I would recommend leaving it on. You want to make sure it soaks as long as
possible and then test it before shipping.
Keysight soaks the 3458A before testing and it makes sense since you want
to calibrate the meters as if they are in use 24/7.
Inspect the supplies and make sure you have no
Don,
I have not negotiated the agreement but am happy with the cost. You will be
too if the meter fails cal and needs repair. I paid for the 5 years of
service up front and the calibration agreement is paid annually. I found it
easier to do it this way. I changed the calibration level to the one
Bill,
The reason why Q14 was dropped was that there was an additional pcb added
to the reference. The A8 preheater did not exist on the older models. It
appears the heating was changed to a split design where the heaters were
divided into two groups.
The control line still goes out P2-1 but
Randy,
There appears to be no MOQ on those bulbs and they have 972 in stock. I
think you can order what you want.
Todd
On Sat, Jan 6, 2018 at 12:34 PM, Randy Evans
wrote:
> Is there a relaible source for the GT-NE6H1925T neons? Mouser has them for
> special order
Randy,
Check that your firmware version on the Prologix is up to date. The latest
USB version is 6.107. You can see it in the Prologix GPIB Configurator
utility that can be downloaded from a link on their website.
http://prologix.biz/resources.html
I have uploaded a pic of my settings shown in
g. All it does is print out "HP3458A" on
> the output console and then just sits there.
>
> Randy
>
> On Sat, Dec 30, 2017 at 7:43 PM, Todd Micallef <tmical...@gmail.com>
> wrote:
>
> > Randy,
> >
> > I just remembered that I changed the dura
Randy,
I just remembered that I changed the duration from 30 minutes to 2 minutes
to speed up the test. I forgot to change it back.
Todd
On Sat, Dec 30, 2017 at 10:27 PM, Todd Micallef <tmical...@gmail.com> wrote:
> Randy,
>
> I just sent it. I haven't modified Martin's cod
gt;
> Sent from my Verizon, Samsung Galaxy smartphone
> Original message ----From: Todd Micallef <tmical...@gmail.com>
> Date: 12/30/17 6:40 PM (GMT-08:00) To: Discussion of precise voltage
> measurement <volt-nuts@febo.com> Subject: Re: [volt-nuts] GPIB Issue wi
ain,
> Randy
>
> Sent from my Verizon, Samsung Galaxy smartphone
> ---- Original message From: Todd Micallef <tmical...@gmail.com>
> Date: 12/30/17 6:13 PM (GMT-08:00) To: Discussion of precise voltage
> measurement <volt-nuts@febo.com> Subject: Re: [volt
ode per your directions (V1.02).
>
> Thanks,
>
> Randy
>
> On Sat, Dec 30, 2017 at 3:55 PM, Todd Micallef <tmical...@gmail.com>
> wrote:
>
> > Randy,
> >
> > I will continue to look for my adapter. I haven't used it for quite a
> > while.
>
ss,TRIGGER_SOURCE);// Hold trigger
> > until meter can be setup
> >EZGPIB_BusWriteData(gpib_address,'INBUF ON');
> >EZGPIB_BusWriteData(gpib_address,AUTO_ZERO);
> >EZGPIB_BusWriteData(gpib_address,OUTPUT_FORMAT);
> > end;
> >
> >
> > Rand
>EZGPIB_BusWriteData(gpib_address,'INBUF ON');
>EZGPIB_BusWriteData(gpib_address,AUTO_ZERO);
>EZGPIB_BusWriteData(gpib_address,OUTPUT_FORMAT);
> end;
>
>
> Randy
>
> On Sat, Dec 30, 2017 at 11:20 AM, Todd Micallef <tmical...@gmail.com>
> wrote:
>
> &
> Todd,
>
> I have version 1.01. Yes, please e-mail me your modified version. Maybe it
> will help.
>
> Thanks,
>
> Randy
>
> On Sat, Dec 30, 2017 at 4:24 AM, Todd Micallef <tmical...@gmail.com>
> wrote:
>
> > Randy,
> >
> > Is there a
Randy,
Is there a version number at the top of the file? I had two versions (1.00
and 1.01). I can't remember all that changed between the two since it has
been a while but I did find an additional timeout code that limited the
wait time for reading data to 1 minute. I am guessing I had a problem
Randy,
I use the GQ-4X from mcumall.com. It works great and I have had no problems
with it. I have bought a few adapters to support PLCC and SOIC packages.
For the NVRAM, don't overlook the STMicroelectronics memory. I have used
the M48Z12 as a direct replacement.with no issues. The only thing I
can't tell by the
photo.
Todd
On Mon, Sep 18, 2017 at 5:44 PM, Dr. David Kirkby (Kirkby Microwave Ltd) <
drkir...@kirkbymicrowave.co.uk> wrote:
> On 18 September 2017 at 18:36, Todd Micallef <tmical...@gmail.com> wrote:
>
> > Dave,
> >
> > Another meter is t
Dave,
Another meter is the Cambridge LOM-510A. I am not sure if it is in your
budget but there has been a review made on EEVBlog. There is one currently
on eBay with a current amplifier that I have never seen before today. It
may be worth reviewing if it meets your needs.
Todd
On Mon, Sep 18,
Igor,
Congratulations on your purchase. Your initial readings on the 10v and
1.018v did not appear too high. It depends on which jumpers were already
selected. I think on all of mine the 40ppm jumper was the only one
installed. The problem with the 1v output might be due to a broken trimmer
pot.
se my friend retires.
>
> Fred
>
> .
>
> On 07/20/2017 04:51 PM, Todd Micallef wrote:
>
>> Fred,
>>
>> Shorted tantalums most likely. I had several on mine that went bad as
>> quickly as I replaced them. I would start with the inguard supply board.
>>
Dave,
I have the same problem. I ordered some connectors from an auction and they
are due to arrive this week. They were ordered as a set since the female
connector was hard to find.
I was going to ask the group if both cables were straight through and if
any of the wires were of any special
I don't have one in my collection but fellow member Dave Partridge had a
similar issue on a 7081. His detailed repair is here
http://www.perdrix.co.uk/Solartron7081/index.htm
Perhaps there are some similarities between the two models that can point
you in the right direction?
If you have a
This is a follow-up to my post last November
https://www.febo.com/pipermail/volt-nuts/2015-November/004562.html
I was able to track down the manufacturer of the keyswitch. It appears that
Lorlin is the manufacturer and the key number is the same.
With the help of another member I was able to
.
On Sun, Jun 26, 2016 at 5:29 PM, Dr. David Kirkby (Kirkby Microwave Ltd) <
drkir...@kirkbymicrowave.co.uk> wrote:
> On 26 June 2016 at 20:46, Todd Micallef <tmical...@gmail.com> wrote:
>
> > David,
> >
> > I have found that many sellers like to remove an
at that TP. Without a schematic, it will take some
> time to determine what is different after the TP652 point - maybe there are
> in fact 2 ADCs, but it's not obvious either way.
>
> Russ
>
> On Sun, Jun 26, 2016 at 11:16 AM, Todd Micallef <tmical...@gmail.com>
> wrote:
&g
David,
I have found that many sellers like to remove any calibration stickers or
tamper evident stickers. Not sure why they do that as a sticker does not
mean the item is calibrated.
If you have 14 days, run it through the paces and perform self-tests. There
are some test scripts here
Andrea,
Thank you for your research. A few years ago I purchased some Pb free
solder from Radio Shack. I recently discovered it is listed as 96/4 solder.
The part number is 640-0025
https://www.radioshack.com/products/lead-free-solder-0-25-oz?variant=5717831877
The MSDS is listed on the
I believe it is listed in the document that comes with Keithley's 2188
connector. There is a brief mention here ...
http://www.signaltestinc.com/v/vspfiles/assets/datasheet/Cnctr-Adpt-Tools_Acc.pdf
Todd
On Sun, Jun 26, 2016 at 12:55 PM, David wrote:
> There was a
I have to review my notes, but I know of the two bad parts that were found, one
was the TF-245. I happened to have a spare on hand. I will look through any
notes to get the other part number.
Todd
Sent from my iPad
> On Jun 26, 2016, at 10:09, Russ Ramirez wrote:
>
>
t; Dave
>
> -Original Message-
> From: volt-nuts [mailto:volt-nuts-boun...@febo.com] On Behalf Of Todd
> Micallef
> Sent: 26 April 2016 21:12
> To: Discussion of precise voltage measurement
> Subject: Re: [volt-nuts] PCB fasteners for Datron meters (David C.
> Partrid
Sure. I have a couple of broken fasteners. What is smallish quantities? I
would be willing to take 10 or more if a minimum quantity is needed.
Todd
On Sun, Apr 24, 2016 at 10:33 AM, David C. Partridge <
david.partri...@perdrix.co.uk
I will get it this evening. Hope to be finished with work within the next
couple of hours.
On Thursday, April 14, 2016, Bert Kachler wrote:
> Hi Todd and ??,
>
> thanks for your answers!
> I think it is a good suggestion to
> try to see first whether the Eproms are intact.
Bert,
I did a test not that long ago for another volt-nut by disconnecting the
cable between the digital and DCV boards. I got the following error:
"9005.9 Pon tru/compl 00173A".
One of mine came with a 9005.7 00150C error and it turned out to be a
disconnected cable on one of the option boards
Try keiconn.com
Kensington Electronics is a distributor, they have a minimum order but you
will probably meet it.
Todd
On Tuesday, March 15, 2016, Pete Lancashire wrote:
> Anyone recommend a North American source ? or have one they are not using ?
>
> And where
I have seen the photos of one in a 3458A go bad. Not a pretty sight but it
looks like the gunk is removable.
The input filters are being changed out on most of my older gear
(Fluke,Keithley,Solartron,Datron) and not all of it is Schaffner. I think
that anything going back that far and is
PeLuLe,
I am currently working on two 1281's. If you have anything you would like
me to try, let me know. I am certain my cal constants are not good on one
of them. The battery was just changed a few days ago and it appears to be
the original from '87. I also had some intermittent NVRAM errors.
inal Message - From: "Todd Micallef" <tmical...@gmail.com>
> To: "Discussion of precise voltage measurement" <volt-nuts@febo.com>
> Sent: Sunday, November 01, 2015 8:23 PM
> Subject: Re: [volt-nuts] Wavetek 1281 - Calibration Key
>
>
> Unfortunat
Unfortunately, when I had my keys made for the 4700, I took the lock out of
a broken 1071 to the locksmith. I originally showed them photos of an
actual key and they looked at me like I was crazy. I could only find two
locksmiths near me for the 4700 and a Valhalla 2724A. Both wanted a lock
before
No problem. I was just looking back at the documents wondering if I missed
anything. The transformer solution is interesting and I see that the 8508A
has a second low voltage transformer (T1). I have no idea if it performs
the same function using a different calibration.
On Sat, Oct 31, 2015 at
Ok, I will look at them again.
Sent from my iPad
> On Oct 30, 2015, at 07:26, Frank Stellmach wrote:
>
> Hello Todd,
>
> many thanks for the reference schematics.
>
> But that's not the SelfCAL module.
>
> This should be capable of doing precise 10:1 ratio
I recently picked up a 1281 that is throwing a couple of errors. While
researching the problem, I looked for the SelfCAL description and did not
see anything conclusive. I will pull out my schematics and see if there are
any references to a particular module. I was under the impression they were
I noticed that some of the vendors listed overall end-end tcr and at least
one that listed the wiper tcr too. Hopefully you can get the datasheets
with both.
Todd
On Saturday, September 26, 2015, David Garrido wrote:
> Thank you for all of the input on this folks.
>
> Does
Dave,
What is the problem with the original reference? I have not heard of too
many failing. There was also a reply from Hank on another message. Perhaps
he can provide the needed parts from a 732A?
Todd
On Mon, Sep 14, 2015 at 6:45 PM, Dave M wrote:
> After making a
Here is a DIY guide to making some lab standards. It is detailed with some
component values.
http://nopr.niscair.res.in/bitstream/123456789/4848/1/JSIR%2065%286%29%20510-513.pdf
On Wed, Aug 19, 2015 at 8:48 PM, Dave M dgmin...@mediacombb.net wrote:
Here's a paragraph from IETLab's web site on
David,
Depending on the time frame needed for these resistors you can order many
(but not all) of the parts from Farnell. I believe the Rhopoint resistors
are the recommended brand. The high value resistors could be Caddock USF
series.
Another alternative would be to contact Edwin Pettis. He
Randy,
I made my own version of the Fluke 5440A-7003 cables here ...
http://us.flukecal.com/products/accessories/test-leads-probes-and-clips/5440a-7003
I bought some Belden 8719 cable, blue Pomona patch cables (B-18-6), and
some gold-plated spade lugs from here
I used a Fluke 5100B and 8920A for my SCAL calibration. I got lucky and it
passed at Agilent. I think if you use a 3325A , you should get results that are
close.
Todd
Sent from my iPad
On May 20, 2015, at 15:46, Electronics and Books via volt-nuts
volt-nuts@febo.com wrote:
It is stated
The 5700A5720A verification test lists a maximum deviation from nominal on
the resistance ranges. Is there any information on how well Fluke's hybrid
resistors are maintaining their specs over the lifetime of the calibrator?
Anyone have access to long term records of the resistance values?
Todd
Frans,
Do you perform an in-house calibration with a 10V reference and 1 ohm 10k
resistors? I have been studying the recommended calibration intervals of
Fluke's 57xx series calibrators and I wonder how well the unit maintains its
stability over long intervals between trips to a cal lab.
Definitely good price. My replacement display board for a Keithley 2400 was
~$480.00 US. I could have replaced a few 3458A boards for a little more.
Agilent typically sent me parts next day for no additional overnight charge.
Keithley/Tek was ~$30 for standard shipping.
The bad thing is that
A Data Proof scanner seems to be the standard of some primary labs. They
don't show up often on the auction site.
This company makes one that fits the front of a 732B. I have never seen it
in use.
http://www.hpd-online.com/reversing_switch.php
There also appears to be a Guildline 9600 up for
There is currently a thread on EEVBlog where one has already been built. I
believe it can be purchased once the final layout is completed.
http://www.eevblog.com/forum/projects/project-kx-diy-calibrator-reference-sourcemeter/
Todd
On Fri, Jan 30, 2015 at 8:24 AM, Dr. David Kirkby (Kirkby
Has anyone had any Gerber file compatibility issues with batch pcb
manufacturers? For example, pad sizes or trace widths not matching what is
on the screen.
Todd
On Mon, Jan 5, 2015 at 4:41 PM, David C. Partridge
david.partri...@perdrix.co.uk wrote:
Lack of forward/back annotation would kill
I believe the 1281 uses the LTZ1000. The datasheet mentions dual references.
My 4920 has both a ltz1000 and a lm399. I think you can find pics of the
1281 dual ref setup on eevblog.
I would not be surprised if the 8508a has a similar configuration.
Todd
On Saturday, December 20, 2014, Richard
Andrea,
If the problem is related to temperature, maybe you can find the offending
part by freezing it?
Todd
On Sat, Dec 13, 2014 at 1:46 PM, Poul-Henning Kamp p...@phk.freebsd.dk
wrote:
In message 20141213183546.ga27...@sol.ermione.com, Andrea Baldoni
writes:
Someone has the
I did the same for my Datron 4700 and Valhalla 2724a. I took a picture of that
ebay listing to the locksmith for a laugh. The lock on the 4700 came out easily
once the outer threaded cap of the lock was removed.
Todd
Sent from my iPad
On Dec 7, 2014, at 11:10, David C. Partridge
I have uploaded the mod file to the ko4bb ftp server. I could not find the
AR or AB manuals so I also added them. Right now they are sitting in the
Recent Uploads directory.
Todd
On Sun, Nov 23, 2014 at 10:31 PM, Dallas Smith dosm...@outlook.com wrote:
Thank you, that will be fine. I'm
Andreas has a good thread on EEVblog
http://www.eevblog.com/forum/projects/t-c-measurements-on-precision-resistors/
I am guessing that you want to calculate the alpha and beta coefficients
over a certain temperature range? I am also interested in doing something
similar, having collected
David,
Congratulations on your new (to you) gear. I had typed a long drawn out
response that was putting me to sleep. I decided to reduce it to a couple
of paragraphs.
The problem with testing / calibrating standards and references is that you
need an even more accurate measurement system, a
panel was removed.
Todd
On Tue, Oct 14, 2014 at 4:52 AM, acb...@gmx.de wrote:
do you know what the part number is for the 7081 connector? I just
purchased a 7071 and would need one.
thanks
Gesendet: Dienstag, 14. Oktober 2014 um 03:26 Uhr
Von: Todd Micallef tmical...@gmail.com
Randy,
I had ordered the connectors from Fluke, but they do not come with the
contacts.
The best place to get one is from Kensington Electronics. They cost me
$9.06 ea. back in March. The shipping costs were not included.
Unfortunately, they have a minimum order and you would need to order about
How about the ones recommended for the 3458a?
http://www.newark.com/agilent-technologies/34137a/test-lead-1kv/dp/01P1238
It is getting harder to find quality leads that fit a standard binding post.
Todd
Sent from my iPad
On Oct 7, 2014, at 16:25, Mark Kahrs mark.ka...@gmail.com wrote:
I wonder if Ametek can be persuaded to release any official calibration /
performance testing procedures if they are no longer supporting the 7081. Does
any one here have any contacts at Ametek?
Todd
Sent from my iPad
On Sep 28, 2014, at 12:14, J. L. Trantham jlt...@att.net wrote:
I know
I wanted to pass on some information that may not have been previously
discussed.
I just finished working on a 732A with minor issues. I replaced the R10 and
R20 pots on the A3 board just like on my other 732A's along with the two
big caps C1 and C3. After performing the same battery charger
I have replaced a fan in one of my 3458A. I used this one ...
http://www.digikey.com/product-detail/en/622L/381-2266-ND/1131198
I contacted ebm-papst with the original part number. They were not able to pull
the original specs to offer a newer sub. I happened to notice it was made in
West
http://www.hparchive.com/appnotes.htm
http://www.analog.com/static/imported-files/application_notes/41727248AN_347.pdf
Sent from my iPad
On Sep 4, 2014, at 9:48, Dave Daniel kc0...@gmail.com wrote:
Do you mean that they are on the web, or that they should be and you haven't
found them,
Mark,
If you need any scans or measurements, let me know. I have one that will be
disassembled for the next 24hrs. I can get this for you tomorrow evening before
I replace the batteries.
Todd
Sent from my iPad
On Aug 25, 2014, at 21:10, Mark Sims hol...@hotmail.com wrote:
I have a 732A
on
the
time - which is correct? For the moment, i am only concerned with
stability. The need for absolute accuracy will come later.
Randy
On Fri, Aug 22, 2014 at 8:06 PM, Todd Micallef tmical...@gmail.com
wrote:
Randy,
You have two possible choices. It can be configured with 4 x
Chris,
Where do you live? I just had some keys made last week for my Datron units.
The first set fit my 4000A and 1071, but they did not fit my 4700. The
4000A key looked more like the pictures online, but the 4700 was completely
different.
Someone may have rekeyed or damaged the lock on the
Randy,
I looked at the 3458a input impedance, and it is difficult to get the meter set
to the FixedZ (10M) mode. However, the 100v range is 10M. It should default to
HiZ on reset.
I forgot to mention to check the outputs and guard resistance to ground. One of
my 732a had some foam baffling
luck.
Joe
-Original Message-
From: volt-nuts-boun...@febo.com [mailto:volt-nuts-boun...@febo.com] On
Behalf Of Todd Micallef
Sent: Saturday, August 23, 2014 4:13 AM
To: Discussion of precise voltage measurement
Subject: Re: [volt-nuts] 732A drift
Randy,
I looked
Never mind. I tested it myself. You should be able to get the in cal led
working without batteries.
Todd
On Sat, Aug 23, 2014 at 7:20 AM, Todd Micallef tmical...@gmail.com wrote:
The output impedance of both the 1v outputs is ~1k ohm. I wonder if Randy
could check the 3458a input impedance
John,
Don't forget to measure your 5V power supply output. I doubt you are having
a problem with it but the NVRAM write is disabled below a certain voltage
(~4.25v)
Todd
On Wed, Aug 20, 2014 at 6:50 PM, acb...@gmx.de wrote:
the easy way to read nvram through usb is a usb-programmer that has
:
Thanks...
I have been looking at the
http://www.ebay.com/itm/Universal-EEPROM-Flash-SPI-BIOS-24-25-BR90-93-USB-Programmer-6000-CHIPS-SP8-B-/371031966902?pt=LH_DefaultDomain_0hash=item56633a84b6
Do you think that would do the chips?
On Wed, Aug 20, 2014 at 4:25 PM, Todd Micallef tmical
Stan,
I have seen this Fluke book mentioned occasionally. I am not sure if has
the tutorials that you seek. It is advertised as specific to dc/ lf ac.
http://www.amazon.com/Calibration-Philosophy-Practice-Fluke-Corporation/dp/0963865005
Todd
On Tue, Aug 19, 2014 at 7:35 PM, Stan Katz
Randy,
Two of mine read in the 3.7-3.8K range, another reads in the 4.2K range. It
is a good sign if the reading dropped in a few hours after turn-on. I
believe the lower thermistor readings indicate the older 732A oven. Some
minor changes were made and both manuals are available on Didier's
Dave,
Thanks for the heads up. I went ahead and verified that all pins matched
the schematic by changing the cam to the four settings and all are as
expected. The new cams have 230vac printed on them and the old one has
220 vac.
I went ahead and installed the old one to prevent any future issues
Randy,
Do you mean that you can select it, but it won't give you a temperature? Or
do you mean that you can't find it in the menu? If you go to the MENU
option, select FULL or something like that. It should enable all the extra
menus.
If you can see it, I believe all you have to do is find it
Randy,
I have had two 3458a that had bad front / rear switches. It was easy to
determine during the calibration of the Cal 0 that failed. A copper short
across both of the inputs, as per the calibration procedure, showed more than
.5 ohm in 2-wire mode on the lowest range. The resistance was
Randy,
I am looking forward to your test results. Does your 845 have the separate
board for the chopper amp as mentioned in the change/errata sheet dated
7/93? I have not seen any further documents from online sources that
references this change. I have two AR models and it appears support for it
Mark,
I am using the STMicroelectronics M48Z12 series in at least one of my
3458A's. I had trouble sourcing one of the two different memory modules
from a reputable seller when I replaced my NVRAM. I purchased a DS1220
previously from Jameco and it died shortly afterward. The datecode on the
Chris,
I believe the current and resistance functions are on the same board. My 4700
is missing one board and it has both missing features.
Todd
Sent from my iPad
On Jul 16, 2014, at 8:48, Chris caal...@gmail.com wrote:
Thank you for the bad news :( I am starting to come to terms with the
Ben,
Do you have the 4921 Current Shunt Adapter option for the 4920? I have not
found any information on it other than what is listed in the 4920 user's
handbook.
It mentions the unit simulates the impedance of a 540B. The guide says that
the instructions are included with the 4921. Do you happen
The remark about calibrating the resistors was merely an example of how I
thought you may have been transferring the calibration of the 10V TVC. My
first thought was that you compared the 10V TVC to the next closest model
at fixed points. The inter-comparison would then continue from that model
to
Willy asked me to post this
I just bought the battery pack on ebay for $700... that was me!
Yes, the transformer and regulator board are OK and I have
them powering the 732b right now!
Can you put this message on volt-nuts for me?
I can't figure out the process..
Thanks!
Willy
Sorry,
Tom beat me to it, I was having trouble with my jems misspelling.
On Wed, Jun 4, 2014 at 6:03 PM, Todd Micallef tmical...@gmail.com wrote:
Willy,
Have you contacted Fluke about getting a replacement transformer? I would
guess it would cost a lot, but one of your other alternatives
Willy,
Have you contacted Fluke about getting a replacement transformer? I would
guess it would cost a lot, but one of your other alternatives is to buy the
external battery pack when it pops up on fleabay. The last one cost about
$700, but I bought mine for less than that. That would give you
86 matches
Mail list logo