Having done field labelling and listings for an NRTL, I would ask your Customer what they need. It could be an internal company requirement for NRTL listing or field labelling. Not necessarily the local jurisdiction requiring it. However, if you really want to know what is locally required, call

Re: [PSES] Harmonised Standards for EMC Directive 2014/30/EU and Low Voltage Directive 2014/35/EU

2015-03-02 Thread Ronald R. Wellman
Not my wording. I recommend you wait until further information is available from the website to make a decision on how to declare to the newer Directives.   When you say, applicable from 20 April 2016 does that the date in which we can start declaring to

Re: [PSES] Mandatory NRTL certification

2012-01-09 Thread Ronald R. Wellman
Hello Rich,   My point exactly about what products are targeted. We have been through these discussions many times in the past and the answer always comes up that your product specific market environment drives third party certification.   Best

RE: DOC Requirement for Year CE Mark is Affixed

2008-06-03 Thread Ronald R. Wellman
According to Annex III, there is no mention as to a date that the DoC is signed by the signatory. Therefore, is the the last two digits of the year in which the CE marking was affixed (for the first time) the same as the date a signatory signed a DoC, if it appears on the DoC? Best regards, Ron

RE: EMC immunity requirements for medical device in Canada?

2008-05-29 Thread Ronald R. Wellman
The following web page lists recognized standards: http://www.hc-sc.gc.ca/dhp-mps/md-im/st ndards-normes/md_rec_stand_im_norm_lst_2_e.html where the following apply to EMC: IEC 60601-1-2 Ed. 2.0, Ed. 2.1 : 2004-11-01 Medical electrical equipment – Part 1-2: General requirements for safety –

RE: EMC immunity requirements for medical device in Canada?

2008-05-29 Thread Ronald R. Wellman
4-6: Testing and measurement techniques - Immunity to conducted disturbances, induced by radio-frequency fields Ron Wellman -Original Message- From: Ronald R. Wellman [mailto:rwell...@wellman.com] Sent: Thursday, May 29, 2008 5:44 AM

RE: Ethics grad work Therac-25

2008-05-12 Thread Ronald R. Wellman
People who are intererested in how management responsibility is viewed by the U.S. government might want to read up on the U.S. Supreme Court ruling of the United States versus Park, 421 U.S. 658 (1975). This is where an executive was found guilty under strict liability for selling unsanitary food

RE: European conductor ampacity table

2003-10-29 Thread Ronald R. Wellman
Hello Sam, I believe that IEC 60335-1 has the information you are looking for. Best regards, Ron Wellman At 02:14 PM 10/29/2003 -0700, POWELL, DOUG wrote: Sam, IEC 60950 (EN 60950) has: Table 3D - Range of conductor sizes to be accepted by terminals Table 3E - Sizes of terminals for mains

RE: Denmark AC outlets

2003-10-28 Thread Ronald R. Wellman
Hello Ian, Be careful in using this web site. It is not up-to-date for some countries, especially Brazil and Thailand. They also specify an Australian plug for Argentina, which is incorrect. Best regards, Ron Wellman At 08:34 AM 10/28/2003 +, Gordon,Ian wrote: All I have found the

RE: IEC60950 - Language of user handbook

2003-10-27 Thread Ronald R. Wellman
Hello Amund, Rather than looking at this from what is required, what about what makes sense? I wouldn't expect someone in Norway to be very pleased to only get documentation in Japanese or Chinese. The end-user would most likely be very dissatisfied and probably return the product. Companies

Re: Chemicals in test equipment

2003-10-23 Thread Ronald R. Wellman
Hello Kim, Depending on whether the chemicals used are carcinogens or reproductive toxicants there are a couple of jurisdictions in the United States that regulate exposure to these chemicals. There is the Environmental Protection Agency (EPA) and the State of California's Safe Drinking Water

Re: Denmark AC outlets

2003-10-22 Thread Ronald R. Wellman
Hello Lou, You mention that In practice, everyone I know sends Schuko plugs to Denmark. Well, that is not necessarily true for manufacturers of test and measurement, and laboratory equipment. I have always specified the Danish plugs for Denmark without any problems for TM equipment. In fact,

Re: Denmark AC outlets

2003-10-22 Thread Ronald R. Wellman
Hello Richard, I would not want a Class 1 product installed without a protective earth ground. I know that we have to continuously deal with installations that do not provide earth grounding but that's the way things are. It is best to ship products with the appropriate power cord for its

RE: centrifuges and the MDD

2003-10-15 Thread Ronald R. Wellman
Hello Paul, I'd say you are okay with just the LVD for general laboratory equipment. This issue of intended use is very similar to other laboratory equipment used to analyze blood and urine. However, I would make a statement that the centrifuges are intended for General Laboratory Use in your

RE: centrifuges and the MDD

2003-10-15 Thread Ronald R. Wellman
Hello Ned, Here is a copy of Rule 3: --- 1.3. Rule 3 All non-invasive devices intended for modifying the biological or chemical composition of blood, other body liquids or other liquids intended for infusion into the body are in Class llb, unless the treatment consists of filtration,

Re: centrifuges and the MDD

2003-10-14 Thread Ronald R. Wellman
Hello Paul, EN 61010-2-020: Safety requirements for electrical equipment for measurement, control, and laboratory use - Particular requirements for laboratory centrifuges, is covered under the LVD and not the MDD. Best regards, Ron Wellman At 05:11 PM 10/14/2003 -0400, Smith, Paul J. wrote:

Re: UL CSA standards date of expiry

2003-08-22 Thread Ronald R. Wellman
Hello Chris, To get the information you want, you need to get the appropriate Announcement Bulletins from UL and Information Letters from CSA for the standards in question. I suggest you contact UL and CSA directly to get this information because it is normally available as part of the

Re: Australian AS/NZS 3123 and IEC 309 Industrial Plug Differences

2003-08-02 Thread Ronald R. Wellman
Hello Peter, I suggest that you visit the Standards Australia web site and read the scope of AS/NZS 3123 at: http://www.standards.com.au/catalogue/script/Details.asp?DocN=AS670712832076 I believe that Note # 7 answers your question. Best regards, Ron Wellman At 02:31 PM 8/1/2003 -0700,

Re: LED lamps

2003-07-24 Thread Ronald R. Wellman
Hello Ian, You have nothing to worry about in the United States. LEDs are not considered laser devices in the USA. However, depending on who the manufacturer of the LEDs is, I would verify with them what information they have available on standards compliance with their LEDs. Best regards,

Scope of the LVD

2003-06-28 Thread Ronald R. Wellman
Hello all, The following document may be dated, but it provides incite into the applicability of the LVD: http://www.wbc-europe.com/prodserv/market_segment/low_voltaged73_23.pdf Best regards, Ron Wellman This message is from the IEEE EMC Society Product Safety Technical Committee emc-pstc

Re: CE without LVD?

2003-06-27 Thread Ronald R. Wellman
Hello David, Your issue is not new. I have been in a similar situation where the same argument was made for a product powered by SELV. However, EN 60950 addresses other safety issues with products other than electrically caused fire and electrical shock. One needs to also address the

Re: Power cord Requirement in Latin America

2003-05-09 Thread Ronald R. Wellman
Hello Kevin, What information are you looking for? Nominal Voltages and Frequency or certification requirements? From the country list you provide, only two of them require approval of power cords: Argentina and most recently Brazil. I suggest that you contact Feller on this because they are

RE: FDA requirements for beauty treatment device

2003-04-11 Thread Ronald R. Wellman
Hello Peter, This may be similar to the product in question: http://www.alliedhealth.net/micro_ablation.htm Best regards, Ron Wellman At 10:12 AM 4/9/2003 -0700, Peter L. Tarver wrote: Nick - this sounds like one of those electric skin/muscle toner devices. While I didn't find specific

Re: FDA requirements for beauty treatment device

2003-04-10 Thread Ronald R. Wellman
Hello Nick, I'd be curious as to what claims the manufacturer makes for this beauty treatment product. That may determine one thing in regards to potential misbranding under FDA regulations but if there are similar devices already on the market in the USA, I would check to see how they are

Re: Non-Compliant Products

2003-04-02 Thread Ronald R. Wellman
Hello Enci, Even though EMC compliance is regulated around the world the majority of times that people are found to be non-compliant is from complaints. In most cases it is competitors blowing the whistle on each other. However, as a product safety professional I am usually not concerned with

Re: California NRTL requirements.

2003-03-25 Thread Ronald R. Wellman
Hello Donald, I assume you are more interested in the City of San Bernadino rather than the county. However, based on your message I cannot tell if you are stating that the City of San Bernadino requires NRTL listing or not. If they do, and they do not have a localized testing service like

Re: Class 2 laser

2003-03-03 Thread Ronald R. Wellman
Hello Richard, Have you taken a look at 21 CFR 1040? The FDA/CDRH has certain compliance requirements for laser systems and laser products. Best regards, Ron Wellman At 03:16 PM 3/3/2003 -0500, richwo...@tycoint.com wrote: Are any national restrictions on the use of a Class 2 laser in areas

Re: (Fwd) RE: IEC 61010 requirements

2003-02-21 Thread Ronald R. Wellman
Hello Jon, Thanks for the clarification. This helps in understanding which safety standards may be applied to a medical device based on its intended use. Therefore, it is very important that we understand what a medical device is versus medical equipment or laboratory equipment, as they are

Re: IEC 61010 requirements

2003-02-20 Thread Ronald R. Wellman
-0800, Mike wrote: My opinion is that unless there are unusual circumstances, an ITE (IEC60950) certified monitor should satisfy ordinary lab needs. Any dissent? Mike Harris/Teccom Co./Hayward CA - Original Message - From: Ronald R. Wellman rwell...@wellman.com To: Han, Delphina d

RE: IEC 61010 requirements

2003-02-20 Thread Ronald R. Wellman
Hello Greg, I have to disagree with your interpretation. There are many in vitro devices that are classified as medical equipment and are subject to pre-market approvals in the US, Japan, and EU. There are strict regulations on labeling products in the US where even advertising is considered

Re: IEC 61010 requirements

2003-02-19 Thread Ronald R. Wellman
Hello Delphina, Laboratory equipment can be considered medical devices if you advertise them as such or claim that they can be used be physicians to form medical opinions. However, most laboratory equipment manufacturers get around this by claiming that their products are intended for

Re: mouse heart monitor

2003-02-05 Thread Ronald R. Wellman
Hello Sam, I am not familiar with EU requirements for veterinary medical devices. However, you may want to see what the FDA has to say about them at: http://www.fda.gov/cvm/index/consumer/regofdevices.htm Best regards, Ron Wellman At 04:01 PM 2/5/2003 -0600, Sam Davis wrote: All, I'm trying

Re: HP 6034L Power supply schematics / repair (was Agilent 6842A)

2003-02-05 Thread Ronald R. Wellman
Hello all; I find it inappropriate to bad mouth any company in this forum. Regardless of who you work or worked for. This forum exists for the exchange of information related to our profession as EMC and Safety engineers. If you prefer to continue this thread, do it privately. Regards, Ron

Re: FCC Importation Issue

2003-02-01 Thread Ronald R. Wellman
This message has been converted via the evaluation version of Transend Migrator. Use beyond the trial period specified in your Software Evaluation Agreement is prohibited. Please contact Transend Corporation at (650) 324-5370 or sales.i...@transend.com to obtain a license suitable for use in a

Re: Translations: Chinese to English

2003-01-30 Thread Ronald R. Wellman
Hello Jeffrey, Try the following: http://www.ostrans.com/ I have used them extensively in the past to translate many Chinese and Korean documents into English. Best regards, Ron Wellman At 11:56 PM 1/29/2003 -0800, Collins, Jeffrey wrote: Group, Can you recommend a company or individual

Re: wire size

2003-01-22 Thread Ronald R. Wellman
Here is what I have: Total Cross Sectional Area (mm)^2(Brown Sharpe) AWG of single conductor 107.1459 84.9487000 67.433300 53.48210 42.40611 33.62362 26.66743 21.14744 16.76395 13.29936 10.54967 8.36748 6.63269 5.260410 4.172811 3.310312 2.624513 2.081614 1.651315 1.309016 1.038717 0.823518

Re: HAR Green / Yellow Wire

2003-01-21 Thread Ronald R. Wellman
Hello Doug, Check out the following standard: IEC Publication 60173, Colors of the cores of flexible cables and cords. The color scheme you are looking for is referenced in this standard. Best regards, Ron Wellman At 03:53 PM 1/20/2003 -0700, POWELL, DOUG wrote: Hello group, In the past I

Re: SV: NRTL in the U.S.

2003-01-17 Thread Ronald R. Wellman
Hello Amund, Unfortunately, your question can be broadly interpreted. As a matter of law the U.S. has agencies (FCC, FDA) that require product approvals depending on their intended use and function. Also, the U.S. has environmental laws that are overseen by the Environmental Protection Agency

RE: NRTL in the U.S.

2003-01-15 Thread Ronald R. Wellman
Hello Sam, Chances are if you are ever involved in a product liability law suit in the United States that the jury will look at you very unfavorably if your product is not NRTL Listed. Also, expect that the NRTL will not help defend you in a law suit even if your product is Listed with them.

RE: NRTL in the U.S.

2003-01-14 Thread Ronald R. Wellman
Hello Peter, This is probably true when an electrician is involved in a product installation. However, the majority of products that we are dealing with do not require an electrician to install. They are either installed by the Customer or a representative of the product manufacturer. The

RE: NRTL in the U.S.

2003-01-14 Thread Ronald R. Wellman
Hello Gregg, Martin is correct that NRTL Listing is not a requirement to sell electronic products in the U.S. Product usage is something entirely different and may require compliance under OSHA regulations. Best regards, Ron Wellman At 09:50 AM 1/14/2003 -0500, Gregg Kervill wrote: Joe,

Re: NRTL in the U.S.

2003-01-14 Thread Ronald R. Wellman
Hello Peter, Those of us who have had to list products in the United States already know this. What is required by law versus what is a customer requirement are two different things. NRTL listing is a Customer requirement as viewed from a product manufacturer. However, there are some cities,

Re: strange symbols (to me anyway)

2003-01-05 Thread Ronald R. Wellman
or iconic. My reference to necessity is only to point out that there is something currently in place that is being used internationally in place of this symbol. Best regards, Ron Wellman At 09:02 PM 1/4/2003 +, John Woodgate wrote: I read in !emc-pstc that Ronald R. Wellman rwell

Re: strange symbols (to me anyway)

2003-01-04 Thread Ronald R. Wellman
Hello Lou and Rich, Thanks for investigating this. It appears that we are living with a very old German standard that can be better replaced by using IEC 60529 for Ingress Protection (IP). Many safety standards are now making reference to IP numbers, so this for indoor use marking is

Re: In Search of Power Cord Test Equipment

2002-12-27 Thread Ronald R. Wellman
You should call a power cord manufacturer to see what they use. Depending on where your vendor is, I would try Electri-cord for North America, Feller for Europe, and Volex for the UK and Asia. At 09:05 AM 12/26/2002 -0500, oover...@lexmark.com wrote: One of our vendors is looking for test