM
To: Discussion of precise time and frequency measurement
Subject: Re: [time-nuts] HP 10811-60165 Double Oven
Hi Ulrich,
Wonderfully, if you google for:
HP 10811-60165 pinouts
you get a bunch of time-nuts pages with the info you want
Hi,
I'm going through my collection of 10811 variants and listing them on
ebay over the next couple months.
Two 10811B are listed now. Watch user ID corbymite.
Cheers,
Corby Dawson
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go
I have acquired one of these double oven oscillators. Seems like a useful
thing. Need info on pin outs, and operating voltages. Also any other useful
information or links to same would be appreciated. Please reply off list with
any files of specs etc to my email - johncr...@aol.com
All help
: [time-nuts] HP 10811 - 60165 Info Request
I have acquired one of these double oven oscillators. Seems like a useful
thing. Need info on pin outs, and operating voltages. Also any other useful
information or links to same would be appreciated. Please reply off list with
any files of specs etc
Thanks to TVB for hosting this manual containing info for most models of the
venerable 10811.
http://www.leapsecond.com/museum/10811a/10811a.pdf
Thomas Knox
To: time-nuts@febo.com
From: johncr...@aol.com
Date: Mon, 10 Mar 2014 14:17:58 -0400
Subject: [time-nuts] HP 10811 - 60165 Info
Power (+12V)
3 - ORG Oven Monitor Return (Com)
4 - YEL Oven Monitor Output
5 - GRN Oven Power (+18-24V)
6 - BLU Oven Return (Com)
Thomas Knox
**
To: time-nuts at febo.com
From: johncroos at aol.com
Date: Mon, 10 Mar 2014 14:17:58 -0400
Subject: [time-nuts] HP 10811 - 60165 Info Request
I've seen a pin-out for the outer-oven 6-position connector (2 heater wires,
2 thermistor wires), but I've not found anything on the pin-out of the other
6-position connector. Has anyone come across the pin-outs for the
10811-60165 connectors?
CS
___
Chris,
The 10811-60165 may be similar to the HP 10811-60158
that uses using the following pin-out:
1 - BRN Oscillator Return (Com)
2 - RED Oscillator Power (+12V)
3 - ORG Oven Monitor Return (Com)
4 - YEL Oven Monitor Output
5 - GRN Oven Power (+18-24V)
6 - BLU Oven Return (Com)
The following
...@febo.com]On
Behalf Of Richard H McCorkle
Sent: Tuesday, February 18, 2014 11:07 AM
To: Discussion of precise time and frequency measurement
Subject: Re: [time-nuts] HP 10811-60165 Double Oven Crystal Oscillator
Pin-Outs
Chris,
The 10811-60165 may be similar to the HP 10811-60158
that uses using
or
somewhere else?
CS
-Original Message-
From: time-nuts-boun...@febo.com [mailto:time-nuts-boun...@febo.com]On
Behalf Of Richard H McCorkle
Sent: Tuesday, February 18, 2014 11:07 AM
To: Discussion of precise time and frequency measurement
Subject: Re: [time-nuts] HP 10811-60165 Double Oven
On 26/07/13 22:16, Bob Camp wrote:
Hi
Well if you get desperate to try it, I have CO2 and a chamber sitting in the
garage. The biggest issue is that if we use too much CO2 for the chamber, the
beer may go flat…
Can't have that in the name of science, we need good beer after the
Hi
On Jul 31, 2013, at 5:23 AM, Magnus Danielson mag...@rubidium.dyndns.org
wrote:
On 26/07/13 22:16, Bob Camp wrote:
Hi
Well if you get desperate to try it, I have CO2 and a chamber sitting in the
garage. The biggest issue is that if we use too much CO2 for the chamber,
the beer may
Hi
Ok, since it's moved *down* and it's about on temperature there are a couple
of likely possibilities:
1) The crystal leaks and there's air in it
2) The trap coil(s) / capacitor(s) have changed value
3) The varicap is shorted / forward biased
Normal factory troubleshoot (after checking
Good comments. Indeed I will order up a new xtal from digi-key. They should
be in stock any day now. I do agree with the comment on the cap in
parallel. I had not thought about the series cap dropping and there is a
.1uf as I recall. Need to look. The varicap seems to work very well so
don't
Hi
HP spec'd the crystals to be at a 1x10^-8 / C slope at the marked temperature
on the crystal. The center of the curve should be in the vicinity of 90C.
You are ~ -4.5 x 10^-6 low, so that would be 10 C away from that point going
cold and a bit further going hot. Since the oven is at
sounds like time to add a R across the temp set R. That will lower the R
and raise the temp. Its also easy to actually do.
Thanks
On Fri, Jul 26, 2013 at 12:08 PM, Bob Camp li...@rtty.us wrote:
Hi
HP spec'd the crystals to be at a 1x10^-8 / C slope at the marked
temperature on the crystal.
Hi Paul,
Did you see this page that talks about tweaking the 10811 oven controller?
http://www.realhamradio.com/z3801a-turning-point.htm
Ed
On 7/26/2013 10:47 AM, paul swed wrote:
sounds like time to add a R across the temp set R. That will lower the R
and raise the temp. Its also easy to
Yes thanks ran across it. Skimmed it. Did not think I would be doing it. :-)
On Fri, Jul 26, 2013 at 1:15 PM, Ed Palmer ed_pal...@sasktel.net wrote:
Hi Paul,
Did you see this page that talks about tweaking the 10811 oven controller?
Ed Palmer wrote:
Hi Paul,
Did you see this page that talks about tweaking the 10811 oven controller?
http://www.realhamradio.com/z3801a-turning-point.htm
Ed
Although this article anecdotally claims success, it overlooks
a number of important issues. First of all, a large proportion
of
Rick
Thanks. But I am far simpler than that. I am trying to get the 10811 back
to 10 Mhz its -45Hz. In the long thread above I go into the troubleshooting
details so far. Because a crystal simply can't go this bad. :-) I hope not
at least.
Regards
Paul.
On Fri, Jul 26, 2013 at 1:38 PM, Rick
Hi
Doing any real adjustment on these OCXO's pretty much requires you have a
temperature chamber and a pretty good standard. A GPSDO will do for the
standard, the chamber is a bit harder to come by.
Bob
On Jul 26, 2013, at 1:38 PM, Rick Karlquist rich...@karlquist.com wrote:
Ed Palmer
So laying a sock over the insulated oven can doesn't count? Its a thick
sock.
Still not at the stage of needing a chamber I am 45 Hz off.
On Fri, Jul 26, 2013 at 2:36 PM, Bob Camp li...@rtty.us wrote:
Hi
Doing any real adjustment on these OCXO's pretty much requires you have a
temperature
Hi
Tossing a nice fluffy towel over it works pretty well. Just the thing if you
are doing a bench turn hunt on an AT based OCXO. It's also a quick way to
answer the age old question is it broke or is it just all the people walking
by while I'm trying to test it.
You need something that will
Bob
I think the power bill exceeds the cost of a 10811. The chamber and CO2.
Going to pass.Though I am sure I can run down to work and throw a unit in
the various walk in units we have. Suspect they will not appreciate me
adjusting the temps though.
Back to work as they say.
Regards
Paul
WB8TSL
Hi
Well if you get desperate to try it, I have CO2 and a chamber sitting in the
garage. The biggest issue is that if we use too much CO2 for the chamber, the
beer may go flat…
Bob
On Jul 26, 2013, at 4:09 PM, paul swed paulsw...@gmail.com wrote:
Bob
I think the power bill exceeds the cost
and frequency measurement time-nuts@febo.com
Sent: Wednesday, July 24, 2013 9:29 PM
Subject: Re: [time-nuts] HP 10811 update
Well learned quite a bit Q4 does have an issue. Further there is a reason
it has such high Beta. Its must deliver 30 ma to the oscillator and buffer.
I was quite
...@gmail.com
To: Discussion of precise time and frequency measurement
time-nuts@febo.com
Sent: Wednesday, July 24, 2013 9:29 PM
Subject: Re: [time-nuts] HP 10811 update
Well learned quite a bit Q4 does have an issue. Further there is a reason
it has such high Beta. Its must deliver 30 ma
To: Discussion of precise time and frequency measurement
Subject: Re: [time-nuts] HP 10811 update
I find it hard to believe that NTE spec the 123AP to replace a transistor
with min beta of 500...
This is quite entertaining:
http://www.vetco.net/catalog/product_info.php?products_id=1856
Look at the parts it's
and frequency measurement
Subject: Re: [time-nuts] HP 10811 update
I find it hard to believe that NTE spec the 123AP to replace a transistor
with min beta of 500...
This is quite entertaining:
http://www.vetco.net/catalog/product_info.php?products_id=1856
Look at the parts it's supposed
paul swed wrote:
I can't believe you found the transistor. When I pulled it out last night,
its actually a MPSA18!!! I had not had time to look it up but figured it
was a ebay leftover hunt. :-)
At that price I may order 20 of them. Like the gain.
Toshiba used to make a transistor with even
measurement
Subject: Re: [time-nuts] HP 10811 update
Well learned quite a bit Q4 does have an issue. Further there is a reason
it
has such high Beta. Its must deliver 30 ma to the oscillator and buffer.
I was quite surprised by this current level. I was guessing the
oscillator
was a few mils
Well lucks not so good do have the voltage closer but as Bob C said. That
ain't the issue.
Darn I hate when he is right. By the way the oscillator draws 26 ma approx
as a reference. Reassembled everything and let the oven heat up it settles
at 81C 15 min after start and within a respectable range
On 7/25/2013 8:32 PM, paul swed wrote:
Well lucks not so good do have the voltage closer but as Bob C said. That
ain't the issue.
this may is a dumb question:
Lower than expected frequency means some extra capacitance in parallel
or not enough in series with the crystal... ?
Would that put
Hello to the group. Numbers of threads running on the HP 10811. Mine is 45
Hz low. The oven is most likely ok. I did get the can open per the various
other threads and the web pix. Started testing the circuits and have found
that the regulator for the oscillator isn't. Its supposed to produce 5.7V
Thanks Charles. I suspect all said the regulated output delivers 1 ma. It
doesn't feed much.
I have the 2n3904s so cheap test to see if that picks up the 45 Hz
On Wed, Jul 24, 2013 at 3:20 PM, Charles P. Steinmetz
charles_steinm...@lavabit.com wrote:
Paul wrote:
Q4 a 2n6429. What
2N6429... $1 each:
http://www.talonix.com/shop/item.aspx?itemid=3522
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the
Hi
The regulator should be fixed, and the OCXO will work better with it repaired.
That said, shifting the regulated voltage from 5.7 to 5.27 volts should not
shift that oscillator 45 Hz.
Bob
On Jul 24, 2013, at 9:00 AM, paul swed paulsw...@gmail.com wrote:
Hello to the group. Numbers of
I can hope
I will embed a k thermocouple also.
On Wed, Jul 24, 2013 at 7:45 PM, Bob Camp li...@rtty.us wrote:
Hi
The regulator should be fixed, and the OCXO will work better with it
repaired. That said, shifting the regulated voltage from 5.7 to 5.27 volts
should not shift that oscillator
Well learned quite a bit Q4 does have an issue. Further there is a reason
it has such high Beta. Its must deliver 30 ma to the oscillator and buffer.
I was quite surprised by this current level. I was guessing the oscillator
was a few mils and the buffer maybe 8.
The 2n3904 simply does not cut
swed
Sent: Thursday, 25 July 2013 11:29 AM
To: Discussion of precise time and frequency measurement
Subject: Re: [time-nuts] HP 10811 update
Well learned quite a bit Q4 does have an issue. Further there is a reason it
has such high Beta. Its must deliver 30 ma to the oscillator and buffer.
I
[mailto:time-nuts-boun...@febo.com] On
Behalf Of paul swed
Sent: Wednesday, July 24, 2013 8:29 PM
To: Discussion of precise time and frequency measurement
Subject: Re: [time-nuts] HP 10811 update
Well learned quite a bit Q4 does have an issue. Further there is a reason it
has such high Beta. Its
and frequency measurement
Subject: Re: [time-nuts] HP 10811 update
Well learned quite a bit Q4 does have an issue. Further there is a reason
it
has such high Beta. Its must deliver 30 ma to the oscillator and buffer.
I was quite surprised by this current level. I was guessing the oscillator
was a few
into the problem any farther.
Anyone have any more info or ideas?
73,
Tom WB6UZZ
--- On Thu, 10/18/12, John Miles jmi...@pop.net wrote:
From: John Miles jmi...@pop.net
Subject: Re: [time-nuts] HP 10811 Tuning Diode
To: 'Discussion of precise time and frequency measurement'
time-nuts
Of Adrian
Sent: Thursday, October 18, 2012 12:19 PM
To: Discussion of precise time and frequency measurement
Subject: [time-nuts] HP 10811 Tuning Diode
Hello,
I found the 10811 in my 8662A to be quite noisy and was able to trace
the problem down to the tuning diode.
Actually, the phase noise started
: time-nuts-boun...@febo.com [mailto:time-nuts-boun...@febo.com] On
Behalf Of John Miles
Sent: Thursday, October 18, 2012 11:55 PM
To: 'Discussion of precise time and frequency measurement'
Subject: Re: [time-nuts] HP 10811 Tuning Diode
That's extremely interesting, Adrian. I've never heard
I wonder if the glass case got a crack in it or if the kovar seal
was failing. Maybe a failed solder joint (which gets fixed when you
install the new diode). I've never heard of a noisy varactor before either.
Rick
___
time-nuts mailing list --
...@febo.com
[mailto:time-nuts-boun...@febo.com] On Behalf Of Rick Karlquist
Sent: Friday, October 19, 2012 11:01
To: Discussion of precise time and frequency measurement
Subject: Re: [time-nuts] HP 10811 Tuning Diode
I wonder if the glass case got a crack in it or if the kovar seal
was failing
Slightly off topic: the oscillator transistor has some 1854-
part number, but is actually a 2N5179 selected for some minimum
fT at 30 mADC collector current. When the oscillator starts up,
before the ALC takes effect, a fairly large DC current can flow
and some 2N5179's don't have enough
I didn't notice any damage, nor was there a visibly bad solder joint.
However, I couldn't reproduce the excess noise with a quick bench test
at 30V / 100k neither.
DC-wise there are a few microvolts only. Same with the SMD diode.
Exept, when the original glass diode is exposed to light. With a
Of Adrian
Sent: Friday, October 19, 2012 4:10 PM
To: rich...@karlquist.com; Discussion of precise time and frequency
measurement
Subject: Re: [time-nuts] HP 10811 Tuning Diode
I didn't notice any damage, nor was there a visibly bad solder joint.
However, I couldn't reproduce the excess noise
As I mentioned, the varactor is equivalent to the Motorola MV1650. I
have a bunch as well left over from building VCXOs. :-)
David
On 10/19/12 3:37 PM, Rick Karlquist wrote:
Slightly off topic: the oscillator transistor has some 1854-
part number, but is actually a 2N5179 selected for
, October 19, 2012 9:10 PM
Subject: Re: [time-nuts] HP 10811 Tuning Diode
I didn't notice any damage, nor was there a visibly bad solder joint.
However, I couldn't reproduce the excess noise with a quick bench test at
30V / 100k neither.
DC-wise there are a few microvolts only. Same with the SMD
[mailto:time-nuts-boun...@febo.com] On Behalf Of Bob Camp
Sent: Friday, October 19, 2012 13:20
To: 'Discussion of precise time and frequency measurement'
Subject: Re: [time-nuts] HP 10811 Tuning Diode
Hi
The normal 10811 package has several openings in it. Light
*might* get
Hello,
I found the 10811 in my 8662A to be quite noisy and was able to trace
the problem down to the tuning diode.
Actually, the phase noise started to jump when the oven heated up. As a
quick solution, I replaced it with a SMD type that I had at hands.
I would appreciate any hint where to
That would be a Motorola MV1650 or equivalent. You may have to select
for 100pf +/- 5% @ 4V as the standard spec is 10% but it probably does
not matter as the tuning ratio will be greater than the 2 spec.
David
On 10/18/12 3:18 PM, Adrian wrote:
Hello,
I found the 10811 in my 8662A to be
Probably unobtainium.
0122-0244 is shown as Motorola SMV315-24 or MSI Electronics SQ529.
For whatever that's worth. Reminds me of a popular audio amplifier
design that was in Popular Electronics using Motorola SS1122 diodes.
Some special non catalog diode. Turns out Dan Myers, the designer,
Subject: [time-nuts] HP 10811 Tuning Diode
Hello,
I found the 10811 in my 8662A to be quite noisy and was able to trace
the problem down to the tuning diode.
Actually, the phase noise started to jump when the oven heated up. As a
quick solution, I replaced it with a SMD type that I had at hands
gotten around to digging into the problem any farther.
Anyone have any more info or ideas?
73,
Tom WB6UZZ
--- On Thu, 10/18/12, John Miles jmi...@pop.net wrote:
From: John Miles jmi...@pop.net
Subject: Re: [time-nuts] HP 10811 Tuning Diode
To: 'Discussion of precise time and frequency
Congratulations on a job well done and a saved prize piece of technology.
John WA4WDL
--
From: Frank Stellmach frank.stellm...@freenet.de
Sent: Saturday, January 07, 2012 9:17 AM
To: time-nuts@febo.com
Subject: [time-nuts] HP 10811-60111
Does your Mode B 10811 use a different xtral or same xtral, different
frequency?
What does Mode B do better, ... and worse?
Why was it not used in the 10811?
When trouble shooting a freq stability problem, usually the more data the
better.
What can really help to narrow down the cause or see
WarrenS wrote:
Does your Mode B 10811 use a different xtral or same xtral, different
frequency?
What does Mode B do better, ... and worse?
Why was it not used in the 10811?
Mode B is the same crystal on a different frequency
(about 10.95 MHz). It does not have a turnover and
the tempco is
On 23/09/11 05:40, Richard (Rick) Karlquist wrote:
On 9/22/2011 5:17 PM, ws at Yahoo wrote:
Within minutes the frequency changed more than the spec
For humidity to get thru something like that it takes weeks or more it
does it at all.
That fast of reaction, Sure sounds like some other effect
I have made plots of the effects of everything I can find that effects the
freq of a HP10811.
Most things are much slower than minutes, more like an hour time constant,
such as anything effecting the outside case's temperature
OR the effect is much faster than minutes, things such as voltage
WarrenS wrote:
I have to wonder if the unit being tested had its high impedance oven
control points lifted off the PCB board and on Teflon standoffs like the
production units?
ws
It was a production unit, no modifications whatsoever.
The oven change is an interesting theory; I never
Magnus Danielson wrote:
On 23/09/11 05:40, Richard (Rick) Karlquist wrote:
Blowing on the case? Did you consider the increased cooling that higher
humidity provides?
Cheers,
Magnus
Against, the frequency change we saw was considerably more
than what we got from changing the temperature
Maybe you have a bad joint in the tuning circuit and the humidity makes
the actual tuning voltage vary.
-John
===
Magnus Danielson wrote:
On 23/09/11 05:40, Richard (Rick) Karlquist wrote:
Blowing on the case? Did you consider the increased cooling that higher
humidity
On 23/09/11 23:25, Rick Karlquist wrote:
Magnus Danielson wrote:
On 23/09/11 05:40, Richard (Rick) Karlquist wrote:
Blowing on the case? Did you consider the increased cooling that higher
humidity provides?
Cheers,
Magnus
Against, the frequency change we saw was considerably more
than
Esteemed Colleagues,
I will attempt to make a coherent technical reply to the remarks made about my
10811 posting.
But first a bit of qualification. My remarks refer to the 10811 and that
vintage of OCXO’s.
Wrote: Doing what you describe will result in a very sensitive humidity sensor,
Perry Sandeen wrote:
Wrote: Doing what you describe will result in a very sensitive humidity
sensor, having eliminated the thermometer effect.
I do not understand. I believed that since the OCXO temperature will be
substationally higher than the surrounding temps, any residual moisture
You had a leak. If epoxy was really as bad as you indicate,
it would not be usable for holding pressure, or mild vacuum,
and yet it is. Somehow, someway you left a big hole in the
bucket.
-Chuck Harris
Rick Karlquist wrote:
Perry Sandeen wrote:
Wrote: Doing what you describe will result in
A problem I've seen when using the hermetic sealed soldered version of the
10811 from a dual oven unit,
is when the case is sealed the osc makes a good barometer because of changes
in its case due to barometric changes.
A 1 inch difference (such as 30 to 29) caused something on the order of
Within minutes the frequency changed more than the spec
For humidity to get thru something like that it takes weeks or more it does
it at all.
That fast of reaction, Sure sounds like some other effect like blowing a
little air on the case or loading the osc output with water in the output
Humidity is a confusing subject to many engineers and scientists.
Unlike most parameters it is a quantity with two input variables,
concentration and temperature. There are many ways to combine
these to give different units.
As a research scientist I spent most of my career working with
composite
On 9/22/2011 5:17 PM, ws at Yahoo wrote:
Within minutes the frequency changed more than the spec
For humidity to get thru something like that it takes weeks or more it
does it at all.
That fast of reaction, Sure sounds like some other effect like blowing a
little air on the case or loading the
List,
Wrote: It was not cheap, it was not high performance, it was not elegant. But
it did get product shipped on time.
“In the land of the blind the one eyed man is king”
Rick,
All your statements are most likely 100 percent accurate. You are the expert.
The problem for the rest of us is
List,
Wrote: I have no problems with my 10811. I just want to experiment with ways
to improve performance, right now I have several that are in the E-13
YES!
Regards,
Perrier
___
time-nuts mailing list -- time-nuts@febo.com
To
I have in the past reversed the varactor, grounding R2 to get a positive
tuning voltage from 1.5 to 12 Volt. Also brought out the 5.7 volt for
reference purposes. That way every thing was positive. Is easy to get to. But
now I have some very good units and I am worried I may change
] On
Behalf Of ewkeh...@aol.com
Sent: Friday, October 29, 2010 5:00 AM
To: time-nuts@febo.com
Subject: [time-nuts] HP 10811 EFC
I have in the past reversed the varactor, grounding R2 to get a positive
tuning voltage from 1.5 to 12 Volt. Also brought out the 5.7 volt for
reference purposes
-boun...@febo.com [mailto:time-nuts-boun...@febo.com] On
Behalf Of ewkeh...@aol.com
Sent: Friday, October 29, 2010 5:00 AM
To: time-nuts@febo.com
Subject: [time-nuts] HP 10811 EFC
I have in the past reversed the varactor, grounding R2 to get a positive
tuning voltage from 1.5 to 12 Volt. Also
I plan on replacing the Xtal Osc. of my HP 5062C with a HP 10811 since I
have a few ones with AV less than 1E-12 from 1 to 100 sec. I would like to
use the +18 Volt in the unit for heater. Has any one experience with running
the 10811 below 20 Volt, any thing I should look out for?
Thanks
I think you should probably be fine at 18V. The oven will run down
to at least 15V and maybe 12V.
Rick Karlquist N6RK
On 10/19/2010 5:27 AM, ewkeh...@aol.com wrote:
I plan on replacing the Xtal Osc. of my HP 5062C with a HP 10811 since I
have a few ones with AV less than 1E-12 from 1 to 100
Thank you Rick, you obviously must know.
Bert Kehren
In a message dated 10/19/2010 12:02:11 P.M. Eastern Daylight Time,
rich...@karlquist.com writes:
I think you should probably be fine at 18V. The oven will run down
to at least 15V and maybe 12V.
Rick Karlquist N6RK
On 10/19/2010 5:27
The 10811-60111 that came in my 5370B counter turned out to have an open
circuit Thermistor in it. This is a non-replaceable part in theory.
However, nothing to lose, I have successfully substituted a Digikey Part
number: 317-1371 100 kOhm 1% Thermistor ( from Cantherm, part number:
1 - 100 of 134 matches
Mail list logo