Hi,

On 1/9/19 3:05 AM, Tom Van Baak wrote:
>> Wouldn't any counter that has 50 picosecond resolution look the same? 
> 
> Nope. If you had a 20 GHz timebase you'd have 50 ps resolution and all your 
> readings would be exact multiples of 50 ps. It would be very clean and 
> simple. Readings of 25 ps or 49 ps or any other non-multiple of 50 ps would 
> never occur. The histogram would be composed of thin vertical lines -- each 
> exactly 50 ps apart.
> 
>> Its data output is quantized by its resolution.  For example, the 5334A 
>> with 2ns resolution will always show bins no less than 2ns apart.
> 
> Right, the 5335A has a 500 MHz clock and so all readings are multiples of 2 
> ns. The counter cannot output a reading that ends in 1 ns or .123 ns. The 
> histogram would be composed of thin vertical lines -- each exactly 2 ns apart.

HP5335A is an interpolating counter. It uses 10 MHz as coarse counting
clock and then interpolate with x200 factor using a time-stretcher
circuit that then counts the error-pulse in 10 MHz clock using TTL. The
time-stretcher has no calibration of the analog, but only uses
calibration pulses to numerically compensate it's slope-offset.

The Philips/Fluke counter PM6654C has a 500 MHz coarse counter and no
interpolator, and indeed achieves 2 ns quantization steps.

The HP5371A, 5372A and 5373A also uses 500 MHz coarse counter, but then
do a x10 improved time-resolution to achieve 200 ps single-shot
resolution. The delay-line approach has a merrit in that it can handle a
very high sample rate, so the 100 ns or 75 ns time between samples gives
the maximum sample rate of 10 MS/s or 13.3 MS/s, but the time-resolution
is suffering. It is however recommended for the interested to dig into
the programmers manual, since it gives quite a bit of insight as to how
all measurements is done and calculated.

The Pendulum/Fluke CNT-80/81/90/91 all use a 100 MHz coarse counter and
then using error-pulse interpolators. The CNT-90/91 uses essentially the
same mechanism, but CNT-91 has improved performance in several aspects
in order to reach the 50 ps resolution.

> Now it gets a bit more complicated with an interpolating counter, like a 
> SR620 or hp5370. In this case there are fractions involved, but they are 
> fixed fractions.
> 
> The numerology [1] for SR620 quantization is something like 90 MHz, 111.11 
> ns, 1/512, for a final result of 21.701... ps.

The coarse counter of SR620 is 90 MHz, and then error-pulse
interpolation from that. The 90 MHz is synthesized using two
frequency trippler circuits after each other as I recall it.

> The numerology [2] for the hp5370 is something like 200 MHz, 5 ns, 257/256, 
> 5.01953125 ns, 1/256, for a final result of 19.607... ps

The coarse counter of HP5370 is indeed 200 MHz, then using a tuned
coax-delay oscillator of 255/256*200 MHz as interpolator creating a
time-stretched approach using a coinsidense-detection. Marvelous
solution for it's time.

> There is determinism or fine quantization in the readings because even though 
> this ~20 ps value is not a nice round number, it is a fixed number.

Yes, for the simplest linear quantization model which suffice here.

> /// Consequently, for any of the 4 counters mentioned above, if you take a 
> noise floor dataset and sort | uniq -c you will see a small number of large 
> peaks. Like a picket fence. A picket fence inside a Gaussian envelope.

Which is to be expected of the noise we have.

> What we see in the TICC is something quite different. It's not a problem, 
> just very different. Why is that?

Well, you could run into an issue if the TIC-chip does recalibrations,
and jumps around over time and temperature. This may or may not be what
you want, depending on what you are doing.

>> You've mentioned this before, and I'm having trouble getting my head 
>> around it.  I may have this all wrong, but isn't the quantization simply 
>> the resolution of the device?  Your histogram shows humps about 50 
>> picoseconds apart, but that's the resolution of the counter.
> 
> Here's what's happening. Unlike the counters mentioned above there is no 
> fixed cycle or fraction in the TICC. The ring counter is free-running and its 
> rate can vary from unit to unit, from temperature to temperature, from 
> reading to reading. That's why the all-important post-calibration step is 
> performed for every reading. This results in readings that are not fixed 
> fractions of 1 ns or 20 ps or 64 ps or 62.5 ps or any other deterministic 
> value. So each vertical "line" of the histogram is no longer a thin line, it 
> is spread out quite a lot.

Ah, yes, that explains it naturally. I have not looked very closely on
that chip.

Cheers,
Magnus

> /// If you take a TICC noise floor dataset and sort | uniq -c you will also 
> see small number of large peaks, but the peaks are fat camel humps instead of 
> thin picket fences. [3] It has a completely different look. Gaussian camel 
> humps within a Gaussian envelope.
> 
> /tvb
> 
> [1] SR620 math: https://www.thinksrs.com/downloads/pdfs/manuals/SR620m.pdf
> [2] hp5370 math: http://www.hpl.hp.com/hpjournal/pdfs/IssuePDFs/1978-08.pdf
> [3] http://leapsecond.com/pages/ticc/

_______________________________________________
time-nuts mailing list -- [email protected]
To unsubscribe, go to 
http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.

Reply via email to