Brian,
Thank you very much for your exhaustive answer. 

I should have been more clear; in our medical devices we usually design the 
SMPS by ourselves, and we don't want to have a separate certification for 
the supply unit, but just one for the system as a whole. 

I was just wondering if, according to IEC 60601-1, we ought to test the SMPS 
for overload and shortcircuit or not, as I don't find such requirements in 
the standard (neither do you...).

As per Douglas Adams, Root Beer and Cheetos ....touchè

Paolo Peruzzi
  




From: "Brian O'Connell" <[email protected]>
To: <[email protected]>
Sent: Fri, 23 Mar 2007 06:41:12 -0800
Subject: RE: SMPS test according to IEC 60601-1

> Mr. Peruzzi,
> 
> With the exception of certain ECMA safety standards and the IEC 
> safety standard for power supplies (61204), I am not aware of any 
IEC/harmonized
> safety standards that contain special clauses for the consideration 
> of the power supply's performance and construction as a special type 
> of component; i.e., there are no "a component power supply shall 
> ..." statements.
> 
> In fact, most safety standards will have a statement in the scope 
> that says that if the power supply is not installed within the end-
> use system, it is not covered; e.g.,
> 
> "power supply systems which are not an integral part of the equipment...".
> 
> AND
> 
> "Where a separate power supply (such as those containing transformer 
> and dc conversion in a separate package and feeding dc directly into 
> the equipment) is required, the supply shall be identified in the 
instructions
> for use as: 1) part of the equipment (the requirements of this standard
> apply to it also) OR 2) part of an ME System (meaning the supply can
> comply to another IEC standard but with appropriate isolation in the 
> ME equipment). The supply must be specified in adequate detail to assure
> continuing compliance to this standard."
> 
> The 3d ed of 60601-1 says this about power supplies:
> 4.8 *Power supply
> 4.8.1 Source of power for ME EQUIPMENT
> ME EQUIPMENT shall either be powered by an INTERNAL ELECTRICAL POWER
> SOURCE, be specified for connection to a separate power supply, or be
> suitable for connection to a SUPPLY MAINS.
> 5.5 Supply voltages, type of current, nature of supply, frequency
> a) Where test results are influenced by deviations of the supply voltage
> from its RATED value, the effect of such deviations shall be taken into
> account. The supply voltage during tests shall be according to 4.8 or
> according to that specified by the MANUFACTURER, whichever is least
> favourable.
> b) ME EQUIPMENT for a.c. only shall be tested with a.c. at RATED frequency
> (if marked) ± 1 Hz between 0 and 100 Hz and ± 1 % above 100 Hz. ME
> EQUIPMENT marked with a RATED frequency range shall be tested at the 
> least favourable frequency within that range. c) ME EQUIPMENT 
> designed for more than one RATED voltage, or for both a.c. and d.c., 
> shall be tested in conditions (described in 5.4) related to the 
> least favourable voltage and nature of supply, for example, number 
> of phases (except for single-phase supply) and type of current. It 
> may be necessary to perform some tests more than once in order to establish
> which supply configuration is least favourable.
> 16.3 *Power supply
> If ME EQUIPMENT is intended to receive its power from other 
> equipment in an ME SYSTEM, the instructions for use shall specify 
> such other equipment sufficiently to ensure compliance with the 
> requirements of this standard
> (see 4.8.1, 5.5 g) and 7.10.2.3).
> 
> Most SMPS are certified as 'components', so you will probably be limited
> to applying the requirements for components of the end-use standard. 
> And this is why the conditions of acceptability and the Type Test 
conditions
> for the SMPS are so very very important.
> 
> As the regulatory honcho for your company, you must insist the the vendor
> provide a CB test report for the unit. And you must study and understand
> whether the vendor's type test conditions can apply to both normal 
> and abnormal operating conditions of your end-use box. Because the 
> new IEC medical standard has complex risk-analysis requirements, you 
> will need a direct safety contact at your power supply vendor.
> 
> No one can hand this to you in neatly packaged standard that also 
> has the added bonus of an additional clause explaining the meaning 
> of life (this was, in fact, clearly defined by Douglas Adams...). 
> Ultimately, I can only offer this advise - Root Beer and Cheetos are 
> very effective compliance-requirement analysis tools.
> 
> luck,
> Brian
> 
> > -----Original Message-----
> > From: [email protected] [mailto:[email protected]]On Behalf Of Paolo
> > Peruzzi
> > Sent: Friday, March 23, 2007 1:32 AM
> > To: [email protected]
> > Subject: RE: SMPS test according to IEC 60601-1
> >
> >
> > Brian,
> > Could you please give me ONE example of a specific
> > requirement on a POWER
> > SUPPLY (not on a single component) in IEC 60601-1?
> >
> > Thanks,
> > Paolo Peruzzi
> >
> > ---------- Original Message -----------
> > From: "Brian O'Connell" <[email protected]>
> > To: <[email protected]>
> > Sent: Thu, 22 Mar 2007 08:10:13 -0800
> > Subject: RE: SMPS test according to IEC 60601-1
> >
> > > Although, the 3d edition of IEC 60601-1 'appears' to relax test
> > > requirements for a component P.S. (and explicityly allows
> > the use of
> > > I.T. approved units), and I recommend that you avoid the temptation
> > > to reduce these tests to that of 'transformer'.
> > >
> > > My employer makes many custom and off-the-shelf power
> > supply models that
> > > are certified to medical standards, and we encourage our
> > customers to
> > > repeat certain SFC tests and abnormal operating conditions in the
> > > end-use installation.
> > >
> > > Also, try to get a copy of the CB Test Report. Carefully review the
> > > tests, test conditions, and what failed as a result of the
> > test. The
> > > performance of the P.S. during abnormal operating conditions should
> > > drive the conditions of acceptability for the end-use installation.
> > >
> > > Requirements for SFC test and abnormal operating conditions will be
> > > dependent on the affect of the end-unit installation has on
> > the component
> > > power supply. And yes there are, in fact, specific requirements in
> > > 601-1 that would apply to a component power supply. Read the
> > > standard again. No, again.
> > >
> > > luck,
> > > Brian
> > >
> > > > -----Original Message-----
> > > > From: [email protected] [mailto:[email protected]]On
> > Behalf Of Paolo
> > > > Peruzzi
> > > > Sent: Thursday, March 22, 2007 7:25 AM
> > > > To: [email protected]
> > > > Subject: SMPS test according to IEC 60601-1
> > > >
> > > > Dear group,
> > > >
> > > > I would like to gather some opinions about safety tests on
> > > > switch mode power
> > > > supplies within medical devices, according to IEC 60601-1.
> > > > My question arises because a test house technician told me he
> > > > used to test
> > > > power supplies as they were transformers, for what concerns
> > > > short circuit
> > > > and overload tests, on each dc output.
> > > > I couldn't find such requirements for power supplies in the
> > > > standard (I
> > > > couldn't find any!!!), but I think it would be at least wise
> > > > to do these
> > > > tests.
> > > > Do you think they should be included as single fault
> > > > condition tests, even
> > > > if not mentioned?
> > > > Any other thoughts?
> > > >
> > > > Thanks,
> > > >
> > > > Paolo Peruzzi
> 
> -
> ----------------------------------------------------------------
> This message is from the IEEE Product Safety Engineering Society
> emc-pstc discussion list.    Website:  http://www.ieee-pses.org/
> 
> To post a message to the list, send your e-mail to [email protected]
> 
> Instructions:  http://listserv.ieee.org/request/user-guide.html
> 
> List rules: http://www.ieee-pses.org/listrules.html
> 
> For help, send mail to the list administrators:
> 
>      Scott Douglas           [email protected]
>      Mike Cantwell           [email protected]
> 
> For policy questions, send mail to:
> 
>      Jim Bacher:             [email protected]
>      David Heald:            [email protected]
> 
> All emc-pstc postings are archived and searchable on the web at:
> 
>     http://www.ieeecommunities.org/emc-pstc


-

This message is from the IEEE Product Safety Engineering Society
emc-pstc discussion list.    Website:  http://www.ieee-pses.org/

To post a message to the list, send your e-mail to [email protected]

Instructions:  http://listserv.ieee.org/request/user-guide.html

List rules: http://www.ieee-pses.org/listrules.html

For help, send mail to the list administrators:

     Scott Douglas           [email protected]
     Mike Cantwell           [email protected]

For policy questions, send mail to:

     Jim Bacher:             [email protected]
     David Heald:            [email protected]

All emc-pstc postings are archived and searchable on the web at:

    http://www.ieeecommunities.org/emc-pstc

______________________________________________________________________
This email has been scanned by the MessageLabs Email Security System.
For more information please visit http://www.messagelabs.com/email 
______________________________________________________________________

Reply via email to