Hi all,
I can suggest a different solution I have tested to have some important
improvement. First, the lower phase shift sensitivity with the use of a low
pass filter in conjunction with two notch filters instead the high Q band pass
filter, second the unity gain to avoid a 13/20 dB amplif
George,
you can replace the input transformer with the mini-circuits model
T2-613-1-KK81 or T662-KK81 for under 2 Dollars. The inductors L1 and L3 are
standards value you can find smd or standard on ebay. About L4 you can put in
place a standard 4.7uH value increasing the series capacitor for
Hi,
This is a good and reasonable description of what we normally need. I agree.
Some amateur like me want some more than the normal, and If you want, more than
the necessary. This happen only for testing ours ability to improve the
performance of our instruments we use as hobbist.
For this
Hi all,
I am looking to understand the real difference between the old and new A3 board
installed in the HP5065A. It is the multiplier chain from 5 to 60 MHz. I am
sure there is a phase noise improvement between the two versions.
So I am looking for the old type A3 schematics. Actually I hav
Hi all,
I am looking for the internal schematic of the 00105-6013 OCXO mounted in the
old series of the HP5065A.
I have two of them not working and I would like to fix it.
thanks,
Luciano
www.timeok.it
Message sent via Atmail Open - http://atmail.org/
Thank you all for the answer:
The old one is in the 105A/B and the 5061A manual.
In the 5065A there is only a the block diagram of the oscillator.
I have found it.
Luciano
On Thu 05/02/15 20:32 , Charles Steinmetz wrote:
> Luciano wrote:
>
> >I am looking for the internal schematic of the 0
Hi all,
I have for sale a HP5370A in working order for 285 Euro+ shipping cost.
If you are interested pse email me directly: tim...@timeok.it
I will give you back the shipping cost to your country.
thanks,
Luciano
Message sent via Atmail Open - http://atmail.org/
___
Dear friends,
The first HP5065A were produced around the late years 60. Over time some flaws,
related to engineering and the choice of components, are highlighted.
The main ones are the following:
- The capacitors used primarily on power lines of +20 Volts tend to lose
the liquid inside a
Thanks Corby,
I will add this info in a document I am preparing with pictures to be publish
on my site www.timeok.it.
Luciano
On Sun 24/05/15 07:31 , wrote:
> Also replace C23 and C24 on the A11 board.
> Even if they read OK, I change them to be safe!
> 20uf 50V Axial.
>
> Cheers,
>
> Co
Hi Magnus,
I have done this experience. You can find my project here:
http://www.timeok.it/files/low_noise_regenerative_divider.pdf
An important goal was to have a range as input level to be used with several
source.
When the oscillation are is engaged, the amplifier reaches immediately th
Hi all,
I have developed an upgrade kit for the HP5065A.
This add one or two 10MHz outputs and can be installed in all the units have
the HP10811 as OCXO.
The main features of this amplifier-separator are:
Very low residual phase noise, typical -150dB @ 1Hz
High reverse isolation typical 110 d
Hi all,
I am looking on Wikipedia a description of the Residual Phase Noise but this
page do not exist.
Can some one of the time-nuts expert write a full description of this physical
aspect for Wikipedia?
thanks,
Luciano
tim...@timeok.it
www.timeok.it
Message sent via Atmail Open - http://a
Dear Magnus,
I am writing a doc on my Distribution Amplifier project and I was looking a
complete, but simple to understand, description about the Phase Noise
introduced, or Added, by a "non generative" devices.
thanks,
Luciano
On Mon 13/07/15 17:37 , Magnus Danielson wrote:
> Dear Luci
Hi Skipp,
I suggest you a simple passive filter with harmonic notch as this:
http://www.timeok.it/files/5_and_10mhz_low_pass_notch_filter.pdf
Obviously you have to made the 10MHz version,
Ciao,
Luciano
On Thu 16/07/15 19:49 , skipp Isaham via time-nuts wrote:
> re: 10MHz Square to Sine Wav
It is the simplest solution because you have not to tune the notch filter but
you have -20dB on 2nd harmonic and -40dB for the third instead -50dB (min for
both) if you made a low-pass with notch filters.
Luciano
On Fri 17/07/15 05:10 , D W wrote:
> Hi,
>
> A while back I ordered an it
I an curious to know the price of this crystal.
Luciano
On Fri 17/07/15 00:02 , KA2WEU--- via time-nuts wrote:
> I am working with this .. amazing device, Ulrich N1UL
>
> ___
> time-nuts mailing list -- time-nuts@febo.com
> To unsubscribe, go
noise.
But before measure using an oscilloscope the output of the GPS OCXO directly on
the output pin to verify if there is a sine-wave or square
Luciano
timeok
Message sent via Atmail Open - http://atmail.org/
___
time-nuts mailing list -- time-nuts@fe
Hi ,
Are available some more kit to add 10MHz output to the HP5065A.
these kit are applicable only if the HP10811 OCXO is installed.
reference: http://www.timeok.it/files/hp5065AoptH10v200.pdf
Email me directly.
tim...@timeok.it
thank you,
Luciano
Message sent via Atmail Open - http://atm
Dear ZZ,
To measure phase noise or Allan deviation or other type of measure you need a
Reference to compare the DUT, device under test.
This reference have to be better performances than the DUT we can suppose have
to be ten time better than the DUT and in case of a GPSDO a good source can be
a
MELAB.
The question I ask you is this: is there a solution to this problem with
features want from me with costs under 1500 dollars/euros?
Luciano
timeok
Message sent via Atmail Open - http://atmail.org/
___
time-nuts mailing list -- time-nuts@febo.c
Hi,
I have seen, but I do not remember where, someone have rebuilt the external
oven controller to complete a stand alone OCXO removed from an HPZ3801A.
Can you hep me to find the document?
Luciano
timeok
Message sent via Atmail Open - http://atmail.org
here the ADEV of the two oscillator. I have added the Super performance of an
HP105B (old oscillator) bought in an Ham fest.
Luciano
On Fri 07/08/15 03:37 , Bob Camp wrote:
> Hi
>
> HP 10811’s vary over almost a 100:1 range in terms of ADEV performance
> at short tau. The standard model is
Hi Corby,
I agree with you that if selected, there may be many differences between
objects of the same type. Unfortunately it is very difficult to know what is
selected and what is not. For example I found a 10811 mounted in a Z3801A who
have the temperature set-point completely wrong. I r
Hi all,
I try to ansver to all you:
Luciano, how was the blue trace taken? Is this from your DMTD project? If so,
it's looking promising.
The green and magenta traces are definitely in the right ballpark for
measurements on a 5370-class counter. At 3E-11 @ t=12s the magenta trace is
optimistic
I am interested in knowing the exact name or number of the colors used by HP
for the HP5065A front panels.
Can some one help?
I know there are three versions:
Single color, Gray
Dual color (similar to cream and brown)
Single color, Cream (very rare, I have one)
thanks,
Luciano
Timeok
;ccid=uC8SbHON&simid=608034311696942306&thid=OIP.Mb82f126c738d9a403e654754b922b332H0&ajaxhist=0
It is possible this is the right color for the old models of HP5065A.
Now I hope other persons can help me to have more information to match with
yours.
Have a good night,
Luciano
Timeok
On T
and the C Field regulator, can be significantly improved.
An old uA723 can be a good solution for regulator and reference for the C Field
current generator.
I think I will work at it , it would be interesting we can exchange more
information. |
Luciano
Timeok
On Tue 18/08/15 22:05 , Poul
Hi ,
I think the Regulator Board should still be questioned.
To calibrate the +20 V,it is necessary to use an Extender Board because the
calibration trimmer is not accessible with the card fitted normally.
I have used the extender HP built for service model 5060-0049. After alignment
I could see
ifier picture is a draft
I would like to have your comments to get ideas.
Luciano
Timeok
Message sent via Atmail Open - http://atmail.org/
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo
, but is not an High stability like the R11.
At the moment I cannot disconnect the HP5065A under test but will be the next
operation I will do.
Luciano
Timeok
On Fri 21/08/15 22:22 , "Poul-Henning Kamp" wrote:
>
>
> I had this morning, and put it into action
ls with a single input channels or 2 x 6
output channels if you want to have a double DA to distribute, for example,
5 and 10 MHz.
Thank you for your comments,
Luciano
Timeok
see: http://www.timeok.it/wp/ [1]
Message sent via Atmail Open - http://atmail.org/
comparative table.ods
Description:
Hi,
I would like to receive some feedback on the Wavecrest 2079 using Timelab.
The main question is: can someone send me some outputs related to the noise
floor of the Wavecrest?
I would like to be reassured that, as written in the manual, the function
double(:meas:data) get the equivalent
want you can fill in your details and send them to me, I'll add them to
the list.
I'll be waiting, thanks,
visit the snlist site: http://www.timeok.it/wp/hp5065a-corner-3/
Luciano
Timeok
Message sent via Atmail Open - http://atmail.org/
__
I remember you my post:
I have done some short time measurements (30 Min.) on the current generator, in
particular on the reference input and feedback input of the differential
amplifier Q6A and .Q6B (base pin to GND).
On the reference input Q6A I have a stability equivalent to more than +/-1nA
Hi Charles,
On Fri 28/08/15 19:53 , Charles Steinmetz wrote:
Charles wrote:
> I'm not aware of any instrument that needs a reference signal of more
> than +13dBm (= 1Vrms), and most will happily accept much less than
> that with no reduction in performance -- so I see no reason why a DA
>
May be some one cannot read the file .ods, so I have converted it to .pdf
Luciano
On Sat 29/08/15 11:24 , tim...@timeok.it wrote:
> Hi Charles,
>
> On Fri 28/08/15 19:53 , Charles Steinmetz wrote:
>
> Charles wrote:
>
> > I'm not aware of any instrument that needs a reference signal of m
Hi Chuck,
can you give me more info about th PHK's changes to the oven?
Luciano
On Mon 31/08/15 14:47 , Chuck Harris wrote:
> Hi Bob,
>
> I haven't seen anyone bashing the designer, or the design of the 5065.
> What I have seen a couple of enthusiasts that love the unit enough that
> they
/Timeok-Time-and-Frequency-House-Standard-v-2.11.pdf
Luciano
On Thu 17/09/15 02:00 , Can Altineller wrote:
> Hello,
>
> I started putting a test setup together when I suspected the DS3231 RTC's
> I
> got from dx.com were fake. So I put 2 mcus, each measuring microseco
Hi all,
I have an HP5065A for sale. It mount the option 01 digital clock.
The unit is refurbished and tested. Capacitors and diodes upgraded.
upgrade info:
http://www.timeok.it/wp/wp-content/uploads/2015/09/First-aid-when-you-buy-an-hp5065A-v-1.2.pdf
No power cable.
I ask 1250.oo Euros.
Shipping
> Hi Jim,
>
> first of all verify all the capacitors installed. Refer to:
>
> http://www.timeok.it/wp/wp-content/uploads/2015/09/First-aid-when-you-buy-a
> n-hp5065A-v-1.2.pdf[1]
>
> Luciano
>
> On Mon 07/12/15 05:30 , "Jim/Anna McIntyre" wrote:
>
> > Greetings all,
> > My name is J
Jim,
> Except...
>
> Microphonics.
>
> What's the norm here? I see it in the 2nd harmonic level on the panel
> meter. It runs about 20 normally, but a pretty light tap (with my
> fingernail) anywhere on the instrument can peg the meter for a few
> seconds. While I had the scope on A8 TP2 / TP3
one of the most important spec is the channel to channel separation. Do you
have done some measurements?
Luciano
www.timeok.it
On Thu 17/12/15 20:44 , Anders Wallin wrote:
> First prototype assembled today, tested with 12 VDC SMPS wall-wart supply
> and with 12 V lead-acid battery.
> Compar
I remember the discussion on the battery pack of the HP5065A option 002.
I have right now the opportunity to have an original pack for service and I
have taken some picture of it.
see:
http://www.timeok.it/wp/wp-content/uploads/2015/08/HP5065A-battey-pack-v1.0.pdf
Luciano
www.timeok.it
Mess
To verify if the where the delay is generated, antenna and splitter or GPS,
simply invert the antenna cable on the GPSs antenna connector and verify if the
delay change as sign.
Luciano
On Tue 22/12/15 02:24 , d...@irtelemetrics.com wrote:
> Hi All,
>
> So I've been playing with some tim
.
pse can I have some info about?
ref: http://www.tapr.org/~n8ur/T2_Mini_Manual.pdf
Luciano
timeok
Message sent via Atmail Open - http://atmail.org/
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin
Have a look on this:
http://www.timeok.it/wp/wp-content/uploads/2015/08/10_to_100_mhz_multiplier1.pdf
Luciano
timeok
On Fri 22/01/16 11:15 , REEVES Paul wrote:
> Why not use something like an HP5254B/C ? They give out 50MHz harmonics up
> to the low Ghz region, all filtered by a nice
Hi all,
On Wed 03/02/16 07:06 , "John Miles" wrote:
> Hi, Timothy --
>
> > I've had an Agilent 53220A for some time, and recently discovered the
> > wonderful TimeLab software, but I've hit a bit of a snag trying to run
> some
> > ADEV measurements on a Rb clock. The problem is that the grap
Hi Attila,
I normally use the HP53132A in Frequency mode with the Timelab. The time
interval mode, I know is more accurate but using the 1PPS as input you lost too
much resolution.
looking :
http://www.timeok.it/wp/wp-content/uploads/2015/09/Timeok-Time-and-Frequency-House-Standard-v-2.1.pdf
Hi all,
Among the many experts and documentation available on the internet my question
might seem trivial .
I'm looking for a written document , not generic information, describing the
procedure , timing and methods of calibration of one or more standard present
in a laboratory frequency stan
Hi Mike,
You have not mentioned the best Rubidium available, the HP5065A. Last week two
of them are sold on ebay.
Second chose can be the Efratom FRK or M100, than all the others.
Luciano
On Fri 11/03/16 12:02 , time...@metachaos.net wrote:
> Hi,
>
> In looking at eBay, I current see quite
Tom,
it is not, see the Farnell code below followed by Amphenol part number:
Farnell code: 1914117 pn 97-3106A18-22S connector
Farnell code: 1654733 pn 9779-513-8 plastic ring
Farnell code: 151628 pn 97-3057-1010-1 cable clamp and bush
see also:
http://canada.newark.com/amphenol-industrial/
Ulrich,
you can buy it from Farnell:
Farnell code: 1914117 pn 97-3106A18-22S connector
Farnell code: 1654733 pn 9779-513-8 plastic ring
Farnell code: 151628 pn 97-3057-1010-1 cable clamp and bush
Luciano
www.timeok.it
On Thu 17/03/16 15:29 , KA2WEU--- via time-nuts wrote:
> The HP 50
Here the italian version:
http://www.crit.rai.it/eletel/2012-1/121-2.pdf
Luciano
www.timeok.it
On Mon 04/04/16 21:28 , Magnus Danielson wrote:
> Hi Hal,
>
> On 04/04/2016 05:39 AM, Hal Murray wrote:
> >
> > Has anybody studied what happens when a GPSDO comes out of holdover? Has
> > anybody
I have an Agilent 34970A data logger. It use the "Bencklink Data Logger3" as
software.
I would like to capture the graph in a printable format instead the original
.csv . A .gif will be good to be inserted in documents.
Can some one help me?
thanks ,
Luciano
www.timeok.it
Message sent via A
Ok Thank you all you, I will try these solutions and I will ask also to
Keysight.
Luciano
www.timeok.it
On Sun 24/04/16 00:10 , jimlux wrote:
> On 4/23/16 12:15 PM, DaveH wrote:
> > .csv (comma seperated values) is a text format, .gif (graphical
> interchange
> > format) is an image format.
Hi all,
if I will use a dual mixer TI meter, suppose to have :
LO 5.000.005 Hz
Ref and Ftest 5 MHz
IF= 5Hz
how do I have to setup Timelab for acquisition?
Sampling interval =
Input frequency =
Scale factor =
Thanks all,
Luciano
timeok
see :
http://www.symmetricom.com/lp/gbu/email/phase-noise-test-probe-landing-page/?emailid=3120A_Launch_Email_Quote
it seem the same KE5FX product.
timeok
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com
Yes, Mr Rex I am guilty. may be other are perfect!
with the compliment of Timeok
Il 2012-12-19 23:26 Rex ha scritto:
--
timeok
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman
Yes Mr Rex,
I am guilty .. not all people are perfect!
best regards form Timeok
Il 2012-12-19 23:26 Rex ha scritto:
This is a sign that people can't digest all the messages on
time-nuts.
There was this...
http://www.mail-archive.com/time-nuts@febo.com/msg53586.html
(thread re
Interesting info about a new developed Maser working at room
temperature:
http://www.nature.com/nature/journal/v488/n7411/full/nature11339.html
timeok
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi
Claude,
your output is a square wave on 50 Ohm as write close the connector. Use
this filter to have a clean sinewave output.
http://www.timeok.it/files/5_and_10mhz_low_pass_notch_filter.pdf
Luciano
www.timeok.it
On Thu 17/04/14 7:32 AM , Claude Fender wrote:Hello List,
I've recently bou
On Tue 3/06/14 1:46 PM , Donal Geoghegan wrote:
-
Message sent via Atmail Open - http://atmail.org/
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
You can find an interesting solution with the INA330. In the data sheet is
described a complete temperature PID controlled .
http://pdf.datasheetcatalog.com/datasheets/270/255641_DS.pdf
Luciano
www.timeok.it
On Mon 21/07/14 4:02 PM , Attila Kinali wrote:Moin Bert,
On Mon, 21 Jul 2014 09:0
Hi all,
here a multiplier from 10 to 100.
http://www.timeok.it/files/10_to_100_mhz_multiplier.pdf
Hoping this can help.
Luciano
On Sun 28/09/14 3:57 PM , "Dr. David Kirkby (Kirkby Microwave
Ltd)" wrote:
> On 28 September 2014 14:49, dlewis wrote:
> > If you can use 90 MHz, .
>
> No,
the instructions there.
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.
--
timeok
___
time
y prized.
/tvb
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.
--
timeok
___
time-nuts maili
e for manuals on these instruments? I checked the
site of KO4BB, unfortunately he doesn't have a PDF of the manual.
Best regards, Jeroen
On 09/17/2012 02:50 PM, Timeok wrote:
Yes Tom you right,
but your example is on a rare Sulzer 2.5 MHz. Another very good
reference can be an HP
frequency
standard?
Thanks,
Ron
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.
For sale MA259 dual oven 5, 1, 0.1 MHz frequency standard. Oscillator
and power supply. Condition: used, working unit. Rack mount.Very good
price.
see the manual to:
http://ko4bb.com/Manuals/Racal/Racal_MA2595Mc-s_Precision_Frequency_Standard.pdf
Contact me directly, I can send picture.
Luc
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.
George and now?
hi!
Luciano
timeok
___
time-nuts
correction: HP53132A
timeok
Il 2012-10-02 13:29 Timeok ha scritto:
test using HP53123A:
The pink line is the noise floor using the TI mode
The blue line is the noise floor using the Frequency mode (gate 2
sec)
The other two line are the same oscillator, an HP105B tested with
both the modes
measurement on the 10MHz for example using Stable
32 etc.
Heavy, large, and power hungry, but quite solid.
Said
On Oct 2, 2012, at 4:29, Timeok wrote:
test using HP53123A:
The pink line is the noise floor using the TI mode
The blue line is the noise floor using the Frequency mode (gate 2
sec)
The
2012-10-02 14:43 Adrian ha scritto:
Timeok schrieb:
test using HP53123A:
The pink line is the noise floor using the TI mode
The blue line is the noise floor using the Frequency mode (gate 2
sec)
The other two line are the same oscillator, an HP105B tested with
both the modes.
As you can see
ith a blocking capacitors in series to avoid
any antenna or active splitter interaction.
Luciano
timeok
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instruc
I agree,
Luciano
timeok
Il 2012-10-09 10:41 Azelio Boriani ha scritto:
Is there any difference between what a GPS receiver can receve via
crosstalk or receive directly from the antenna? In my opinion
crosstalk is
absolutely less than the last argument about GPS antenna splitters.
On Tue, Oct
as mechanical realization ad impedance
matching.
That's all.
All the people in time-nuts community want to improve day by day the hw
and sw they have at home.
yes, yesterday night I have bring a bier. Cirio!
Luciano
timeok
Il 2012-10-09 15:15 paul swed ha scritto:
Lots of comments. I
For sale a working Sulzer frequency standard 5A 5 MHz oscillator
complete with power supply, rack mount.
I ask 180 Euro. Location Italy. Example shipping cost: to Germany 40
Euro (via ground 10-15 Kg)no tracking nr. Other options available.
Email me directly to: tim...@timeok.it
Luciano
_
Hi George,
I like , I want to buy it
I live in Italy and I collect and made by myself some equipment of time
and frequency standard.
Please can you quote me the shipping cost to Italy?
thankyou
Lucianpo
timeok-it
Il 2012-10-15 21:57 George Race ha scritto:
I still have the HP 115BR Clock. It
-bin/mailman/listinfo/time-nuts
and follow the instructions there.
--
timeok
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.
, 2012, at 8:27 AM, Timeok wrote:
Hi all,
to avoid more display "consumption" during long acquisitions, there
is an HPIB command to blank the display on the counter. I have asked
to KE5FX to add this feature in Timelab during the acquisition setup.
Hope he have time to do this.
Luc
Is the project a complete dual mixer + TIC ?
What about the SW? may I use the Timelab to have a strip chart like
Adev? If not what SW I can use?
thanks,
Luciano
timeok
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://w
e question I have are:
What is the project plan about the paper production to describe
or/and
sell the HW?
Is the project a complete dual mixer + TIC ?
What about the SW? may I use the Timelab to have a strip chart like
Adev? If not what SW I
Hi Corby,
very interesting. I have some questions?
1) it is difficult to read the frequency value on the plot. Can you specify?
2) how do you set the offset of the op-amp trimmer ?
3) do you think is possible to use the -20V on the HP5065A instead the -12V
from the DC-DC converter?
4) what frequ
Hi all,
I have posted here my experience with the installation of the band pass filter
in front of the lamp of the HP5065A.
http://www.timeok.it/wp/wp-content/uploads/2015/08/HP5065A-Super-upgrade-v-1.4.pdf
Luciano
tim...@timeok.it
Message sent via Atmail Open - http://atmail.org/
___
See also:
http://www.timeok.it/wp-content/uploads/2015/08/high-performance-frequency-doublerv1-31.pdf
Luciano
www.timeok.it
From: "time-nuts" time-nuts-boun...@febo.com
To: time-nuts@febo.com
Cc:
Date: Sat, 9 Jul 2016 17:57:57 -0400
Subject: Re: [time-nuts] Does doubling a frequency alter the
Hi,
I agree with you. Some HP spectrum analyzer have internal 50 and 100 Hz spurs
down to -80dBc due the fan as described in the manual. It can be the spurs are
from the test set and/or connection cables. If your oscillator is battery
powered you can be sure the spurs are not from the source.
I think that these pages can help you:
http://www.timeok.it/wp-content/uploads/2015/09/First-aid-when-you-buy-an-hp5065A-v-1.2.pdf
http://www.timeok.it/wp-content/uploads/2015/09/First-aid-II-when-you-buy-an-hp5065A-v-1.0.pdf
Luciano
From "time-nuts" time-nuts-boun...@febo.com
To "time-
Hi,
I have updated the paper adding the schematic diagram and in addition some
updating for the comparative table.
http://www.timeok.it/wp-content/uploads/2015/08/Frequency-Distribution-Design-Basic-Module-v-4.1.pdf
This study come from a development of a Bruce Griffith idea.
rega
Sorry for the broadcast.
Jim, N1JR, I want to answer to your direct email about the distribution
amplifier but there are some problem in your email basket. please verify.
Luciano
www.timeok.it
From "time-nuts" time-nuts-boun...@febo.com
To time-nuts@febo.com
Cc
Date Sat
Hi John,
you can visit the pages:
http://www.timeok.it/hp5065a-corner-3/
regards,
Luciano
timeok
From "time-nuts" time-nuts-boun...@febo.com
To "Discussion of precise time and frequency measurement" time-nuts@febo.com
Cc
Date Sat, 29 Oct
timeok
From "time-nuts" time-nuts-boun...@febo.com
To "Discussion of precise time and frequency measurement" time-nuts@febo.com
Cc
Date Thu, 10 Nov 2016 11:20:37 +1100
Subject [time-nuts] Sapphire oscillators
Anyone got any comment
Hi Li,
can you share more info about your project?
Luciano
www.timeok.it
From "time-nuts" time-nuts-boun...@febo.com
To "Discussion of precise time and frequency measurement" time-nuts@febo.com
Cc
Date Sat, 26 Nov 2016 13:16:34 +0800
Subject Re: [time-nuts] New Timesta
Hi John,
I have planned to buy two TICC.
An interesting feature would be to be able to do two simultaneous acquisitions,
and Timelab as real time display,using the two indipendent input channels and
the 10Mhz clock as single reference.
Luciano
www.timeok.it
From "time-nuts" time-nuts-boun
Hi John,
a question: In the draft operating manual is write is possible to use 2, 4
,6 etc. input using multiple (1,2,3) TICC.
How are connected the extra TICC boards?
Is it possible a single Arduino drive multiple boards or each board is
connected to an Arduino processor?
thanks,
multiple
boards to be synchronized but we haven't implemented the full system yet
-- in part because until now there are only 4 working TICCs in the
world, and they are in 3 locations!
John
----
On 12/8/2016 8:09 AM, timeok wrote:
>
> Hi John,
> a question: In
other devices in the signal
chain.
Luciano
Timeok
-
Message sent via Atmail Open - http://atmail.org/
Links:
--
[1]
http://webmail.timeok.it/parse.php?redirect=https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
___
tim
Hi,
I have the complete manual of the HP115BR. Pse contact me directly to send
you the file.
If you need I can suggest some service activity normally are needed by this
old clock.
Luciano,
tim...@timeok.it
From "time-nuts" time-nuts-boun...@febo.com
To time-nuts@febo.com
Hi,
the thought of being able to work on building a H Maser has always accompanied
me in recent years.
I fully understand the many difficulties of this project and also the necessity
of a work team.
Maybe a Passive Maser would be easiest to implement, but I do not know in
detail the processes
o
Bob, Dave,
The most big problem is to find a group of relatively close people with the
will to spend many hours, months for this project.
Here in Italy I found some people who work in public university who are
available to give external help, for example they were
given a Pirani sensor and th
Hi Corby,
I an looking your Pressure Compensation for the HP5065A ad I have a question
about the preliminary adjustment of R3.
Can you give me the alignment procedure you have used?. I suppose you have
used the Pressure chamber looking at the frequency offset.
thanks,
Luciano
1 - 100 of 123 matches