Although, the 3d edition of IEC 60601-1 'appears' to relax test
requirements for a component P.S. (and explicityly allows the use of I.T.
approved units), and I recommend that you avoid the temptation to reduce
these tests to that of 'transformer'.

My employer makes many custom and off-the-shelf power supply models that
are certified to medical standards, and we encourage our customers to
repeat certain SFC tests and abnormal operating conditions in the end-use
installation.

Also, try to get a copy of the CB Test Report. Carefully review the tests,
test conditions, and what failed as a result of the test. The performance
of the P.S. during abnormal operating conditions should drive the
conditions of acceptability for the end-use installation.

Requirements for SFC test and abnormal operating conditions will be
dependent on the affect of the end-unit installation has on the component
power supply. And yes there are, in fact, specific requirements in 601-1
that would apply to a component power supply. Read the standard again. No,
again.

luck,
Brian

> -----Original Message-----
> From: emc-p...@ieee.org [mailto:emc-p...@ieee.org]On Behalf Of Paolo
> Peruzzi
> Sent: Thursday, March 22, 2007 7:25 AM
> To: emc-p...@ieee.org
> Subject: SMPS test according to IEC 60601-1
>
> Dear group,
>
> I would like to gather some opinions about safety tests on
> switch mode power
> supplies within medical devices, according to IEC 60601-1.
> My question arises because a test house technician told me he
> used to test
> power supplies as they were transformers, for what concerns
> short circuit
> and overload tests, on each dc output.
> I couldn't find such requirements for power supplies in the
> standard (I
> couldn't find any!!!), but I think it would be at least wise
> to do these
> tests.
> Do you think they should be included as single fault
> condition tests, even
> if not mentioned?
> Any other thoughts?
>
> Thanks,
>
> Paolo Peruzzi
> Regulatory Medical R&D
> El.En. S.p.A.
> Via Baldanzese, 17
> 50041 Calenzano (FI)
> Italy

-

This message is from the IEEE Product Safety Engineering Society
emc-pstc discussion list.    Website:  http://www.ieee-pses.org/

To post a message to the list, send your e-mail to emc-p...@ieee.org

Instructions:  http://listserv.ieee.org/request/user-guide.html

List rules: http://www.ieee-pses.org/listrules.html

For help, send mail to the list administrators:

     Scott Douglas           emcp...@ptcnh.net
     Mike Cantwell           mcantw...@ieee.org

For policy questions, send mail to:

     Jim Bacher:             j.bac...@ieee.org
     David Heald:            emc-p...@daveheald.com

All emc-pstc postings are archived and searchable on the web at:

    http://www.ieeecommunities.org/emc-pstc

______________________________________________________________________
This email has been scanned by the MessageLabs Email Security System.
For more information please visit http://www.messagelabs.com/email 
______________________________________________________________________

Reply via email to