If it is not rack mounted, then they are over testing.
Ned Devine
Entela Inc
From: Peter L. Tarver [mailto:peter.tar...@sanmina-sci.com]
Sent: Tuesday, November 25, 2003 12:32 PM
To: PSTC
Cc: Robert Johnson
Subject: RE: Blocked ventillation testing
Bob -
You're quite right that the proposed
To: Ned Devine; Emc-Pstc
Subject: Re: Grounding
Ned,
I had this discussion with a TUV engineer way back in time. It is not in
the UL or IEC standards specifically AFAIK. What the TUV guy asked me to
do was consider the life of the product. The door hinge may have good
(low) resistance today
it said that
in the standard (IEC 60950-1 or IEC 60601-1). I looked and could not find it.
Am I just missing it, or is it not in the standards?
There is a reference in CSA standards to C22.2 No. 0.4 that has the
requirement, but I can not find a similar reference in UL or IEC standards.
Ned
Hi,
A headhunter contacted me about this position. If interested, please
contact Stratos at sdal...@prausa.com
Ned Devine
Entela Inc
Are you an EMC engineer looking for a company that will truly appreciate
your
skills? Do you want to work on next generation products you'll see
FYI
I just received notice of NARTE’s Product Safety Certification program. For
details, see www.narte.org http://www.narte.org/
Ned Devine
Entela, Inc.
3033 Madison Ave. SE
Grand Rapids, MI 49548
Phone: 616 248 9671
Fax: 616 574 9752
e-mail: ndev...@entela.com
www.entela.com
Paul,
If the centrifuge is used for blood, then it is covered under the MDD.
It is classified under Rule 3 and is a Class IIa device. See the MEDDEV
Guidelines for the Classification of Medical Device. Centrifugation of
blood to prepare it for transfusion or autotransfusion.
Ned Devine
interpolate. See recommendation 27.
If you should have any questions or comments, please feel free to contact me.
Sincerely,
Ned Devine
Entela, Inc.
3033 Madison Ave. SE
Grand Rapids, MI 49548
Phone: 616 248 9671
Fax: 616 574 9752
e-mail: ndev...@entela.com
www.entela.com
Entela, Inc. A Certified
locating the engineer in one of these offices.
Entela is an OSHA approved NRTL; accredited by the SCC, for electrical
product safety; and a participating National Certification Body (NCB)
within the IECEE/CB scheme.
If you are interested, or would like more information, please contact
me.
Ned Devine
Hi,
In the past I have used Henkel Sicomet 77 to adhere thermocouples. I
just tried to reorder some and was told it has been discontinued.
Does anyone have a recommendation on a replacement?
Thanks
Ned Devine
Entela, Inc.
3033 Madison Ave. SE
Grand Rapids, MI 49548
Phone: 616 248 9671
Hi,
Yes it is required for some of the directives. For the MDD, the requirement
is in Article 14, paragraph 2. For the IVDD it is in Article 10, paragraph
3.
If you should have any questions or comments, please feel free to contact
me.
Sincerely,
Ned Devine
Entela, Inc.
3033 Madison Ave. SE
scope to humans. The scope for
electrical laboratory equipment includes ...measures, indicates, monitors
or analyses substances, or is used to prepare materials, This sounds
like the purpose of your device.
Ned Devine
Entela, Inc.
3033 Madison Ave. SE
Grand Rapids, MI 49548
Phone: 616 248
Hi,
What is the software for? It is possible that it will fall under the
Medical Device Directive or IVDD (and FDA) if it is used in the medical
field.
If you should have any questions or comments, please feel free to contact
me.
Sincerely,
Ned Devine
Entela, Inc.
3033 Madison Ave. SE
Grand
Hi,
In the USA the FDA has a guidance document that some of the branch's (i.e.
Anesthesiology and Respiratory) use. See
http://www.fda.gov/cdrh/ode/638.pdf
This includes both EMI, and mechanical and environmental requirements.
Ned Devine
Entela, Inc.
3033 Madison Ave. SE
Grand Rapids, MI
of two layers can withstand
the dielectric
strength test for REINFORCED INSULATION.
Ned Devine
Program Manager
Entela, Inc.
3033 Madison Ave. SE
Grand Rapids, MI 49548
1 616 248 9671 Phone
1 616 574 9752 Fax
ndev...@entela.com e-mail
Entela, Inc. A Certified Woman Owned Business
www.entela.com
Hi,
Got this from a headhunter. The job is in Pennsylvania. If interested,
please contact him directly.
Ned Devine
Program Manager
Entela, Inc.
3033 Madison Ave. SE
Grand Rapids, MI 49548
1 616 248 9671 Phone
1 616 574 9752 Fax
ndev...@entela.com e-mail
Entela, Inc. A Certified Woman Owned
.
Ned Devine
Program Manager
Entela, Inc.
3033 Madison Ave. SE
Grand Rapids, MI 49548
1 616 248 9671 Phone
1 616 574 9752 Fax
ndev...@entela.com e-mail
Entela, Inc. A Certified Woman Owned Business
www.entela.com
-Original Message-
From: Brooks, Barbara [mailto:bbro...@hnt.wylelabs.com
Hi,
Thanks to all who responded. This seems to be the answer I was looking for.
Time to get out the physics book and do a little research.
Ned
Ned Devine
Program Manager
Entela, Inc.
3033 Madison Ave. SE
Grand Rapids, MI 49548
1 616 248 9671 Phone
1 616 574 9752 Fax
ndev...@entela.com e-mail
are made of brass.
Thanks
Ned
Ned Devine
Program Manager
Entela, Inc.
3033 Madison Ave. SE
Grand Rapids, MI 49548
1 616 248 9671 Phone
1 616 574 9752 Fax
ndev...@entela.com e-mail
Entela, Inc. A Certified Woman Owned Business
www.entela.com
61010-1, Second Edition. I also noted references to UL
61010B-1 and UL 61010C-1. I am assuming that UL is splitting out
laboratory, control and measuring equipment into three standards.
Ned
Ned Devine
Program Manager
Entela, Inc.
3033 Madison Ave. SE
Grand Rapids, MI 49548
1 616 248 9671
to write software.
We have also had a problem with data dropping (1 or 2 readings out of a
scan), but only at high speed and to a notebook PC. We assumed it was a
function of the RS-232 bus.
Ned
Ned Devine
Program Manager
Entela, Inc.
3033 Madison Ave. SE
Grand Rapids, MI 49548
1 616 248
Hi,
Yes. Triple checked.
Ned Devine
Program Manager
Entela, Inc.
3033 Madison Ave. SE
Grand Rapids, MI 49548
1 616 248 9671 Phone
1 616 574 9752 Fax
ndev...@entela.com e-mail
Entela, Inc. A Certified Woman Owned Business
www.entela.com
-Original Message-
From: plaw...@west.net
being
to IEC 60521 in which 400 A/m is required.
For the Second Edition of IEC 60601-1-2, Clause 36.202.8.1, the limit is 3
A/m.
Ned Devine
Program Manager
Entela, Inc.
3033 Madison Ave. SE
Grand Rapids, MI 49548
1 616 248 9671 Phone
1 616 574 9752 Fax
ndev...@entela.com e-mail
Entela, Inc
Hi,
He needs Field Evaluation. The investigation and cost is based on an unit
price. Have him contact me.
Ned Devine
Program Manager
Entela, Inc.
3033 Madison Ave. SE
Grand Rapids, MI 49548
1 616 248 9671 Phone
1 616 574 9752 Fax
ndev...@entela.com e-mail
-Original Message-
From
and address of the
organization responsible for the product may be used.
Ned Devine
Program Manager
Entela, Inc.
3033 Madison Ave. SE
Grand Rapids, MI 49548
1 616 248 9671 Phone
1 616 574 9752 Fax
ndev...@entela.com e-mail
-Original Message-
From: Peter Tarver [mailto:peter.tar
of lower limits?
The device in question meets IEC 60601-1-2 First Edition, but will not pass
the requirements in the Second Edition. Can the justification be that
since there are no known risks on the device, the current level of immunity
is adequate?
Thanks
Ned
Ned Devine
Program Manager
Entela
Nick,
The FDA classification for hearing aids, air-conduction, are at
http://www.accessdata.fda.gov/scripts/cdrh/cfdocs/cfPCD/classification.cfm?I
D=943
They are Class I, Exempt devices.
Ned
Ned Devine
Program Manager III
Entela, Inc.
3033 Madison Ave. SE
Grand Rapids, MI 49548
616 248
. The requirements are Basic Insulation,
Dielectric is 1,500 Vac. Creepage is 4.0 mm, Air Clearance is 2.5 mm. You
can use anything the meets these requirements. See also Clause 17.a) for
separation requirements.
Ned
Ned Devine
Program Manager III
Entela, Inc.
3033 Madison Ave. SE
Grand Rapids, MI
that you comply with IEC 60601-1-2.
Ned
Ned Devine
Program Manager III
Entela, Inc.
3033 Madison Ave. SE
Grand Rapids, MI 49548
616 248 9671 Phone
616 574 9752 Fax
ndev...@entela.com e-mail
-Original Message-
From: am...@westin-emission.no [mailto:am...@westin-emission.no]
Sent
Peter,
Be careful, some agencies have taken the 1.0 mm thickness requirement for
transformers (57.9.4.e, first dash) and applied it to potted components the
need reinforced insulation. IEC 60950, I believe only requires 0.4 mm for
reinforced.
Ned Devine
Program Manager III
Entela, Inc.
3033
and 2-20. 2-19 and 2-20 both require the unit
to pass at 3 V/m and to pass or fail safe at 10 V/m.
Ned Devine
Program Manager III
Entela, Inc.
3033 Madison Ave. SE
Grand Rapids, MI 49548
616 248 9671 Phone
616 574 9752 Fax
ndev...@entela.com e-mail
-Original Message-
From: Ralph
the marketing guys :-)
I know this is a lot of information, so I am expecting to pay for it. If
you can help, please contact me.
Thanks
Ned
Ned Devine
Program Manager III
Entela, Inc.
3033 Madison Ave. SE
Grand Rapids, MI 49548
616 248 9671 Phone
616 574 9752 Fax
ndev...@entela.com e-mail
Hi,
My guess is that it has to do with the Machinery Directive. See Article 4,
paragraph 2 and point B of Annex II. This deals with machinery ...that it
is intended to be incorporated into machinery or assembled with other
machinery to constitute machinery covered by this Directive...
Ned
lifetime is up?
Ned Devine
Ned Devine
Program Manager III
Entela, Inc.
3033 Madison Ave. SE
Grand Rapids, MI 49548
616 248 9671 Phone
616 574 9752 Fax
ndev...@entela.com e-mail
-Original Message-
From: Dick Grobner [mailto:dick.grob...@medgraph.com]
Sent: Monday, June 25, 2001 7:02 PM
be able to switch to the ohm
meter.
Ned
Ned Devine
Program Manager III
Entela, Inc.
3033 Madison Ave. SE
Grand Rapids, MI 49548
616 248 9671 Phone
616 574 9752 Fax
ndev...@entela.com e-mail
---
This message is from the IEEE EMC Society Product Safety
this information.
If you place the device on the market or into service, it must have the CE
marking. There is an Annex in the MDD (X) that talks about Clinical
Evaluations. You might be able to use this.
Ned
Ned Devine
Program Manager III
Entela, Inc.
3033 Madison Ave. SE
Grand Rapids, MI 49548
, this is only OK if you can be sure that you know which will be the
neutral.
Ned
Ned Devine
Program Manager III
Entela, Inc.
3033 Madison Ave. SE
Grand Rapids, MI 49548
616 248 9671 Phone
616 574 9752 Fax
ndev...@entela.com e-mail
-Original Message-
From: geor...@lexmark.com [mailto:geor
of the receptacle. Receptacles so identified shall be used only with
grounding conductors that are isolated in accordance with Section
250-146(d). Isolated ground receptacles installed in nonmetallic boxes shall
be covered with a nonmetallic faceplate.
Ned
Ned Devine
Program Manager III
Entela
article 4, paragraph 4
Member States may require the information to be supplied pursuant to Annex
I, part B, section 8 to be in their official language(s) when a device
reaches the final user.
Ned
Ned Devine
Program Manager III
Entela, Inc.
3033 Madison Ave. SE
Grand Rapids, MI 49548
616 248 9671
contractual access to that information.
Ned
Ned Devine
Program Manager III
Entela, Inc.
3033 Madison Ave. SE
Grand Rapids, MI 49548
Phone 616 248 9671
Fax 616 574 9752
e-mail ndev...@entela.com
-Original Message-
From: wo...@sensormatic.com [mailto:wo...@sensormatic.com]
Sent: Tuesday
these
municipalities.
Ned Devine
Entela, Inc.
Program Manager III
Phone 616 248 9671
Fax 616 574 9752
e-mail ndev...@entela.com
-Original Message-
From: Veit, Andy [mailto:andy.v...@mts.com]
Sent: Friday, January 26, 2001 7:06 AM
To: 'Rich Nute'
Cc: 'emc-p...@majordomo.ieee.org'
Subject
) you want to put it out.
You don't want the fire department to use their hoses to put it out.
Ned Devine
Entela, Inc.
Program Manager III
Phone 616 248 9671
Fax 616 574 9752
e-mail ndev...@entela.com
-Original Message-
From: Ken Javor [mailto:ken.ja...@emccompliance.com]
Sent: Friday
) parts in medical equipment? Yes!
Do I also want to limit the patient leakage current to as low as 10 micro
amps? Yes! Do I need these levels for my computer? No.
Ned Devine
Entela, Inc.
Program Manager III
Phone 616 248 9671
Fax 616 574 9752
e-mail ndev...@entela.com
-Original Message
have
both the responsibility and authority and unless there was fraud involved,
he is going to be liable.
Ned Devine
Entela, Inc.
Program Manager III
Phone 616 248 9671
Fax 616 574 9752
e-mail ndev...@entela.com
-Original Message-
From: Mark Schmidt [mailto:mschm...@xrite.com]
Sent
to
their being placed on the market under his own name. This subparagraph does
not apply to the person who, while not a manufacturer within the meaning of
the first subparagraph, assembles or adapts devices already on the market to
their intended purpose for an individual patient;
Ned Devine
. For
example, see UL 1950, Third Edition, Clause 5.4.4 c).
Ned Devine
Entela, Inc.
Program Manager III
Phone 616 248 9671
Fax 616 574 9752
e-mail ndev...@entela.com
-Original Message-
From: Rich Nute [mailto:ri...@sdd.hp.com]
Sent: Wednesday, November 29, 2000 7:38 PM
To: matsu
anyone have any comments or ideas?
Thanks
Ned Devine
Entela, Inc.
Program Manager III
Phone 616 248 9671
Fax 616 574 9752
e-mail ndev...@entela.com
---
This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list
one?
- Are there any labs in the USA that do this type of testing?
Thanks
Ned Devine
Entela, Inc.
3033 Madison Ave. S.E.
Grand Rapids, MI 49548
Program Manager III
Phone 616 248 9671
Fax 616 574 9752
e-mail ndev...@entela.com
of approximately 0,7 kg/l.
Our Chemist researched it and found it is Hexane, 95+%. The Aldrich Cat.
No. is 20,875-2. The CAS number is 110-5A-3.
The Isopropyl Alcohol is 100% (not rubbing alcohol) and is a very aggressive
solvent. It is at least as harsh as the Hexane.
Ned Devine
Entela, Inc
is Classified to UL 2601-1. This is mostly because UL does not require you
to meet all of the requirements. They say the FDA (US Government agency)
covers them.
Ned Devine
Entela, Inc.
Program Manager III
Phone 616 248 9671
Fax 616 574 9752
e-mail ndev...@entela.com
-Original Message
Brent's example of adding an alternate plastic. If no tests are
necessary, then it would not be an amendment. If there was testing, then it
would be an amendment.
Ned Devine
Entela, Inc.
Program Manager III
Phone 616 248 9671
Fax 616 574 9752
e-mail ndev...@entela.com
-Original Message
for accessible
parts. You will need to verify that the circuit meets the requirements of
6.3.2.1, 6.3.2.2 and 6.3.2.3 during a single fault.
Ned Devine
Entela, Inc.
Program Manager III
Phone 616 248 9671
Fax 616 574 9752
e-mail ndev...@entela.com
-Original Message-
From: Loop, Robert
60664.
Maybe - They now require a risk management process. See Clause 4.6 in the
draft. I suppose you could use the risk assessment to prove that the
spacing you wanted to use were OK.
Ned Devine
Entela, Inc.
Program Manager III
Phone 616 248 9671
Fax 616 574 9752
e-mail ndev...@entela.com
connect is Type BF. The patient leakage current limit is 100 uA in normal
condition and 500 uA in single fault condition.
3. Call UL. Maybe I missed some out in UL544.
Ned Devine
Entela, Inc.
Program Manager III
Phone 616 248 9671
Fax 616 574 9752
e-mail ndev...@entela.com
. The NB will tell you what they
want to see.
Ned Devine
Entela, Inc.
Program Manager III
Phone 616 248 9671
Fax 616 574 9752
e-mail ndev...@entela.com
-Original Message-
From: plaw...@west.net [mailto:plaw...@west.net]
Sent: Monday, March 20, 2000 12:46 PM
To: EMC-PSTC
Subject
. If the USA, hospital grade power cords have clear
ends. The nurse saw the Black, White and Green wires in the IEC 320
connector and plugged in the leads. The infant died. I believe that the
FDA had reports of 8 similar incidences.
Ned Devine
Entela, Inc.
Program Manager III
Phone 616 248 9671
Fax
Hi,
I have always interpreted the requirement to include slots. My thinking was
that what ever could bridge a groove could also bridge a slot.
Ned Devine
Entela, Inc.
Program Manager III
Phone 616 248 9671
Fax 616 574 9752
e-mail ndev...@entela.com
-Original Message-
From: Dirk
Hi Mark,
No. The MDD, Article 14 address this. You only have to register with the
CO of the country where you have your authorized representative. But, I
have heard that some countries have passed national laws requiring you to
register all devices.
Ned Devine
Entela, Inc.
Program Manager
Hi,
I was quoting from the new edition, 3rd, of IEC 60950.
But, it does bring up the question of when the EN, UL/CSA versions will be
issued. Anyone have the proposed dates?
Ned Devine
Entela, Inc.
Program Manager III
Phone 616 248 9671
Fax 616 574 9752
e-mail ndev...@entela.com
Hi,
You might be in trouble, Clause 1.5.6 of IEC 60950, 3rd edition requires Y.
See Clause below.
Ned Devine
Entela, Inc.
Program Manager III
Phone 616 248 9671
Fax 616 574 9752
e-mail ndev...@entela.com
1.5.6 Capacitors in primary circuits
A capacitor connected between two line
Hi,
Yes, you need three or more layers of pre-preg in between the SELV and
MAINS traces.
Ned Devine
Entela, Inc.
Program Manager III
Phone 616 248 9671
Fax 616 574 9752
e-mail ndev...@entela.com
-Original Message-
From: Andrews, Kurt [mailto:kandr...@tracewell.com]
Sent: Tuesday
. You also would not have to do any type or routine testing.
Ned Devine
Entela, Inc.
Program Manager III
Phone 616 248 9671
Fax 616 574 9752
e-mail ndev...@entela.com
-Original Message-
From: Andrews, Kurt [mailto:kandr...@tracewell.com]
Sent: Tuesday, January 11, 2000 9:29 AM
To: Ned
Hi,
Very interesting. UL 1950, Clause 2.9.6 and IEC 60950, 3rd edition, Clause
2.10.5.3 both state that the distance between two adjacent tracks on the
same layer of a PCB is treated as distance through insulation.
Ned Devine
Entela, Inc.
Program Manager III
Phone 616 248 9671
Fax 616 574
(i.e. MDD, Annex II, paragraph 3.1, third -). I am assuming
that the other directives have a similar requirement.
Ned Devine
Entela, Inc.
Program Manager III
Phone 616 248 9671
Fax 616 574 9752
e-mail ndev...@entela.com
-Original Message-
From: wo...@sensormatic.com [mailto:wo
Chubb Instrumentation, model JCI 140 Static Monitor. Does anyone have any
experiences with this meter or others?
Ned Devine
Entela, Inc.
Program Manager III
Phone 616 248 9671
Fax 616 574 9752
e-mail ndev...@entela.com
-
This message is coming from the emc-pstc discussion list.
To cancel
Hi,
For a list of Accredited certification organizations go to
http://www.scc.ca/certific/colist.html
For the scope of Accredited certification organizations go to
http://www.scc.ca/search-front/indexacc.html
Ned Devine
Entela, Inc.
Program Manager III
Phone 616 248 9671
Fax 616 574 9752
on burning. The fire went up and out the top ventilation opening
and into the building ceiling. At about this time, the fire department
showed up and put the fire out. As with most fires, the majority of damage
was from the water and the smoke.
Ned Devine
Entela, Inc.
Program Manager III
Phone
; Fax 800-962-9591
Technical Service: 800-231-8327; Fax (414) 287-4079
E-Mail: aldr...@sial.com
Ned Devine
Entela, Inc.
Program Manager III
Phone 616 248 9671
Fax 616 574 9752
e-mail ndev...@entela.com
-Original Message-
From: Ehler, Kyle [mailto:kyle.eh...@lsil.com]
Sent: Wednesday
Hi,
I would think a thermocouple would be OK. It should be close to +/- 1 deg.
C. If you need better accuracy, try a RTD. It should be +/- 0.1 deg. C.
Ned Devine
Entela, Inc.
Program Manager III
Phone 616 248 9671
Fax 616 574 9752
e-mail ndev...@entela.com
-Original Message-
From
Hi,
See Annex A.5, Hot Flaming Oil test.
I have never done this test, but I have the ladle. Good luck.
Ned Devine
Entela, Inc.
Program Manager III
Phone 616 248 9671
Fax 616 574 9752
e-mail ndev...@entela.com
-Original Message-
From: John Juhasz [ mailto:jjuh
Hello,
This is what we use. We have a number of power supplies from 50 amps to
1000 amps. We use a calibrated current shunt to verify the output current.
Ned Devine
Entela, Inc.
Program Manager III
Phone 616 248 9671
Fax 616 574 9752
e-mail ndev...@entela.com
-Original Message
Hi,
That reference is to the old CB Bulletin 89AI, page 117. It has been
superseded by CB Bulletin, 94AI, page 108. That deviation has been removed.
Ned Devine
Entela, Inc.
Program Manager III
Phone 616 248 9671
Fax 616 574 9752
e-mail ndev...@entela.com
-Original Message-
From
Hi,
See below for answers to your questions. In summary, IEC 60601-1, Clause
56.11 does not allow this type if construction.
Ned Devine
Entela, Inc.
Program Manager III
Phone 616 248 9671
Fax 616 574 9752
e-mail ndev...@entela.com
Dear All,
One of the sections in the medical equipment
or the lifetime of the
product.
Ned Devine
Entela, Inc.
Program Manager III
Phone 616 248 9671
Fax 616 574 9752
e-mail ndev...@entela.com
-Original Message-
From: Ehler, Kyle [mailto:kyle.eh...@lsil.com]
Sent: Thursday, July 22, 1999 2:32 PM
To: Scott Douglas; 'EMC-PSTC'
Subject: RE: Record
Hi,
Does anyone know what is the directive for fire extinguishers. I checked
the simple pressure vessel directive, but it excludes fire extinguishers.
If it make any difference, this particular fire extinguisher is for use in
the engine compartment of a boat.
Thanks
Ned Devine
Entela, Inc
of the blades or towards the body of the
plug-in? Also, is the engagement face where the blades enter the body or
where the body rests on the wall. For North America this is the same, but
for some European outlets there is a difference.
Ned Devine
Entela, Inc.
Program Manager III
Phone 616 248
for a
thermostat to get stuck in the closed position and then start working?
Maybe after being subjected to a mechnical shock? The ones I have checked
seem to be working. e.g. are in the normally closed position and when
heated open and then close when cooled.
Ned Devine
ndev...@entela.com
Phone 616 248
Hi,
The Machinery Directive (and many others) was amended by the CE marking
directive 93/68/EEC (OJ L220, 8/30/93, page 8) to remove the year of
manufacture from the CE marking. The same Directive also amended, Annex
I, Clause 1.7.3 to require - the year of construction. to be on the unit.
Ned
, (and
assumes all responsibilities except design) do they become the
manufacturer?
Answer: If you name is not on the unit, you are not the manufacture and
have no responsibilities under the MDD.
Ned Devine
Entela, Inc.
Program Manager III
Toll Free: 1 800 888 3787 Ext. 671
Direct: 1 616 248
Hi,
At a former employer (EU Notified Body based in Germany), we were told to
ignore normal LED's, unless when looking into them they looked
Dazzling. The rational was that the requirement was going to be changed
to exempt them and there was no sense in enforcing it.
Ned Devine
Entela, Inc
Hi,
I am looking for labs that can perform some or all of the testing required
per the telecommunications terminal equipment directive.
Please respond to me.
Thanks
Ned Devine
Entela, Inc.
Program Manager III
Toll Free: 1 800 888 3787 Ext. 671
Direct: 1 616 248 9671
Fax
wheatstone bridge like the Biddle.
Thanks
Ned Devine
Entela, Inc.
Toll Free: 1 800 888 3787 Ext. 671
Direct: 1 616 248 9671
Fax:1 616 574 9752
e-mail: ndev...@entela.com
-
This message is coming from the emc-pstc discussion list.
To cancel your
NOT
concerned with shock and vibration during shipping, just during use.
What, if anything, do PC makers do for operational vibration?
Any help would be appreciated.
Ned Devine
Entela, Inc.
Program Manager III
Toll Free: 1 800 888 3787 Ext. 671
Direct: 1 616 248 9671
Fax:1 616 574
Hi,
No. The 4th amendment to IEC 950 has circuits and component values for
the 10/700 and 1.2/50 impulses.
Ned Devine
Entela, Inc.
Program Manager III
Toll Free: 1 800 888 3787 Ext. 671
Direct: 1 616 248 9671
Fax:1 616 574 9752
e-mail: ndev...@entela.com
Hi,
The 1.2/50 us impulse circuit is described in IEC 950, 4 th Amendment,
Annex N. It has values for the components. If you don't have the
standard, give me a call.
Ned Devine
Entela, Inc.
Program Manager III
Toll Free: 1 800 888 3787 Ext. 671
Direct: 1 616 248 9671
Fax
at the top vent for the chamber
Lesson No. 4, use 360 Deg. sprinkler heads in the chamber.
Lesson No. 5, Murphy was an optimist.
Ned Devine
Entela, Inc.
Program Manager III
Toll Free: 1 800 888 3787 Ext. 671
Direct: 1 616 248 9671
Fax:1 616 248 0591
e-mail
: Jon D Curtis[SMTP:j...@world.std.com]
Sent: Friday, July 10, 1998 10:37 AM
To: Chris Dupres
Cc: Ned Devine; emc-pstc
Subject: Re: isopropyl alcohol
What if anything are laboratories doing to calibrate chemical
solutions?
Seems to me that as an ISO Guide 25 laboratory
Hi,
I had someone ask me about the UK electricity at work act of 1989. He
said it has something to do with leakage current.
Has anyone heard of this? If so, what is it?
Ned Devine
Entela, Inc.
Toll Free: 1 800 888 3787 Ext. 671
Direct: 1 616 248 9671
Fax:1 616 248
Hello,
I am looking for the CB Scheme national deviations to IEC 950. I
believe that they are specified in CB Bulletin 89A I. Can anybody
verify this? Also, does anyone know if the CB Bulletin's are on the
web?
Thanks
Ned Devine
Entela, Inc.
Program Manager III
Toll Free: 1 800 888
Mel,
Make sure that the treadmill is not part of a medical Stress System.
Typically, a Stress System uses a treadmill to stress the heart.
Ned Devine
Entela, Inc.
Toll Free: 1 800 888 3787 Ext. 671
Direct: 1 616 248 9671
Fax:1 616 248 0591
e-mail: ndev
sound level meter. But, the
manufacture states that it does not meet the requirements of IEC 1010-1.
Ned Devine
Entela, Inc.
Toll Free: 1 800 888 3787
Direct: 1 616 248 9671
Fax:1 616 248 0591
e-mail: ndev...@entela.com
Hello,
I believe that the EU has restrictions on the use of cadmium (e.g.
cadmium plated connectors). Does any one know where this is documented?
Ned Devine
Entela, Inc.
Toll Free: 1 800 888 3787
Direct: 1 616 248 9671
Fax: 1 616 248 0591
e-mail: ndev...@entela.com
confusion, we do not allow products to be marked device
A, made by B, or device A made by B for C, or device A made for B, or
device A distributed by B, or similar wording.
Ned Devine
TUV Product Service
ndev...@tuvps.com
(f) 'manufacturer' means the natural or legal person with
responsibility
it to
replace your IEC 320 attachment plug with a IEC 309 attachment plug.
Ned Devine
TUV Product Service
ndev...@tuvps.com
612 638 0246
-
Original Text
From: ray_russ...@leco.com, on 04/17/1997 9:05 AM:
To: emc-p...@ieee.org
Dear Friends,
We have a large piece of lab equipment
Hello,
I don't know if it is on the WEB, but I have an electronic copy of the MDD.
It is about 180Kb. I have not check to see if it is 100% accurate. So
use at your own risk. Let me know if you would like a copy.
Ned Devine
TUV Product Service
ndev...@tuvps.com
-
Original
Hello,
In general, we require that all non-operator replaceable fuse(s) be
specified in the service manual. e.g F1 - Type T, 3 A, 250 V.
Ned Devine
TUV Product Service
ndev...@tuvps.com
-
Original Text
From: Randy Stephens steph...@vivanet.com, on 03/11/1997 10:53 AM:
To: emc-p
Hi,
Per the Low Voltage Directive, Article 10, TUV Product Service's number is
5.
Ned Devine
TUV PS
ndev...@tuvps.com
-
Original Text
From: George Alspaugh geor...@lexmark.com, on 02/25/1997 9:31 AM:
To: emc-pstc%ieee.org emc-p...@ieee.org
We have used TUV Rheinland