RE: Surge immunity

1999-03-08 Thread Robert F. Martin ITS/QS-Box
The performance criterion for surge in light industrial immunity is B. This 'allows' some degradation of performance within specified tolerance. For a power supply, this may imply that you can monitor the output using a meter. However, because a power supply is used for some other device, it

RE: Room grounding

1998-10-09 Thread Robert F. Martin ITS/QS-Box
'Single point ground' in this situation is used for the same purpose as we would in electronics design. When using a shielded room for making low frequency measurements (particularly MIL-STD), multiple ground connections can make significant power frequency and harmonic noise. Generally this

RE: SLIM Committee

1998-10-08 Thread Robert F. Martin ITS/QS-Box
I am not sure of the appropriate web site, but the SLIM committee has issued its final report on the EMC directive. It includes 20 specific recommendations. VERY briefly, it encourages the Commission to evaluate existing standards and set up review processes for new ones, with a view towards

RE: Lab qualification

1998-10-06 Thread Robert F. Martin ITS/QS-Box
I am not sure what level of detail that you are looking for, but you might try to look at EN45001/2, and the associated guide for EMC labs that was published in the UK. Another avenue would be to model your format after the NVLAP program. They have a Handbook 150 (general) and 150-x (for

RE: teleco 48 volt specifications

1998-10-02 Thread Robert F. Martin ITS/QS-Box
While I haven't read a copy, you might try the BELLCORE document GR-513-CORE LSSGR, Section 13: Power. Bellcore can be contacted at 1(800)521-2673 for more information. Bob Martin Sr. Technical Manager Intertek Testing Services (978)263-2662 fax(978)263-7086 r...@itsqs.com The opinions

RE: NIS-81

1998-09-29 Thread Robert F. Martin ITS/QS-Box
NIS 81 is published by NAMAS in the UK. I am not aware of any availability on the web (but I haven't looked for it there). A companion/alternate document in the US is NIST tech note 1297. There is also an ISO document which addresses the topic. ETSI ETR 028 addresses MU as well. ANSI C63 has

RE: 2.4 GHz LAN Approbations

1998-09-23 Thread Robert F. Martin ITS/QS-Box
ITS has done several research projects regarding 2.4GHz LAN products. In the US, the products require CERTIFICATION of compliance with parts 15.109, 15.209, 15.107, 15.249. Safety requirements are per UL 1950, as you mention. In Europe, radio type approval, and EMC type examination is

RE: Radiation Monitors

1998-09-22 Thread Robert F. Martin ITS/QS-Box
There are any number of different radiation monitors available. Some are for rad-haz. Others are for EMC instrumentation. Holoday Instruments makes and markets the rad-haz monitors. They also OEM the probes for other companies (e.g.- Amplifier Research). They are located in Minnesota -

RE: USA-EU MOU

1998-09-04 Thread Robert F. Martin ITS/QS-Box
You may have heard by now, but this topic was discussed at the EMC Symposium. As with most government issues, there was no firm date applied to implementation. In Art Wall's words, we are close. There is still debate about approving US entities as CABs, and about European submittals for the

RE: Re[2]: Inconsistency of EN 50082-1 with 1000-4 series

1998-09-04 Thread Robert F. Martin ITS/QS-Box
Performance criteria is one of the hardest things to handle when doing the generic immunity evaluations. The standard is extremely vague, and does open the manufacturer to essentially define his own criteria. EN 50082-x does, however, put caveats on Category B, which causes the most

RE:

1998-09-02 Thread Robert F. Martin ITS/QS-Box
The standards organization is ABNT Comitê Brasileiro de Eletricidade - COBEI Associação Brasileira de Normas Técnicas - ABNT Rua Líbero Badaró, 496 - 3.o andar Caixa Postal 8363 São Paulo, SP, CEP 01008 Brasil The agency regulating radio/telecommunication is Anatel Agencia Nacional de

RE: Korea standard for drives

1998-09-01 Thread Robert F. Martin ITS/QS-Box
In general, Korea has statutory requirements which are identical to IEC/CISPR. For specifics, depending on the type of equipment, you can contact the Radio Research Laboratory in Korea. Korea will accept CB reports for safety, and these must be submitted to KITECH-KTL for a Korean report which

RE: Korea standard for drives

1998-09-01 Thread Robert F. Martin ITS/QS-Box
In general, Korea has statutory requirements which are identical to IEC/CISPR. For specifics, depending on the type of equipment, you can contact the Radio Research Laboratory (the controlling agency) in Korea (+82-0343-52-4001). Another government lab that can be used for testing is Haitong

RE: Conducted measurements on DC supplies

1998-06-30 Thread Robert F. Martin ITS/QS-Box
Like most compliance questions, there are a few different answers here, and more questions. 1) Are the measurements being made for a converter, or for your radio? 2) If the product is a radio, it has to meet other requirements besides EN55022 (e.g.- ETSI). In these cases, there are measurements

RE: CB scheme

1998-06-23 Thread Robert F. Martin ITS/QS-Box
Russia requires that you obtain a GOST-R mark. There are a few organizations that can obtain the mark for you. Just having the CB report will not get you in to the country. You must arrange for the GOST-R mark, through one of the approved agencies. The process can be very cumbersome. Having a CB

RE: Defibrillators

1998-04-21 Thread Robert F. Martin ITS/QS-Box
Doug If you haven't done so already, you might try ANSI/AAMI standard DF-39 Automatic External Defibrillators and remote control defibrillators. I used it a long while ago when testing the surge immunity of an implanted device. I don't remember the design specifics of the standard, because I was

RE: Harmonics

1998-04-20 Thread Robert F. Martin ITS/QS-Box
In my excitement I mis-typed the standard! It should, of course, be EN61000-3-2, not EN61000-3-1, for harmonics. Bob -- From: Robert F. Martin ITS/QS-Box To: emc-p...@ieee.org Subject: Harmonics List-Post: emc-pstc@listserv.ieee.org Date: Monday, April 20, 1998 7:55AM To (hopefully

Harmonics

1998-04-20 Thread Robert F. Martin ITS/QS-Box
To (hopefully!) put this to bed: The European Commission published a complete list of standards and amendments harmonized under the EMC Directive. The list includes references to eventual superseded standards and, when relevant, the date of cessation of presumption of conformity of the superseded

RE: Differences in performance criteria

1998-04-16 Thread Robert F. Martin ITS/QS-Box
If your intent is to satisfy the requirements of the EU for CE marking, then your report should refer to the performance criteria specified in EN 50082-1 (or a product specific standard as applicable). If your intent is only to do 'diagnostic' tests in accordance with a basic standard, then you

RE: Cable lengths needed

1998-04-16 Thread Robert F. Martin ITS/QS-Box
It has been my philosophy is such cases, that a 3 meter length should be provided. During radiated emissions and immunity tests, a minimum of 1 meter should be exposed. In the case of immunity, the 1 meter should be in the 'uniform' field area (even if you are not requiring the field uniformity

RE: EMC Directive and Small DC motors

1998-04-07 Thread Robert F. Martin ITS/QS-Box
While it seems simple enough, the product is more than just a component. It can be considered an apparatus by EU standards because it has an intrinsic function. If the pump was restricted to a specific application, and would not work without it, you might get away without testing it. Because that

RE: RUSSIA

1998-03-06 Thread Robert F. Martin ITS/QS-Box
Russian certification is done through GOSSTANDART. The GOST certificate is based on testing which is similar to the rest of the world (e.g. IEC 950, CISPR22, etc.). CE marking does not however satisfy the requirements. There are a few companies that can offer the GOST certificate. One of them is

EMI room filters

1998-03-06 Thread Robert F. Martin ITS/QS-Box
I am looking for 10 - 12 pre-owned filters for shielded enclosures. The filters should be: 277VAC 60 - 100 A 100dB DC-18GHz If you have any of these items that you are trying to get rid of, please contact me directly. Bob Martin ITS - Northeast Boxborough MA (978)263-2662

RE: CompactPCI

1998-01-26 Thread Robert F. Martin ITS/QS-Box
The military world has always specified that functional ground and chassis ground be isolated. The minimum resistance there is 1Mohm. The best solution tends to be the use of a chassis ground layer (instead of, or in addition to, circuit ground planes) on each board. This becomes the high

HALT testing

1998-01-19 Thread Robert F. Martin ITS/QS-Box
Can anyone enlighten me on the requirements for HALT testing? Bob Martin Sr. Technical Manager ITS- Northeast r...@itsqs.com P.S. - Thanks to everyone for recommending Werlatone directional couplers. (They should pay a finder's fee, there were so many recommendations!) In fact, I have several of

RE: GR-1089 Above 10GHz

1998-01-06 Thread Robert F. Martin ITS/QS-Box
My experience has been that the requirement is limited to 10GHz. This includes some major (and very critical!) telecom clients who operate at ~1.3GHz. Bob Martin Sr Technical Manager ITS - Northeast The opinions expressed are my own and not necessarily those of my employer -- From: Paul

RE: Chart of EMC standards?

1997-12-12 Thread Robert F. Martin ITS/QS-Box
Another source of standards information can be found on the web at DGIII's reference to harmonized standards www2.echo.lu/nasd Be careful of the date of the update (which is displayed on the page). They have a tendency to let it fall behind. At the moment it is fairly current. Also be careful of

Re: GOST Certificate

1997-12-10 Thread Robert F. Martin ITS/QS-Box
FYI I have attached an e-mail from one of our co-workers indicating that the flow for GOST certificates has become a problem. There is no sole accredited agency. At the moment, it depends largely on a company's ability to get test data approved by the Russian authorities. Intertek Testing Services

RE: CE Mark product standards for purchased equipment

1997-03-12 Thread Robert F. Martin ITS/QS-Box
. -- From: Tony Fredriksson To: Robert F. Martin ITS/QS-Box; emc-p...@ieee.org; jlovell%mfg.trim...@ngmsmtp.trimble.com Subject: RE: CE Mark product standards for purchased equipment List-Post: emc-pstc@listserv.ieee.org Date: Tuesday, March 11, 1997 11:36AM Hi, This is very intriguing. Let me