Brian,
Could you please give me ONE example of a specific requirement on a POWER 
SUPPLY (not on a single component) in IEC 60601-1?

Thanks,
Paolo Peruzzi


From: "Brian O'Connell" <[email protected]>
To: <[email protected]>
Sent: Thu, 22 Mar 2007 08:10:13 -0800
Subject: RE: SMPS test according to IEC 60601-1

> Although, the 3d edition of IEC 60601-1 'appears' to relax test
> requirements for a component P.S. (and explicityly allows the use of 
> I.T. approved units), and I recommend that you avoid the temptation 
> to reduce these tests to that of 'transformer'.
> 
> My employer makes many custom and off-the-shelf power supply models that
> are certified to medical standards, and we encourage our customers to
> repeat certain SFC tests and abnormal operating conditions in the 
> end-use installation.
> 
> Also, try to get a copy of the CB Test Report. Carefully review the 
> tests, test conditions, and what failed as a result of the test. The 
> performance of the P.S. during abnormal operating conditions should 
> drive the conditions of acceptability for the end-use installation.
> 
> Requirements for SFC test and abnormal operating conditions will be
> dependent on the affect of the end-unit installation has on the component
> power supply. And yes there are, in fact, specific requirements in 
> 601-1 that would apply to a component power supply. Read the 
> standard again. No, again.
> 
> luck,
> Brian
> 
> > -----Original Message-----
> > From: [email protected] [mailto:[email protected]]On Behalf Of Paolo
> > Peruzzi
> > Sent: Thursday, March 22, 2007 7:25 AM
> > To: [email protected]
> > Subject: SMPS test according to IEC 60601-1
> >
> > Dear group,
> >
> > I would like to gather some opinions about safety tests on
> > switch mode power
> > supplies within medical devices, according to IEC 60601-1.
> > My question arises because a test house technician told me he
> > used to test
> > power supplies as they were transformers, for what concerns
> > short circuit
> > and overload tests, on each dc output.
> > I couldn't find such requirements for power supplies in the
> > standard (I
> > couldn't find any!!!), but I think it would be at least wise
> > to do these
> > tests.
> > Do you think they should be included as single fault
> > condition tests, even
> > if not mentioned?
> > Any other thoughts?
> >
> > Thanks,
> >
> > Paolo Peruzzi
> > Regulatory Medical R&D
> > El.En. S.p.A.
> > Via Baldanzese, 17
> > 50041 Calenzano (FI)
> > Italy
> 
> -
> ----------------------------------------------------------------
> This message is from the IEEE Product Safety Engineering Society
> emc-pstc discussion list.    Website:  http://www.ieee-pses.org/
> 
> To post a message to the list, send your e-mail to [email protected]
> 
> Instructions:  http://listserv.ieee.org/request/user-guide.html
> 
> List rules: http://www.ieee-pses.org/listrules.html
> 
> For help, send mail to the list administrators:
> 
>      Scott Douglas           [email protected]
>      Mike Cantwell           [email protected]
> 
> For policy questions, send mail to:
> 
>      Jim Bacher:             [email protected]
>      David Heald:            [email protected]
> 
> All emc-pstc postings are archived and searchable on the web at:
> 
>     http://www.ieeecommunities.org/emc-pstc


-

This message is from the IEEE Product Safety Engineering Society
emc-pstc discussion list.    Website:  http://www.ieee-pses.org/

To post a message to the list, send your e-mail to [email protected]

Instructions:  http://listserv.ieee.org/request/user-guide.html

List rules: http://www.ieee-pses.org/listrules.html

For help, send mail to the list administrators:

     Scott Douglas           [email protected]
     Mike Cantwell           [email protected]

For policy questions, send mail to:

     Jim Bacher:             [email protected]
     David Heald:            [email protected]

All emc-pstc postings are archived and searchable on the web at:

    http://www.ieeecommunities.org/emc-pstc

______________________________________________________________________
This email has been scanned by the MessageLabs Email Security System.
For more information please visit http://www.messagelabs.com/email 
______________________________________________________________________

Reply via email to