Re: [PSES] Active Implantable Medical Devices
Hi Grace For your point 2 Under the RE-D for EMC compliance, you are not required to use a notified body if a particular standard is not listed in the OJEU. See Articles 3(1)(b) and 17(2) of the RE-D http://eur-lex.europa.eu/legal-content/EN/TXT/HTML/?uri=CELEX:32014L0053=EN Kind regards Kevin From: Grace Lin [mailto:graceli...@gmail.com] Sent: Monday, April 03, 2017 1:35 AM To: EMC-PSTC@LISTSERV.IEEE.ORG Subject: [PSES] Active Implantable Medical Devices Dear Members, For active implantable medical devices to comply with the RED requirements: 1. Do all active implantable medical devices require Notified Body’s involvement per RED? From online search (https://ec.europa.eu/growth/single-market/ce-marking/manufacturers_en<https://urldefense.proofpoint.com/v2/url?u=https-3A__ec.europa.eu_growth_single-2Dmarket_ce-2Dmarking_manufacturers-5Fen=DwMFaQ=0YGvTs3tT-VMy8_v51yLDw=opngXt0GJI9CAIpmFDu4mUv_xoHzGtzX8PosGAmQ4SY=PpPM367w6MVG8MaOZz-CW6nLegeKiGFCMhme5b5amOc=TPWHe5tJnrMBRWXN9ELlsMoCHqbLg-YRimZSeaMeTYA=>), it seems the answer is yes. Could you please confirm? 2. The 301 489 series EMC standards covering RED have not been published/harmonized. For those manufacturers that do not want to wait for the harmonized standards to be published, should the manufacturers proceed with the DRAFT version of the standards (and send to NB)? 3. Do you have any idea when 301 489 series standards covering RED will be published? The current status is “On Approval”. Thank you very much for your time and I look forward to hearing from you. Best regards, Grace Lin - This message is from the IEEE Product Safety Engineering Society emc-pstc discussion list. To post a message to the list, send your e-mail to <emc-p...@ieee.org<mailto:emc-p...@ieee.org>> All emc-pstc postings are archived and searchable on the web at: http://www.ieee-pses.org/emc-pstc.html<https://urldefense.proofpoint.com/v2/url?u=http-3A__www.ieee-2Dpses.org_emc-2Dpstc.html=DwMFaQ=0YGvTs3tT-VMy8_v51yLDw=opngXt0GJI9CAIpmFDu4mUv_xoHzGtzX8PosGAmQ4SY=PpPM367w6MVG8MaOZz-CW6nLegeKiGFCMhme5b5amOc=9wIifSrMBQWt7oys7eMYVu-bgKZKi-UxbKFKgGICuHI=> Attachments are not permitted but the IEEE PSES Online Communities site at http://product-compliance.oc.ieee.org/<https://urldefense.proofpoint.com/v2/url?u=http-3A__product-2Dcompliance.oc.ieee.org_=DwMFaQ=0YGvTs3tT-VMy8_v51yLDw=opngXt0GJI9CAIpmFDu4mUv_xoHzGtzX8PosGAmQ4SY=PpPM367w6MVG8MaOZz-CW6nLegeKiGFCMhme5b5amOc=S6DIee8tRdQWWYR2X8m4RwaztNCaJPlNz9uw6U3k4oQ=> can be used for graphics (in well-used formats), large files, etc. Website: http://www.ieee-pses.org/<https://urldefense.proofpoint.com/v2/url?u=http-3A__www.ieee-2Dpses.org_=DwMFaQ=0YGvTs3tT-VMy8_v51yLDw=opngXt0GJI9CAIpmFDu4mUv_xoHzGtzX8PosGAmQ4SY=PpPM367w6MVG8MaOZz-CW6nLegeKiGFCMhme5b5amOc=_Tm2Y92Q97A-4Tb83PbWpjVqYDp4T2Dw7Ib48CFzUDk=> Instructions: http://www.ieee-pses.org/list.html (including how to unsubscribe)<https://urldefense.proofpoint.com/v2/url?u=http-3A__www.ieee-2Dpses.org_list.html=DwMFaQ=0YGvTs3tT-VMy8_v51yLDw=opngXt0GJI9CAIpmFDu4mUv_xoHzGtzX8PosGAmQ4SY=PpPM367w6MVG8MaOZz-CW6nLegeKiGFCMhme5b5amOc=8ZE2rsf6J8pQlmHN2MzEek8h9GKFB2Cer9qvRuKIH5U=> List rules: http://www.ieee-pses.org/listrules.html<https://urldefense.proofpoint.com/v2/url?u=http-3A__www.ieee-2Dpses.org_listrules.html=DwMFaQ=0YGvTs3tT-VMy8_v51yLDw=opngXt0GJI9CAIpmFDu4mUv_xoHzGtzX8PosGAmQ4SY=PpPM367w6MVG8MaOZz-CW6nLegeKiGFCMhme5b5amOc=bbl6sSX1ZsmA9t4akpnv99GR0D9OEE-0U5NYKGkGhzI=> For help, send mail to the list administrators: Scott Douglas <sdoug...@ieee.org<mailto:sdoug...@ieee.org>> Mike Cantwell <mcantw...@ieee.org<mailto:mcantw...@ieee.org>> For policy questions, send mail to: Jim Bacher <j.bac...@ieee.org<mailto:j.bac...@ieee.org>> David Heald <dhe...@gmail.com<mailto:dhe...@gmail.com>> This e-mail contains privileged and confidential information intended for the use of the addressees named above. If you are not the intended recipient of this e-mail, you are hereby notified that you must not disseminate, copy or take any action in respect of any information contained in it. If you have received this e-mail in error, please notify the sender immediately by e-mail and immediately destroy this e-mail and its attachments. - This message is from the IEEE Product Safety Engineering Society emc-pstc discussion list. To post a message to the list, send your e-mail to <emc-p...@ieee.org> All emc-pstc postings are archived and searchable on the web at: http://www.ieee-pses.org/emc-pstc.html Attachments are not permitted but the IEEE PSES Online Communities site at http://product-compliance.oc.ieee.org/ can be used for graphics (in well-used formats), large files, etc.
Re: [PSES] Active Implantable Medical Devices
ETSI EN 301 489 series addresses 3.1 of the RED, not 3.2, so use of the draft versions should be acceptable without a NB. The first sentence of the scope of ETSI EN 301 489-1: The present document covers the essential requirements of article 3.1(b) of Directive 2014/53/EU [i.1] and article 6 of Directive 2014/30/EU [i.2] for radio equipment and associated ancillary equipment, excluding broadcast receivers, in respect of ElectroMagnetic Compatibility (EMC). As for DC magnetic fields, those are being covered in other standards than the ETSI. No communication, so not really an RED problem. Instead, look at AAMI PC69, EN 45502-2-1, or ISO 14708-3. All test implantables to 50 mT at DC. Thanks, David From: Grace Lin [mailto:graceli...@gmail.com] Sent: 03 April 2017 06:35 To: EMC-PSTC@LISTSERV.IEEE.ORG<mailto:EMC-PSTC@LISTSERV.IEEE.ORG> Subject: [PSES] Active Implantable Medical Devices Dear Members, For active implantable medical devices to comply with the RED requirements: 1. Do all active implantable medical devices require Notified Body’s involvement per RED? From online search (https://ec.europa.eu/growth/single-market/ce-marking/manufacturers_en), it seems the answer is yes. Could you please confirm? 2. The 301 489 series EMC standards covering RED have not been published/harmonized. For those manufacturers that do not want to wait for the harmonized standards to be published, should the manufacturers proceed with the DRAFT version of the standards (and send to NB)? 3. Do you have any idea when 301 489 series standards covering RED will be published? The current status is “On Approval”. Thank you very much for your time and I look forward to hearing from you. Best regards, Grace Lin - This message is from the IEEE Product Safety Engineering Society emc-pstc discussion list. To post a message to the list, send your e-mail to <emc-p...@ieee.org<mailto:emc-p...@ieee.org>> All emc-pstc postings are archived and searchable on the web at: http://www.ieee-pses.org/emc-pstc.html Attachments are not permitted but the IEEE PSES Online Communities site at http://product-compliance.oc.ieee.org/ can be used for graphics (in well-used formats), large files, etc. Website: http://www.ieee-pses.org/ Instructions: http://www.ieee-pses.org/list.html (including how to unsubscribe)<http://www.ieee-pses.org/list.html> List rules: http://www.ieee-pses.org/listrules.html For help, send mail to the list administrators: Scott Douglas <sdoug...@ieee.org<mailto:sdoug...@ieee.org>> Mike Cantwell <mcantw...@ieee.org<mailto:mcantw...@ieee.org>> For policy questions, send mail to: Jim Bacher <j.bac...@ieee.org<mailto:j.bac...@ieee.org>> David Heald <dhe...@gmail.com<mailto:dhe...@gmail.com>> - This message is from the IEEE Product Safety Engineering Society emc-pstc discussion list. To post a message to the list, send your e-mail to <emc-p...@ieee.org<mailto:emc-p...@ieee.org>> All emc-pstc postings are archived and searchable on the web at: http://www.ieee-pses.org/emc-pstc.html Attachments are not permitted but the IEEE PSES Online Communities site at http://product-compliance.oc.ieee.org/ can be used for graphics (in well-used formats), large files, etc. Website: http://www.ieee-pses.org/ Instructions: http://www.ieee-pses.org/list.html (including how to unsubscribe)<http://www.ieee-pses.org/list.html> List rules: http://www.ieee-pses.org/listrules.html For help, send mail to the list administrators: Scott Douglas <sdoug...@ieee.org<mailto:sdoug...@ieee.org>> Mike Cantwell <mcantw...@ieee.org<mailto:mcantw...@ieee.org>> For policy questions, send mail to: Jim Bacher <j.bac...@ieee.org<mailto:j.bac...@ieee.org>> David Heald <dhe...@gmail.com<mailto:dhe...@gmail.com>> - This message is from the IEEE Product Safety Engineering Society emc-pstc discussion list. To post a message to the list, send your e-mail to <emc-p...@ieee.org> All emc-pstc postings are archived and searchable on the web at: http://www.ieee-pses.org/emc-pstc.html Attachments are not permitted but the IEEE PSES Online Communities site at http://product-compliance.oc.ieee.org/ can be used for graphics (in well-used formats), large files, etc. Website: http://www.ieee-pses.org/ Instructions: http://www.ieee-pses.org/list.html (including how to unsubscribe) List rules: http://www.ieee-pses.org/listrules.html For help, send mail to the list administrators: Scott Douglas <sdoug...@ieee.org> Mike Cantwell <mcantw...@ieee.org> For policy questions, send mail to: Jim Bacher: <j.bac...@ieee.org> David Heald: <dhe...@gmail.com>
Re: [PSES] Active Implantable Medical Devices
Grace I pressed “send” too early: For devices operating in the 2483.5 to 2500 MHz band, ETSI EN 301 489-35 V2.1.1<http://www.etsi.org/deliver/etsi_en/301400_301499/30148935/02.01.01_60/en_30148935v020101p.pdf> has been published by ETSI, and it is very likely that this will be listed in the OJ in April or May. However, a revision is already underway with Draft ETSI EN 301 489-35 V2.2.0<http://www.etsi.org/deliver/etsi_en/301400_301499/30148935/02.02.00_20/en_30148935v020200a.pdf> due to become V2.2.1 before being listed in the OJ next year. It would make sense to include this latest draft in your testing, so that you could list V2.1.1 on you DoC and update to v2.2.1 in due course with little, or perhaps no, additional testing. Regards Charlie Charlie Blackham Sulis Consultants Ltd Tel: +44 (0)7946 624317 Web: www.sulisconsultants.com<https://outlook.hslive.net/owa/redir.aspx?C=02be3bf3e3a544d1bdf7b6c99fbd12f5=http%3a%2f%2fwww.sulisconsultants.com%2f> Registered in England and Wales, number 05466247 From: Charlie Blackham [mailto:char...@sulisconsultants.com] Sent: 03 April 2017 13:18 To: EMC-PSTC@LISTSERV.IEEE.ORG Subject: Re: [PSES] Active Implantable Medical Devices Grace 1. The RED only requires the use of a Notified Body where Harmonised Standards have not been fully applied for article 3.2 (Radio Spectrum), as per article 17(3). Medical Device Directive requirements for Notified Bodies apply separately. 2. Whilst latest draft standards probably represent “state of the art”, they may change, and the manufacturer is free to choose the most appropriate standard, without having to consult a RED Notified Body 3. The progress of all ETSI standards being developed for RED can be checked at: https://portal.etsi.org/webapp/WorkProgram/WorkItemPlan.asp?titleType=all=HIGHVERSION_ALL==TRUE=2014%2F53%2FEU_SUB_TB=True_MOVED_ON=_FLG=N_BOOLEAN=OR_BOOLEAN=OR_BOOLEAN=OR_OUTDATED==Search=FALSE=FALSE_TYPE=RPLAN=ALL regards Charlie Charlie Blackham Sulis Consultants Ltd Tel: +44 (0)7946 624317 Web: www.sulisconsultants.com<https://outlook.hslive.net/owa/redir.aspx?C=02be3bf3e3a544d1bdf7b6c99fbd12f5=http%3a%2f%2fwww.sulisconsultants.com%2f> Registered in England and Wales, number 05466247 From: Grace Lin [mailto:graceli...@gmail.com] Sent: 03 April 2017 06:35 To: EMC-PSTC@LISTSERV.IEEE.ORG<mailto:EMC-PSTC@LISTSERV.IEEE.ORG> Subject: [PSES] Active Implantable Medical Devices Dear Members, For active implantable medical devices to comply with the RED requirements: 1. Do all active implantable medical devices require Notified Body’s involvement per RED? From online search (https://ec.europa.eu/growth/single-market/ce-marking/manufacturers_en), it seems the answer is yes. Could you please confirm? 2. The 301 489 series EMC standards covering RED have not been published/harmonized. For those manufacturers that do not want to wait for the harmonized standards to be published, should the manufacturers proceed with the DRAFT version of the standards (and send to NB)? 3. Do you have any idea when 301 489 series standards covering RED will be published? The current status is “On Approval”. Thank you very much for your time and I look forward to hearing from you. Best regards, Grace Lin - This message is from the IEEE Product Safety Engineering Society emc-pstc discussion list. To post a message to the list, send your e-mail to <emc-p...@ieee.org<mailto:emc-p...@ieee.org>> All emc-pstc postings are archived and searchable on the web at: http://www.ieee-pses.org/emc-pstc.html Attachments are not permitted but the IEEE PSES Online Communities site at http://product-compliance.oc.ieee.org/ can be used for graphics (in well-used formats), large files, etc. Website: http://www.ieee-pses.org/ Instructions: http://www.ieee-pses.org/list.html (including how to unsubscribe)<http://www.ieee-pses.org/list.html> List rules: http://www.ieee-pses.org/listrules.html For help, send mail to the list administrators: Scott Douglas <sdoug...@ieee.org<mailto:sdoug...@ieee.org>> Mike Cantwell <mcantw...@ieee.org<mailto:mcantw...@ieee.org>> For policy questions, send mail to: Jim Bacher <j.bac...@ieee.org<mailto:j.bac...@ieee.org>> David Heald <dhe...@gmail.com<mailto:dhe...@gmail.com>> - This message is from the IEEE Product Safety Engineering Society emc-pstc discussion list. To post a message to the list, send your e-mail to <emc-p...@ieee.org<mailto:emc-p...@ieee.org>> All emc-pstc postings are archived and searchable on the web at: http://www.ieee-pses.org/emc-pstc.html Attachments are not permitted but the IEEE PSES Online Communities site at http://product-compliance.oc.ieee.org/ can be used for graphics (in well-
Re: [PSES] Active Implantable Medical Devices
Grace 1. The RED only requires the use of a Notified Body where Harmonised Standards have not been fully applied for article 3.2 (Radio Spectrum), as per article 17(3). Medical Device Directive requirements for Notified Bodies apply separately. 2. Whilst latest draft standards probably represent “state of the art”, they may change, and the manufacturer is free to choose the most appropriate standard, without having to consult a RED Notified Body 3. The progress of all ETSI standards being developed for RED can be checked at: https://portal.etsi.org/webapp/WorkProgram/WorkItemPlan.asp?titleType=all=HIGHVERSION_ALL==TRUE=2014%2F53%2FEU_SUB_TB=True_MOVED_ON=_FLG=N_BOOLEAN=OR_BOOLEAN=OR_BOOLEAN=OR_OUTDATED==Search=FALSE=FALSE_TYPE=RPLAN=ALL regards Charlie Charlie Blackham Sulis Consultants Ltd Tel: +44 (0)7946 624317 Web: www.sulisconsultants.com<https://outlook.hslive.net/owa/redir.aspx?C=02be3bf3e3a544d1bdf7b6c99fbd12f5=http%3a%2f%2fwww.sulisconsultants.com%2f> Registered in England and Wales, number 05466247 From: Grace Lin [mailto:graceli...@gmail.com] Sent: 03 April 2017 06:35 To: EMC-PSTC@LISTSERV.IEEE.ORG Subject: [PSES] Active Implantable Medical Devices Dear Members, For active implantable medical devices to comply with the RED requirements: 1. Do all active implantable medical devices require Notified Body’s involvement per RED? From online search (https://ec.europa.eu/growth/single-market/ce-marking/manufacturers_en), it seems the answer is yes. Could you please confirm? 2. The 301 489 series EMC standards covering RED have not been published/harmonized. For those manufacturers that do not want to wait for the harmonized standards to be published, should the manufacturers proceed with the DRAFT version of the standards (and send to NB)? 3. Do you have any idea when 301 489 series standards covering RED will be published? The current status is “On Approval”. Thank you very much for your time and I look forward to hearing from you. Best regards, Grace Lin - This message is from the IEEE Product Safety Engineering Society emc-pstc discussion list. To post a message to the list, send your e-mail to <emc-p...@ieee.org<mailto:emc-p...@ieee.org>> All emc-pstc postings are archived and searchable on the web at: http://www.ieee-pses.org/emc-pstc.html Attachments are not permitted but the IEEE PSES Online Communities site at http://product-compliance.oc.ieee.org/ can be used for graphics (in well-used formats), large files, etc. Website: http://www.ieee-pses.org/ Instructions: http://www.ieee-pses.org/list.html (including how to unsubscribe)<http://www.ieee-pses.org/list.html> List rules: http://www.ieee-pses.org/listrules.html For help, send mail to the list administrators: Scott Douglas <sdoug...@ieee.org<mailto:sdoug...@ieee.org>> Mike Cantwell <mcantw...@ieee.org<mailto:mcantw...@ieee.org>> For policy questions, send mail to: Jim Bacher <j.bac...@ieee.org<mailto:j.bac...@ieee.org>> David Heald <dhe...@gmail.com<mailto:dhe...@gmail.com>> - This message is from the IEEE Product Safety Engineering Society emc-pstc discussion list. To post a message to the list, send your e-mail to <emc-p...@ieee.org> All emc-pstc postings are archived and searchable on the web at: http://www.ieee-pses.org/emc-pstc.html Attachments are not permitted but the IEEE PSES Online Communities site at http://product-compliance.oc.ieee.org/ can be used for graphics (in well-used formats), large files, etc. Website: http://www.ieee-pses.org/ Instructions: http://www.ieee-pses.org/list.html (including how to unsubscribe) List rules: http://www.ieee-pses.org/listrules.html For help, send mail to the list administrators: Scott Douglas <sdoug...@ieee.org> Mike Cantwell <mcantw...@ieee.org> For policy questions, send mail to: Jim Bacher: <j.bac...@ieee.org> David Heald: <dhe...@gmail.com>
Re: [PSES] Active Implantable Medical Devices
John TC 106 is for human exposure I was thinking of EMC issues - seems a gap here. Andy -Original Message- From: John Woodgate [mailto:jmw1...@btinternet.com] Sent: 03 April 2017 09:45 To: McCallum, Andy <andy.mccal...@mottmac.com>; EMC-PSTC@LISTSERV.IEEE.ORG Subject: RE: [PSES] Active Implantable Medical Devices DC magnetic fields are outside the scope of ETSI, and fall within that of IEC TC106. Remember that the question was about the RED, and no-one as far as I know, uses DC magnetic fields for radio communication. I might except Gaia, because a very broad definition of radio might include communication between the geomagnetic field and compass needles. (;-) But that's not strictly DC, as it is well-known to vary, albeit rather slowly. With best wishes DESIGN IT IN! OOO – Own Opinions Only www.jmwa.demon.co.uk J M Woodgate and Associates Rayleigh England Sylvae in aeternum manent. -Original Message- From: McCallum, Andy [mailto:andy.mccal...@mottmac.com] Sent: Monday, April 3, 2017 9:23 AM To: EMC-PSTC@LISTSERV.IEEE.ORG Subject: Re: [PSES] Active Implantable Medical Devices Hi Just had a brief scan of the standard and it doesn’t appear to cover immunity to DC Magnetic fields. I would have thought that many patients fitted with these devices would be having on going medical treatment such that they may require MRI scans, where they could be subject to magnetic fields of greater than 7T. Also wouldn't it be sensible to ensure that all devices met the ICNIRP exposure levels ? Regards Andy -Original Message- From: Dürrer Bernd [mailto:bernd.duer...@wilo.com] Sent: 03 April 2017 08:37 To: EMC-PSTC@LISTSERV.IEEE.ORG Subject: [PSES] AW: [PSES] Active Implantable Medical Devices Dear Grace, you can check the status and scheduled publication date (incl. citation in the OJ) of ETSI standards at https://portal.etsi.org/webapp/WorkProgram/SimpleSearch/QueryForm.asp For the 301 489 series standards, click on "EN - European standard (Telecommunication series)" in the Type field and enter "301 489" (incl. the blank in the middle) as number. The search will return a table of work items: The "Status" column contains hyperlinks to the schedule of the standard, that includes the target date for citation in the OJ at the bottom of the table. Kind regards, Bernd Von: Grace Lin [mailto:graceli...@gmail.com] Gesendet: Montag, 3. April 2017 07:35 An: EMC-PSTC@LISTSERV.IEEE.ORG Betreff: [PSES] Active Implantable Medical Devices Dear Members, For active implantable medical devices to comply with the RED requirements: 1. Do all active implantable medical devices require Notified Body’s involvement per RED? From online search (https://ec.europa.eu/growth/single-market/ce-marking/manufacturers_en), it seems the answer is yes. Could you please confirm? 2. The 301 489 series EMC standards covering RED have not been published/harmonized. For those manufacturers that do not want to wait for the harmonized standards to be published, should the manufacturers proceed with the DRAFT version of the standards (and send to NB)? 3. Do you have any idea when 301 489 series standards covering RED will be published? The current status is “On Approval”. Thank you very much for your time and I look forward to hearing from you. Best regards, Grace Lin - This message is from the IEEE Product Safety Engineering Society emc-pstc discussion list. To post a message to the list, send your e-mail to <mailto:emc-p...@ieee.org> All emc-pstc postings are archived and searchable on the web at: http://www.ieee-pses.org/emc-pstc.html Attachments are not permitted but the IEEE PSES Online Communities site at http://product-compliance.oc.ieee.org/ can be used for graphics (in well-used formats), large files, etc. Website: http://www.ieee-pses.org/ Instructions: http://www.ieee-pses.org/list.html List rules: http://www.ieee-pses.org/listrules.html For help, send mail to the list administrators: Scott Douglas <mailto:sdoug...@ieee.org> Mike Cantwell <mailto:mcantw...@ieee.org> For policy questions, send mail to: Jim Bacher <mailto:j.bac...@ieee.org> David Heald <mailto:dhe...@gmail.com> WILO SE Nortkirchenstrasse 100, 44263 Dortmund Amtsgericht Dortmund, HRB 21356 www.wilo.com Vorstand/Executive Board: Oliver Hermes (Vorsitzender/Chairman), Dr. Markus Beukenberg, Carsten Krumm, Eric Lachambre, Mathias Weyers Aufsichtsratsvorsitzender/Chairman of the Supervisory Board: Prof. Dr. Norbert Wieselhuber Dieses Dokument ist vertraulich zu behandeln. Die Weitergabe sowie Vervielfaeltigung, Verwertung und Mitteilung seines Inhalts ist nur mit unserer ausdruecklichen Genehmigung gestattet. Alle Rechte vorbehalten, insbesondere fuer den Fall der Schutzrechtsanmeldung. This document has to be treated
Re: [PSES] AW: [PSES] Active Implantable Medical Devices
It's true that there is such information, but I would treat it with caution. There can/will be delays in ETSI and even more in the citation process. Current hot topics are 'reasonably foreseeable' and 'dated references'. With best wishes DESIGN IT IN! OOO – Own Opinions Only www.jmwa.demon.co.uk J M Woodgate and Associates Rayleigh England Sylvae in aeternum manent. -Original Message- From: Dürrer Bernd [mailto:bernd.duer...@wilo.com] Sent: Monday, April 3, 2017 8:37 AM To: EMC-PSTC@LISTSERV.IEEE.ORG Subject: [PSES] AW: [PSES] Active Implantable Medical Devices Dear Grace, you can check the status and scheduled publication date (incl. citation in the OJ) of ETSI standards at https://portal.etsi.org/webapp/WorkProgram/SimpleSearch/QueryForm.asp For the 301 489 series standards, click on "EN - European standard (Telecommunication series)" in the Type field and enter "301 489" (incl. the blank in the middle) as number. The search will return a table of work items: The "Status" column contains hyperlinks to the schedule of the standard, that includes the target date for citation in the OJ at the bottom of the table. Kind regards, Bernd Von: Grace Lin [mailto:graceli...@gmail.com] Gesendet: Montag, 3. April 2017 07:35 An: EMC-PSTC@LISTSERV.IEEE.ORG Betreff: [PSES] Active Implantable Medical Devices Dear Members, For active implantable medical devices to comply with the RED requirements: 1. Do all active implantable medical devices require Notified Body’s involvement per RED? From online search (https://ec.europa.eu/growth/single-market/ce-marking/manufacturers_en), it seems the answer is yes. Could you please confirm? 2. The 301 489 series EMC standards covering RED have not been published/harmonized. For those manufacturers that do not want to wait for the harmonized standards to be published, should the manufacturers proceed with the DRAFT version of the standards (and send to NB)? 3. Do you have any idea when 301 489 series standards covering RED will be published? The current status is “On Approval”. Thank you very much for your time and I look forward to hearing from you. Best regards, Grace Lin - This message is from the IEEE Product Safety Engineering Society emc-pstc discussion list. To post a message to the list, send your e-mail to <mailto:emc-p...@ieee.org> All emc-pstc postings are archived and searchable on the web at: http://www.ieee-pses.org/emc-pstc.html Attachments are not permitted but the IEEE PSES Online Communities site at http://product-compliance.oc.ieee.org/ can be used for graphics (in well-used formats), large files, etc. Website: http://www.ieee-pses.org/ Instructions: http://www.ieee-pses.org/list.html List rules: http://www.ieee-pses.org/listrules.html For help, send mail to the list administrators: Scott Douglas <mailto:sdoug...@ieee.org> Mike Cantwell <mailto:mcantw...@ieee.org> For policy questions, send mail to: Jim Bacher <mailto:j.bac...@ieee.org> David Heald <mailto:dhe...@gmail.com> WILO SE Nortkirchenstrasse 100, 44263 Dortmund Amtsgericht Dortmund, HRB 21356 www.wilo.com Vorstand/Executive Board: Oliver Hermes (Vorsitzender/Chairman), Dr. Markus Beukenberg, Carsten Krumm, Eric Lachambre, Mathias Weyers Aufsichtsratsvorsitzender/Chairman of the Supervisory Board: Prof. Dr. Norbert Wieselhuber Dieses Dokument ist vertraulich zu behandeln. Die Weitergabe sowie Vervielfaeltigung, Verwertung und Mitteilung seines Inhalts ist nur mit unserer ausdruecklichen Genehmigung gestattet. Alle Rechte vorbehalten, insbesondere fuer den Fall der Schutzrechtsanmeldung. This document has to be treated confidentially. Its contents are not to be passed on, duplicated, exploited or disclosed without our expressed permission. All rights reserved, especially the right to apply for protective rights. - This message is from the IEEE Product Safety Engineering Society emc-pstc discussion list. To post a message to the list, send your e-mail to <emc-p...@ieee.org> All emc-pstc postings are archived and searchable on the web at: http://www.ieee-pses.org/emc-pstc.html Attachments are not permitted but the IEEE PSES Online Communities site at http://product-compliance.oc.ieee.org/ can be used for graphics (in well-used formats), large files, etc. Website: http://www.ieee-pses.org/ Instructions: http://www.ieee-pses.org/list.html (including how to unsubscribe) List rules: http://www.ieee-pses.org/listrules.html For help, send mail to the list administrators: Scott Douglas <sdoug...@ieee.org> Mike Cantwell <mcantw...@ieee.org> For policy questions, send mail to: Jim Bacher: <j.bac...@ieee.org> David Heald: <dhe...@gmail.com> - ---
Re: [PSES] Active Implantable Medical Devices
OK. Neither CISPR or IEC TC77 regard the DC magnetic field as a significant EMI issue. In fact, it's really outside the scope of CISPR, although CISPR/B doesn't take much notice of that. The subject might be within the scope of IEC 60601-1-2, but there is no indication that it is planned to introduce requirements. With best wishes DESIGN IT IN! OOO – Own Opinions Only www.jmwa.demon.co.uk J M Woodgate and Associates Rayleigh England Sylvae in aeternum manent. -Original Message- From: McCallum, Andy [mailto:andy.mccal...@mottmac.com] Sent: Monday, April 3, 2017 10:00 AM To: John Woodgate <jmw1...@btinternet.com>; EMC-PSTC@LISTSERV.IEEE.ORG Subject: RE: [PSES] Active Implantable Medical Devices John TC 106 is for human exposure I was thinking of EMC issues - seems a gap here. Andy -Original Message- From: John Woodgate [mailto:jmw1...@btinternet.com] Sent: 03 April 2017 09:45 To: McCallum, Andy <andy.mccal...@mottmac.com>; EMC-PSTC@LISTSERV.IEEE.ORG Subject: RE: [PSES] Active Implantable Medical Devices DC magnetic fields are outside the scope of ETSI, and fall within that of IEC TC106. Remember that the question was about the RED, and no-one as far as I know, uses DC magnetic fields for radio communication. I might except Gaia, because a very broad definition of radio might include communication between the geomagnetic field and compass needles. (;-) But that's not strictly DC, as it is well-known to vary, albeit rather slowly. With best wishes DESIGN IT IN! OOO – Own Opinions Only www.jmwa.demon.co.uk J M Woodgate and Associates Rayleigh England Sylvae in aeternum manent. -Original Message- From: McCallum, Andy [mailto:andy.mccal...@mottmac.com] Sent: Monday, April 3, 2017 9:23 AM To: EMC-PSTC@LISTSERV.IEEE.ORG Subject: Re: [PSES] Active Implantable Medical Devices Hi Just had a brief scan of the standard and it doesn’t appear to cover immunity to DC Magnetic fields. I would have thought that many patients fitted with these devices would be having on going medical treatment such that they may require MRI scans, where they could be subject to magnetic fields of greater than 7T. Also wouldn't it be sensible to ensure that all devices met the ICNIRP exposure levels ? Regards Andy -Original Message- From: Dürrer Bernd [mailto:bernd.duer...@wilo.com] Sent: 03 April 2017 08:37 To: EMC-PSTC@LISTSERV.IEEE.ORG Subject: [PSES] AW: [PSES] Active Implantable Medical Devices Dear Grace, you can check the status and scheduled publication date (incl. citation in the OJ) of ETSI standards at https://portal.etsi.org/webapp/WorkProgram/SimpleSearch/QueryForm.asp For the 301 489 series standards, click on "EN - European standard (Telecommunication series)" in the Type field and enter "301 489" (incl. the blank in the middle) as number. The search will return a table of work items: The "Status" column contains hyperlinks to the schedule of the standard, that includes the target date for citation in the OJ at the bottom of the table. Kind regards, Bernd Von: Grace Lin [mailto:graceli...@gmail.com] Gesendet: Montag, 3. April 2017 07:35 An: EMC-PSTC@LISTSERV.IEEE.ORG Betreff: [PSES] Active Implantable Medical Devices Dear Members, For active implantable medical devices to comply with the RED requirements: 1. Do all active implantable medical devices require Notified Body’s involvement per RED? From online search (https://ec.europa.eu/growth/single-market/ce-marking/manufacturers_en), it seems the answer is yes. Could you please confirm? 2. The 301 489 series EMC standards covering RED have not been published/harmonized. For those manufacturers that do not want to wait for the harmonized standards to be published, should the manufacturers proceed with the DRAFT version of the standards (and send to NB)? 3. Do you have any idea when 301 489 series standards covering RED will be published? The current status is “On Approval”. Thank you very much for your time and I look forward to hearing from you. Best regards, Grace Lin - This message is from the IEEE Product Safety Engineering Society emc-pstc discussion list. To post a message to the list, send your e-mail to <mailto:emc-p...@ieee.org> All emc-pstc postings are archived and searchable on the web at: http://www.ieee-pses.org/emc-pstc.html Attachments are not permitted but the IEEE PSES Online Communities site at http://product-compliance.oc.ieee.org/ can be used for graphics (in well-used formats), large files, etc. Website: http://www.ieee-pses.org/ Instructions: http://www.ieee-pses.org/list.html List rules: http://www.ieee-pses.org/listrules.html For help, send mail to the list administrators: Scott Douglas <mailto:sdoug...@ieee.org> Mike Cantwell <mailto:mcantw...@ieee.org> For policy questions, send mail to: Jim Bacher <
Re: [PSES] Active Implantable Medical Devices
DC magnetic fields are outside the scope of ETSI, and fall within that of IEC TC106. Remember that the question was about the RED, and no-one as far as I know, uses DC magnetic fields for radio communication. I might except Gaia, because a very broad definition of radio might include communication between the geomagnetic field and compass needles. (;-) But that's not strictly DC, as it is well-known to vary, albeit rather slowly. With best wishes DESIGN IT IN! OOO – Own Opinions Only www.jmwa.demon.co.uk J M Woodgate and Associates Rayleigh England Sylvae in aeternum manent. -Original Message- From: McCallum, Andy [mailto:andy.mccal...@mottmac.com] Sent: Monday, April 3, 2017 9:23 AM To: EMC-PSTC@LISTSERV.IEEE.ORG Subject: Re: [PSES] Active Implantable Medical Devices Hi Just had a brief scan of the standard and it doesn’t appear to cover immunity to DC Magnetic fields. I would have thought that many patients fitted with these devices would be having on going medical treatment such that they may require MRI scans, where they could be subject to magnetic fields of greater than 7T. Also wouldn't it be sensible to ensure that all devices met the ICNIRP exposure levels ? Regards Andy -Original Message- From: Dürrer Bernd [mailto:bernd.duer...@wilo.com] Sent: 03 April 2017 08:37 To: EMC-PSTC@LISTSERV.IEEE.ORG Subject: [PSES] AW: [PSES] Active Implantable Medical Devices Dear Grace, you can check the status and scheduled publication date (incl. citation in the OJ) of ETSI standards at https://portal.etsi.org/webapp/WorkProgram/SimpleSearch/QueryForm.asp For the 301 489 series standards, click on "EN - European standard (Telecommunication series)" in the Type field and enter "301 489" (incl. the blank in the middle) as number. The search will return a table of work items: The "Status" column contains hyperlinks to the schedule of the standard, that includes the target date for citation in the OJ at the bottom of the table. Kind regards, Bernd Von: Grace Lin [mailto:graceli...@gmail.com] Gesendet: Montag, 3. April 2017 07:35 An: EMC-PSTC@LISTSERV.IEEE.ORG Betreff: [PSES] Active Implantable Medical Devices Dear Members, For active implantable medical devices to comply with the RED requirements: 1. Do all active implantable medical devices require Notified Body’s involvement per RED? From online search (https://ec.europa.eu/growth/single-market/ce-marking/manufacturers_en), it seems the answer is yes. Could you please confirm? 2. The 301 489 series EMC standards covering RED have not been published/harmonized. For those manufacturers that do not want to wait for the harmonized standards to be published, should the manufacturers proceed with the DRAFT version of the standards (and send to NB)? 3. Do you have any idea when 301 489 series standards covering RED will be published? The current status is “On Approval”. Thank you very much for your time and I look forward to hearing from you. Best regards, Grace Lin - This message is from the IEEE Product Safety Engineering Society emc-pstc discussion list. To post a message to the list, send your e-mail to <mailto:emc-p...@ieee.org> All emc-pstc postings are archived and searchable on the web at: http://www.ieee-pses.org/emc-pstc.html Attachments are not permitted but the IEEE PSES Online Communities site at http://product-compliance.oc.ieee.org/ can be used for graphics (in well-used formats), large files, etc. Website: http://www.ieee-pses.org/ Instructions: http://www.ieee-pses.org/list.html List rules: http://www.ieee-pses.org/listrules.html For help, send mail to the list administrators: Scott Douglas <mailto:sdoug...@ieee.org> Mike Cantwell <mailto:mcantw...@ieee.org> For policy questions, send mail to: Jim Bacher <mailto:j.bac...@ieee.org> David Heald <mailto:dhe...@gmail.com> WILO SE Nortkirchenstrasse 100, 44263 Dortmund Amtsgericht Dortmund, HRB 21356 www.wilo.com Vorstand/Executive Board: Oliver Hermes (Vorsitzender/Chairman), Dr. Markus Beukenberg, Carsten Krumm, Eric Lachambre, Mathias Weyers Aufsichtsratsvorsitzender/Chairman of the Supervisory Board: Prof. Dr. Norbert Wieselhuber Dieses Dokument ist vertraulich zu behandeln. Die Weitergabe sowie Vervielfaeltigung, Verwertung und Mitteilung seines Inhalts ist nur mit unserer ausdruecklichen Genehmigung gestattet. Alle Rechte vorbehalten, insbesondere fuer den Fall der Schutzrechtsanmeldung. This document has to be treated confidentially. Its contents are not to be passed on, duplicated, exploited or disclosed without our expressed permission. All rights reserved, especially the right to apply for protective rights. - This message is from the IEEE Product Safety Engineering Society em
Re: [PSES] Active Implantable Medical Devices
Hi Just had a brief scan of the standard and it doesn’t appear to cover immunity to DC Magnetic fields. I would have thought that many patients fitted with these devices would be having on going medical treatment such that they may require MRI scans, where they could be subject to magnetic fields of greater than 7T. Also wouldn't it be sensible to ensure that all devices met the ICNIRP exposure levels ? Regards Andy -Original Message- From: Dürrer Bernd [mailto:bernd.duer...@wilo.com] Sent: 03 April 2017 08:37 To: EMC-PSTC@LISTSERV.IEEE.ORG Subject: [PSES] AW: [PSES] Active Implantable Medical Devices Dear Grace, you can check the status and scheduled publication date (incl. citation in the OJ) of ETSI standards at https://portal.etsi.org/webapp/WorkProgram/SimpleSearch/QueryForm.asp For the 301 489 series standards, click on "EN - European standard (Telecommunication series)" in the Type field and enter "301 489" (incl. the blank in the middle) as number. The search will return a table of work items: The "Status" column contains hyperlinks to the schedule of the standard, that includes the target date for citation in the OJ at the bottom of the table. Kind regards, Bernd Von: Grace Lin [mailto:graceli...@gmail.com] Gesendet: Montag, 3. April 2017 07:35 An: EMC-PSTC@LISTSERV.IEEE.ORG Betreff: [PSES] Active Implantable Medical Devices Dear Members, For active implantable medical devices to comply with the RED requirements: 1. Do all active implantable medical devices require Notified Body’s involvement per RED? From online search (https://ec.europa.eu/growth/single-market/ce-marking/manufacturers_en), it seems the answer is yes. Could you please confirm? 2. The 301 489 series EMC standards covering RED have not been published/harmonized. For those manufacturers that do not want to wait for the harmonized standards to be published, should the manufacturers proceed with the DRAFT version of the standards (and send to NB)? 3. Do you have any idea when 301 489 series standards covering RED will be published? The current status is “On Approval”. Thank you very much for your time and I look forward to hearing from you. Best regards, Grace Lin - This message is from the IEEE Product Safety Engineering Society emc-pstc discussion list. To post a message to the list, send your e-mail to <mailto:emc-p...@ieee.org> All emc-pstc postings are archived and searchable on the web at: http://www.ieee-pses.org/emc-pstc.html Attachments are not permitted but the IEEE PSES Online Communities site at http://product-compliance.oc.ieee.org/ can be used for graphics (in well-used formats), large files, etc. Website: http://www.ieee-pses.org/ Instructions: http://www.ieee-pses.org/list.html List rules: http://www.ieee-pses.org/listrules.html For help, send mail to the list administrators: Scott Douglas <mailto:sdoug...@ieee.org> Mike Cantwell <mailto:mcantw...@ieee.org> For policy questions, send mail to: Jim Bacher <mailto:j.bac...@ieee.org> David Heald <mailto:dhe...@gmail.com> WILO SE Nortkirchenstrasse 100, 44263 Dortmund Amtsgericht Dortmund, HRB 21356 www.wilo.com Vorstand/Executive Board: Oliver Hermes (Vorsitzender/Chairman), Dr. Markus Beukenberg, Carsten Krumm, Eric Lachambre, Mathias Weyers Aufsichtsratsvorsitzender/Chairman of the Supervisory Board: Prof. Dr. Norbert Wieselhuber Dieses Dokument ist vertraulich zu behandeln. Die Weitergabe sowie Vervielfaeltigung, Verwertung und Mitteilung seines Inhalts ist nur mit unserer ausdruecklichen Genehmigung gestattet. Alle Rechte vorbehalten, insbesondere fuer den Fall der Schutzrechtsanmeldung. This document has to be treated confidentially. Its contents are not to be passed on, duplicated, exploited or disclosed without our expressed permission. All rights reserved, especially the right to apply for protective rights. - This message is from the IEEE Product Safety Engineering Society emc-pstc discussion list. To post a message to the list, send your e-mail to <emc-p...@ieee.org> All emc-pstc postings are archived and searchable on the web at: http://www.ieee-pses.org/emc-pstc.html Attachments are not permitted but the IEEE PSES Online Communities site at http://product-compliance.oc.ieee.org/ can be used for graphics (in well-used formats), large files, etc. Website: http://www.ieee-pses.org/ Instructions: http://www.ieee-pses.org/list.html (including how to unsubscribe) List rules: http://www.ieee-pses.org/listrules.html For help, send mail to the list administrators: Scott Douglas <sdoug...@ieee.org> Mike Cantwell <mcantw...@ieee.org> For policy questions, send mail to: Jim Bacher: <j.bac...@ieee.org> David Heald: <dhe...@gmai
[PSES] AW: [PSES] Active Implantable Medical Devices
Dear Grace, you can check the status and scheduled publication date (incl. citation in the OJ) of ETSI standards at https://portal.etsi.org/webapp/WorkProgram/SimpleSearch/QueryForm.asp For the 301 489 series standards, click on "EN - European standard (Telecommunication series)" in the Type field and enter "301 489" (incl. the blank in the middle) as number. The search will return a table of work items: The "Status" column contains hyperlinks to the schedule of the standard, that includes the target date for citation in the OJ at the bottom of the table. Kind regards, Bernd Von: Grace Lin [mailto:graceli...@gmail.com] Gesendet: Montag, 3. April 2017 07:35 An: EMC-PSTC@LISTSERV.IEEE.ORG Betreff: [PSES] Active Implantable Medical Devices Dear Members, For active implantable medical devices to comply with the RED requirements: 1. Do all active implantable medical devices require Notified Body’s involvement per RED? From online search (https://ec.europa.eu/growth/single-market/ce-marking/manufacturers_en), it seems the answer is yes. Could you please confirm? 2. The 301 489 series EMC standards covering RED have not been published/harmonized. For those manufacturers that do not want to wait for the harmonized standards to be published, should the manufacturers proceed with the DRAFT version of the standards (and send to NB)? 3. Do you have any idea when 301 489 series standards covering RED will be published? The current status is “On Approval”. Thank you very much for your time and I look forward to hearing from you. Best regards, Grace Lin - This message is from the IEEE Product Safety Engineering Society emc-pstc discussion list. To post a message to the list, send your e-mail to <mailto:emc-p...@ieee.org> All emc-pstc postings are archived and searchable on the web at: http://www.ieee-pses.org/emc-pstc.html Attachments are not permitted but the IEEE PSES Online Communities site at http://product-compliance.oc.ieee.org/ can be used for graphics (in well-used formats), large files, etc. Website: http://www.ieee-pses.org/ Instructions: http://www.ieee-pses.org/list.html List rules: http://www.ieee-pses.org/listrules.html For help, send mail to the list administrators: Scott Douglas <mailto:sdoug...@ieee.org> Mike Cantwell <mailto:mcantw...@ieee.org> For policy questions, send mail to: Jim Bacher <mailto:j.bac...@ieee.org> David Heald <mailto:dhe...@gmail.com> WILO SE Nortkirchenstrasse 100, 44263 Dortmund Amtsgericht Dortmund, HRB 21356 www.wilo.com Vorstand/Executive Board: Oliver Hermes (Vorsitzender/Chairman), Dr. Markus Beukenberg, Carsten Krumm, Eric Lachambre, Mathias Weyers Aufsichtsratsvorsitzender/Chairman of the Supervisory Board: Prof. Dr. Norbert Wieselhuber Dieses Dokument ist vertraulich zu behandeln. Die Weitergabe sowie Vervielfaeltigung, Verwertung und Mitteilung seines Inhalts ist nur mit unserer ausdruecklichen Genehmigung gestattet. Alle Rechte vorbehalten, insbesondere fuer den Fall der Schutzrechtsanmeldung. This document has to be treated confidentially. Its contents are not to be passed on, duplicated, exploited or disclosed without our expressed permission. All rights reserved, especially the right to apply for protective rights. - This message is from the IEEE Product Safety Engineering Society emc-pstc discussion list. To post a message to the list, send your e-mail to <emc-p...@ieee.org> All emc-pstc postings are archived and searchable on the web at: http://www.ieee-pses.org/emc-pstc.html Attachments are not permitted but the IEEE PSES Online Communities site at http://product-compliance.oc.ieee.org/ can be used for graphics (in well-used formats), large files, etc. Website: http://www.ieee-pses.org/ Instructions: http://www.ieee-pses.org/list.html (including how to unsubscribe) List rules: http://www.ieee-pses.org/listrules.html For help, send mail to the list administrators: Scott Douglas <sdoug...@ieee.org> Mike Cantwell <mcantw...@ieee.org> For policy questions, send mail to: Jim Bacher: <j.bac...@ieee.org> David Heald: <dhe...@gmail.com>
Re: [PSES] Active Implantable Medical Devices
This one: ETSI DRAFT FOR APPROVAL - EN 301 489-35 ElectroMagnetic Compatibility (EMC) standard for radio equipment and services; Part 35: Specific requirements for Low Power Active Medical Implants (LP-AMI) operating in the 2 483,5 MHz to 2 500 MHz bands; Harmonised Standard covering the essential requirements of article 3.1(b) of Directive 2014/53/EU (AP) is out for national comment. I guess the ETSI closing date is around June 30, but that could mean that it is not published until September or later. You would have to ask the NB if they will work with a draft. With best wishes DESIGN IT IN! OOO – Own Opinions Only <http://www.jmwa.demon.co.uk/> www.jmwa.demon.co.uk J M Woodgate and Associates Rayleigh England Sylvae in aeternum manent. From: Grace Lin [mailto:graceli...@gmail.com] Sent: Monday, April 3, 2017 6:35 AM To: EMC-PSTC@LISTSERV.IEEE.ORG Subject: [PSES] Active Implantable Medical Devices Dear Members, For active implantable medical devices to comply with the RED requirements: 1. Do all active implantable medical devices require Notified Body’s involvement per RED? From online search (https://ec.europa.eu/growth/single-market/ce-marking/manufacturers_en), it seems the answer is yes. Could you please confirm? 2. The 301 489 series EMC standards covering RED have not been published/harmonized. For those manufacturers that do not want to wait for the harmonized standards to be published, should the manufacturers proceed with the DRAFT version of the standards (and send to NB)? 3. Do you have any idea when 301 489 series standards covering RED will be published? The current status is “On Approval”. Thank you very much for your time and I look forward to hearing from you. Best regards, Grace Lin - This message is from the IEEE Product Safety Engineering Society emc-pstc discussion list. To post a message to the list, send your e-mail to <emc-p...@ieee.org <mailto:emc-p...@ieee.org> > All emc-pstc postings are archived and searchable on the web at: http://www.ieee-pses.org/emc-pstc.html Attachments are not permitted but the IEEE PSES Online Communities site at http://product-compliance.oc.ieee.org/ can be used for graphics (in well-used formats), large files, etc. Website: http://www.ieee-pses.org/ Instructions: http://www.ieee-pses.org/list.html (including how to unsubscribe) <http://www.ieee-pses.org/list.html> List rules: http://www.ieee-pses.org/listrules.html For help, send mail to the list administrators: Scott Douglas <sdoug...@ieee.org <mailto:sdoug...@ieee.org> > Mike Cantwell <mcantw...@ieee.org <mailto:mcantw...@ieee.org> > For policy questions, send mail to: Jim Bacher <j.bac...@ieee.org <mailto:j.bac...@ieee.org> > David Heald <dhe...@gmail.com <mailto:dhe...@gmail.com> > - This message is from the IEEE Product Safety Engineering Society emc-pstc discussion list. To post a message to the list, send your e-mail to <emc-p...@ieee.org> All emc-pstc postings are archived and searchable on the web at: http://www.ieee-pses.org/emc-pstc.html Attachments are not permitted but the IEEE PSES Online Communities site at http://product-compliance.oc.ieee.org/ can be used for graphics (in well-used formats), large files, etc. Website: http://www.ieee-pses.org/ Instructions: http://www.ieee-pses.org/list.html (including how to unsubscribe) List rules: http://www.ieee-pses.org/listrules.html For help, send mail to the list administrators: Scott Douglas <sdoug...@ieee.org> Mike Cantwell <mcantw...@ieee.org> For policy questions, send mail to: Jim Bacher: <j.bac...@ieee.org> David Heald: <dhe...@gmail.com>
[PSES] Active Implantable Medical Devices
Dear Members, For active implantable medical devices to comply with the RED requirements: 1. Do all active implantable medical devices require Notified Body’s involvement per RED? From online search ( https://ec.europa.eu/growth/single-market/ce-marking/manufacturers_en), it seems the answer is yes. Could you please confirm? 2. The 301 489 series EMC standards covering RED have not been published/harmonized. For those manufacturers that do not want to wait for the harmonized standards to be published, should the manufacturers proceed with the DRAFT version of the standards (and send to NB)? 3. Do you have any idea when 301 489 series standards covering RED will be published? The current status is “On Approval”. Thank you very much for your time and I look forward to hearing from you. Best regards, Grace Lin - This message is from the IEEE Product Safety Engineering Society emc-pstc discussion list. To post a message to the list, send your e-mail to <emc-p...@ieee.org> All emc-pstc postings are archived and searchable on the web at: http://www.ieee-pses.org/emc-pstc.html Attachments are not permitted but the IEEE PSES Online Communities site at http://product-compliance.oc.ieee.org/ can be used for graphics (in well-used formats), large files, etc. Website: http://www.ieee-pses.org/ Instructions: http://www.ieee-pses.org/list.html (including how to unsubscribe) List rules: http://www.ieee-pses.org/listrules.html For help, send mail to the list administrators: Scott Douglas <sdoug...@ieee.org> Mike Cantwell <mcantw...@ieee.org> For policy questions, send mail to: Jim Bacher: <j.bac...@ieee.org> David Heald: <dhe...@gmail.com>
Re: [PSES] Interpretation for RoHS2 requirement (exemption of medical devices) of CE Marked device under MDD & transfer or ownership but same product
Hi Leo: It seems to me that all you need to do is change the name on the DOCs. Good luck, have a good Thanksgiving, and best regards, Rich - This message is from the IEEE Product Safety Engineering Society emc-pstc discussion list. To post a message to the list, send your e-mail toAll emc-pstc postings are archived and searchable on the web at: http://www.ieee-pses.org/emc-pstc.html Attachments are not permitted but the IEEE PSES Online Communities site at http://product-compliance.oc.ieee.org/ can be used for graphics (in well-used formats), large files, etc. Website: http://www.ieee-pses.org/ Instructions: http://www.ieee-pses.org/list.html (including how to unsubscribe) List rules: http://www.ieee-pses.org/listrules.html For help, send mail to the list administrators: Scott Douglas Mike Cantwell For policy questions, send mail to: Jim Bacher: David Heald:
[PSES] Interpretation for RoHS2 requirement (exemption of medical devices) of CE Marked device under MDD & transfer or ownership but same product
Hello All, I got an interesting set of questions: If the product, even though the same device, but co. Name has changed with a transfer of assets but with some of the same c-level executives, if they will need to consider the product now needs to meet the RoHS2 directive. See below discussion and questions. They are producing the same medical device as pre 2014 requirement for RoHS2 exemption for medical devices but a new CE Mark as the co. Name has changed slightly. They still need to be able to sell the same product. See detailed questions below. Any specific guidance you can point too in this situation. I can’t find anything specific to this situation which wouldn’t say they are placing on the market newly or recently which means meeting the RoHS2 as a medical device because past the date of the medical device exemption transition of placing on the market (as a new co. Name with a new CE Mark) but I am hoping not needed as a huge set-back for them otherwise. Questions: 1) So, before the co. name change they originally “placed the product on the market” (CE Marked under MDD but as exempted Med Equip from the RoHS2) before the July 2014 exemption date. So, not CE Marked under the RoHS2. They are exempted with this original co. Name for the same product up til July 2019. No question this is ok. 2) They say the product came off the market in April 2015 but if the same device goes back on the market after that would it have to go thru “placing on the market” again as it already happened back before July 2014 (so originally placed on the market)? The question is 2 parted a) if the same co. Name as originally and b) with a new co. Name with the following background info. For them it would put them back a couple years of work to get a newly qualified device that also meets RoHS2 requirements: “Would the newly formed company STILL qualify? Here is the situation: Up Med Corp. (’s the same for both company names) was purchased, and everything became the property of the new owner with the one exception of the UBI number (business license number for US). Patents, product, model numbers, etc. were all purchased AS IS and will be used as such (so, we really will have the same exact model number). The newly formed company is named Up Med Technology Inc., and along with this also came the rights to have the same nomenclature to be able to identify ourselves as “Up Med”. Switching the UBI has caused us to need to get a new CE Mark under our new business name, however, we were given some help/streamlining from the NB since really the only changes are 1) Name 2) Address 3) Executive Management (and even that has 2 original members in it). We are working on getting our 13485 ISO Cert (Stage II audit coming early December).” 3) The existing product (mentioned in item 2) also has some parts that are end of life so they need to replace those parts but the part # or model # of the overall product will not change. The parts they are replacing with may be RoHS compliant so, will this impact the RoHS2 requirements for the product? Will it need to meet the ROHS2 requirements or can it still be exempt? To me this is closer to the requirement, but they would prefer not mention in their justification to the MDD NB nothing about parts being RoHS2 compliant in their justification. Thx much, Leonard (Leo) Eisner, P.E. Principal Consultant, Eisner Safety Consultants Phone: (503) 244-6151 Mobile: (503) 709-8328 Email: l...@eisnersafety.com Website: www.EisnerSafety.com *** Internet E-mail Confidentiality Disclaimer *** This e-mail message may contain privileged or confidential information. If you are not the intended recipient, you may not disclose, use, disseminate, distribute, copy or rely upon this message or attachment in any way. If you received this e-mail message in error, please return by forwarding the message and its attachments to the sender. Eisner Safety Consultants do not accept liability for any errors, omissions, corruption or virus in the contents of this message or any attachments that arise as a result of e-mail transmission. *** - This message is from the IEEE Product Safety Engineering Society emc-pstc discussion list. To post a message to the list, send your e-mail to <emc-p...@ieee.org> All emc-pstc postings are archived and searchable on the web at: http://www.ieee-pses.org/emc-pstc.html Attachments are not permitted but the IEEE PSES Online Communities site at http://product-compliance.oc.ieee.org/ can be used for graphics (in well-used formats), large files, etc. Website: http://www.ieee-pses.org/ Instructions: http://www.ieee-pses.org/list.html (including how to unsubscribe) List rules: http://www.ieee-pses.org/listrules.html Fo
RE: [PSES] Medical devices and Israel
A small correction to Peter answer: The Standards Institution of Israel (SII) would the only body which issue a LETER recognized by Israeli Customs. Tests for import products in Israel can be conducted by any of the Designated Laboratories. This designation is done by the Israeli Standardization Commissioner which belong to the Israeli Ministry of Industry, Trade and Labor. For Medical Electrical Equipment in Israel, IEC 60601-1 (with small deviation-see below) and IEC 60601-1-2 (without deviations) are the accepted standards. 4.7 A. Equipment that is to be connected to the mains, shall be intended for one of the following voltages and frequencies: Nominal frequency of 50Hz. Nominal voltage of 230V for portable and hand-held equipment. Nominal voltage 230V, for one phase equipment with input power not exceeding 4kVA. Nominal voltage 400V, for multiphase equipment. B. It is allowed to connect to the mains other equipment, with the following ratings: One phase equipment, for the range of 220 to 240 Volts. Multiphase equipment, for the range of 380 to 440 Volts. The equipment shall comply with all the requirements of the standard, while being connected to the above mentioned mains frequencies and voltages, as if it was marked for 50Hz and 230 or 400 Volts. 36 Change clause 36. - EMC – to: The equipment shall comply with the requirements of SI 1011 part 1.2 (IEC 60601-1-2). The steps which need to followed to enter in the Israeli Market are as follow: The product (Medical Electrical Equipment) which will enter (first time) in Israel shall be evaluated (full tested or assessed based on a CB Certificate+TR) according 601-1 – based on this evaluation (which can be done by Israeli Designated Laboratories: SII, ITL-Israel Testing Laboratories and Hermon Laboratories) SII shall issue the Letter for Customs and build a “prototype” file. Based on this file, for each future shipping, a sample named “shin” will be assessed by SII for conformance with the “prototype”. Our lab can assist all which are interested on Israeli market. I hope that the above clarify the issue. Steli Loznen, M.Sc., SM-IEEE Q.A Certification Manager I.T.L (Product Testing) Ltd. Convener IEC/TC62/SC62A/WG17 1 Bat Sheva St., POB 87 Lod 71100, Israel V: +972-(0)8-9153100 Ext.203 F: +972-(0)8-9153101 M: +972-(0)54-7245794 st...@itl.co.il http://www.itl.co.il From: peter merguerian [mailto:pmerguerian2...@yahoo.com] Sent: Tuesday, December 08, 2009 6:37 AM To: EMC-PSTC@LISTSERV.IEEE.ORG Subject: Re: [PSES] Medical devices and Israel Derek, Do you have the harmonized tariff code? The Standards Institution of Israel (SII) would the the only laboratory in Israel able to test your product for import. I can help you gain the SII approval if you wish if you serne me the product brochure together with any test reports you might have for safety and emc Thanks, Peter --- On Mon, 12/7/09, Derek Walton lfresea...@aol.com wrote: From: Derek Walton lfresea...@aol.com Subject: Medical devices and Israel To: IEEE EMC SAFETY PSTC emc-p...@ieee.org Date: Monday, December 7, 2009, 2:34 PM Hi folks, A while back I tested a light box that goes on the wall of a Dr's office. It's used for showing X-Ray films. I was asked to perform testing to EN60601-2, which we did. I have been asked to produce an EMC Certificate, but I must have missed where the requirements for this came from. Can anyone enlighten me please. Additionally, this morning, I'm informed that to import into Israel, the test has to be performed by a Certified Electronics Laboratory: what is one of those? Many thanks, Derfel Walton L F Research - This message is from the IEEE Product Safety Engineering Society emc-pstc discussion list. To post a message to the list, send your e-mail to emc-p...@ieee.org http://us.mc1120.ma l.yahoo.com/mc/compose?to=emc-p...@ieee.org All emc-pstc postings are archived and searchable on the web at: http://www.ieeecommunities.org/emc-pstc Graphics (in well-used formats), large files, etc. can be posted to that URL. Website: http://www.ieee-pses.org/ Instructions: http://listserv.ieee.org/request/user-guide.html List rules: http://www.ieee-pses.org/listrules.html For help, send mail to the list administrators: Scott Douglas emcp...@socal.rr.com h tp://us.mc1120.mail.yahoo.com/mc/compose?to=emcp...@socal.rr.com Mike Cantwell mcantw...@ieee.org htt ://us.mc1120.mail.yahoo.com/mc/compose?to=mcantw...@ieee.org For policy questions, send mail to: Jim Bacher: j.bac...@ieee.org http: /us.mc1120.mail.yahoo.com/mc/compose
Re: Medical devices and Israel
Derek, Do you have the harmonized tariff code? The Standards Institution of Israel (SII) would the the only laboratory in Israel able to test your product for import. I can help you gain the SII approval if you wish if you serne me the product brochure together with any test reports you might have for safety and emc Thanks, Peter --- On Mon, 12/7/09, Derek Walton lfresea...@aol.com wrote: From: Derek Walton lfresea...@aol.com Subject: Medical devices and Israel To: IEEE EMC SAFETY PSTC emc-p...@ieee.org Date: Monday, December 7, 2009, 2:34 PM Hi folks, A while back I tested a light box that goes on the wall of a Dr's office. It's used for showing X-Ray films. I was asked to perform testing to EN60601-2, which we did. I have been asked to produce an EMC Certificate, but I must have missed where the requirements for this came from. Can anyone enlighten me please. Additionally, this morning, I'm informed that to import into Israel, the test has to be performed by a Certified Electronics Laboratory: what is one of those? Many thanks, Derfel Walton L F Research - This message is from the IEEE Product Safety Engineering Society emc-pstc discussion list. To post a message to the list, send your e-mail to emc-p...@ieee.org http://us.mc1120.ma l.yahoo.com/mc/compose?to=emc-p...@ieee.org All emc-pstc postings are archived and searchable on the web at: http://www.ieeecommunities.org/emc-pstc Graphics (in well-used formats), large files, etc. can be posted to that URL. Website: http://www.ieee-pses.org/ Instructions: http://listserv.ieee.org/request/user-guide.html List rules: http://www.ieee-pses.org/listrules.html For help, send mail to the list administrators: Scott Douglas emcp...@socal.rr.com h tp://us.mc1120.mail.yahoo.com/mc/compose?to=emcp...@socal.rr.com Mike Cantwell mcantw...@ieee.org htt ://us.mc1120.mail.yahoo.com/mc/compose?to=mcantw...@ieee.org For policy questions, send mail to: Jim Bacher: j.bac...@ieee.org http: /us.mc1120.mail.yahoo.com/mc/compose?to=j.bac...@ieee.org David Heald: dhe...@gmail.com http:/ us.mc1120.mail.yahoo.com/mc/compose?to=dhe...@gmail.com - This message is from the IEEE Product Safety Engineering Society emc-pstc discussion list. To post a message to the list, send your e-mail to emc-p...@ieee.org All emc-pstc postings are archived and searchable on the web at http://www.ieeecommunities.org/emc-pstc Graphics (in well-used formats), large files, etc. can be posted to that URL. Website: http://www.ieee-pses.org/ Instructions: http://listserv.ieee.org/request/user-guide.html List rules: http://www.ieee-pses.org/listrules.html For help, send mail to the list administrators: Scott Douglas emcp...@socal.rr.com Mike Cantwell mcantw...@ieee.org For policy questions, send mail to: Jim Bacher j.bac...@ieee.org David Heald dhe...@gmail.com
Medical devices and Israel
Hi folks, A while back I tested a light box that goes on the wall of a Dr's office. It's used for showing X-Ray films. I was asked to perform testing to EN60601-2, which we did. I have been asked to produce an EMC Certificate, but I must have missed where the requirements for this came from. Can anyone enlighten me please. Additionally, this morning, I'm informed that to import into Israel, the test has to be performed by a Certified Electronics Laboratory: what is one of those? Many thanks, Derfel Walton L F Research - This message is from the IEEE Product Safety Engineering Society emc-pstc discussion list. To post a message to the list, send your e-mail to emc-p...@ieee.org All emc-pstc postings are archived and searchable on the web at: http://www.ieeecommunities.org/emc-pstc Graphics (in well-used formats), large files, etc. can be posted to that URL. Website: http://www.ieee-pses.org/ Instructions: http://listserv.ieee.org/request/user-guide.html List rules: http://www.ieee-pses.org/listrules.html For help, send mail to the list administrators: Scott Douglas emcp...@socal.rr.com Mike Cantwell mcantw...@ieee.org For policy questions, send mail to: Jim Bacher: j.bac...@ieee.org David Heald: dhe...@gmail.com
New EMC Standard for Medical Devices in Japan
Group, Thanks for the replies but does anyone have a pulse on when the new EMC Standard for Medical Devices will be published for Japan. JIS T 0601-X-X ? Has anyone seen a draft version of this? Regards, Jeff Collins Compliance Reliability Consultant Alcon Surgical Labs - This message is from the IEEE Product Safety Engineering Society emc-pstc discussion list. Website: http://www.ieee-pses.org/ To post a message to the list, send your e-mail to emc-p...@ieee.org Instructions: http://listserv.ieee.org/request/user-guide.html List rules: http://www.ieee-pses.org/listrules.html For help, send mail to the list administrators: Scott Douglas emcp...@ptcnh.net Mike Cantwell mcantw...@ieee.org For policy questions, send mail to: Jim Bacher: j.bac...@ieee.org David Heald: emc-p...@daveheald.com All emc-pstc postings are archived and searchable on the web at: http://www.ieeecommunities.org/emc-pstc __ This e-mail has been scanned by MCI Managed Email Content Service, using Skeptic(tm) technology powered by MessageLabs. For more information on MCI's Managed Email Content Service, visit http://www.mci.com. __
Re: JIS EMC standards: EMC requirements for Medical Devices in Japan
On Thu, 11 Oct 2007 06:28:30 -0700 (PDT), jeff collins jeffcollin...@yahoo.com wrote: Does anyone have any current information regarding EMC requirements for Medical Devices in Japan? JIS T 0601-1-2:2002 (corresponds to IEC 60601-1-2 ed.1) is still current, although IEC 60601-1-2 ed.2 will also be accepted. Regards, Tom Tomonori Sato vef00...@nifty.ne.jp URL: http://homepage3.nifty.com/tsato/ - This message is from the IEEE Product Safety Engineering Society emc-pstc discussion list.Website: http://www.ieee-pses.org/ To post a message to the list, send your e-mail to emc-p...@ieee.org Instructions: http://listserv.ieee.org/request/user-guide.html List rules: http://www.ieee-pses.org/listrules.html For help, send mail to the list administrators: Scott Douglas emcp...@ptcnh.net Mike Cantwell mcantw...@ieee.org For policy questions, send mail to: Jim Bacher: j.bac...@ieee.org David Heald:emc-p...@daveheald.com All emc-pstc postings are archived and searchable on the web at: http://www.ieeecommunities.org/emc-pstc __ This e-mail has been scanned by MCI Managed Email Content Service, using Skeptic(tm) technology powered by MessageLabs. For more information on MCI's Managed Email Content Service, visit http://www.mci.com. __
JIS EMC standards: EMC requirements for Medical Devices in Japan
Dear List Members, Does anyone have any current information regarding EMC requirements for Medical Devices in Japan? Regards, Jeff Collins Compliance Reliability Consultant Alcon Surgical Labs - This message is from the IEEE Product Safety Engineering Society emc-pstc discussion list. Website: http://www.ieee-pses.org/ To post a message to the list, send your e-mail to emc-p...@ieee.org Instructions: http://listserv.ieee.org/request/user-guide.html List rules: http://www.ieee-pses.org/listrules.html For help, send mail to the list administrators: Scott Douglas emcp...@ptcnh.net Mike Cantwell mcantw...@ieee.org For policy questions, send mail to: Jim Bacher: j.bac...@ieee.org David Heald: emc-p...@daveheald.com All emc-pstc postings are archived and searchable on the web at: http://www.ieeecommunities.org/emc-pstc __ This e-mail has been scanned by MCI Managed Email Content Service, using Skeptic(tm) technology powered by MessageLabs. For more information on MCI's Managed Email Content Service, visit http://www.mci.com. __
RE: CSA requirements for medical devices
Hello Paolo, We have three labs with medical groups to assist you. Gianluca at our HQ may be a good start: Gianluca Arcari gianluca.arc...@csa-international.org (416) 747-4000 Or you may try Michel Brossoit michel.bross...@csa-international.org (514) 428-2411 And our International Technical Contact for your area is Lu Soska who can be very helpful luanne.so...@csa-international.org (216) 524-4990 Best regards, Jason L. Chesley CSA International 2210 Justin Trail Alpharetta, GA 30004 P: (678) 992-0134 F: (770) 500-3948 jason.ches...@csa-international.org www.csa-international.org From: emc-p...@ieee.org [mailto:emc-p...@ieee.org] On Behalf Of Paolo Peruzzi Sent: Wednesday, April 26, 2006 5:25 AM To: emc-p...@ieee.org Subject: CSA requirements for medical devices Dear all, I've been asked to make some modification on a CE marked medical device in order to make it compliant with CSA requirements. Some modifications concern non-combustible enclosure and a dedicated ground point. So I'm a little confused, cause I thought I had to simply follow the last edition of CSA deviations to IEC 60601-1, but I haven't found anything like that in it. So where can I find all the needed technical requirements when I'm going to sell a medical device in the canadian market? Thanks in advance, Paolo Peruzzi Regulatory Medical RD El.En. S.p.A. Via Baldanzese, 17 50041 Calenzano (FI) Italy Tel. +39 055 8826807 FAX +39 055 8832884 http://www.elengroup.com/ - This message is from the IEEE Product Safety Engineering Society emc-pstc discussion list.Website: http://www.ieee-pses.org/ To post a message to the list, send your e-mail to emc-p...@ieee.org Instructions: http://listserv.ieee.org/request/user-guide.html List rules: http://www.ieee-pses.org/listrules.html For help, send mail to the list administrators: Scott Douglas emcp...@ptcnh.net Mike Cantwell mcantw...@ieee.org For policy questions, send mail to: Jim Bacher: j.bac...@ieee.org David Heald:emc-p...@daveheald.com All emc-pstc postings are archived and searchable on the web at: http://www.ieeecommunities.org/emc-pstc This message is intended only for the use of the person or organization to which it is addressed, and may contain information that is privileged, confidential and exempt from disclosure under applicable law. If the reader of this message is not the intended recipient, or responsible for delivering the message to the intended recipient, you are hereby notified that any dissemination, distribution or copying of this communication is strictly prohibited. If you have received this communication in error, please notify the sender immediately by email and delete the original message immediately. The sender, its subsidiaries and affiliates, do not accept liability for any errors, omissions, corruption or virus in the contents of this message or any attachments that arise as a result of e-mail transmission. Thank you. - This message is from the IEEE Product Safety Engineering Society emc-pstc discussion list.Website: http://www.ieee-pses.org/ To post a message to the list, send your e-mail to emc-p...@ieee.org Instructions: http://listserv.ieee.org/request/user-guide.html List rules: http://www.ieee-pses.org/listrules.html For help, send mail to the list administrators: Scott Douglas emcp...@ptcnh.net Mike Cantwell mcantw...@ieee.org For policy questions, send mail to: Jim Bacher: j.bac...@ieee.org David Heald:emc-p...@daveheald.com All emc-pstc postings are archived and searchable on the web at: http://www.ieeecommunities.org/emc-pstc
Re: CSA requirements for medical devices
If no success with CSA talk with Global Advantage in Toronto. email ga...@globaladvantage.ca Regards: Kevin Keegan Senior Associate KES Associates 1 Stonecroft Terrace Kanata, Ontario Canada K2K 2V1 Tel: 613-592-0820 Email: kkee...@kesandassociates.com Web: http://www.kesandassociates.com Regulatory Approval Forum Chairperson www.raft-global.org On 26 Apr 2006 at 10:36, John Woodgate wrote: In message 2006042207.m77...@elen.it, dated Wed, 26 Apr 2006, Paolo Peruzzi standa...@elen.it writes I've been asked to make some modification on a CE marked medical device in order to make it compliant with CSA requirements. Some modifications concern non-combustible enclosure and a dedicated ground point. Who is asking for these changes? So I'm a little confused, cause I thought I had to simply follow the last edition of CSA deviations to IEC 60601-1, but I haven't found anything like that in it. So where can I find all the needed technical requirements when I'm going to sell a medical device in the canadian market? Once again, first try asking CSA. If it works, you have the answer with the greatest reliability. Particularly if CSA says you don't need to make those changes. -- OOO - Own Opinions Only. Try www.jmwa.demon.co.uk and www.isce.org.uk 2006 is YMMVI- Your mileage may vary immensely. John Woodgate, J M Woodgate and Associates, Rayleigh, Essex UK - This message is from the IEEE Product Safety Engineering Society emc-pstc discussion list.Website: http://www.ieee-pses.org/ To post a message to the list, send your e-mail to emc-p...@ieee.org Instructions: http://listserv.ieee.org/request/user-guide.html List rules: http://www.ieee-pses.org/listrules.html For help, send mail to the list administrators: Scott Douglas emcp...@ptcnh.net Mike Cantwell mcantw...@ieee.org For policy questions, send mail to: Jim Bacher: j.bac...@ieee.org David Heald:emc-p...@daveheald.com All emc-pstc postings are archived and searchable on the web at: http://www.ieeecommunities.org/emc-pstc -- No virus found in this incoming message. Checked by AVG Free Edition. Version: 7.1.385 / Virus Database: 268.4.6/324 - Release Date: 4/25/2006 IMPORTANT NOTICE: This email is intended only for the individual or entity to which it is specifically addressed and should not be read by, or delivered to, any other person. Such material may contain privileged or confidential information, the disclosure or other use of which by other than the intended recipient may result in the breach of certain laws or the infringement of rights of third parties. If you have received this email in error, please delete it immediately and notify the sender. I thank you in advance for your co-operation and assistance. - This message is from the IEEE Product Safety Engineering Society emc-pstc discussion list. Website: http://www.ieee-pses.org/ To post a message to the list, send your e-mail to emc-p...@ieee.org Instructions: http://listserv.ieee.org/request/user-guide.html List rules: http://www.ieee-pses.org/listrules.html For help, send mail to the list administrators: Scott Douglas emcp...@ptcnh.net Mike Cantwell mcantw...@ieee.org For policy questions, send mail to: Jim Bacher: j.bac...@ieee.org David Heald: emc-p...@daveheald.com All emc-pstc postings are archived and searchable on the web at: http://www.ieeecommunities.org/emc-pstc
Re: CSA requirements for medical devices
In message 2006042207.m77...@elen.it, dated Wed, 26 Apr 2006, Paolo Peruzzi standa...@elen.it writes I've been asked to make some modification on a CE marked medical device in order to make it compliant with CSA requirements. Some modifications concern non-combustible enclosure and a dedicated ground point. Who is asking for these changes? So I'm a little confused, cause I thought I had to simply follow the last edition of CSA deviations to IEC 60601-1, but I haven't found anything like that in it. So where can I find all the needed technical requirements when I'm going to sell a medical device in the canadian market? Once again, first try asking CSA. If it works, you have the answer with the greatest reliability. Particularly if CSA says you don't need to make those changes. -- OOO - Own Opinions Only. Try www.jmwa.demon.co.uk and www.isce.org.uk 2006 is YMMVI- Your mileage may vary immensely. John Woodgate, J M Woodgate and Associates, Rayleigh, Essex UK - This message is from the IEEE Product Safety Engineering Society emc-pstc discussion list.Website: http://www.ieee-pses.org/ To post a message to the list, send your e-mail to emc-p...@ieee.org Instructions: http://listserv.ieee.org/request/user-guide.html List rules: http://www.ieee-pses.org/listrules.html For help, send mail to the list administrators: Scott Douglas emcp...@ptcnh.net Mike Cantwell mcantw...@ieee.org For policy questions, send mail to: Jim Bacher: j.bac...@ieee.org David Heald:emc-p...@daveheald.com All emc-pstc postings are archived and searchable on the web at: http://www.ieeecommunities.org/emc-pstc
CSA requirements for medical devices
Dear all, I've been asked to make some modification on a CE marked medical device in order to make it compliant with CSA requirements. Some modifications concern non-combustible enclosure and a dedicated ground point. So I'm a little confused, cause I thought I had to simply follow the last edition of CSA deviations to IEC 60601-1, but I haven't found anything like that in it. So where can I find all the needed technical requirements when I'm going to sell a medical device in the canadian market? Thanks in advance, Paolo Peruzzi Regulatory Medical RD El.En. S.p.A. Via Baldanzese, 17 50041 Calenzano (FI) Italy Tel. +39 055 8826807 FAX +39 055 8832884 http://www.elengroup.com/ - This message is from the IEEE Product Safety Engineering Society emc-pstc discussion list.Website: http://www.ieee-pses.org/ To post a message to the list, send your e-mail to emc-p...@ieee.org Instructions: http://listserv.ieee.org/request/user-guide.html List rules: http://www.ieee-pses.org/listrules.html For help, send mail to the list administrators: Scott Douglas emcp...@ptcnh.net Mike Cantwell mcantw...@ieee.org For policy questions, send mail to: Jim Bacher: j.bac...@ieee.org David Heald:emc-p...@daveheald.com All emc-pstc postings are archived and searchable on the web at: http://www.ieeecommunities.org/emc-pstc
wireless in medical devices
Hi all, could anybody tell me if there are particular restrictions in the use of a wireless kit in medical devices not for data transfer but for remote control and activation? I'd like to know in particular if, for european market, it would be sufficient to certify the system according to MD directive and RTTE directive. Thanks, Paolo Peruzzi Regulatory medical RD El.En. S.p.A. Via Baldanzese, 17 50041 Calenzano (FI) Italy Tel. +39 055 8826807 FAX +39 055 8832884 http://www.elengroup.com/ This message is from the IEEE Product Safety Engineering Society emc-pstc discussion list.Website: http://www.ieee-pses.org/ To post a message to the list, send your e-mail to emc-p...@ieee.org Instructions: http://listserv.ieee.org/listserv/request/user-guide.html List rules: http://www.ieee-pses.org/listrules.html For help, send mail to the list administrators: Scott Douglas emcp...@ptcnh.net Mike Cantwellmcantw...@ieee.org For policy questions, send mail to: Richard Nute: ri...@ieee.org Jim Bacher: j.bac...@ieee.org All emc-pstc postings are archived and searchable on the web at: http://www.ieeecommunities.org/emc-pstc
[Fwd: canadian requirements for medical devices]
Paulo, 1. Your approval to ISO 13485 must be by a Notified Body or other registrar which is accredited by the Canadian Medical Devices Conformity Assessment System (CMDCAS) 2. The Canadian Medical Device Regulations are available at: http://laws.justice.gc.ca/en/f-27/sor-98-282/text.html 3. Guidance on how to apply for a medical device licence is available at: http://www.hc-sc.gc.ca/hpfb-dgpsa/tpd-dpt/lcn_nw4_e.html 4. The frequently asked questions at: http://www.hc-sc.gc.ca/hpfb-dgpsa/tpd-dpt/quesans9_e.html may also be useful. I hope this helps. Regards, Jon Griver http://www.601help.com The Medical Device Designers' Guide to IEC 60601-1 Subject: canadian requirements for medical devices From:P. Peruzzi standa...@elen.it List-Post: emc-pstc@listserv.ieee.org Date:Tue, September 14, 2004 3:42 am To: emc-p...@ieee.org Hi group, I'm not familiar with canadian regulation and I've collected not very clear answers on the matter; I'd like to know what exactly I have to do in order to sell a medical device in Canada. What I've understood is that my company should have a quality system in compliance with Canadian version of 13485 (and it already has), and we shall apply the medical devices canadian directive If it's correct, where can I find such a directive? And what about technical standards? I know only 601-1 deviations. Is there a sort of harmonized standards list like fo EU MDD? thanks in advance, Paolo Peruzzi El.En. S.p.A. Via Baldanzese, 17 50041 Calenzano (FI) Italy Tel. +39 055 8826807 FAX +39 055 8832884 http://www.elengroup.com/ This message is from the IEEE Product Safety Engineering Society emc-pstc discussion list.Website: http://www.ieee-pses.org/ To post a message to the list, send your e-mail to emc-p...@ieee.org Instructions: http://listserv.ieee.org/listserv/request/user-guide.html List rules: http://www.ieee-pses.org/listrules.html For help, send mail to the list administrators: Ron Pickard: emc-p...@hypercom.com Dave Heald: emc_p...@symbol.com For policy questions, send mail to: Richard Nute: ri...@ieee.org Jim Bacher: j.bac...@ieee.org All emc-pstc postings are archived and searchable on the web at: http://www.ieeecommunities.org/emc-pstc This message is from the IEEE Product Safety Engineering Society emc-pstc discussion list.Website: http://www.ieee-pses.org/ To post a message to the list, send your e-mail to emc-p...@ieee.org Instructions: http://listserv.ieee.org/listserv/request/user-guide.html List rules: http://www.ieee-pses.org/listrules.html For help, send mail to the list administrators: Ron Pickard: emc-p...@hypercom.com Dave Heald: emc_p...@symbol.com For policy questions, send mail to: Richard Nute: ri...@ieee.org Jim Bacher: j.bac...@ieee.org All emc-pstc postings are archived and searchable on the web at: http://www.ieeecommunities.org/emc-pstc
canadian requirements for medical devices
Hi group, I'm not familiar with canadian regulation and I've collected not very clear answers on the matter; I'd like to know what exactly I have to do in order to sell a medical device in Canada. What I've understood is that my company should have a quality system in compliance with Canadian version of 13485 (and it already has), and we shall apply the medical devices canadian directive If it's correct, where can I find such a directive? And what about technical standards? I know only 601-1 deviations. Is there a sort of harmonized standards list like fo EU MDD? thanks in advance, Paolo Peruzzi El.En. S.p.A. Via Baldanzese, 17 50041 Calenzano (FI) Italy Tel. +39 055 8826807 FAX +39 055 8832884 http://www.elengroup.com/ This message is from the IEEE Product Safety Engineering Society emc-pstc discussion list.Website: http://www.ieee-pses.org/ To post a message to the list, send your e-mail to emc-p...@ieee.org Instructions: http://listserv.ieee.org/listserv/request/user-guide.html List rules: http://www.ieee-pses.org/listrules.html For help, send mail to the list administrators: Ron Pickard: emc-p...@hypercom.com Dave Heald: emc_p...@symbol.com For policy questions, send mail to: Richard Nute: ri...@ieee.org Jim Bacher: j.bac...@ieee.org All emc-pstc postings are archived and searchable on the web at: http://www.ieeecommunities.org/emc-pstc
Re: Medical devices and voltage dip testing
I read in !emc-pstc that Pat Lawler pat.law...@verizon.net wrote (in 20040429172618.kxjh10678.out004.verizon@outgoing.verizon.net) about 'Medical devices and voltage dip testing' on Thu, 29 Apr 2004: EN60601-1-2:2001 (EMC requirements for general medical devices) requires various voltage dip tests at the extremes of the rated input voltage range. If a product is rated for 100-240Vac operation, this means test levels like 40Vac for 100ms (100Vac/50Hz nominal input, 60% dip for 5 cycles). The performance criteria makes this especially difficult - The system shall provide essential performance. This looks like a major design change for many of our power supplies. What has been the experience of list members designing power supplies or systems to meet this requirement? Have you found it to be a minor design challenge, or are you still figuring out how to meet the test requirement? It looks very much as though the 'world-wide' power supplies weren't properly taken into account by TC 62 or SC62A. I suggest you contact people concerned with those committees and ask them. There is a list of officers available form the public part of the IEC web site: http://www/iec.ch -- Regards, John Woodgate, OOO - Own Opinions Only. The good news is that nothing is compulsory. The bad news is that everything is prohibited. http://www.jmwa.demon.co.uk Also see http://www.isce.org.uk This message is from the IEEE EMC Society Product Safety Technical Committee emc-pstc discussion list. Visit our web site at: http://www.ieee-pses.org/ To cancel your subscription, send mail to: majord...@ieee.org with the single line: unsubscribe emc-pstc For help, send mail to the list administrators: Ron Pickard: emc-p...@hypercom.com Dave Heald: emc_p...@symbol.com For policy questions, send mail to: Richard Nute: ri...@ieee.org Jim Bacher: j.bac...@ieee.org All emc-pstc postings are archived and searchable on the web at: http://www.ieeecommunities.org/emc-pstc
RE: Medical devices and voltage dip testing
Brent DeWitt wrote: that in the CDV (not law yet) of 1st amendment to 60601-1-2 we have changed essential performance from a defined term to an undefined term due to the instability of IEC 60601-1 , which is still at the CDV stage. My _personal_ opinion as a representative (and only that), is that: Thou shall not fool the customer. The 1st Amendment states quite clearly that the manufacturer will list which functions are essential functions in the user documentation. This is a hot potato, indeed. However, whenever a standard is made stricter, complaints may be expected based on the idea that equipment which has worked well enough for some considerable time is now judged unworthy of approval, and the expense incurred complying with new requirements. No one minds a new standard if all the old equipment passes! The question which provoked this thread related to power quality. UPS's and batteries ARE one answer to that question. But those in the business of writing standards might bear in mind the old adage: Be careful what you ask for; you might get it. I'll shut up now. Disclaimer: My opinions, NOT those of my agency or where I work. Cortland Richmond This message is from the IEEE EMC Society Product Safety Technical Committee emc-pstc discussion list. Visit our web site at: http://www.ieee-pses.org/ To cancel your subscription, send mail to: majord...@ieee.org with the single line: unsubscribe emc-pstc For help, send mail to the list administrators: Ron Pickard: emc-p...@hypercom.com Dave Heald: emc_p...@symbol.com For policy questions, send mail to: Richard Nute: ri...@ieee.org Jim Bacher: j.bac...@ieee.org All emc-pstc postings are archived and searchable on the web at: http://www.ieeecommunities.org/emc-pstc
RE: Medical devices and voltage dip testing
First, I need to disclose my bias. I am a member of the committee that wrote 60601-1-2 2001. Cortland is correct in bringing up the issue of essential performance. Note that in the CDV (not law yet) of 1st amendment to 60601-1-2 we have changed essential performance from a defined term to an undefined term due to the instability of IEC 60601-1 , which is still at the CDV stage. My _personal_ opinion as a representative (and only that), is that: Thou shall not fool the customer. The 1st Amendment states quite clearly that the manufacturer will list which functions are essential functions in the user documentation. Note that anything I've said is ONLY MY opinion! With respect, Brent DeWitt -Original Message- From: owner-emc-p...@majordomo.ieee.org [mailto:owner-emc-p...@majordomo.ieee.org]On Behalf Of Cortland Richmond Sent: Thursday, April 29, 2004 8:03 PM To: ieee pstc list Subject: Re: Medical devices and voltage dip testing Pat Lawler posted: If a product is rated for 100-240Vac operation, this means test levels like 40Vac for 100ms (100Vac/50Hz nominal input, 60% dip for 5 cycles). The performance criteria makes this especially difficult - The system shall provide essential performance. Do not forget that, in large part, the *manufacturer* must say what essential performance is. Cortland Disclaimer: My opinions; NOT those of my agency or where I work. --- This message is from the IEEE EMC Society Product Safety Technical Committee emc-pstc discussion list. Visit our web site at: http://www.ieee-pses.org/ To cancel your subscription, send mail to: majord...@ieee.org with the single line: unsubscribe emc-pstc For help, send mail to the list administrators: Ron Pickard: emc-p...@hypercom.com Dave Heald: emc_p...@symbol.com For policy questions, send mail to: Richard Nute: ri...@ieee.org Jim Bacher: j.bac...@ieee.org All emc-pstc postings are archived and searchable on the web at: http://www.ieeecommunities.org/emc-pstc This message is from the IEEE EMC Society Product Safety Technical Committee emc-pstc discussion list. Visit our web site at: http://www.ieee-pses.org/ To cancel your subscription, send mail to: majord...@ieee.org with the single line: unsubscribe emc-pstc For help, send mail to the list administrators: Ron Pickard: emc-p...@hypercom.com Dave Heald: emc_p...@symbol.com For policy questions, send mail to: Richard Nute: ri...@ieee.org Jim Bacher: j.bac...@ieee.org All emc-pstc postings are archived and searchable on the web at: http://www.ieeecommunities.org/emc-pstc
Re: Medical devices and voltage dip testing
Pat Lawler posted: If a product is rated for 100-240Vac operation, this means test levels like 40Vac for 100ms (100Vac/50Hz nominal input, 60% dip for 5 cycles). The performance criteria makes this especially difficult - The system shall provide essential performance. Do not forget that, in large part, the *manufacturer* must say what essential performance is. Cortland Disclaimer: My opinions; NOT those of my agency or where I work. This message is from the IEEE EMC Society Product Safety Technical Committee emc-pstc discussion list. Visit our web site at: http://www.ieee-pses.org/ To cancel your subscription, send mail to: majord...@ieee.org with the single line: unsubscribe emc-pstc For help, send mail to the list administrators: Ron Pickard: emc-p...@hypercom.com Dave Heald: emc_p...@symbol.com For policy questions, send mail to: Richard Nute: ri...@ieee.org Jim Bacher: j.bac...@ieee.org All emc-pstc postings are archived and searchable on the web at: http://www.ieeecommunities.org/emc-pstc
Re: Medical devices and voltage dip testing
Hi Pat: If a product is rated for 100-240Vac operation, this means test levels like 40Vac for 100ms (100Vac/50Hz nominal input, 60% dip for 5 cycles). The performance criteria makes this especially difficult - The system shall provide essential performance. This looks like a major design change for many of our power supplies. My health insurance premiums are going to go up! The way I read the requirement, the power supply output will need to be maintained through this dip. Most power supplies cannot sustain the output with input that low for that long a period. The requirement does indeed imply a MAJOR design change for typical power supplies. The 100 ms time period (5 cycles) is really tough. For a switcher, it means some really big capacitors in both the rectifier and the output. It may also mean the switcher will need to work at even lower voltages. The power supply will be bigger and more expensive. Good luck, Rich This message is from the IEEE EMC Society Product Safety Technical Committee emc-pstc discussion list. Visit our web site at: http://www.ieee-pses.org/ To cancel your subscription, send mail to: majord...@ieee.org with the single line: unsubscribe emc-pstc For help, send mail to the list administrators: Ron Pickard: emc-p...@hypercom.com Dave Heald: emc_p...@symbol.com For policy questions, send mail to: Richard Nute: ri...@ieee.org Jim Bacher: j.bac...@ieee.org All emc-pstc postings are archived and searchable on the web at: http://www.ieeecommunities.org/emc-pstc
Medical devices and voltage dip testing
EN60601-1-2:2001 (EMC requirements for general medical devices) requires various voltage dip tests at the extremes of the rated input voltage range. If a product is rated for 100-240Vac operation, this means test levels like 40Vac for 100ms (100Vac/50Hz nominal input, 60% dip for 5 cycles). The performance criteria makes this especially difficult - The system shall provide essential performance. This looks like a major design change for many of our power supplies. What has been the experience of list members designing power supplies or systems to meet this requirement? Have you found it to be a minor design challenge, or are you still figuring out how to meet the test requirement? --- Pat Lawler pat.law...@verizon.net This message is from the IEEE EMC Society Product Safety Technical Committee emc-pstc discussion list. Visit our web site at: http://www.ieee-pses.org/ To cancel your subscription, send mail to: majord...@ieee.org with the single line: unsubscribe emc-pstc For help, send mail to the list administrators: Ron Pickard: emc-p...@hypercom.com Dave Heald: emc_p...@symbol.com For policy questions, send mail to: Richard Nute: ri...@ieee.org Jim Bacher: j.bac...@ieee.org All emc-pstc postings are archived and searchable on the web at: http://www.ieeecommunities.org/emc-pstc
RE: PVC in Medical Devices
Peter, I don't know much about Medical, but I have heard from some of our major customers that PVC use needs to be reduced because of chemicals covered under WEEE. Josh From: peter.valent...@us.datex-ohmeda.com [mailto:peter.valent...@us.datex-ohmeda.com] Sent: Wednesday, October 15, 2003 8:51 AM To: emc-p...@majordomo.ieee.org Subject: PVC in Medical Devices Are there formal requirements requiring that PVC (Poly Vinyl Chloride) NOT be used in medical device materials? If so, what are they called, and can there be exceptions to these rules. From my understanding, there are only informal ones that medical device manufacturers should adhere to for example when selling in Europe. The insulating shrink material used on Nickel Metal Hydride batteries consists of PVC. Almost all battery suppliers use the same shrink material sold in many parts of the world. It would appear that this may be a concern. Thanks Peter Valentyik This message is from the IEEE EMC Society Product Safety Technical Committee emc-pstc discussion list. Visit our web site at: http://www.ewh.ieee.org/soc/emcs/pstc/ To cancel your subscription, send mail to: majord...@ieee.org with the single line: unsubscribe emc-pstc For help, send mail to the list administrators: Ron Pickard: emc-p...@hypercom.com Dave Heald: emc_p...@symbol.com For policy questions, send mail to: Richard Nute: ri...@ieee.org Jim Bacher: j.bac...@ieee.org Archive is being moved, we will announce when it is back on-line. All emc-pstc postings are archived and searchable on the web at: http://www.ieeecommunities.org/emc-pstc This message is from the IEEE EMC Society Product Safety Technical Committee emc-pstc discussion list. Visit our web site at: http://www.ewh.ieee.org/soc/emcs/pstc/ To cancel your subscription, send mail to: majord...@ieee.org with the single line: unsubscribe emc-pstc For help, send mail to the list administrators: Ron Pickard: emc-p...@hypercom.com Dave Heald: emc_p...@symbol.com For policy questions, send mail to: Richard Nute: ri...@ieee.org Jim Bacher: j.bac...@ieee.org Archive is being moved, we will announce when it is back on-line. All emc-pstc postings are archived and searchable on the web at: http://www.ieeecommunities.org/emc-pstc
PVC in Medical Devices
Are there formal requirements requiring that PVC (Poly Vinyl Chloride) NOT be used in medical device materials? If so, what are they called, and can there be exceptions to these rules. From my understanding, there are only informal ones that medical device manufacturers should adhere to for example when selling in Europe. The insulating shrink material used on Nickel Metal Hydride batteries consists of PVC. Almost all battery suppliers use the same shrink material sold in many parts of the world. It would appear that this may be a concern. Thanks Peter Valentyik This message is from the IEEE EMC Society Product Safety Technical Committee emc-pstc discussion list. Visit our web site at: http://www.ewh.ieee.org/soc/emcs/pstc/ To cancel your subscription, send mail to: majord...@ieee.org with the single line: unsubscribe emc-pstc For help, send mail to the list administrators: Ron Pickard: emc-p...@hypercom.com Dave Heald: emc_p...@symbol.com For policy questions, send mail to: Richard Nute: ri...@ieee.org Jim Bacher: j.bac...@ieee.org Archive is being moved, we will announce when it is back on-line. All emc-pstc postings are archived and searchable on the web at: http://www.ieeecommunities.org/emc-pstc
Medical devices
Hi all Can some one send me a copy of CCA 227-7 Rotine tests for medical equipment, or point me to a downloadeable fil. Best regards, Kim Boll Jensen Bolls Rådgivning Hyacintvej 6 DK-3660 Stenløse Tlf.: 48 18 35 66 Fax: 48 18 35 30 Mobil: 22 99 69 91 E-mail: k...@bolls.dk web: www.bolls.dk This message is from the IEEE EMC Society Product Safety Technical Committee emc-pstc discussion list. Visit our web site at: http://www.ewh.ieee.org/soc/emcs/pstc/ To cancel your subscription, send mail to: majord...@ieee.org with the single line: unsubscribe emc-pstc For help, send mail to the list administrators: Ron Pickard: emc-p...@hypercom.com Dave Heald: emc_p...@symbol.com For policy questions, send mail to: Richard Nute: ri...@ieee.org Jim Bacher: j.bac...@ieee.org Archive is being moved, we will announce when it is back on-line. All emc-pstc postings are archived and searchable on the web at: http://www.ieeecommunities.org/emc-pstc
RE: RTTE for Medical Devices?
Del, I'm not an RTTE expert, but from the MDD side under the 2nd edition of 60601-1-2 (2001), the function of the card and it's communications would have to be evaluated. If it falls under the ESSENTIAL FUNCTION of the medical device, it would be subject to the requirements of 60601-1-2. These requirements are somewhat more stringent than the RTTE directive. That probably doesn't help much, but it's a bit more information. Regards, Brent DeWitt From: owner-emc-p...@majordomo.ieee.org [mailto:owner-emc-p...@majordomo.ieee.org]On Behalf Of D.Han Sent: Friday, July 18, 2003 4:20 PM To: emc-p...@majordomo.ieee.org Subject: RTTE for Medical Devices? Hi All, I would appreciate your thoughts on the following scenario: Company A designs a medical product that incorporates a wireless device, lets say, a wireless PCI card. The wireless card is manufactured by Company B and has been evaluated to the RTTE directive and thus CE-marked. This wireless card is installed in the end medical product, unmodified and according to manufacturers instructions. Company As name goes onto the end product, but Company Bs name, markings, labeling etc. remain on the wireless card. Would Company A need to consider RTTE (in addition to the MDD) for their end product? If yes, what additional testing would this encompass? Thanks! Del _ Do you Yahoo!? SBC http://pa.yahoo.com/*http://rd.ya oo.com/evt=1207/*http://promo.yahoo.com/sbc/ Yahoo! DSL - Now only $29.95 per month!
RTTE for Medical Devices?
Hi All, I would appreciate your thoughts on the following scenario: Company A designs a medical product that incorporates a wireless device, let’s say, a wireless PCI card. The wireless card is manufactured by Company B and has been evaluated to the RTTE directive and thus CE-marked. This wireless card is installed in the end medical product, unmodified and according to manufacturer’s instructions. Company A’s name goes onto the end product, but Company B’s name, markings, labeling etc. remain on the wireless card. Would Company A need to consider RTTE (in addition to the MDD) for their end product? If yes, what additional testing would this encompass? Thanks! Del _ Do you Yahoo!? SBC Yahoo! DSL http://pa.yahoo.com/*ht p://rd.yahoo.com/evt=1207/*http://promo.yahoo.com/sbc/ - Now only $29.95 per month!
RE: Medical Devices Environmental Requirements
Joel Hello! I do not know the use of your medical device really does. But there are various types of standards which you can use. We recently evaluated a medical device which was worn on the wrist and which transmitted radio signals to a base station much like a cellular telephone. For environmental tests, we used the European ETSI standard ETS 300 019-2-7, Environmental conditions and environmental tests for telecommunication equipment. Part 2-7 Specification of environmental tests Portable and non-stationary uise To satisfy the FDA, you could also apply MIL-STD-810E which specifies Mechanical Shock, Vibration and Altitude Testing. All the Best from the same country. This e-mail message may contain privileged or confidential information. If you are not the intended recipient, you may not disclose, use, disseminate, distribute, copy or rely upon this message or attachment in any way. If you received this e-mail message in error, please return by forwarding the message and its attachments to the sender. PETER S. MERGUERIAN Technical Director I.T.L. (Product Testing) Ltd. 26 Hacharoshet St., POB 211 Or Yehuda 60251, Israel Tel: + 972-(0)3-5339022 Fax: + 972-(0)3-5339019 Mobile: + 972-(0)54-838175 http://www.itl.co.il http://www.i-spec.com -Original Message- From: Mandel, Joel [mailto:joel_man...@adc.com] Sent: Tuesday, November 05, 2002 6:43 PM To: 'emc-p...@majordomo.ieee.org' Subject: Medical Devices Environmental Requirements Hi All Can anybody help me out with Environmental testing requirement/Standards (Temp,Vibration) for medical devices? Thank You Joel Mandel Reliability Dept ADC Israel --- This message is from the IEEE EMC Society Product Safety Technical Committee emc-pstc discussion list. Visit our web site at: http://www.ewh.ieee.org/soc/emcs/pstc/ To cancel your subscription, send mail to: majord...@ieee.org with the single line: unsubscribe emc-pstc For help, send mail to the list administrators: Ron Pickard: emc-p...@hypercom.com Dave Heald: davehe...@attbi.com For policy questions, send mail to: Richard Nute: ri...@ieee.org Jim Bacher: j.bac...@ieee.org All emc-pstc postings are archived and searchable on the web at: http://ieeepstc.mindcruiser.com/ Click on browse and then emc-pstc mailing list --- This message is from the IEEE EMC Society Product Safety Technical Committee emc-pstc discussion list. Visit our web site at: http://www.ewh.ieee.org/soc/emcs/pstc/ To cancel your subscription, send mail to: majord...@ieee.org with the single line: unsubscribe emc-pstc For help, send mail to the list administrators: Ron Pickard: emc-p...@hypercom.com Dave Heald: davehe...@attbi.com For policy questions, send mail to: Richard Nute: ri...@ieee.org Jim Bacher: j.bac...@ieee.org All emc-pstc postings are archived and searchable on the web at: http://ieeepstc.mindcruiser.com/ Click on browse and then emc-pstc mailing list
Re: Medical Devices Environmental Requirements
Have you tried http://europa.eu.int/comm/enterprise/medical_devices/index.htm Cheers Alan E Hutley EMC Compliance Journal www.compliance-club.com - Original Message - From: Jim Conrad jc...@shore.net To: Naftali Shani nsh...@catena.com; 'Mandel, Joel' joel_man...@adc.com; emc-p...@majordomo.ieee.org Sent: Tuesday, November 05, 2002 8:03 PM Subject: RE: Medical Devices Environmental Requirements Have you checked 60601-1? Jim -Original Message- From: owner-emc-p...@majordomo.ieee.org [mailto:owner-emc-p...@majordomo.ieee.org]On Behalf Of Naftali Shani Sent: Tuesday, November 05, 2002 1:27 PM To: 'Mandel, Joel'; 'emc-p...@majordomo.ieee.org' Subject: RE: Medical Devices Environmental Requirements Are there any customers-specific requirements? In the telecom industry, NEBS/ETSI (Telcordia GR-63 ETS 300 019) come to mind. Regards, Naftali Shani, Catena Networks (www.catena.com) 307 Legget Drive, Ottawa, Ontario, Canada K2K 3C8 613.599.6430/866.2CATENA (X.8277); C 295.7042; F 599.0445 E-mail: nsh...@catena.com -Original Message- From: Mandel, Joel [mailto:joel_man...@adc.com] Sent: Tuesday, November 05, 2002 11:43 AM To: 'emc-p...@majordomo.ieee.org' Subject:Medical Devices Environmental Requirements Hi All Can anybody help me out with Environmental testing requirement/Standards (Temp,Vibration) for medical devices? Thank You Joel Mandel Reliability Dept ADC Israel --- This message is from the IEEE EMC Society Product Safety Technical Committee emc-pstc discussion list. Visit our web site at: http://www.ewh.ieee.org/soc/emcs/pstc/ To cancel your subscription, send mail to: majord...@ieee.org with the single line: unsubscribe emc-pstc For help, send mail to the list administrators: Ron Pickard: emc-p...@hypercom.com Dave Heald: davehe...@attbi.com For policy questions, send mail to: Richard Nute: ri...@ieee.org Jim Bacher: j.bac...@ieee.org All emc-pstc postings are archived and searchable on the web at: http://ieeepstc.mindcruiser.com/ Click on browse and then emc-pstc mailing list --- This message is from the IEEE EMC Society Product Safety Technical Committee emc-pstc discussion list. Visit our web site at: http://www.ewh.ieee.org/soc/emcs/pstc/ To cancel your subscription, send mail to: majord...@ieee.org with the single line: unsubscribe emc-pstc For help, send mail to the list administrators: Ron Pickard: emc-p...@hypercom.com Dave Heald: davehe...@attbi.com For policy questions, send mail to: Richard Nute: ri...@ieee.org Jim Bacher: j.bac...@ieee.org All emc-pstc postings are archived and searchable on the web at: http://ieeepstc.mindcruiser.com/ Click on browse and then emc-pstc mailing list --- This message is from the IEEE EMC Society Product Safety Technical Committee emc-pstc discussion list. Visit our web site at: http://www.ewh.ieee.org/soc/emcs/pstc/ To cancel your subscription, send mail to: majord...@ieee.org with the single line: unsubscribe emc-pstc For help, send mail to the list administrators: Ron Pickard: emc-p...@hypercom.com Dave Heald: davehe...@attbi.com For policy questions, send mail to: Richard Nute: ri...@ieee.org Jim Bacher: j.bac...@ieee.org All emc-pstc postings are archived and searchable on the web at: http://ieeepstc.mindcruiser.com/ Click on browse and then emc-pstc mailing list --- This message is from the IEEE EMC Society Product Safety Technical Committee emc-pstc discussion list. Visit our web site at: http://www.ewh.ieee.org/soc/emcs/pstc/ To cancel your subscription, send mail to: majord...@ieee.org with the single line: unsubscribe emc-pstc For help, send mail to the list administrators: Ron Pickard: emc-p...@hypercom.com Dave Heald: davehe...@attbi.com For policy questions, send mail to: Richard Nute: ri...@ieee.org Jim Bacher: j.bac...@ieee.org All emc-pstc postings are archived and searchable on the web at: http://ieeepstc.mindcruiser.com/ Click on browse and then emc-pstc mailing list
RE: Medical Devices Environmental Requirements
Hi, In the USA the FDA has a guidance document that some of the branch's (i.e. Anesthesiology and Respiratory) use. See http://www.fda.gov/cdrh/ode/638.pdf This includes both EMI, and mechanical and environmental requirements. Ned Devine Entela, Inc. 3033 Madison Ave. SE Grand Rapids, MI 49548 Phone: 616 248 9671 Fax: 616 574 9752 e-mail: ndev...@entela.com www.entela.com Entela, Inc. A Certified Woman Owned Business -Original Message- From: Mandel, Joel [mailto:joel_man...@adc.com] Sent: Tuesday, November 05, 2002 11:43 AM To: 'emc-p...@majordomo.ieee.org' Subject: Medical Devices Environmental Requirements Hi All Can anybody help me out with Environmental testing requirement/Standards (Temp,Vibration) for medical devices? Thank You Joel Mandel Reliability Dept ADC Israel --- This message is from the IEEE EMC Society Product Safety Technical Committee emc-pstc discussion list. Visit our web site at: http://www.ewh.ieee.org/soc/emcs/pstc/ To cancel your subscription, send mail to: majord...@ieee.org with the single line: unsubscribe emc-pstc For help, send mail to the list administrators: Ron Pickard: emc-p...@hypercom.com Dave Heald: davehe...@attbi.com For policy questions, send mail to: Richard Nute: ri...@ieee.org Jim Bacher: j.bac...@ieee.org All emc-pstc postings are archived and searchable on the web at: http://ieeepstc.mindcruiser.com/ Click on browse and then emc-pstc mailing list --- This message is from the IEEE EMC Society Product Safety Technical Committee emc-pstc discussion list. Visit our web site at: http://www.ewh.ieee.org/soc/emcs/pstc/ To cancel your subscription, send mail to: majord...@ieee.org with the single line: unsubscribe emc-pstc For help, send mail to the list administrators: Ron Pickard: emc-p...@hypercom.com Dave Heald: davehe...@attbi.com For policy questions, send mail to: Richard Nute: ri...@ieee.org Jim Bacher: j.bac...@ieee.org All emc-pstc postings are archived and searchable on the web at: http://ieeepstc.mindcruiser.com/ Click on browse and then emc-pstc mailing list
RE: Medical Devices Environmental Requirements
Have you checked 60601-1? Jim -Original Message- From: owner-emc-p...@majordomo.ieee.org [mailto:owner-emc-p...@majordomo.ieee.org]On Behalf Of Naftali Shani Sent: Tuesday, November 05, 2002 1:27 PM To: 'Mandel, Joel'; 'emc-p...@majordomo.ieee.org' Subject: RE: Medical Devices Environmental Requirements Are there any customers-specific requirements? In the telecom industry, NEBS/ETSI (Telcordia GR-63 ETS 300 019) come to mind. Regards, Naftali Shani, Catena Networks (www.catena.com) 307 Legget Drive, Ottawa, Ontario, Canada K2K 3C8 613.599.6430/866.2CATENA (X.8277); C 295.7042; F 599.0445 E-mail: nsh...@catena.com -Original Message- From: Mandel, Joel [mailto:joel_man...@adc.com] Sent: Tuesday, November 05, 2002 11:43 AM To: 'emc-p...@majordomo.ieee.org' Subject:Medical Devices Environmental Requirements Hi All Can anybody help me out with Environmental testing requirement/Standards (Temp,Vibration) for medical devices? Thank You Joel Mandel Reliability Dept ADC Israel --- This message is from the IEEE EMC Society Product Safety Technical Committee emc-pstc discussion list. Visit our web site at: http://www.ewh.ieee.org/soc/emcs/pstc/ To cancel your subscription, send mail to: majord...@ieee.org with the single line: unsubscribe emc-pstc For help, send mail to the list administrators: Ron Pickard: emc-p...@hypercom.com Dave Heald: davehe...@attbi.com For policy questions, send mail to: Richard Nute: ri...@ieee.org Jim Bacher: j.bac...@ieee.org All emc-pstc postings are archived and searchable on the web at: http://ieeepstc.mindcruiser.com/ Click on browse and then emc-pstc mailing list --- This message is from the IEEE EMC Society Product Safety Technical Committee emc-pstc discussion list. Visit our web site at: http://www.ewh.ieee.org/soc/emcs/pstc/ To cancel your subscription, send mail to: majord...@ieee.org with the single line: unsubscribe emc-pstc For help, send mail to the list administrators: Ron Pickard: emc-p...@hypercom.com Dave Heald: davehe...@attbi.com For policy questions, send mail to: Richard Nute: ri...@ieee.org Jim Bacher: j.bac...@ieee.org All emc-pstc postings are archived and searchable on the web at: http://ieeepstc.mindcruiser.com/ Click on browse and then emc-pstc mailing list --- This message is from the IEEE EMC Society Product Safety Technical Committee emc-pstc discussion list. Visit our web site at: http://www.ewh.ieee.org/soc/emcs/pstc/ To cancel your subscription, send mail to: majord...@ieee.org with the single line: unsubscribe emc-pstc For help, send mail to the list administrators: Ron Pickard: emc-p...@hypercom.com Dave Heald: davehe...@attbi.com For policy questions, send mail to: Richard Nute: ri...@ieee.org Jim Bacher: j.bac...@ieee.org All emc-pstc postings are archived and searchable on the web at: http://ieeepstc.mindcruiser.com/ Click on browse and then emc-pstc mailing list
(Fwd) Medical Devices Environmental Requirements
Joel, To the best of my knowledge, there are no specific environmental standards for medical equipment. IEC 60601-1, (the product safety standard), clause 10.1, states Equipment shall be capable, while packed for transport and storage, of being exposed to environmental conditions as stated by the manufacturer. Clause 10.2.1 specifies ambient temperature, relative humidity and atmospheric pressure ranges for operation in normal use. Particular types of medical equipment, which are required to comply with one of the IEC 60601-2-XX standards, may be subject to additional or different environmental requirements. Regards, Jon Griver http://www.601help.com The Medical Device Developer's Guide to IEC 60601-1 Hi All Can anybody help me out with Environmental testing requirement/Standards (Temp,Vibration) for medical devices? Thank You Joel Mandel Reliability Dept ADC Israel --- This message is from the IEEE EMC Society Product Safety Technical Committee emc-pstc discussion list. Visit our web site at: http://www.ewh.ieee.org/soc/emcs/pstc/ To cancel your subscription, send mail to: majord...@ieee.org with the single line: unsubscribe emc-pstc For help, send mail to the list administrators: Ron Pickard: emc-p...@hypercom.com Dave Heald: davehe...@attbi.com For policy questions, send mail to: Richard Nute: ri...@ieee.org Jim Bacher: j.bac...@ieee.org All emc-pstc postings are archived and searchable on the web at: http://ieeepstc.mindcruiser.com/ Click on browse and then emc-pstc mailing list
RE: Medical Devices Environmental Requirements
Are there any customers-specific requirements? In the telecom industry, NEBS/ETSI (Telcordia GR-63 ETS 300 019) come to mind. Regards, Naftali Shani, Catena Networks (www.catena.com) 307 Legget Drive, Ottawa, Ontario, Canada K2K 3C8 613.599.6430/866.2CATENA (X.8277); C 295.7042; F 599.0445 E-mail: nsh...@catena.com -Original Message- From: Mandel, Joel [mailto:joel_man...@adc.com] Sent: Tuesday, November 05, 2002 11:43 AM To: 'emc-p...@majordomo.ieee.org' Subject:Medical Devices Environmental Requirements Hi All Can anybody help me out with Environmental testing requirement/Standards (Temp,Vibration) for medical devices? Thank You Joel Mandel Reliability Dept ADC Israel --- This message is from the IEEE EMC Society Product Safety Technical Committee emc-pstc discussion list. Visit our web site at: http://www.ewh.ieee.org/soc/emcs/pstc/ To cancel your subscription, send mail to: majord...@ieee.org with the single line: unsubscribe emc-pstc For help, send mail to the list administrators: Ron Pickard: emc-p...@hypercom.com Dave Heald: davehe...@attbi.com For policy questions, send mail to: Richard Nute: ri...@ieee.org Jim Bacher: j.bac...@ieee.org All emc-pstc postings are archived and searchable on the web at: http://ieeepstc.mindcruiser.com/ Click on browse and then emc-pstc mailing list --- This message is from the IEEE EMC Society Product Safety Technical Committee emc-pstc discussion list. Visit our web site at: http://www.ewh.ieee.org/soc/emcs/pstc/ To cancel your subscription, send mail to: majord...@ieee.org with the single line: unsubscribe emc-pstc For help, send mail to the list administrators: Ron Pickard: emc-p...@hypercom.com Dave Heald: davehe...@attbi.com For policy questions, send mail to: Richard Nute: ri...@ieee.org Jim Bacher: j.bac...@ieee.org All emc-pstc postings are archived and searchable on the web at: http://ieeepstc.mindcruiser.com/ Click on browse and then emc-pstc mailing list
Medical Devices Environmental Requirements
Hi All Can anybody help me out with Environmental testing requirement/Standards (Temp,Vibration) for medical devices? Thank You Joel Mandel Reliability Dept ADC Israel --- This message is from the IEEE EMC Society Product Safety Technical Committee emc-pstc discussion list. Visit our web site at: http://www.ewh.ieee.org/soc/emcs/pstc/ To cancel your subscription, send mail to: majord...@ieee.org with the single line: unsubscribe emc-pstc For help, send mail to the list administrators: Ron Pickard: emc-p...@hypercom.com Dave Heald: davehe...@attbi.com For policy questions, send mail to: Richard Nute: ri...@ieee.org Jim Bacher: j.bac...@ieee.org All emc-pstc postings are archived and searchable on the web at: http://ieeepstc.mindcruiser.com/ Click on browse and then emc-pstc mailing list
RE: Distributors of Medical Devices in Europe
Hi Neil, Thanks for your response. Let me see if I can clarify. The product we are talking about is Class 1 device. We have offices and service centers in the UK, Germany and France. In addition to this we use a network of distributors. These distributors for this device in question are medical device distributors. Currently we manufacture product here in the US but that could change in the future. To date we have product that fall into ITE, Laboratory equipment and Medical categories. We currently have DOC for this device in question and it conforms with the MDD. According to Article 14, COUNCIL DIRECTIVE 93/42/EEC it indicates informing the competent authorities in that member state. Thus my confusion, I did not concern myself with this issue until recently, when the question was asked, Are there any requirements for European Distributors of Medical Devices . Article 14 Article 14 Registration of persons responsible for placing devices on the market 1. Any manufacturer who, under his own name, places devices on the market in accordance with the procedures referred to in Article 11 (5) and (6) and any other natural or legal person engaged in the activities referred to in Article 12 shall inform the competent authorities of the Member State in which he has his registered place of business of the address of the registered place of business and the description of the devices concerned. 2. Where a manufacturer who places devices referred to in paragraph 1 on the market under his own name does not have a registered place of business in a Member State, he shall designate the person(s) responsible for marketing them who is (are) established in the Community. These persons shall inform the competent authorities of the Member State in which they have their registered place of business of the address of the registered place of business and the category of devices concerned. 3. The Member States shall on request inform the other Member States and the Commission of the details referred to in paragraphs 1 and 2. Thank you, Mark Schmidt Regulatory Compliance X-Rite Incorporated Grandville, MI USA (616) 257 2469 mschm...@xrite.com -Original Message- From: Barker, Neil [mailto:neil.bar...@e2vtechnologies.com] Sent: Wednesday, September 18, 2002 11:29 AM To: Mark Schmidt; emc-p...@majordomo.ieee.org Subject:RE: Distributors of Medical Devices in Europe File: Neil R. Barker (E-mail).vcf In general a Class 1 Medical Device requires only a manufacturer's Declaration of Conformity before it is placed on the market. However, in the context of the Medical Devices Directive the manufacturer is not the person or entity who makes the product but the person or entity who places the product on the EU market in their own name. Hence a manufacturer can be one person in an office who contracts with others to design and manufacture a product which is then placed on the market under a label, Brand X, that belongs to the one-person company. It is the responsibility of that one-person company to issue the Declaration of Conformity, and it is their responsibility to ensure that all the requirements of the MDD are met. In the case of Distributors, it is the responsibility of the original importer into the EU, but not of further distibution thereafter. It is worth reading the Directive itself. It is not that large a document and is quite readable. Best regards, Neil R. Barker Compliance Engineering Manager E2V Technologies Waterhouse Lane Chelmsford Essex CM1 2QU U.K. Tel: +44 (01245) 453616 Fax: +44 (01245) 453410 E-mail: neil.bar...@e2vtechnologies.com -Original Message- From: Mark Schmidt [mailto:mschm...@xrite.com] Sent: 18 September 2002 13:11 To: emc-p...@majordomo.ieee.org Subject: Distributors of Medical Devices in Europe Hello Group, Are there any requirements for European Distributors of Medical Devices, i.e. license or registration. The Device in question is a Class 1 device. Thank you. Mark Schmidt Regulatory Compliance X-Rite Incorporated Grandville, MI USA (616) 257 2469 mschm...@xrite.com --- This message is from the IEEE EMC Society Product Safety Technical Committee emc-pstc discussion list. Visit our web site at: http://www.ewh.ieee.org/soc/emcs/pstc/ To cancel your subscription, send mail to: majord...@ieee.org with the single line: unsubscribe emc-pstc For help, send mail to the list administrators: Ron Pickard: emc-p...@hypercom.com Dave Heald: davehe...@attbi.com For policy questions, send mail to: Richard Nute: ri...@ieee.org Jim Bacher: j.bac...@ieee.org All emc-pstc postings are archived and searchable on the web at: http://ieeepstc.mindcruiser.com/ Click on browse and then emc-pstc mailing list
Re: Distributors of Medical Devices in Europe
Mark, I assume that you mean Class 1 according to the EU's Medical Devices Directive, and not the FDA classification. Although the principle is the same, i.e. Class 1 is low-risk, the classification method is not identical. The MDD deals with the device rather than the distributor. Compliance of the device with the MDD is mandatory. You should probably check the Health Ministry or 'Competant Authority' (as defined in the MDD), to check whether there are requirements on the distributor. As this subject is not covered by the MDD, the regulations are likely to be different in each country of the EU. You can contact the UK Competent Authority via their website at: http://www.medical-devices.gov.uk/ Distributors would certainly be obligated to report incidents involving injury or death to the authorities. I hope this helps. Jon Griver http://www.601help.com The Medical Device Developer's Guide to IEC 60601-1 On 18 Sep 2002 at 8:11, Mark Schmidt wrote: Hello Group, Are there any requirements for European Distributors of Medical Devices, i.e. license or registration. The Device in question is a Class 1 device. Thank you. Mark Schmidt Regulatory Compliance X-Rite Incorporated Grandville, MI USA (616) 257 2469 mschm...@xrite.com --- This message is from the IEEE EMC Society Product Safety Technical Committee emc-pstc discussion list. Visit our web site at: http://www.ewh.ieee.org/soc/emcs/pstc/ To cancel your subscription, send mail to: majord...@ieee.org with the single line: unsubscribe emc-pstc For help, send mail to the list administrators: Ron Pickard: emc-p...@hypercom.com Dave Heald: davehe...@attbi.com For policy questions, send mail to: Richard Nute: ri...@ieee.org Jim Bacher: j.bac...@ieee.org All emc-pstc postings are archived and searchable on the web at: http://ieeepstc.mindcruiser.com/ Click on browse and then emc-pstc mailing list --- This message is from the IEEE EMC Society Product Safety Technical Committee emc-pstc discussion list. Visit our web site at: http://www.ewh.ieee.org/soc/emcs/pstc/ To cancel your subscription, send mail to: majord...@ieee.org with the single line: unsubscribe emc-pstc For help, send mail to the list administrators: Ron Pickard: emc-p...@hypercom.com Dave Heald: davehe...@attbi.com For policy questions, send mail to: Richard Nute: ri...@ieee.org Jim Bacher: j.bac...@ieee.org All emc-pstc postings are archived and searchable on the web at: http://ieeepstc.mindcruiser.com/ Click on browse and then emc-pstc mailing list
Distributors of Medical Devices in Europe
Hello Group, Are there any requirements for European Distributors of Medical Devices, i.e. license or registration. The Device in question is a Class 1 device. Thank you. Mark Schmidt Regulatory Compliance X-Rite Incorporated Grandville, MI USA (616) 257 2469 mschm...@xrite.com --- This message is from the IEEE EMC Society Product Safety Technical Committee emc-pstc discussion list. Visit our web site at: http://www.ewh.ieee.org/soc/emcs/pstc/ To cancel your subscription, send mail to: majord...@ieee.org with the single line: unsubscribe emc-pstc For help, send mail to the list administrators: Ron Pickard: emc-p...@hypercom.com Dave Heald: davehe...@attbi.com For policy questions, send mail to: Richard Nute: ri...@ieee.org Jim Bacher: j.bac...@ieee.org All emc-pstc postings are archived and searchable on the web at: http://ieeepstc.mindcruiser.com/ Click on browse and then emc-pstc mailing list
(Fwd) Medical devices, current limit
Kim, This type of product may be covered by IEC 60601-2-10, PARTICULAR REQUIREMENTS FOR THE SAFETY OF NERVE AND MUSCLE STIMULATORS, or possibly, IEC 60601-2-14, PARTICULAR REQUIREMENTS FOR THE SAFETY OF ELECTROCONVULSIVE THERAPY EQUIPMENT I don't have a copy of these standards, but they should relate to this issue. With regard to a general discusion of this subject, have a look at the Rationale section (Appendix A) of IEC 60601-1. The rationale on sub-clause 19.3 gives information, a graph and references on this subject. Regards, Jon Griver http://www.601help.com THe Medical Device Developer's Guide to IEC 60601-1 --- Forwarded message follows --- Hi all I have a medical device under EU directive 93/42/EEC. It is a class IIa product and sends some small electrical pulses through the body for treatments. Does anyone know of any limit for the current ? (voltage is generated from battery and is max. 16Vdc) Does anyone have the current/frequency diagram over max. current the the body can withstand without hazards ? (It is normally used for mains leakage current protection). Best regards, Kim Boll Jensen Bolls Raadgivning Denmark --- End of forwarded message --- --- This message is from the IEEE EMC Society Product Safety Technical Committee emc-pstc discussion list. Visit our web site at: http://www.ewh.ieee.org/soc/emcs/pstc/ To cancel your subscription, send mail to: majord...@ieee.org with the single line: unsubscribe emc-pstc For help, send mail to the list administrators: Ron Pickard: emc-p...@hypercom.com Dave Heald: davehe...@attbi.com For policy questions, send mail to: Richard Nute: ri...@ieee.org Jim Bacher: j.bac...@ieee.org All emc-pstc postings are archived and searchable on the web at: http://ieeepstc.mindcruiser.com/ Click on browse and then emc-pstc mailing list
RE: Noise Levels for Household Appliances / Medical Devices
John, I am intersted in acoustic noise levels. This e-mail message may contain privileged or confidential information. If you are not the intended recipient, you may not disclose, use, disseminate, distribute, copy or rely upon this message or attachment in any way. If you received this e-mail message in error, please return by forwarding the message and its attachments to the sender. PETER S. MERGUERIAN Technical Director I.T.L. (Product Testing) Ltd. 26 Hacharoshet St., POB 211 Or Yehuda 60251, Israel Tel: + 972-(0)3-5339022 Fax: + 972-(0)3-5339019 Mobile: + 972-(0)54-838175 http://www.itl.co.il http://www.i-spec.com -Original Message- From: John Shinn [mailto:john.sh...@sanmina-sci.com] Sent: Wednesday, August 14, 2002 6:53 PM To: 'Peter Merguerian'; 'EMC-PSTC (E-mail) ' Subject: RE: Noise Levels for Household Appliances / Medical Devices Are you talking electrical (EMC) noise or Acoustic? John Shinn -Original Message- From: owner-emc-p...@majordomo.ieee.org [mailto:owner-emc-p...@majordomo.ieee.org]On Behalf Of Peter Merguerian Sent: Wednesday, August 14, 2002 3:09 AM To: EMC-PSTC (E-mail) Subject: Noise Levels for Household Appliances / Medical Devices Dear All, Any recommended European or North American noise levels for household appliances such as massagers? What are the applicable standards? Any recommended European or North American noise levels for medical equipment, such as massagers used by professionals in a therapy center? What are the applicable standards? This e-mail message may contain privileged or confidential information. If you are not the intended recipient, you may not disclose, use, disseminate, distribute, copy or rely upon this message or attachment in any way. If you received this e-mail message in error, please return by forwarding the message and its attachments to the sender. PETER S. MERGUERIAN Technical Director I.T.L. (Product Testing) Ltd. 26 Hacharoshet St., POB 211 Or Yehuda 60251, Israel Tel: + 972-(0)3-5339022 Fax: + 972-(0)3-5339019 Mobile: + 972-(0)54-838175 http://www.itl.co.il http://www.i-spec.com --- This message is from the IEEE EMC Society Product Safety Technical Committee emc-pstc discussion list. Visit our web site at: http://www.ewh.ieee.org/soc/emcs/pstc/ To cancel your subscription, send mail to: majord...@ieee.org with the single line: unsubscribe emc-pstc For help, send mail to the list administrators: Ron Pickard: emc-p...@hypercom.com Dave Heald: davehe...@attbi.com For policy questions, send mail to: Richard Nute: ri...@ieee.org Jim Bacher: j.bac...@ieee.org All emc-pstc postings are archived and searchable on the web at: http://ieeepstc.mindcruiser.com/ Click on browse and then emc-pstc mailing list --- This message is from the IEEE EMC Society Product Safety Technical Committee emc-pstc discussion list. Visit our web site at: http://www.ewh.ieee.org/soc/emcs/pstc/ To cancel your subscription, send mail to: majord...@ieee.org with the single line: unsubscribe emc-pstc For help, send mail to the list administrators: Ron Pickard: emc-p...@hypercom.com Dave Heald: davehe...@attbi.com For policy questions, send mail to: Richard Nute: ri...@ieee.org Jim Bacher: j.bac...@ieee.org All emc-pstc postings are archived and searchable on the web at: http://ieeepstc.mindcruiser.com/ Click on browse and then emc-pstc mailing list
Medical devices, current limit
Hi all I have a medical device under EU directive 93/42/EEC. It is a class IIa product and sends some small electrical pulses through the body for treatments. Does anyone know of any limit for the current ? (voltage is generated from battery and is max. 16Vdc) Does anyone have the current/frequency diagram over max. current the the body can withstand without hazards ? (It is normally used for mains leakage current protection). Best regards, Kim Boll Jensen Bolls Raadgivning Denmark attachment: kimboll.vcf
Re: Noise Levels for Household Appliances / Medical Devices
I don't think there are mandatory noise limits for these appliances, but directive 86/594/EEC says that if you do declare a noise value for an appliance, you have to use the right test methods and it has to be accurate. Try looking for EN 60704 in the BSI on-line catalogue, and Statutory Instrument 1990:161 at www.hmso.gov.uk. Rgds Nick. At 12:09 +0200 14/8/02, Peter Merguerian wrote: Dear All, Any recommended European or North American noise levels for household appliances such as massagers? What are the applicable standards? Any recommended European or North American noise levels for medical equipment, such as massagers used by professionals in a therapy center? What are the applicable standards? This e-mail message may contain privileged or confidential information. If you are not the intended recipient, you may not disclose, use, disseminate, distribute, copy or rely upon this message or attachment in any way. If you received this e-mail message in error, please return by forwarding the message and its attachments to the sender. PETER S. MERGUERIAN Technical Director I.T.L. (Product Testing) Ltd. 26 Hacharoshet St., POB 211 Or Yehuda 60251, Israel Tel: + 972-(0)3-5339022 Fax: + 972-(0)3-5339019 Mobile: + 972-(0)54-838175 http://www.itl.co.il http://www.i-spec.com --- This message is from the IEEE EMC Society Product Safety Technical Committee emc-pstc discussion list. Visit our web site at: http://www.ewh.ieee.org/soc/emcs/pstc/ To cancel your subscription, send mail to: majord...@ieee.org with the single line: unsubscribe emc-pstc For help, send mail to the list administrators: Ron Pickard: emc-p...@hypercom.com Dave Heald: davehe...@attbi.com For policy questions, send mail to: Richard Nute: ri...@ieee.org Jim Bacher: j.bac...@ieee.org All emc-pstc postings are archived and searchable on the web at: http://ieeepstc.mindcruiser.com/ Click on browse and then emc-pstc mailing list
RE: Noise Levels for Household Appliances / Medical Devices
Are you talking electrical (EMC) noise or Acoustic? John Shinn -Original Message- From: owner-emc-p...@majordomo.ieee.org [mailto:owner-emc-p...@majordomo.ieee.org]On Behalf Of Peter Merguerian Sent: Wednesday, August 14, 2002 3:09 AM To: EMC-PSTC (E-mail) Subject: Noise Levels for Household Appliances / Medical Devices Dear All, Any recommended European or North American noise levels for household appliances such as massagers? What are the applicable standards? Any recommended European or North American noise levels for medical equipment, such as massagers used by professionals in a therapy center? What are the applicable standards? This e-mail message may contain privileged or confidential information. If you are not the intended recipient, you may not disclose, use, disseminate, distribute, copy or rely upon this message or attachment in any way. If you received this e-mail message in error, please return by forwarding the message and its attachments to the sender. PETER S. MERGUERIAN Technical Director I.T.L. (Product Testing) Ltd. 26 Hacharoshet St., POB 211 Or Yehuda 60251, Israel Tel: + 972-(0)3-5339022 Fax: + 972-(0)3-5339019 Mobile: + 972-(0)54-838175 http://www.itl.co.il http://www.i-spec.com --- This message is from the IEEE EMC Society Product Safety Technical Committee emc-pstc discussion list. Visit our web site at: http://www.ewh.ieee.org/soc/emcs/pstc/ To cancel your subscription, send mail to: majord...@ieee.org with the single line: unsubscribe emc-pstc For help, send mail to the list administrators: Ron Pickard: emc-p...@hypercom.com Dave Heald: davehe...@attbi.com For policy questions, send mail to: Richard Nute: ri...@ieee.org Jim Bacher: j.bac...@ieee.org All emc-pstc postings are archived and searchable on the web at: http://ieeepstc.mindcruiser.com/ Click on browse and then emc-pstc mailing list --- This message is from the IEEE EMC Society Product Safety Technical Committee emc-pstc discussion list. Visit our web site at: http://www.ewh.ieee.org/soc/emcs/pstc/ To cancel your subscription, send mail to: majord...@ieee.org with the single line: unsubscribe emc-pstc For help, send mail to the list administrators: Ron Pickard: emc-p...@hypercom.com Dave Heald: davehe...@attbi.com For policy questions, send mail to: Richard Nute: ri...@ieee.org Jim Bacher: j.bac...@ieee.org All emc-pstc postings are archived and searchable on the web at: http://ieeepstc.mindcruiser.com/ Click on browse and then emc-pstc mailing list
RE: Noise Levels for Household Appliances / Medical Devices
Peter, I don't have a clue to what levels we are exposed at home (never measured) or what standard/s do apply, but for telecom equipment in North America and Europe I have some references: 1. For Europe, ETS 300 753 (my version is from October 1997) uses the range of 5-5.5 bels for close proximity in an office environment - contact me off-line for further details 2. In North America, Telcordia GR-63 uses the range of 60-65 dBA in a 'similar' environment Regards, Naftali Shani, Catena Networks (www.catena.com) 307 Legget Drive, Ottawa, Ontario, Canada K2K 3C8 613.599.6430/866.2CATENA (X.8277); C 295.7042; F 599.0445 E-mail: nsh...@catena.com -Original Message- From: Peter Merguerian [mailto:pmerguer...@itl.co.il] Sent: Wednesday, August 14, 2002 6:09 AM To: EMC-PSTC (E-mail) Subject: Noise Levels for Household Appliances / Medical Devices Dear All, Any recommended European or North American noise levels for household appliances such as massagers? What are the applicable standards? Any recommended European or North American noise levels for medical equipment, such as massagers used by professionals in a therapy center? What are the applicable standards? This e-mail message may contain privileged or confidential information. If you are not the intended recipient, you may not disclose, use, disseminate, distribute, copy or rely upon this message or attachment in any way. If you received this e-mail message in error, please return by forwarding the message and its attachments to the sender. PETER S. MERGUERIAN Technical Director I.T.L. (Product Testing) Ltd. 26 Hacharoshet St., POB 211 Or Yehuda 60251, Israel Tel: + 972-(0)3-5339022 Fax: + 972-(0)3-5339019 Mobile: + 972-(0)54-838175 http://www.itl.co.il http://www.i-spec.com --- This message is from the IEEE EMC Society Product Safety Technical Committee emc-pstc discussion list. Visit our web site at: http://www.ewh.ieee.org/soc/emcs/pstc/ To cancel your subscription, send mail to: majord...@ieee.org with the single line: unsubscribe emc-pstc For help, send mail to the list administrators: Ron Pickard: emc-p...@hypercom.com Dave Heald: davehe...@attbi.com For policy questions, send mail to: Richard Nute: ri...@ieee.org Jim Bacher: j.bac...@ieee.org All emc-pstc postings are archived and searchable on the web at: http://ieeepstc.mindcruiser.com/ Click on browse and then emc-pstc mailing list --- This message is from the IEEE EMC Society Product Safety Technical Committee emc-pstc discussion list. Visit our web site at: http://www.ewh.ieee.org/soc/emcs/pstc/ To cancel your subscription, send mail to: majord...@ieee.org with the single line: unsubscribe emc-pstc For help, send mail to the list administrators: Ron Pickard: emc-p...@hypercom.com Dave Heald: davehe...@attbi.com For policy questions, send mail to: Richard Nute: ri...@ieee.org Jim Bacher: j.bac...@ieee.org All emc-pstc postings are archived and searchable on the web at: http://ieeepstc.mindcruiser.com/ Click on browse and then emc-pstc mailing list
Noise Levels for Household Appliances / Medical Devices
Dear All, Any recommended European or North American noise levels for household appliances such as massagers? What are the applicable standards? Any recommended European or North American noise levels for medical equipment, such as massagers used by professionals in a therapy center? What are the applicable standards? This e-mail message may contain privileged or confidential information. If you are not the intended recipient, you may not disclose, use, disseminate, distribute, copy or rely upon this message or attachment in any way. If you received this e-mail message in error, please return by forwarding the message and its attachments to the sender. PETER S. MERGUERIAN Technical Director I.T.L. (Product Testing) Ltd. 26 Hacharoshet St., POB 211 Or Yehuda 60251, Israel Tel: + 972-(0)3-5339022 Fax: + 972-(0)3-5339019 Mobile: + 972-(0)54-838175 http://www.itl.co.il http://www.i-spec.com --- This message is from the IEEE EMC Society Product Safety Technical Committee emc-pstc discussion list. Visit our web site at: http://www.ewh.ieee.org/soc/emcs/pstc/ To cancel your subscription, send mail to: majord...@ieee.org with the single line: unsubscribe emc-pstc For help, send mail to the list administrators: Ron Pickard: emc-p...@hypercom.com Dave Heald: davehe...@attbi.com For policy questions, send mail to: Richard Nute: ri...@ieee.org Jim Bacher: j.bac...@ieee.org All emc-pstc postings are archived and searchable on the web at: http://ieeepstc.mindcruiser.com/ Click on browse and then emc-pstc mailing list
RE: Medical Devices in Japan and Australia
Mike, For Australia, some of the standards you may wish to check include: Safety - AS/NZS 3200.X.X suite of standards (Approval and Test Specifications - Medical electrical equipment). These are based on the IEC 601-X-X suite of standards. EMC - AS/NZS 2064 (equivalent to CISPR 11 and EN55011) Best regards, Kevin Richardson Stanimore Pty Limited Compliance Advice Solutions for Technology (including Australian Agent Services) (Legislation/Regulations/Standards) Ph: 02-4329-4070 (Int'l: +61-2-4329-4070) Fax: 02-4328-5639 (Int'l: +61-2-4328-5639) Mobile: 04-1224-1620 (Int'l: +61-4-1224-1620) Email:k...@compuserve.com (kevin.richard...@ieee.org - alternate email) The material transmitted in this message or contained in attachments to this message may contain confidential and/or privileged material and is intended only for the addressee. Any use of or reliance upon this material by persons or entities other than the addressee is prohibited. If you receive this in error, please notify the sender and destroy any copies of the material immediately. -Original Message- From: owner-emc-p...@majordomo.ieee.org [mailto:owner-emc-p...@majordomo.ieee.org]On Behalf Of Mike Cantwell Sent: Friday, 25 January 2002 6:24 AM To: emc-pstc (E-mail) Subject: Medical Devices in Japan and Australia Does anyone know: 1) Standards for medical devices in Japan and Australia? 2) Approval procedure for medical devices in Japan and Australia? Any help/direction would be appreciated. Thanks, Mike Cantwell --- This message is from the IEEE EMC Society Product Safety Technical Committee emc-pstc discussion list. Visit our web site at: http://www.ewh.ieee.org/soc/emcs/pstc/ To cancel your subscription, send mail to: majord...@ieee.org with the single line: unsubscribe emc-pstc For help, send mail to the list administrators: Michael Garretson:pstc_ad...@garretson.org Dave Healddavehe...@mediaone.net For policy questions, send mail to: Richard Nute: ri...@ieee.org Jim Bacher: j.bac...@ieee.org All emc-pstc postings are archived and searchable on the web at: No longer online until our new server is brought online and the old messages are imported into the new server. --- This message is from the IEEE EMC Society Product Safety Technical Committee emc-pstc discussion list. Visit our web site at: http://www.ewh.ieee.org/soc/emcs/pstc/ To cancel your subscription, send mail to: majord...@ieee.org with the single line: unsubscribe emc-pstc For help, send mail to the list administrators: Michael Garretson:pstc_ad...@garretson.org Dave Healddavehe...@mediaone.net For policy questions, send mail to: Richard Nute: ri...@ieee.org Jim Bacher: j.bac...@ieee.org All emc-pstc postings are archived and searchable on the web at: No longer online until our new server is brought online and the old messages are imported into the new server.
Medical Devices in Japan and Australia
Does anyone know: 1) Standards for medical devices in Japan and Australia? 2) Approval procedure for medical devices in Japan and Australia? Any help/direction would be appreciated. Thanks, Mike Cantwell --- This message is from the IEEE EMC Society Product Safety Technical Committee emc-pstc discussion list. Visit our web site at: http://www.ewh.ieee.org/soc/emcs/pstc/ To cancel your subscription, send mail to: majord...@ieee.org with the single line: unsubscribe emc-pstc For help, send mail to the list administrators: Michael Garretson:pstc_ad...@garretson.org Dave Healddavehe...@mediaone.net For policy questions, send mail to: Richard Nute: ri...@ieee.org Jim Bacher: j.bac...@ieee.org All emc-pstc postings are archived and searchable on the web at: No longer online until our new server is brought online and the old messages are imported into the new server.
RE: Medical devices in Asia
An excellent resource, John. Thank you. Regards, Peter L. Tarver, PE Product Safety Manager Sanmina-SCI Homologation Services peter.tar...@sanmina-sci.com -Original Message- From: Jon Griver Sent: Tuesday, January 22, 2002 9:59 PM Peter, An excellent place to start your search is: http://www.ita.doc.gov/td/mdequip/regulations.html which is the US Department of Commerce site, giving information on the medical device requirements for many countries. Good Luck Jon Griver www.601help.com The Medical Device Developer's Guide to IEC 60601-1 All - I have been asked to look into medical device requirements in Asia. This would include Pacific Rim and other Asian countries (the request was not specific, despite a little prodding). I am only beginning my research and would appreciate any pointers you might be able to provide that will shorten the time it takes me in discovery. Thank you. Regards, Peter L. Tarver, PE Product Safety Manager Sanmina-SCI Homologation Services peter.tar...@sanmina-sci.com --- This message is from the IEEE EMC Society Product Safety Technical Committee emc-pstc discussion list. Visit our web site at: http://www.ewh.ieee.org/soc/emcs/pstc/ To cancel your subscription, send mail to: majord...@ieee.org with the single line: unsubscribe emc-pstc For help, send mail to the list administrators: Michael Garretson:pstc_ad...@garretson.org Dave Healddavehe...@mediaone.net For policy questions, send mail to: Richard Nute: ri...@ieee.org Jim Bacher: j.bac...@ieee.org All emc-pstc postings are archived and searchable on the web at: No longer online until our new server is brought online and the old messages are imported into the new server.
Re: Medical devices in Asia
Peter, An excellent place to start your search is: http://www.ita.doc.gov/td/mdequip/regulations.html which is the US Department of Commerce site, giving information on the medical device requirements for many countries. Good Luck Jon Griver www.601help.com The Medical Device Developer's Guide to IEC 60601-1 All - I have been asked to look into medical device requirements in Asia. This would include Pacific Rim and other Asian countries (the request was not specific, despite a little prodding). I am only beginning my research and would appreciate any pointers you might be able to provide that will shorten the time it takes me in discovery. Thank you. Regards, Peter L. Tarver, PE Product Safety Manager Sanmina-SCI Homologation Services peter.tar...@sanmina-sci.com --- This message is from the IEEE EMC Society Product Safety Technical Committee emc-pstc discussion list. Visit our web site at: http://www.ewh.ieee.org/soc/emcs/pstc/ To cancel your subscription, send mail to: majord...@ieee.org with the single line: unsubscribe emc-pstc For help, send mail to the list administrators: Michael Garretson:pstc_ad...@garretson.org Dave Healddavehe...@mediaone.net For policy questions, send mail to: Richard Nute: ri...@ieee.org Jim Bacher: j.bac...@ieee.org All emc-pstc postings are archived and searchable on the web at: No longer online until our new server is brought online and the old messages are imported into the new server. --- This message is from the IEEE EMC Society Product Safety Technical Committee emc-pstc discussion list. Visit our web site at: http://www.ewh.ieee.org/soc/emcs/pstc/ To cancel your subscription, send mail to: majord...@ieee.org with the single line: unsubscribe emc-pstc For help, send mail to the list administrators: Michael Garretson:pstc_ad...@garretson.org Dave Healddavehe...@mediaone.net For policy questions, send mail to: Richard Nute: ri...@ieee.org Jim Bacher: j.bac...@ieee.org All emc-pstc postings are archived and searchable on the web at: No longer online until our new server is brought online and the old messages are imported into the new server.
Medical devices in Asia
All - I have been asked to look into medical device requirements in Asia. This would include Pacific Rim and other Asian countries (the request was not specific, despite a little prodding). I am only beginning my research and would appreciate any pointers you might be able to provide that will shorten the time it takes me in discovery. Thank you. Regards, Peter L. Tarver, PE Product Safety Manager Sanmina-SCI Homologation Services peter.tar...@sanmina-sci.com --- This message is from the IEEE EMC Society Product Safety Technical Committee emc-pstc discussion list. Visit our web site at: http://www.ewh.ieee.org/soc/emcs/pstc/ To cancel your subscription, send mail to: majord...@ieee.org with the single line: unsubscribe emc-pstc For help, send mail to the list administrators: Michael Garretson:pstc_ad...@garretson.org Dave Healddavehe...@mediaone.net For policy questions, send mail to: Richard Nute: ri...@ieee.org Jim Bacher: j.bac...@ieee.org All emc-pstc postings are archived and searchable on the web at: No longer online until our new server is brought online and the old messages are imported into the new server.
Re: Medical devices radiated immunity
I read in !emc-pstc that Paolo Peruzzi paolo.peru...@esaote.com wrote (in ofb1e46450.bdd89c4b-onc1256b1a.00494...@esaote.com) about 'Medical devices radiated immunity', on Thu, 6 Dec 2001: The second edition of IEC 60601-1-2 (EMC for medical electrical equipments) sets out a radiated immunity test between 80 MHz and 2.5 GHz. For test methods it refers to IEC 61000-4-3, which covers frequencies up to 2 GHz. So what about the test method? Can you follow the procedure described in Annex J of new 61000-3-4 61000-4-3 ( FDIS version) as indipendent windows method up to 2.5 GHz? Not until the FDIS has passed its vote (closing date 2002-02-01), and not in Europe until the new edition has been notified in the Official Journal. But since there is an obvious error in 60601-1-2, in calling up 61000-4-3 for measurements above 2 GHz, which it does not cover, I think you can make a case in your technical file for using the method. -- Regards, John Woodgate, OOO - Own Opinions Only. http://www.jmwa.demon.co.uk After swimming across the Hellespont, I felt like a Hero. --- This message is from the IEEE EMC Society Product Safety Technical Committee emc-pstc discussion list. Visit our web site at: http://www.ewh.ieee.org/soc/emcs/pstc/ To cancel your subscription, send mail to: majord...@ieee.org with the single line: unsubscribe emc-pstc For help, send mail to the list administrators: Michael Garretson:pstc_ad...@garretson.org Dave Healddavehe...@mediaone.net For policy questions, send mail to: Richard Nute: ri...@ieee.org Jim Bacher: j.bac...@ieee.org All emc-pstc postings are archived and searchable on the web at: No longer online until our new server is brought online and the old messages are imported into the new server.
Medical devices radiated immunity
Hi all, The second edition of IEC 60601-1-2 (EMC for medical electrical equipments) sets out a radiated immunity test between 80 MHz and 2.5 GHz. For test methods it refers to IEC 61000-4-3, which covers frequencies up to 2 GHz. So what about the test method? Can you follow the procedure described in Annex J of new 61000-3-4 ( FDIS version) as indipendent windows method up to 2.5 GHz? Thanks in advance, p. - ESAOTE S.p.A. Paolo Peruzzi Research Product DevelopmentDesign Quality Control Via di Caciolle,15tel:+39.055.4229306 I- 50127 Florence fax:+39.055.4223305 e-mail: paolo.peru...@esaote.com --- This message is from the IEEE EMC Society Product Safety Technical Committee emc-pstc discussion list. Visit our web site at: http://www.ewh.ieee.org/soc/emcs/pstc/ To cancel your subscription, send mail to: majord...@ieee.org with the single line: unsubscribe emc-pstc For help, send mail to the list administrators: Michael Garretson:pstc_ad...@garretson.org Dave Healddavehe...@mediaone.net For policy questions, send mail to: Richard Nute: ri...@ieee.org Jim Bacher: j.bac...@ieee.org All emc-pstc postings are archived and searchable on the web at: No longer online until our new server is brought online and the old messages are imported into the new server.
Medical Devices - another question
So how does one break into the medical device field for a medical device manufacturer? I was recently downsized and all of the compliance jobs in the Northeast seem to be for medical device manufacturers. I have 15+ years of product safety, EMI, etc, but no medical device experience. Opinions appreciated. --- This message is from the IEEE EMC Society Product Safety Technical Committee emc-pstc discussion list. Visit our web site at: http://www.ewh.ieee.org/soc/emcs/pstc/ To cancel your subscription, send mail to: majord...@ieee.org with the single line: unsubscribe emc-pstc For help, send mail to the list administrators: Michael Garretson:pstc_ad...@garretson.org Dave Healddavehe...@mediaone.net For policy questions, send mail to: Richard Nute: ri...@ieee.org Jim Bacher: j.bac...@ieee.org All emc-pstc postings are archived and searchable on the web at: No longer online until our new server is brought online and the old messages are imported into the new server. --- This message is from the IEEE EMC Society Product Safety Technical Committee emc-pstc discussion list. Visit our web site at: http://www.ewh.ieee.org/soc/emcs/pstc/ To cancel your subscription, send mail to: majord...@ieee.org with the single line: unsubscribe emc-pstc For help, send mail to the list administrators: Michael Garretson:pstc_ad...@garretson.org Dave Healddavehe...@mediaone.net For policy questions, send mail to: Richard Nute: ri...@ieee.org Jim Bacher: j.bac...@ieee.org All emc-pstc postings are archived and searchable on the web at: No longer online until our new server is brought online and the old messages are imported into the new server.
RE: Susceptibility level of medical devices (incubator) - urgent!
Hi, Per the FDA guidance on EMC http://www.fda.gov/cdrh/ode/638.pdf The limit is 3 V/m. See the page numbered 25 of the document. Note: This is for general medical equipment! There is a guidance document for incubators http://www.fda.gov/cdrh/ode/incubator.html It references IEC 60601-2-19 and 2-20. 2-19 and 2-20 both require the unit to pass at 3 V/m and to pass or fail safe at 10 V/m. Ned Devine Program Manager III Entela, Inc. 3033 Madison Ave. SE Grand Rapids, MI 49548 616 248 9671 Phone 616 574 9752 Fax ndev...@entela.com e-mail -Original Message- From: Ralph Cameron [mailto:ral...@igs.net] Sent: Monday, August 27, 2001 5:33 PM To: Antonio Sarolic; emc-p...@majordomo.ieee.org Subject: Re: Susceptibility level of medical devices (incubator) - urgent! Antonio: I would say many incubators were manufactured when there were no concerns and the susceptibility was a function of the design and very variable. I have heard they vary from 0.1V/m to 10V/m but don't think there is any set standard. The medical gurus prefer to isolate their equipment by physical barriers and hope that something will not affect them In Canada Health Canada could answer thequestion . In the U.S. I believe its the Food and Drug Administration. Ralph Cameron EMC Consulting and Suppression of Consumer Electronics (After sale). - Original Message - From: Antonio Sarolic antonio.saro...@fer.hr To: emc-p...@majordomo.ieee.org Sent: Monday, August 27, 2001 11:59 AM Subject: Susceptibility level of medical devices (incubator) - urgent! Hi I need urgent info on susceptibility level of medical electronic devices, especially baby incubators. Can anyone confirm if it is 1V/m (according to EU standards)? The EMI source is the GSM BS antenna (900MHz). Thanks very much. Antonio Antonio Sarolic, M.S.E.E. Faculty of Electrical Engineering and Computing Dept. of Radiocommunications and Microwave Engineering Unska 3, HR-1 Zagreb, CROATIA tel. +385 1 61 29 789, fax. +385 1 61 29 717 E-mail: antonio.saro...@fer.hr --- This message is from the IEEE EMC Society Product Safety Technical Committee emc-pstc discussion list. Visit our web site at: http://www.ewh.ieee.org/soc/emcs/pstc/ To cancel your subscription, send mail to: majord...@ieee.org with the single line: unsubscribe emc-pstc For help, send mail to the list administrators: Michael Garretson:pstc_ad...@garretson.org Dave Healddavehe...@mediaone.net For policy questions, send mail to: Richard Nute: ri...@ieee.org Jim Bacher: j.bac...@ieee.org All emc-pstc postings are archived and searchable on the web at: No longer online until our new server is brought online and the old messages are imported into the new server. --- This message is from the IEEE EMC Society Product Safety Technical Committee emc-pstc discussion list. Visit our web site at: http://www.ewh.ieee.org/soc/emcs/pstc/ To cancel your subscription, send mail to: majord...@ieee.org with the single line: unsubscribe emc-pstc For help, send mail to the list administrators: Michael Garretson:pstc_ad...@garretson.org Dave Healddavehe...@mediaone.net For policy questions, send mail to: Richard Nute: ri...@ieee.org Jim Bacher: j.bac...@ieee.org All emc-pstc postings are archived and searchable on the web at: No longer online until our new server is brought online and the old messages are imported into the new server. --- This message is from the IEEE EMC Society Product Safety Technical Committee emc-pstc discussion list. Visit our web site at: http://www.ewh.ieee.org/soc/emcs/pstc/ To cancel your subscription, send mail to: majord...@ieee.org with the single line: unsubscribe emc-pstc For help, send mail to the list administrators: Michael Garretson:pstc_ad...@garretson.org Dave Healddavehe...@mediaone.net For policy questions, send mail to: Richard Nute: ri...@ieee.org Jim Bacher: j.bac...@ieee.org All emc-pstc postings are archived and searchable on the web at: No longer online until our new server is brought online and the old messages are imported into the new server.
Re: Susceptibility level of medical devices (incubator) - urgent!
Antonio: I would say many incubators were manufactured when there were no concerns and the susceptibility was a function of the design and very variable. I have heard they vary from 0.1V/m to 10V/m but don't think there is any set standard. The medical gurus prefer to isolate their equipment by physical barriers and hope that something will not affect them In Canada Health Canada could answer thequestion . In the U.S. I believe its the Food and Drug Administration. Ralph Cameron EMC Consulting and Suppression of Consumer Electronics (After sale). - Original Message - From: Antonio Sarolic antonio.saro...@fer.hr To: emc-p...@majordomo.ieee.org Sent: Monday, August 27, 2001 11:59 AM Subject: Susceptibility level of medical devices (incubator) - urgent! Hi I need urgent info on susceptibility level of medical electronic devices, especially baby incubators. Can anyone confirm if it is 1V/m (according to EU standards)? The EMI source is the GSM BS antenna (900MHz). Thanks very much. Antonio Antonio Sarolic, M.S.E.E. Faculty of Electrical Engineering and Computing Dept. of Radiocommunications and Microwave Engineering Unska 3, HR-1 Zagreb, CROATIA tel. +385 1 61 29 789, fax. +385 1 61 29 717 E-mail: antonio.saro...@fer.hr --- This message is from the IEEE EMC Society Product Safety Technical Committee emc-pstc discussion list. Visit our web site at: http://www.ewh.ieee.org/soc/emcs/pstc/ To cancel your subscription, send mail to: majord...@ieee.org with the single line: unsubscribe emc-pstc For help, send mail to the list administrators: Michael Garretson:pstc_ad...@garretson.org Dave Healddavehe...@mediaone.net For policy questions, send mail to: Richard Nute: ri...@ieee.org Jim Bacher: j.bac...@ieee.org All emc-pstc postings are archived and searchable on the web at: No longer online until our new server is brought online and the old messages are imported into the new server. --- This message is from the IEEE EMC Society Product Safety Technical Committee emc-pstc discussion list. Visit our web site at: http://www.ewh.ieee.org/soc/emcs/pstc/ To cancel your subscription, send mail to: majord...@ieee.org with the single line: unsubscribe emc-pstc For help, send mail to the list administrators: Michael Garretson:pstc_ad...@garretson.org Dave Healddavehe...@mediaone.net For policy questions, send mail to: Richard Nute: ri...@ieee.org Jim Bacher: j.bac...@ieee.org All emc-pstc postings are archived and searchable on the web at: No longer online until our new server is brought online and the old messages are imported into the new server.
Susceptibility level of medical devices (incubator) - urgent!
Hi I need urgent info on susceptibility level of medical electronic devices, especially baby incubators. Can anyone confirm if it is 1V/m (according to EU standards)? The EMI source is the GSM BS antenna (900MHz). Thanks very much. Antonio Antonio Sarolic, M.S.E.E. Faculty of Electrical Engineering and Computing Dept. of Radiocommunications and Microwave Engineering Unska 3, HR-1 Zagreb, CROATIA tel. +385 1 61 29 789, fax. +385 1 61 29 717 E-mail: antonio.saro...@fer.hr --- This message is from the IEEE EMC Society Product Safety Technical Committee emc-pstc discussion list. Visit our web site at: http://www.ewh.ieee.org/soc/emcs/pstc/ To cancel your subscription, send mail to: majord...@ieee.org with the single line: unsubscribe emc-pstc For help, send mail to the list administrators: Michael Garretson:pstc_ad...@garretson.org Dave Healddavehe...@mediaone.net For policy questions, send mail to: Richard Nute: ri...@ieee.org Jim Bacher: j.bac...@ieee.org All emc-pstc postings are archived and searchable on the web at: No longer online until our new server is brought online and the old messages are imported into the new server.
Re: Medical devices in Japan
Thanks to everyone who replied on this thread. I've forwarded the responses to the original enquirer and doubtless they will contact anyone who offered further help as they feel appropriate. Regards Nick. At 03:02 +0100 1/8/2001, Nick Williams wrote: I've been asked if I know what the regime is for medical devices in Japan. I don't know anything about this, but if there's anyone out there who can give me a brief idea or possibly even some URLs to look at, I'd be grateful. Anyone in the business of providing consultancy in this area please let me know and I'll likely put the enquirer in touch direct. Thanks and regards Nick. --- This message is from the IEEE EMC Society Product Safety Technical Committee emc-pstc discussion list. Visit our web site at: http://www.ewh.ieee.org/soc/emcs/pstc/ To cancel your subscription, send mail to: majord...@ieee.org with the single line: unsubscribe emc-pstc For help, send mail to the list administrators: Michael Garretson:pstc_ad...@garretson.org Dave Healddavehe...@mediaone.net For policy questions, send mail to: Richard Nute: ri...@ieee.org Jim Bacher: j.bac...@ieee.org All emc-pstc postings are archived and searchable on the web at: http://www.rcic.com/ click on Virtual Conference Hall,
RE: Medical devices in Japan
Hi Nick, To assist you in determining the applicable standards you should contact a representative from JAAME or MHLW at the following address: Japan External Trade Organization (JETRO) New York Health, Labor and Welfare Department McGraw- Hill Building 44th Floor 1221 Avenue of the Americas New York, New York 10020-1060 (212) 997 6471 or 212 997 6472 Fax 212 869 4750 This url doesn't have much info but here it is: http://www.jaame.or.jp/english/index.html Good luck, I wish I could be of more assistance. Mark -Original Message- From: Nick Williams [mailto:nick.willi...@conformance.co.uk] Sent: Tuesday, July 31, 2001 10:02 PM To: emc-p...@majordomo.ieee.org Subject:Medical devices in Japan I've been asked if I know what the regime is for medical devices in Japan. I don't know anything about this, but if there's anyone out there who can give me a brief idea or possibly even some URLs to look at, I'd be grateful. Anyone in the business of providing consultancy in this area please let me know and I'll likely put the enquirer in touch direct. Thanks and regards Nick. --- This message is from the IEEE EMC Society Product Safety Technical Committee emc-pstc discussion list. Visit our web site at: http://www.ewh.ieee.org/soc/emcs/pstc/ To cancel your subscription, send mail to: majord...@ieee.org with the single line: unsubscribe emc-pstc For help, send mail to the list administrators: Michael Garretson:pstc_ad...@garretson.org Dave Healddavehe...@mediaone.net For policy questions, send mail to: Richard Nute: ri...@ieee.org Jim Bacher: j.bac...@ieee.org All emc-pstc postings are archived and searchable on the web at: http://www.rcic.com/ click on Virtual Conference Hall, --- This message is from the IEEE EMC Society Product Safety Technical Committee emc-pstc discussion list. Visit our web site at: http://www.ewh.ieee.org/soc/emcs/pstc/ To cancel your subscription, send mail to: majord...@ieee.org with the single line: unsubscribe emc-pstc For help, send mail to the list administrators: Michael Garretson:pstc_ad...@garretson.org Dave Healddavehe...@mediaone.net For policy questions, send mail to: Richard Nute: ri...@ieee.org Jim Bacher: j.bac...@ieee.org All emc-pstc postings are archived and searchable on the web at: http://www.rcic.com/ click on Virtual Conference Hall,
RE: Medical devices in Japan
I too am not familiar with Japan, but I have a few tidbits to share. JAMEI (Japan Association of Medical Equipment Industry) Equipment appears to be approved under the leadership of the Ministry of Health and Welfare. Richard Woods -- From: Nick Williams [SMTP:nick.willi...@conformance.co.uk] Sent: Tuesday, July 31, 2001 10:02 PM To: emc-p...@majordomo.ieee.org Subject: Medical devices in Japan I've been asked if I know what the regime is for medical devices in Japan. I don't know anything about this, but if there's anyone out there who can give me a brief idea or possibly even some URLs to look at, I'd be grateful. Anyone in the business of providing consultancy in this area please let me know and I'll likely put the enquirer in touch direct. Thanks and regards Nick. --- This message is from the IEEE EMC Society Product Safety Technical Committee emc-pstc discussion list. Visit our web site at: http://www.ewh.ieee.org/soc/emcs/pstc/ To cancel your subscription, send mail to: majord...@ieee.org with the single line: unsubscribe emc-pstc For help, send mail to the list administrators: Michael Garretson:pstc_ad...@garretson.org Dave Healddavehe...@mediaone.net For policy questions, send mail to: Richard Nute: ri...@ieee.org Jim Bacher: j.bac...@ieee.org All emc-pstc postings are archived and searchable on the web at: http://www.rcic.com/ click on Virtual Conference Hall, --- This message is from the IEEE EMC Society Product Safety Technical Committee emc-pstc discussion list. Visit our web site at: http://www.ewh.ieee.org/soc/emcs/pstc/ To cancel your subscription, send mail to: majord...@ieee.org with the single line: unsubscribe emc-pstc For help, send mail to the list administrators: Michael Garretson:pstc_ad...@garretson.org Dave Healddavehe...@mediaone.net For policy questions, send mail to: Richard Nute: ri...@ieee.org Jim Bacher: j.bac...@ieee.org All emc-pstc postings are archived and searchable on the web at: http://www.rcic.com/ click on Virtual Conference Hall,
RE: Medical devices in Japan
Nick, I have a contact in Japan if you need it. I do know that Japan just approved the 2nd Edition of IEC 60601-1-2 but what I am not sure of is if JTA requires compliance or what the transition date might be. If you would like me to look into this, contact me directly. Best regards, Jim Jim Conrad Conrad EMC Consulting P.O. Box 25 Hamilton, MA 01936 USA Phone: 978-468-3909 E-Mail: jc...@shore.net -Original Message- From: owner-emc-p...@majordomo.ieee.org [mailto:owner-emc-p...@majordomo.ieee.org]On Behalf Of Nick Williams Sent: Tuesday, July 31, 2001 10:02 PM To: emc-p...@majordomo.ieee.org Subject: Medical devices in Japan I've been asked if I know what the regime is for medical devices in Japan. I don't know anything about this, but if there's anyone out there who can give me a brief idea or possibly even some URLs to look at, I'd be grateful. Anyone in the business of providing consultancy in this area please let me know and I'll likely put the enquirer in touch direct. Thanks and regards Nick. --- This message is from the IEEE EMC Society Product Safety Technical Committee emc-pstc discussion list. Visit our web site at: http://www.ewh.ieee.org/soc/emcs/pstc/ To cancel your subscription, send mail to: majord...@ieee.org with the single line: unsubscribe emc-pstc For help, send mail to the list administrators: Michael Garretson:pstc_ad...@garretson.org Dave Healddavehe...@mediaone.net For policy questions, send mail to: Richard Nute: ri...@ieee.org Jim Bacher: j.bac...@ieee.org All emc-pstc postings are archived and searchable on the web at: http://www.rcic.com/ click on Virtual Conference Hall, --- This message is from the IEEE EMC Society Product Safety Technical Committee emc-pstc discussion list. Visit our web site at: http://www.ewh.ieee.org/soc/emcs/pstc/ To cancel your subscription, send mail to: majord...@ieee.org with the single line: unsubscribe emc-pstc For help, send mail to the list administrators: Michael Garretson:pstc_ad...@garretson.org Dave Healddavehe...@mediaone.net For policy questions, send mail to: Richard Nute: ri...@ieee.org Jim Bacher: j.bac...@ieee.org All emc-pstc postings are archived and searchable on the web at: http://www.rcic.com/ click on Virtual Conference Hall,
Medical devices in Japan
I've been asked if I know what the regime is for medical devices in Japan. I don't know anything about this, but if there's anyone out there who can give me a brief idea or possibly even some URLs to look at, I'd be grateful. Anyone in the business of providing consultancy in this area please let me know and I'll likely put the enquirer in touch direct. Thanks and regards Nick. --- This message is from the IEEE EMC Society Product Safety Technical Committee emc-pstc discussion list. Visit our web site at: http://www.ewh.ieee.org/soc/emcs/pstc/ To cancel your subscription, send mail to: majord...@ieee.org with the single line: unsubscribe emc-pstc For help, send mail to the list administrators: Michael Garretson:pstc_ad...@garretson.org Dave Healddavehe...@mediaone.net For policy questions, send mail to: Richard Nute: ri...@ieee.org Jim Bacher: j.bac...@ieee.org All emc-pstc postings are archived and searchable on the web at: http://www.rcic.com/ click on Virtual Conference Hall,
RE: Serial Number Format - Medical Devices
Hi Dick, Section (e) deals with items that have an shelf life (i.e. sterile products). Having the actual month and year makes perfect sense. Active devices usually are not date sensitive, except for usable lifetime (i.e. 10 years) as defined by the manufacture. The year of manufacture only, needs to be on the label per 13.3.l. It was my understanding that the actual year should be on the label. Symbol 6.16 of ISO 15223 can be used or the year (i.e. 01) has to be an integral part of the serial number. If it was a code in the serial number, how would the user know when the usable lifetime is up? Ned Devine Ned Devine Program Manager III Entela, Inc. 3033 Madison Ave. SE Grand Rapids, MI 49548 616 248 9671 Phone 616 574 9752 Fax ndev...@entela.com e-mail -Original Message- From: Dick Grobner [mailto:dick.grob...@medgraph.com] Sent: Monday, June 25, 2001 7:02 PM To: IEEE EMC-PSTC E-Mail Forum (E-mail) Subject: Serial Number Format - Medical Devices A question to the group: Referencing the Medical Device Directive, Annex I, paragraph 13.3 The label must bear the following particulars: (l), year of manufacture for active devices other than those covered by (e). This indication may be included in the batch or serial number. One of my colleagues (in another company) was told by their Notified Body (BSI) that it has been interpreted (by whom?) that this means the month and year needs to be embedded into the actual serial number, such as XXZZ1 whereas XX: month and ZZ: year (or similar). We have never encountered this with our Notified Body (TUV). Our S/N's cross back to a SN log which will reveal the month and year of manufacture. Who is right? Does anyone know where this official interpitation is coming from? The rational behind it? Thank You Dick Grobner Medical Graphics Corporation 350 Oak Grove Parkway St Paul MN 55127 651-766-3395 651-766-3389 (fax) dick.grob...@medgraph.com --- This message is from the IEEE EMC Society Product Safety Technical Committee emc-pstc discussion list. Visit our web site at: http://www.ewh.ieee.org/soc/emcs/pstc/ To cancel your subscription, send mail to: majord...@ieee.org with the single line: unsubscribe emc-pstc For help, send mail to the list administrators: Michael Garretson:pstc_ad...@garretson.org Dave Healddavehe...@mediaone.net For policy questions, send mail to: Richard Nute: ri...@ieee.org Jim Bacher: j.bac...@ieee.org All emc-pstc postings are archived and searchable on the web at: http://www.rcic.com/ click on Virtual Conference Hall, --- This message is from the IEEE EMC Society Product Safety Technical Committee emc-pstc discussion list. Visit our web site at: http://www.ewh.ieee.org/soc/emcs/pstc/ To cancel your subscription, send mail to: majord...@ieee.org with the single line: unsubscribe emc-pstc For help, send mail to the list administrators: Michael Garretson:pstc_ad...@garretson.org Dave Healddavehe...@mediaone.net For policy questions, send mail to: Richard Nute: ri...@ieee.org Jim Bacher: j.bac...@ieee.org All emc-pstc postings are archived and searchable on the web at: http://www.rcic.com/ click on Virtual Conference Hall,
Re: Serial Number Format - Medical Devices
At 06:55 +0100 26/6/2001, John Woodgate wrote: 36BDBCA75E0FD411A80100104B93ABF202C2C24B@mgcmail, Dick Grobner dick.grob...@medgraph.com inimitably wrote: Referencing the Medical Device Directive, Annex I, paragraph 13.3 The label must bear the following particulars: (l), year of manufacture for active devices other than those covered by (e). This indication may be included in the batch or serial number. One of my colleagues (in another company) was told by their Notified Body (BSI) that it has been interpreted (by whom?) that this means the month and year needs to be embedded into the actual serial number, such as XXZZ1 whereas XX: month and ZZ: year (or similar). We have never encountered this with our Notified Body (TUV). Our S/N's cross back to a SN log which will reveal the month and year of manufacture. Who is right? Does anyone know where this official interpitation is coming from? The rational behind it? The words 'MAY be included in the serial number' do not mean 'SHALL be included in the serial number'. I suspect that BSI Testing is being more Royalist than the King. BUT, I don't think that what you are doing is correct, either. The words 'must bear year of manufacture' mean that the year must be stated overtly, not just be deducible from your serial number records. -- Dick, I'd agree exactly with John on this. The directive requires the date to be in a legible form, and provides for the manufacturer to use the serial number as a way of doing this so that manufacturers who use a suitable serial number system do not have to put both date and serial number on if they don't want to. You may care to refer to EN1041:1998 and specifically clause 4.1.10. Rgds Nick. --- This message is from the IEEE EMC Society Product Safety Technical Committee emc-pstc discussion list. Visit our web site at: http://www.ewh.ieee.org/soc/emcs/pstc/ To cancel your subscription, send mail to: majord...@ieee.org with the single line: unsubscribe emc-pstc For help, send mail to the list administrators: Michael Garretson:pstc_ad...@garretson.org Dave Healddavehe...@mediaone.net For policy questions, send mail to: Richard Nute: ri...@ieee.org Jim Bacher: j.bac...@ieee.org All emc-pstc postings are archived and searchable on the web at: http://www.rcic.com/ click on Virtual Conference Hall,
Re: Serial Number Format - Medical Devices
I'd say both are probably right. I've never done medical. But, I have run into this type of thing when I was doing ISO-9001 as a lead auditor. Critical words for interpretation are This indication may be included ... It specifically says *may* which means either *you can* or *you don't have to* depending upon some circumstances. In this instance, I do not know what those circumstances are. It does not say shall which means *must do*. The use of the word *may* leaves it wide open to interpretation. The use of the word *shall* leaves no room for interpretation. Thus, the reason why I said both interpretations are probably right. Now, an attempt to appeal their interpretation could be made. But, if they are requiring others to do this, I doubt they'll back down. In any event, the requirement is to be able to trace from the label of the device to the date of mfr for that device. I suppose one could appeal in the case of: 1) a date was embedded in the serial number and 2) that date code was not of some obvious construction (i.e. encoded/encrypted somehow), then you'd have to look up the serial anyway. Therefore, what's the need of a date code embedded in the serial number when you have to look it up? Also, the word indication may be interpreted any number of ways also. If you want to disect the wording, I'm sure it'll just get worse. Regards, Doug McKean --- This message is from the IEEE EMC Society Product Safety Technical Committee emc-pstc discussion list. Visit our web site at: http://www.ewh.ieee.org/soc/emcs/pstc/ To cancel your subscription, send mail to: majord...@ieee.org with the single line: unsubscribe emc-pstc For help, send mail to the list administrators: Michael Garretson:pstc_ad...@garretson.org Dave Healddavehe...@mediaone.net For policy questions, send mail to: Richard Nute: ri...@ieee.org Jim Bacher: j.bac...@ieee.org All emc-pstc postings are archived and searchable on the web at: http://www.rcic.com/ click on Virtual Conference Hall,
Re: Serial Number Format - Medical Devices
36BDBCA75E0FD411A80100104B93ABF202C2C24B@mgcmail, Dick Grobner dick.grob...@medgraph.com inimitably wrote: Referencing the Medical Device Directive, Annex I, paragraph 13.3 The label must bear the following particulars: (l), year of manufacture for active devices other than those covered by (e). This indication may be included in the batch or serial number. One of my colleagues (in another company) was told by their Notified Body (BSI) that it has been interpreted (by whom?) that this means the month and year needs to be embedded into the actual serial number, such as XXZZ1 whereas XX: month and ZZ: year (or similar). We have never encountered this with our Notified Body (TUV). Our S/N's cross back to a SN log which will reveal the month and year of manufacture. Who is right? Does anyone know where this official interpitation is coming from? The rational behind it? The words 'MAY be included in the serial number' do not mean 'SHALL be included in the serial number'. I suspect that BSI Testing is being more Royalist than the King. BUT, I don't think that what you are doing is correct, either. The words 'must bear year of manufacture' mean that the year must be stated overtly, not just be deducible from your serial number records. -- Regards, John Woodgate, OOO - Own Opinions Only. Phone +44 (0)1268 747839 Fax +44 (0)1268 777124. http://www.jmwa.demon.co.uk Why not call a vertically- applied manulo-pedally-operated quasi-planar chernozem-penetrating and excavating implement a SPADE? --- This message is from the IEEE EMC Society Product Safety Technical Committee emc-pstc discussion list. Visit our web site at: http://www.ewh.ieee.org/soc/emcs/pstc/ To cancel your subscription, send mail to: majord...@ieee.org with the single line: unsubscribe emc-pstc For help, send mail to the list administrators: Michael Garretson:pstc_ad...@garretson.org Dave Healddavehe...@mediaone.net For policy questions, send mail to: Richard Nute: ri...@ieee.org Jim Bacher: j.bac...@ieee.org All emc-pstc postings are archived and searchable on the web at: http://www.rcic.com/ click on Virtual Conference Hall,
RE: Production Line Test Requirements - Medical Devices
Dear Jon, Dick and All Memebers, I believe the UL method for Production-Line Grounding Continuity Test is just to verify that someone did not forget to connect the earthing lead from the appliance inlet or power corrd to the chassis. Therefore an ohmmeter, buzzer or the like is all that is considered necessary. The feeding of an ac or dc current between the protective earth terminal and other parts of the unit likely to be energized by a single insulation fault is not necessary during the production-line tests because these tests have already been conducted by the lab during the type tests. UL and other NRTLs want to make sure that assembler did not forget to connect the main earthing lead. Also, if you check a UL or NRTL descriptive report used for the quarterly folow-up inspections, you will note that the construction of the product has been described in detail, including but not limited to earthing connections, enclosure, covers, screws and washers, etc. European test labs for follow-up purposes, use a Critical Components List but do not describe the construction of the product in detail. Hence, they require the feeding on an ac or dc low voltage current source into the earthing path for each production unit. Regards PETER S. MERGUERIAN Technical Director I.T.L. (Product Testing) Ltd. 26 Hacharoshet St., POB 211 Or Yehuda 60251, Israel Tel: + 972-(0)3-5339022 Fax: + 972-(0)3-5339019 Mobile: + 972-(0)54-838175 -Original Message- From: Jon Griver [mailto:jo...@medson.com] Sent: Tuesday, May 22, 2001 9:06 AM To: Dick Grobner Cc: emc-p...@majordomo.ieee.org Subject: RE: Production Line Test Requirements - Medical Devices Dick, This is the situation as I understand it. Please note that a few years ago I did factory inspections for TUV Rheinland, so I'm speaking from a position of some knowledge (though a bit outdated). UL includes its production test requirements in its standards. It traditionally has only required a check of earth continuity under low current conditions. The European test labs work on the basis of extracting production tests from the standards. They require these tests to be performed as a condition of using their mark on the product. It is the test lab's decision which production tests are to be performed and it is not mandated by any standard. Traditionally, the European test labs require the earth continuity test to be carried out at high current conditions, as in the standard. The logic behind this is that if there is a fault to earth, it will cause a high current which may cause any weak links in the earth circuit to heat up and possibly char and increase in resistance or even burn out completely. (We've all seen charring around sockets). Thus this test is considered more effective in identifying loose connections, etc. due to production mistakes than the UL test, which might miss them. Presumably if you are selling your medical product in the EU, you have a Notified Body (unless your product is Group 1). You should agree your production test plan with them. Be aware that they may insist on the high current earth continuity test. Good Luck, Jon Griver Medson Ltd -Original Message- From: Dick Grobner [mailto:dick.grob...@medgraph.com] Sent: Monday, May 21, 2001 18:23 To: 'Jon Griver' Cc: IEEE EMC-PSTC E-Mail Forum (E-mail) Subject: RE: Production Line Test Requirements - Medical Devices Jon Thx for the info, however, I re-checked the two known US standards (UL2601, Appendix D - Manufacturer's Responsibilities, Construction Considerations and Requirements for Factory Tests) (NFPA 99, Grounding Circuit Continuity - Measurement of Resistance). UL defines Production Line Grounding Continuity Test Equipment as: Any suitable continuity indication device (such as an ohmmeter, a battery and buzzer combination, or the like) may be used to determine compliance with Grounding Continuity Test requirements. The NFPA std is even less definitive on this matter. EN60601-1, Appendix B - Testing During Manufacture and/or Installation, Not Used. See rational to sub-clause 4.1. Which states - Tests described in this standard are type tests. I checked all of the amendments including EN60601-1 for medical systems, and I do not see anything changing this. My questions remains, where does it state to use the 10-25Amp, 50 or 60Hz @ 6Volt for 5 seconds test? Is this a carry over from another EN standard? Why wouldn't a simple continuity tests as defined by UL be appropriate from equipment coming off of the production line (assuring that the protective earth circuit in intake)? Appreciate your feedback Jon! thanks again. Does anyone else on this forum have any input - would appreciate it. --- This message is from the IEEE EMC Society Product Safety Technical Committee emc-pstc discussion list. Visit our web site at: http://www.ewh.ieee.org/soc/emcs/pstc/ To cancel your
RE: Production Line Test Requirements - Medical Devices
Dick, This is the situation as I understand it. Please note that a few years ago I did factory inspections for TUV Rheinland, so I'm speaking from a position of some knowledge (though a bit outdated). UL includes its production test requirements in its standards. It traditionally has only required a check of earth continuity under low current conditions. The European test labs work on the basis of extracting production tests from the standards. They require these tests to be performed as a condition of using their mark on the product. It is the test lab's decision which production tests are to be performed and it is not mandated by any standard. Traditionally, the European test labs require the earth continuity test to be carried out at high current conditions, as in the standard. The logic behind this is that if there is a fault to earth, it will cause a high current which may cause any weak links in the earth circuit to heat up and possibly char and increase in resistance or even burn out completely. (We've all seen charring around sockets). Thus this test is considered more effective in identifying loose connections, etc. due to production mistakes than the UL test, which might miss them. Presumably if you are selling your medical product in the EU, you have a Notified Body (unless your product is Group 1). You should agree your production test plan with them. Be aware that they may insist on the high current earth continuity test. Good Luck, Jon Griver Medson Ltd -Original Message- From: Dick Grobner [mailto:dick.grob...@medgraph.com] Sent: Monday, May 21, 2001 18:23 To: 'Jon Griver' Cc: IEEE EMC-PSTC E-Mail Forum (E-mail) Subject: RE: Production Line Test Requirements - Medical Devices Jon Thx for the info, however, I re-checked the two known US standards (UL2601, Appendix D - Manufacturer's Responsibilities, Construction Considerations and Requirements for Factory Tests) (NFPA 99, Grounding Circuit Continuity - Measurement of Resistance). UL defines Production Line Grounding Continuity Test Equipment as: Any suitable continuity indication device (such as an ohmmeter, a battery and buzzer combination, or the like) may be used to determine compliance with Grounding Continuity Test requirements. The NFPA std is even less definitive on this matter. EN60601-1, Appendix B - Testing During Manufacture and/or Installation, Not Used. See rational to sub-clause 4.1. Which states - Tests described in this standard are type tests. I checked all of the amendments including EN60601-1 for medical systems, and I do not see anything changing this. My questions remains, where does it state to use the 10-25Amp, 50 or 60Hz @ 6Volt for 5 seconds test? Is this a carry over from another EN standard? Why wouldn't a simple continuity tests as defined by UL be appropriate from equipment coming off of the production line (assuring that the protective earth circuit in intake)? Appreciate your feedback Jon! thanks again. Does anyone else on this forum have any input - would appreciate it. --- This message is from the IEEE EMC Society Product Safety Technical Committee emc-pstc discussion list. Visit our web site at: http://www.ewh.ieee.org/soc/emcs/pstc/ To cancel your subscription, send mail to: majord...@ieee.org with the single line: unsubscribe emc-pstc For help, send mail to the list administrators: Michael Garretson:pstc_ad...@garretson.org Dave Healddavehe...@mediaone.net For policy questions, send mail to: Richard Nute: ri...@ieee.org Jim Bacher: j.bac...@ieee.org All emc-pstc postings are archived and searchable on the web at: http://www.rcic.com/ click on Virtual Conference Hall,
Re: Production Line Test Requirements - Medical Devices
200105212113.oaa14...@epgc196.sdd.hp.com, Rich Nute ri...@sdd.hp.com inimitably wrote: Yes, for one strand. No, for five strands. I agree with what you say. In Europe, the PEC within equipment may be 16/0.2, so the possibility of the connection being reduced to one or two strands is, in one sense, 1 in 560, while for your 36 strand wire it is very much lower. -- Regards, John Woodgate, OOO - Own Opinions Only. Phone +44 (0)1268 747839 Fax +44 (0)1268 777124. http://www.jmwa.demon.co.uk Why not call a vertically- applied manulo-pedally-operated quasi-planar chernozem-penetrating and excavating implement a SPADE? --- This message is from the IEEE EMC Society Product Safety Technical Committee emc-pstc discussion list. Visit our web site at: http://www.ewh.ieee.org/soc/emcs/pstc/ To cancel your subscription, send mail to: majord...@ieee.org with the single line: unsubscribe emc-pstc For help, send mail to the list administrators: Michael Garretson:pstc_ad...@garretson.org Dave Healddavehe...@mediaone.net For policy questions, send mail to: Richard Nute: ri...@ieee.org Jim Bacher: j.bac...@ieee.org All emc-pstc postings are archived and searchable on the web at: http://www.rcic.com/ click on Virtual Conference Hall,
Re: Production Line Test Requirements - Medical Devices
Hi John: Because continuity at low current does not ensure that the protective circuit will carry a large fault current - it might be 'hanging on by one strand'. Yes, for one strand. No, for five strands. Some years ago, I did some experiments on what problems the 25-amp test would detect. I simulated broken strands by cutting them one at a time. With five strands intact, the circuit passed the 25-amp, 2-minute test. It failed at 4 strands and 1 minute. (The tested wire was 18 AWG comprised of 36 strands of 34 AWG.) The ability of a few strands to carry the 25-amp current depends on the free length of the those few strands, which in turn determines the heat-sinking provided to those strands. The free length was on the order of 3 mm. My experiment assumed the problem was caused by an incorrectly set wire stripper, that cut a number of strands. So there was a very small free length of strands. I published this study in the Product Safety Newsletter, Vol. 10, No.1, January-March, 1997. Best regards, Rich --- This message is from the IEEE EMC Society Product Safety Technical Committee emc-pstc discussion list. Visit our web site at: http://www.ewh.ieee.org/soc/emcs/pstc/ To cancel your subscription, send mail to: majord...@ieee.org with the single line: unsubscribe emc-pstc For help, send mail to the list administrators: Michael Garretson:pstc_ad...@garretson.org Dave Healddavehe...@mediaone.net For policy questions, send mail to: Richard Nute: ri...@ieee.org Jim Bacher: j.bac...@ieee.org All emc-pstc postings are archived and searchable on the web at: http://www.rcic.com/ click on Virtual Conference Hall,
Re: Production Line Test Requirements - Medical Devices
36BDBCA75E0FD411A80100104B93ABF202C2C079@MGCMAIL, Dick Grobner dick.grob...@medgraph.com inimitably wrote: Why wouldn't a simple continuity tests as defined by UL be appropriate from equipment coming off of the production line (assuring that the protective earth circuit in intake)? Because continuity at low current does not ensure that the protective circuit will carry a large fault current - it might be 'hanging on by one strand'. -- Regards, John Woodgate, OOO - Own Opinions Only. Phone +44 (0)1268 747839 Fax +44 (0)1268 777124. http://www.jmwa.demon.co.uk Why not call a vertically- applied manulo-pedally-operated quasi-planar chernozem-penetrating and excavating implement a SPADE? --- This message is from the IEEE EMC Society Product Safety Technical Committee emc-pstc discussion list. Visit our web site at: http://www.ewh.ieee.org/soc/emcs/pstc/ To cancel your subscription, send mail to: majord...@ieee.org with the single line: unsubscribe emc-pstc For help, send mail to the list administrators: Michael Garretson:pstc_ad...@garretson.org Dave Healddavehe...@mediaone.net For policy questions, send mail to: Richard Nute: ri...@ieee.org Jim Bacher: j.bac...@ieee.org All emc-pstc postings are archived and searchable on the web at: http://www.rcic.com/ click on Virtual Conference Hall,
RE: Production Line Test Requirements - Medical Devices
Hi Dick, The requirement for using 10 to 25 Amps comes from the test house. It is their mark. If you want to use it, they can require you to do anything they want. Is it fair, no. Does the requirement make engineering sense, no (I believe that Rich Nute did an article on how this requirement did not identify anymore failures than the simple ohm meter did). Do you still have to do it, yes! If you are just declaring conformity for the MDD, use your risk analysis (i.e. EN 1441) to show that the risk of using a ohm meter is an acceptable one. Your NB might review it, but you should be able to switch to the ohm meter. Ned Ned Devine Program Manager III Entela, Inc. 3033 Madison Ave. SE Grand Rapids, MI 49548 616 248 9671 Phone 616 574 9752 Fax ndev...@entela.com e-mail --- This message is from the IEEE EMC Society Product Safety Technical Committee emc-pstc discussion list. Visit our web site at: http://www.ewh.ieee.org/soc/emcs/pstc/ To cancel your subscription, send mail to: majord...@ieee.org with the single line: unsubscribe emc-pstc For help, send mail to the list administrators: Michael Garretson:pstc_ad...@garretson.org Dave Healddavehe...@mediaone.net For policy questions, send mail to: Richard Nute: ri...@ieee.org Jim Bacher: j.bac...@ieee.org All emc-pstc postings are archived and searchable on the web at: http://www.rcic.com/ click on Virtual Conference Hall,
RE: Production Line Test Requirements - Medical Devices
Jon Thx for the info, however, I re-checked the two known US standards (UL2601, Appendix D - Manufacturer's Responsibilities, Construction Considerations and Requirements for Factory Tests) (NFPA 99, Grounding Circuit Continuity - Measurement of Resistance). UL defines Production Line Grounding Continuity Test Equipment as: Any suitable continuity indication device (such as an ohmmeter, a battery and buzzer combination, or the like) may be used to determine compliance with Grounding Continuity Test requirements. The NFPA std is even less definitive on this matter. EN60601-1, Appendix B - Testing During Manufacture and/or Installation, Not Used. See rational to sub-clause 4.1. Which states - Tests described in this standard are type tests. I checked all of the amendments including EN60601-1 for medical systems, and I do not see anything changing this. My questions remains, where does it state to use the 10-25Amp, 50 or 60Hz @ 6Volt for 5 seconds test? Is this a carry over from another EN standard? Why wouldn't a simple continuity tests as defined by UL be appropriate from equipment coming off of the production line (assuring that the protective earth circuit in intake)? Appreciate your feedback Jon! thanks again. Does anyone else on this forum have any input - would appreciate it. -Original Message- From: Jon Griver [mailto:jo...@medson.com] Sent: Monday, May 21, 2001 9:16 AM To: Dick Grobner Subject: RE: Production Line Test Requirements - Medical Devices Dick, The requirements for production tests of medical equipment generally are: 1. Hipot at 1500AC - between live and neutral connected together and earth. 2. Earth continuity at 25A between earth pin on plug and enclosure (choose a point that may be problematic if there are manufacturing problems) 3. Earth and/or patient leakage. Whether you do both, one or neither of these tests depends on the product. As the production tests are intended to find production problems (as opposed to type tests which are intended to find design problems), choose the tests that you think make sense from the production engineering point of view. Document the reasoning behind your choice. IT equipment only requires Hipot and earth continuity tests. Hope this helps, Jon -Original Message- From: Dick Grobner [mailto:dick.grob...@medgraph.com] Sent: Monday, May 21, 2001 16:36 To: 'Jon Griver' Subject: RE: Production Line Test Requirements - Medical Devices Jon Thanks, you have confirmed what I thought I knew. We just had our annual ISO 9001 audit (2 weeks ago). One issue that one of the two auditors raised was why are you not doing the ground integrity test (25 Amp test) on 100% of your production units? I asked do you mean ground continuity (Ohm test) and he said no, the ground integrity test. That's when I started digging into the standards. UL (and the former ETL) states high potential and ground continuity tests on all production units. EN60601-1 states that all tests within this document are type tests. The auditor gave me no reference to any EN, etc. when I asked. So, I have a suspicion that this is his wish and not stated in any EN (at least that I know of so far). SO - I will continue to pursue with other outside sources (but not this one auditor!) Thanks for the reply back! PS - does ITE equipment require this test (ground integrity ((25 Amp)) test? -Original Message- From: Jon Griver [mailto:jo...@medson.com] Sent: Sunday, May 20, 2001 1:34 AM To: emc-p...@majordomo.ieee.org Subject: FW: Production Line Test Requirements - Medical Devices Dick, To the best of my knowledge there is no EN standard or guidance on production tests for medical equipment (There is a standard, EN 50116, for IT equipment). Your best bet is to discuss this with your Notified Body and to come to a mutually acceptable set of tests. Regards, Jon Griver Medson Ltd. Good Day Everyone Question I have - Does anyone out there know if an EN standard or guidance document exists that deals with production line test requirements of finished medical devices. Reading in EN60601-1 it states that the test (ground resistance, high potential, etc.) are type test only (See appendix B, and then paragraph 4.1). Thus - they are not identified as production line tests. I know that the NRTL's in the USA (UL/ETL) specifically call out what tests are to be performed after the production build of equipment. Any input would be appreciated. Thx Dick Grobner Medical Graphics Corporation 350 Oak Grove Parkway St Paul MN 55127 651-766-3395 651-766-3389 (fax) dick.grob...@medgraph.com --- This message is from the IEEE EMC Society Product Safety Technical Committee emc-pstc discussion list. Visit our web site at: http://www.ewh.ieee.org/soc/emcs/pstc/ To cancel your subscription, send mail to: majord...@ieee.org
FW: Production Line Test Requirements - Medical Devices
Dick, To the best of my knowledge there is no EN standard or guidance on production tests for medical equipment (There is a standard, EN 50116, for IT equipment). Your best bet is to discuss this with your Notified Body and to come to a mutually acceptable set of tests. Regards, Jon Griver Medson Ltd. Good Day Everyone Question I have - Does anyone out there know if an EN standard or guidance document exists that deals with production line test requirements of finished medical devices. Reading in EN60601-1 it states that the test (ground resistance, high potential, etc.) are type test only (See appendix B, and then paragraph 4.1). Thus - they are not identified as production line tests. I know that the NRTL's in the USA (UL/ETL) specifically call out what tests are to be performed after the production build of equipment. Any input would be appreciated. Thx Dick Grobner Medical Graphics Corporation 350 Oak Grove Parkway St Paul MN 55127 651-766-3395 651-766-3389 (fax) dick.grob...@medgraph.com --- This message is from the IEEE EMC Society Product Safety Technical Committee emc-pstc discussion list. Visit our web site at: http://www.ewh.ieee.org/soc/emcs/pstc/ To cancel your subscription, send mail to: majord...@ieee.org with the single line: unsubscribe emc-pstc For help, send mail to the list administrators: Michael Garretson:pstc_ad...@garretson.org Dave Healddavehe...@mediaone.net For policy questions, send mail to: Richard Nute: ri...@ieee.org Jim Bacher: j.bac...@ieee.org All emc-pstc postings are archived and searchable on the web at: http://www.rcic.com/ click on Virtual Conference Hall, --- This message is from the IEEE EMC Society Product Safety Technical Committee emc-pstc discussion list. Visit our web site at: http://www.ewh.ieee.org/soc/emcs/pstc/ To cancel your subscription, send mail to: majord...@ieee.org with the single line: unsubscribe emc-pstc For help, send mail to the list administrators: Michael Garretson:pstc_ad...@garretson.org Dave Healddavehe...@mediaone.net For policy questions, send mail to: Richard Nute: ri...@ieee.org Jim Bacher: j.bac...@ieee.org All emc-pstc postings are archived and searchable on the web at: http://www.rcic.com/ click on Virtual Conference Hall,
Production Line Test Requirements - Medical Devices
Good Day Everyone Question I have - Does anyone out there know if an EN standard or guidance document exists that deals with production line test requirements of finished medical devices. Reading in EN60601-1 it states that the test (ground resistance, high potential, etc.) are type test only (See appendix B, and then paragraph 4.1). Thus - they are not identified as production line tests. I know that the NRTL's in the USA (UL/ETL) specifically call out what tests are to be performed after the production build of equipment. Any input would be appreciated. Thx Dick Grobner Medical Graphics Corporation 350 Oak Grove Parkway St Paul MN 55127 651-766-3395 651-766-3389 (fax) dick.grob...@medgraph.com --- This message is from the IEEE EMC Society Product Safety Technical Committee emc-pstc discussion list. Visit our web site at: http://www.ewh.ieee.org/soc/emcs/pstc/ To cancel your subscription, send mail to: majord...@ieee.org with the single line: unsubscribe emc-pstc For help, send mail to the list administrators: Michael Garretson:pstc_ad...@garretson.org Dave Healddavehe...@mediaone.net For policy questions, send mail to: Richard Nute: ri...@ieee.org Jim Bacher: j.bac...@ieee.org All emc-pstc postings are archived and searchable on the web at: http://www.rcic.com/ click on Virtual Conference Hall,
FW: Medical devices
The device will need CE, thus the typical route is a tech file and a DoC. Refer to the Medical Device Directive (MDD) or the In-Vitro directive. From the little info you provided I would guess the Medical Directive is the one. Reading the scope of both should assist you in selecting the appropriate directive. Annex I in the directive provides the essential requirements you need to be compliant with. You also need to (first thing) classify your device, article 9 of the MDD will assist here. Article 11 of the MDD addresses conformity assessment procedure, this identifies which of the other Annexes you may follow to allow CE marking of the device. This is somewhat condensed, best advise - obtain the directives, read through them (more than once is highly recommended) and comprehend the contents. Good Luck! If you wish you can contact me directly with any other questions, I may have the right answer! -Original Message- From: k...@i-data.com [mailto:k...@i-data.com] Sent: Wednesday, January 17, 2001 2:13 AM To: emc-p...@majordomo.ieee.org Subject: Medical devices Hi all Can some one give me links to information regarding EU Medical Devices directive. My problem is to define the requirement for the following situation: A manufacturer is making an add-on device for another company's medical device and the total equipment is CE approved without involving the manufacturer. Now the manufacturer want to CE mark this add-on device and sell it to other companies as a CE approved device still for medical equipment. What is the requirements for this manufacturer, they have never been into CE markin before. The equipment is used in laboratoriums to move small glases with different materials for testing in another test device. best regards, Kim Jensen --- This message is from the IEEE EMC Society Product Safety Technical Committee emc-pstc discussion list. To cancel your subscription, send mail to: majord...@ieee.org with the single line: unsubscribe emc-pstc For help, send mail to the list administrators: Jim Bacher: jim_bac...@mail.monarch.com Michael Garretson:pstc_ad...@garretson.org For policy questions, send mail to: Richard Nute: ri...@ieee.org --- This message is from the IEEE EMC Society Product Safety Technical Committee emc-pstc discussion list. To cancel your subscription, send mail to: majord...@ieee.org with the single line: unsubscribe emc-pstc For help, send mail to the list administrators: Jim Bacher: jim_bac...@mail.monarch.com Michael Garretson:pstc_ad...@garretson.org For policy questions, send mail to: Richard Nute: ri...@ieee.org
FW: Medical devices
Kim, Check out http://europa.eu.int/comm/enterprise/newapproach/standardization/harmstds/re flist.html From that page you can get to both the In Vitro Diagnostic Medical Devices Directive and the Medical Devices Directive, as well as the Harmonised Standards for these Directives. It sounds like your product falls under the In Vitro Diagnostic Directive, but check the scope. As an accessory, your product still has to fully comply with the directive. Best Wishes, Jon Griver Medson Ltd. VP Quality Assurance email: jo...@medson.com Hi all Can some one give me links to information regarding EU Medical Devices directive. My problem is to define the requirement for the following situation: A manufacturer is making an add-on device for another company's medical device and the total equipment is CE approved without involving the manufacturer. Now the manufacturer want to CE mark this add-on device and sell it to other companies as a CE approved device still for medical equipment. What is the requirements for this manufacturer, they have never been into CE markin before. The equipment is used in laboratoriums to move small glases with different materials for testing in another test device. best regards, Kim Jensen --- This message is from the IEEE EMC Society Product Safety Technical Committee emc-pstc discussion list. To cancel your subscription, send mail to: majord...@ieee.org with the single line: unsubscribe emc-pstc For help, send mail to the list administrators: Jim Bacher: jim_bac...@mail.monarch.com Michael Garretson:pstc_ad...@garretson.org For policy questions, send mail to: Richard Nute: ri...@ieee.org
Medical devices
Hi all Can some one give me links to information regarding EU Medical Devices directive. My problem is to define the requirement for the following situation: A manufacturer is making an add-on device for another company's medical device and the total equipment is CE approved without involving the manufacturer. Now the manufacturer want to CE mark this add-on device and sell it to other companies as a CE approved device still for medical equipment. What is the requirements for this manufacturer, they have never been into CE markin before. The equipment is used in laboratoriums to move small glases with different materials for testing in another test device. best regards, Kim Jensen --- This message is from the IEEE EMC Society Product Safety Technical Committee emc-pstc discussion list. To cancel your subscription, send mail to: majord...@ieee.org with the single line: unsubscribe emc-pstc For help, send mail to the list administrators: Jim Bacher: jim_bac...@mail.monarch.com Michael Garretson:pstc_ad...@garretson.org For policy questions, send mail to: Richard Nute: ri...@ieee.org
RE: EMC on Medical Devices?
David, 1. For the EU, the EMC standard is EN 60601-1-2. This standard calls out CISPR 11 for radiated emissions, and various obsolete IEC 801 standards for immunity. In addition, check out Clause 36 of the particular safety standard EN 60601-2-XX (if there is one) for your specific product type. For medical equipment, EMC is considered a safety issue. 2. For the US, medical equipment is exempted from FCC Part 15. However medical equipment which uses EM radiation (e.g. microwave therapy equipment), and ultrasound equipment, is subject to FCC Part 18. In addition, the FDA may require immunity testing, depending on the device type, as part of the regulatory approval process. 3. I do not know the situation for Asia Pacific. It is likely that the technical requirements will not be different from the EU and US. However, the regulatory process of demonstrating compliance may differ from country to country. Hope this helps. Jon Griver Medson Ltd. VP Quality Assurance Subject: EMC on Medical Devices? Looking for a list of standards that apply to EMC on medical devices and the markets which they apply ( EU, US, and Asia Pacific). Thanks Regards, David Spencer --- This message is from the IEEE EMC Society Product Safety Technical Committee emc-pstc discussion list. To cancel your subscription, send mail to: majord...@ieee.org with the single line: unsubscribe emc-pstc For help, send mail to the list administrators: Jim Bacher: jim_bac...@mail.monarch.com Michael Garretson:pstc_ad...@garretson.org For policy questions, send mail to: Richard Nute: ri...@ieee.org
EMC on Medical Devices?
Looking for a list of standards that apply to EMC on medical devices and the markets which they apply ( EU, US, and Asia Pacific). Thanks Regards, David Spencer --- This message is from the IEEE EMC Society Product Safety Technical Committee emc-pstc discussion list. To cancel your subscription, send mail to: majord...@ieee.org with the single line: unsubscribe emc-pstc For help, send mail to the list administrators: Jim Bacher: jim_bac...@mail.monarch.com Michael Garretson:pstc_ad...@garretson.org For policy questions, send mail to: Richard Nute: ri...@ieee.org
RE: EMC for medical devices
To further clarify: IEC 60601-1-2 is the EMC requirements. For reference, the CDV of the second edition has just been sent to the full committee for voting. If all goes well, we will see it become official this calendar year. Best regards, Brent DeWitt Datex-Ohmeda IEC SC62A WG-13 member -Original Message- From: owner-emc-p...@ieee.org [mailto:owner-emc-p...@ieee.org]On Behalf Of Chris Duprés Sent: Wednesday, January 26, 2000 3:51 PM To: Kevin Newland Cc: emc-pstc Subject: EMC for medical devices Hi Kevin. You asked: 1-What are the immunity standards used in Europe for medical devices? I think you'll find that the IEC601 covers all the EMC requirements as well. Sorry, can't help with US requirements. Chris Dupres Surrey, UK. - This message is coming from the emc-pstc discussion list. To cancel your subscription, send mail to majord...@ieee.org with the single line: unsubscribe emc-pstc (without the quotes). For help, send mail to ed.pr...@cubic.com, jim_bac...@monarch.com, ri...@sdd.hp.com, or roger.volgst...@compaq.com (the list administrators). - This message is coming from the emc-pstc discussion list. To cancel your subscription, send mail to majord...@ieee.org with the single line: unsubscribe emc-pstc (without the quotes). For help, send mail to ed.pr...@cubic.com, jim_bac...@monarch.com, ri...@sdd.hp.com, or roger.volgst...@compaq.com (the list administrators).
EMC for medical devices
Hi Kevin. You asked: 1-What are the immunity standards used in Europe for medical devices? I think you'll find that the IEC601 covers all the EMC requirements as well. Sorry, can't help with US requirements. Chris Dupres Surrey, UK. - This message is coming from the emc-pstc discussion list. To cancel your subscription, send mail to majord...@ieee.org with the single line: unsubscribe emc-pstc (without the quotes). For help, send mail to ed.pr...@cubic.com, jim_bac...@monarch.com, ri...@sdd.hp.com, or roger.volgst...@compaq.com (the list administrators).
FW: EMC for medical devices
-Original Message- From: Dick Grobner Sent: Wednesday, January 26, 2000 11:19 AM To: 'Kevin Newland' Subject: RE: EMC for medical devices Europe is EN 60601-1-2 In 1996 - for the US - it was the FDA's draft document entitled Reviewers Guidance for Premarket Notification Submissions which was in line with the requirements of 60601-1-2. However it does include a few additional immunity tests which we have successfully waved with the FDA on past submissions. I'm not aware of any other US requirements regarding EMC medical equip. -Original Message- From: Kevin Newland [mailto:kevin_newl...@yahoo.com] Sent: Tuesday, January 25, 2000 4:09 PM To: emc-p...@majordomo.ieee.org Subject: EMC for medical devices Hello All, Could someone let me know the answer to the following questions please: 1-What are the immunity standards used in Europe for medical devices? 2-What EMC standards is used in US/Canada for medical devices? The product in question is approved for safety to IEC601-1. Thank you Kevin __ Do You Yahoo!? Talk to your friends online with Yahoo! Messenger. http://im.yahoo.com - This message is coming from the emc-pstc discussion list. To cancel your subscription, send mail to majord...@ieee.org with the single line: unsubscribe emc-pstc (without the quotes). For help, send mail to ed.pr...@cubic.com, jim_bac...@monarch.com, ri...@sdd.hp.com, or roger.volgst...@compaq.com (the list administrators). - This message is coming from the emc-pstc discussion list. To cancel your subscription, send mail to majord...@ieee.org with the single line: unsubscribe emc-pstc (without the quotes). For help, send mail to ed.pr...@cubic.com, jim_bac...@monarch.com, ri...@sdd.hp.com, or roger.volgst...@compaq.com (the list administrators).
EMC for medical devices
Hello All, Could someone let me know the answer to the following questions please: 1-What are the immunity standards used in Europe for medical devices? 2-What EMC standards is used in US/Canada for medical devices? The product in question is approved for safety to IEC601-1. Thank you Kevin __ Do You Yahoo!? Talk to your friends online with Yahoo! Messenger. http://im.yahoo.com - This message is coming from the emc-pstc discussion list. To cancel your subscription, send mail to majord...@ieee.org with the single line: unsubscribe emc-pstc (without the quotes). For help, send mail to ed.pr...@cubic.com, jim_bac...@monarch.com, ri...@sdd.hp.com, or roger.volgst...@compaq.com (the list administrators).
Medical devices lab in Italy
Hi Everyone I have a client that wishes to have a Class I medical device tested in Italy, since our own lab can not assist at present. Can anybody suggest a lab or two that could possibly help for testing under the MDD ? Thanks, Lizette de Vries-Venter SENIOR ENGINEER Division of Electronics and Appliances SABS Tel : +27 12 428 6990 Fax : +27 12 428 6523 Email : dvve...@sabs.co.za - This message is coming from the emc-pstc discussion list. To cancel your subscription, send mail to majord...@ieee.org with the single line: unsubscribe emc-pstc (without the quotes). For help, send mail to ed.pr...@cubic.com, j...@gwmail.monarch.com, ri...@sdd.hp.com, or roger.volgst...@compaq.com (the list administrators).
Brazil approval of medical devices
Does anybody out of there know which are the steps to be taken in order to get the approval for a medical device to be sold in Brazil? Do they recognize international standards as IEC 60601-1 and family? What is the relevant law? Which are the Notified Bodies for medical devices? Thanks for any help. m.p. - ESAOTE S.p.A. Massimo Polignano Research Product Development Design Quality Control Via di Caciolle,15 tel:+39.055.4229402 I- 50127 Florence fax:+39.055.4223305 e-mail: regr...@esaote.com - This message is coming from the emc-pstc discussion list. To cancel your subscription, send mail to majord...@ieee.org with the single line: unsubscribe emc-pstc (without the quotes). For help, send mail to ed.pr...@cubic.com, j...@gwmail.monarch.com, ri...@sdd.hp.com, or roger.volgst...@compaq.com (the list administrators).
RE: Medical Devices in Hyperbaric Chamber
At 12:16 PM 11/24/98 -0800, ed.pr...@cubic.com wrote: Massimo: Further complicating the issue is that the Apollo 1 fire ocurred with a partial pressure O2 rich environment. My guess(!) is that a 1ATM O2 rich environment would be even more hazardous. Maybe you can find some hints on this via NASA or a hyperbaric chamber manufacturer? Just for the record, Apollo 1 had a pure O2 atmosphere. The disaster halted that practice. And I believe medical oxygen tanks run about 40% O2. I remember during dive training we were told medical O2 tanks weren't 100% O2. - This message is coming from the emc-pstc discussion list. To cancel your subscription, send mail to majord...@ieee.org with the single line: unsubscribe emc-pstc (without the quotes). For help, send mail to ed.pr...@cubic.com, j...@gwmail.monarch.com, ri...@sdd.hp.com, or roger.volgst...@compaq.com (the list administrators).
Re: Medical Devices in Hyperbaric Chamber
1) In a high-pressure atmosphere, convection cooling of power devices should be OK. 2) Corona/voltage breakdown through the air would happen at a lower voltage level. These are based on the idea that there's little convection/corona in vacuum environments. As others have indicated, the gas in the chamber could play an important part of the product safety as well. -- Patrick Lawler plaw...@west.net - This message is coming from the emc-pstc discussion list. To cancel your subscription, send mail to majord...@ieee.org with the single line: unsubscribe emc-pstc (without the quotes). For help, send mail to ed.pr...@cubic.com, j...@gwmail.monarch.com, ri...@sdd.hp.com, or roger.volgst...@compaq.com (the list administrators).
Medical Devices in Hyperbaric Chamber
A distributors of ours asks us about the suitability of our ECG device for use into a Hyperbaric Chamber. Does anybody know any reference safety standard for medical devices in hyperbaric chamber? Are the environmental conditions similar to those in presence of a flammable anaesthetic mixture with air or with oxygen or nitrous oxide (IEC 601-1)? Thanks for any help. m.p. --- ESAOTE S.p.A. Massimo Polignano Research Product Development Regulatory Affairs Via di Caciolle,15 tel:+39.055.4229402 I- 50127 Florencefax:+39.055.4223305 e-mail: regr...@esaote.com - This message is coming from the emc-pstc discussion list. To cancel your subscription, send mail to majord...@ieee.org with the single line: unsubscribe emc-pstc (without the quotes). For help, send mail to ed.pr...@cubic.com, j...@gwmail.monarch.com, ri...@sdd.hp.com, or roger.volgst...@compaq.com (the list administrators).
RE: Medical Devices in Hyperbaric Chamber
Hello Massimo, The use of devices in oxygen enriched environments is not a trivial question. Many materials that are fire retardant at 21% O2 are highly combustible in enriched environments.If your product has not been specifically designed and evaluated for use in such environments, one would be taking a great risk in using them there. Every year there are fires reported in hyperbaric chambers. Remember the Apollo One fire in 1967? IEC 60601-1 (second edition) does not have appropriate requirements for oxygen enriched environments. Work on the third edition is planned to address this issue. There is an excellent article on oxygen environments in the current issue of the ASTM Standardization News (November 1998) and the work of ASTM Committee G-4, Compatibility and Sensitivity of Materials in Oxygen-Enriched Atmospheres. You can contact ASTM at http://www.astm.org Best regards, Dave Osborn Hewlett-Packard Company Medical Products Group dosb...@hp.com ___ I am currently testing a new e-mail client called Outlook. If present, please disregard the winmail.dat attachment. ___ -Original Message- From: Non-HP-regrsfi /HP-ColSprings,mimegw3/dd.HPMEXT1=regr...@esaote.com Sent: Tuesday, November 24, 1998 3:24 AM To: Non-HP-emc-pstc /HP-ColSprings,mimegw3/dd.HPMEXT1=emc-p...@majordomo.ieee.org Cc: Non-HP-regrsfi /HP-ColSprings,mimegw3/dd.HPMEXT1=regr...@esaote.com Subject: Medical Devices in Hyperbaric Chamber A distributors of ours asks us about the suitability of our ECG device for use into a Hyperbaric Chamber. Does anybody know any reference safety standard for medical devices in hyperbaric chamber? Are the environmental conditions similar to those in presence of a flammable anaesthetic mixture with air or with oxygen or nitrous oxide (IEC 601-1)? Thanks for any help. m.p. --- ESAOTE S.p.A. Massimo Polignano Research Product Development Regulatory Affairs Via di Caciolle,15 tel:+39.055.4229402 I- 50127 Florencefax:+39.055.4223305 e-mail: regr...@esaote.com - This message is coming from the emc-pstc discussion list. To cancel your subscription, send mail to majord...@ieee.org with the single line: unsubscribe emc-pstc (without the quotes). For help, send mail to ed.pr...@cubic.com, j...@gwmail.monarch.com, ri...@sdd.hp.com, or roger.volgst...@compaq.com (the list administrators). File: Medical Devices in Hyperbaric Chamber.TXT
RE: Medical Devices in Hyperbaric Chamber
Massimo: Further complicating the issue is that the Apollo 1 fire ocurred with a partial pressure O2 rich environment. My guess(!) is that a 1ATM O2 rich environment would be even more hazardous. Maybe you can find some hints on this via NASA or a hyperbaric chamber manufacturer? Ed From: dave_osb...@hp-waltham-om3.om.hp.com Subject: RE: Medical Devices in Hyperbaric Chamber Date: Tue, 24 Nov 1998 11:13:10 -0500 To: regr...@esaote.com Cc: emc-p...@ieee.org Hello Massimo, The use of devices in oxygen enriched environments is not a trivial question. Many materials that are fire retardant at 21% O2 are highly combustible in enriched environments.If your product has not been specifically designed and evaluated for use in such environments, one would be taking a great risk in using them there. Every year there are fires reported in hyperbaric chambers. Remember the Apollo One fire in 1967? IEC 60601-1 (second edition) does not have appropriate requirements for oxygen enriched environments. Work on the third edition is planned to address this issue. There is an excellent article on oxygen environments in the current issue of the ASTM Standardization News (November 1998) and the work of ASTM Committee G-4, Compatibility and Sensitivity of Materials in Oxygen-Enriched Atmospheres. You can contact ASTM at http://www.astm.org Best regards, Dave Osborn Hewlett-Packard Company Medical Products Group dosb...@hp.com -- Ed Price ed.pr...@cubic.com Electromagnetic Compatibility Lab Cubic Defense Systems San Diego, CA. USA 619-505-2780 List-Post: emc-pstc@listserv.ieee.org Date: 11/24/1998 Time: 12:16:35 -- - This message is coming from the emc-pstc discussion list. To cancel your subscription, send mail to majord...@ieee.org with the single line: unsubscribe emc-pstc (without the quotes). For help, send mail to ed.pr...@cubic.com, j...@gwmail.monarch.com, ri...@sdd.hp.com, or roger.volgst...@compaq.com (the list administrators).